I tried to do a multiple channel with PCI 6229 Council 2090 BNC data acquisition work. After you have created the virtual channel and start the DAQ task when I acquire the signal, I get a strange behavior. I give a square input signal of 1 volt and zero volts.

When I use the 2D U16 Nchnl Nsamp: I get peaks in amplitude constant for the zero volt part, where there are reading the 1 volt portion correctly.

Whereas when I use the Nsamp of Nchnl of Wfm D 1: I get about the signal that is entered into the system with a behavior of square wave and a suitable voltage range.

Please suggest as to what is the difference between the two conditions, and perhaps why am I cleats for the zero volt part.

Concerning

Amanmeet Garg

Graduate student

Simon Fraser University

Burnaby, Canada.

Thanks Dennis,

Your reference is a good information. I was able to solve my problem by keeping the acquisition as a signed integer of 16 bit, as I am able the negative tensions as well.

therefore in the region of zero volts /-EV system was him giving a value of 65535 in case of unsigned 16-bit sampling.

I'm balance correction and external filtering in MATLAB and get reasonable results.

Concerning

Amanmeet Garg

Tags: NI Software

Similar Questions

  • Collection of samples for 1 d digital Wfm N Chan NSamp

    Hello

    I'm stuck in obtaining multiple samples.

    The system configuration:

    NOR SMU-1062 q

    NEITHER SMU-6341

    SCB-68 M series

    I use Digital 1 d Wfm N Chan NSample for data output. I used the digital waveform data table.

    I checked each lines with DMM, and I could only check the last sample. (dt = 4S or more).

    Data table:

    Example 0: 010 1010

    Example 1: 001 0100

    Example 2: 100 1010

    Results: line 0 (PIN 52) (1), line 1 (0), line 2 (0), Line3 (1), Line4 (0), Line5 (1), Line6 (0) - (another Question: it looks like reverse the order for me...)

    You have any suggestions to check all the samples?

    Allow me to attach my file vi.

    Thank you

    Insuk

    Hello

    I found the solution.

    I have not configured the "DAQ - mx sample clock.

    Happy Thanksgiving day!

    Thank you

    Insuk

  • With the help of an output to trigger a task WITHOUT additional cables?

    Hello

    I have a PXI-6229 (Multifunction DAQ) and a PXI-6723 (AO, with a few lines of DIO). My channel mappings are not set in stone yet, so I can use either module.

    For a test system, I use producing 2 continuous and periodicals of digital signals in the background (via 2 channels). The 2 waveforms have different profiles, but they are synchronized by being linked to the same task (Digital 1 Wfm d NChan NSamp).

    Model has: 0 0 1 0 1 0 1 0

    Model b: 1 0 0 0 0 0 0 0

    On the one hand the test steps, I need to read a waveform of voltage over (via a channel of AI). When I get to this step, I want to start reading at one of the rising edges of the B model. How can I accomplish this?

    A method I can think of is:

    • Duplicate the model B on another channel
    • Connect to a PFI line on my controller
    • Use the PFI input to trigger the task to HAVE it.

    Is it possible to do without an extra cable? Thanks in advance!

    Hi JKSH,

    First of all, I want to say that your proposed method is a completely logical and it will work. However, this method includes a more physical routing so the trigger can be delayed.

    You can use the export DAQmx (VI) Signal to route your DIO (with pattern B) line to another exit of Terminal such as a trigger. You can then use this signal newly routed to trigger your channel HERE. Here is the documentation for this specific VI:

    http://zone.NI.com/reference/en-XX/help/370469AA-01/lvdaqmx/mxexportsignal/

  • Error-50352 to DAQmx analog writing

    Hello

    I get the following error message when you try to write data points about 20 million to a Council of PXI - 6713:

    Error-50352 occurred at DAQmx Write (Wfm Analog 1Chan NSamp) .vi:1

    Possible reasons:

    The requested memory could not be allocated.

    Task name: _unnamedTask<1>

    I don't understand the error memory. When this happens, use LabVIEW is, according to the Task Manager Windows, only 750 MB of memory, perhaps 950 MB max, and I ran things successfully when LV was used 1.6 GB of memory.  I have increased my settings of virtual memory to 3 GB for Windows Vista 32-bit. I have 4 GB of memory total, and I have a core duo 1.8 Mhz CPU.

    My VI is attached. I read in a waveform of singles, do a conversion in rooms double and try to write to the Commission. I use DAQmx calendar sample clock instead of "Using the waveform", because using the sample clock results in one less copy of the data. I read the data in singles because that is how I write in advance, and it is much faster to read in simple it is double.

    Is this something that will have to be split into several Scriptures? If so, how does that affect performance? I have a very limited time)<1s) to="" write="" this="" data="">

    Thanks for any help and please let me know if I need to provide more details.

    Hi Djaunl,

    I ran your program with a DDHN file that I created and got the same error your saw.  I then ran the program with highest point running on - the light bulb it the top of the block diagram. I found that the error occurs on the second loop.  This leads me to believe that it was something to do with the implementation of the task.  When I replaced the clear DAQmx with DAQmx stop and this solved the problem.  The reason that this solves the problem, it is the DAQmx clear clears the task each time that the loop runs, thus freeing up your calendar information.  Whe you switch to stop DAQmx, task configuration information are not totally clear and the error is resolved.

  • DAQmx Read 1 d vs 2D DBL WFM

    Another that the convenience of the format of the data acquired, are there benefits (e.g., memory, speed, etc.) usage

    DAQmx readout 1 d Wfm NChan NSamp

    -Or-

    DAQmx read analog 2D DBL NChan NSamp

    I'm a life-long in multiple buffer
    channels using an architecture of producer/consumer. I am not concerned by information in absolute time available in the format of the wave form; It is convenient, but I can get what I need also.  In post processing of the data, the format is a stack or face.  If I collect the Wfm 1 d data I end up stripping to the Y-data for certain analyses, but eventually the construction of waveforms for the analysis of the other if I collect a given 2D DBL.

    -CTF

    I did a quick comparative analysis and it seems that the WFM D 1 is slower and takes more memory than the 2D DBL.  The overall impact of these differences depends on the size of your data blocks.

  • Get the Type (U8, U16... I64) controls within a Cluster

    Hello

    I have a cluster with multiple controls.

    I would get the type of each control inside. To get the class is ok. It's simple and they are all digital. But I need to know if they are U8, U16, U32... I64.

    No idea how to do this?

    Paul

    I got it.

    I had to convert the Refernce element to digital, and then use the BitAccurateDataSize property.

    I lost several hours to find the solution and in 5 minutes after I had posted my problem on the forum I solved it .

  • Open image greyscale U16 problem

    Hi all

    The attached picture was taken by a camera scientific and registered as gray U16 tiff file. This can be confirmed using ImageJ, a popular freeware in the field of microscopy. But when I try to use NI Vision LV2012 screws to open it, I found LV always recoganizes as image I16 and gives a bad display. Could someone help em to deal with this problem?

    Thank you!

    Xsword

    Hello

    I think that there is something wrong with your tiff file. Because when I save it with LabVIEW in U16, I can reopen it perfectly (see attached file). What program did you use to record?

    (If you use IMAQdx to retrieve your images, I know there are a signedeness of pixels that can affect this.) See attributes of your camera under MAX.)

  • Bayer U16 2D table

    Hello

    I use a pilot of the custom camera that gives me image data as Bayer formed in a U16 2D table data. I wonder if I can use the function "IMAQ Bayer to RGB" with it.

    I would be very grateful for the advice.

    Thank you

    D. Ackermann

    Yes, this VI allows you to do exactly that:
    http://zone.NI.com/reference/en-XX/help/370281U-01/imaqvision/imaq_bayer_to_rgb/

    You must use ArrayToImage on the image source first, of course.

    Eric

  • MODBUS U16 to simple, back from setting in the form

    Hello everyone, I am working with a couple inverters Kollmorgen that give and receive data via MODBUS. Using the LabView user's library, I had no problem in the reading and writing of data on their part (assuming I do not more often than 10 times per second). However, I am having some problems with the data format. Kollmorgen has not settled exactly according to the standard and therefore some values are 32-bit registers, some are 64-bit and to initiate, some are signed and some are not; they are all however, delivered in 16-bit of size segments by the function 'read record keeping. Right now I'm only interested in the 32-bit signed value variety.

    While I am able to read and I am sure I read the values I want (I can change the value of Kollmorgen workbench and LabView reflects the change of value), the output does not much sense. For positive values, it is fairly intuitive, like the first U16 (I read an array of 2 values U16) is always a zero, and the other is a x 1000 of what positive real value is. So if I read for example a current 3.5 amps, the first U16 is a zero, and the second is 3500; If I read 1.23 amperes is zero and 1230, and so on.

    However, when reading negative values, I failed to see a pattern, so that it almost always go to a pretty high value (65 324 or something) in the two registers in low as - number 2. I already tried several data manipulation of the methods listed here, including:

    -Conversion of type one in 2 U16 table, then return via join numbers

    -Convert the number in binary, binary table then back to number

    -Force in single, then using convert in I32 and divide those

    So far, no luck, anyone had a similar experience or has worked with these players in the past? Also can't find much information on the documentation of Kollmorgen and their website States that the MSB is sent as the first registry on 64-bit size records. No doubt the positive data format gives a clue, but so far no chance on guessing it.

    Convert an I32 U16 table.  Divide by 1000.

    In the opposite direction.  Take your floating point number.  Multiply by 1000.  Constrain to a U32.  Convert to an array of U16.

    EDIT: Don't mess with the fixed Point.  This is for very special cases, usually involving the FPGA.

  • How to type work of casting for string to U16

    Hello

    Sorry for typing on the same subject, but I still don't understant what do happen when I type the 2-byte in U16 distribution chain. Could you please explain me this?

    Kind regards

    __behemot_

    Hello

    All ASCII characters are represented by 8 bits in memory. If you imagine these 8 bits, you can concatenate the. Who will be the 16-bit long integer (in this case, when you have a chain of 2 elements) and of course not signed, because there is no defined sign in ASCII values.

    So the two integers and character are represented in the same way in the memory, but LabVIEW knows what type is stored in variable. If you working with ASCII characters you'd have numbers, you must convert it in full then you can multiply it and etc. For example, if you want to change 'a' to 'A', you can convert it to integer U8 you must subtract 20 from her and typecast to character (string 1 element).

    I hope that this will help you understand the type cast. If you have any other questions, please write again.

    Kind regards

    Peter

  • FPGA U16

    Hello

    I have a controller PXI with a PXI-7833R and a PIN of SCB-68 shielded. The entrance of 7833R is - 10 v to 10 v with 16 bit resoluation. Is it possible for me to acquire data from 0V to 10V with the U16 data format? If possible, how to configure the software and hardware.

    Thank you very much.

    Hello Xiao,

    It is not possible to adjust the parameters for the series R ADC.  They are set to measure at 09:50, you cannot rotate to measure 0-10 with full resolution.  The only options you have change is to say measurement mode, reference signle finished, etc.

    So you must either front-end packaging to enjoy more bits, or just the resolution you have.

    Hope this helps,

    Adam

  • ASCII table U16

    How can I convert an ascii string, in a table of U16

    The string I have is: 0123400000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000000

    It must be poured into a table, and the table should read:

    [0] 8961

    [1] 64

    [2] 0

    [3] 0

    [4] 0

    [5] 0 etc.

    Anyone have any ideas?

    Now that I got the road work, back to this...

    Here is a quick and dirty solution (assuming we have a string whose length is divisible by 4):

  • Cannot open RAW Image (binary) u16 in Vision

    Have another question Image U16.  Trying to open a raw U16 image.  Opens in ImageJ with the following options:

    I duplicated these settings using this simple code

    But I get the error: "error 1074396074 occurred at ReadFile IMAQ.

    Possible reasons:

    IMAQ Vision: inconsistent image size. »

    Image should look like:

    I write the file "Martha-1849-02-001-6 - view0000_raw_grey_32bit.raw" ftp://ftp.ni.com/incoming if you want to try.

    Thank you

    Don

    The help says "If you read a non-standard file format, you must set the size of the Image to match the contents of the file before using IMAQ ReadFile".

    Use IMAQ SetImageSize to set the size of 3888 x 3072 and it works.

    -Christophe

  • U16 anomaly of image in the container Image Imaq Vision

    Someone would have seen this type of anomaly and know a work around.  What's weird, is that I have not seen in constantly.  Image appears correctly in ImageJ.

    People-

    For some reason, u16 TIF images are sometimes not read correctly and it is clearly a Vision problem of OR focus on. The solution is below not unexpectedly Andrey which is on the Mount Rushmore of Vision expert programming.

    http://forums.NI.com/T5/machine-vision/open-greyscale-U16-image-problem/m-p/2414022#M38645

    Sincerely,

    Don

  • MAX cannot image format u16 error output 0xBFF69037

    I have a GC660m AVT Prosilica, and when you select the pixel format 'mono12' type of output image can be I16. Since this is a 12-bit camera, it does not really matter. But I still wonder why MAX can't output format U16 images? The error code indicates "No decoder available for the selected pixel format", which seems to indicate a software problem. It's MAX? I use LV 2013, MAX 5.5.0.

    Thank you

    Hey Ishi,

    You can get images of U16 of IMAQdx. The only problem is that it is not integrated with the new "Output Type" attribute that has been recently added to IMAQdx. This is because the change in I16/U16 was designed a long time ago and continues to be carried out by a slightly different mechanism. To obtain images of U16 you would keep the output type as I16/Auto and then go to-> show all attributes, the attributes of camera-> Acquisition attributes-> Pixel Signedness and change 'Auto' to 'Unsigned '. By default, IMAQdx can't always gives than i16 images automatically because very few cameras use the full range and 16-bit in the Vision libraries image U16's a little improved, there is still not universal support as it is for I16. We have a problem open, created to allow the output Type change the signedness directly that you expect.

Maybe you are looking for