Calibration of pressure sensor

I have two pressure omegadyne transducers. I tried to calibrate the transducers via linear equations. The problem is that when I put the second slope of line transducer and the y-intercept, the first (slope of the line and intercept) changes automatically to the second parameter. Transducers are mounted on two different channels. I can't have two different equations for the two transducers. I appreciate it if anyone has a solution to this.

Concerning

Ali,

When you're in the DAQ Assistant, on the configuration tab, in the input parameters of voltage is a Custom Scaling pull down.  If you who pull down, you can select Create new and which will allow you to install a custom scale for each transducer if you wish.  Just assign the custom scale for the appropriate channel once you have created the custom scale.

If you use the same transducers of beach and they have different slopes... I refer to them and buy some finding themselves in a few percentage points of each other on the range.  No one should sell you two producers of trans who have the same range that do not use the same slope... the y originally I see who need some tweaking, but not the slope!  .. just in my opinion.  I sent a lot of return of material that does not perform or has not been linear in its response.  ... again, just my opinion.

Enjoy!

Chad

Tags: NI Hardware

Similar Questions

  • Certificate of scaling of pressure sensor

    Hello

    My pressure sensor has the following calibration certificate:

    (lb/PO2) pressure unit of data (mVdc)

    ---------------------                   ---------------------------

    0                             -0.275

    100 49.824

    200 99.774

    100 49.844

    0                              -0.264

    Balance - 0.275s mVdc

    Sensitivity: 100.049 mVdc

    Calibration factors:

    Sensitivity = 20.01 mV/V

    the = 5 VDC voltage will give a result of 0 to 100 mV

    I'll use DAQ support to take the measure of pressure.

    the question is: what should be the two electric and two physical quantities (IN mV/V units) that I need to enter using DAQ Assistant when you descale the pressure transducer by the method of the colon. Once again, please specify the values in mV/V units.

    Thank you

    wisjaf12,

    You can choose one of the two points at your convenience.

    For example: First Value: 49.824 second value: 99.774. (Electricity). First value: 100 second value: 200. (Physics)

    CarmenC.

  • NEITHER USB-6008 connect to thermocuples and pressure sensors, control valve

    I am endevoring to build a gasification plant biomass for bench scale test process control plans. NEITHER USB-6008/6009 will be adapted for use as a data acquisition. I'll take RTDS, thermocouples and pressure sensors. I don't want to use industrial automation controllers.  It is also possible to use the channel of analog output for sending signals to a control valve position (using sufficient current/voltage between the two drivers).

    (1) OK.  I just wanted to be sure that you were aware of the potential dangers.

    (2) an RTD is a resistance that has small changes in resistance per degree of temperature change. To measure that you have need of a current source and a sufficient resolution in order to detect small changes.  At 25 degrees C a typical RTD is 109,73 ohms and resistance ohms 0.38 per degree changes.  If you had 1 my crossing this RTD voltage through it would be 109,7 mV and the voltage change of 0.38 mV by degree.

    The resolution of the 6008 on the most sensitive range is 0.49 mV > 1 degree.  The accuracy of the 6008 is 1.5 mV typical.

    For a Type K thermocouple, voltage at 25 degrees is 1.407 mV and change by degree is 39 µV.  Millivolt solving half of the 6008 translates into about 12 degrees.

    If you need a source of excitement for RTD and a kind of amplification for thermocouples and RTD before she would make any sense to try to use USB-6008.

    (3) I have not used anything except LabVIEW with DAQ devices and drivers.  I think DAQmx can be used with MATLAB and other languages.

    (4) the 6008 is the low range made by NOR. You will need to go to a more expensive camera or add signals conditioning circuits.  Talk to your representative OR assistance in the choice of a suitable device.

    Lynn

  • Pressure sensor ground loop problem

    Hello

    I have an interesting problem, I hope someone can help me with.

    I'm measuring pressure with a pressure sensor.  We see a lot of noise with the measure, and I believe that we have a ground loop.  I have checked the noise is 60 Hz.  What is interesting is that the case of the pressure sensor is connected to the Earth.  When we put on the sensor of metal pressure in the booth, we see the noise.  When disconnect us the stand noise pressure sensor disappears.  The stand is connected to the Earth.  I'm feeding the sensor with a 24V power supply which is also connected to the Earth.  I think that this is the ground loop.

    My problem is that the 24V power supply should be grounded for safety and the I cannot isolate the ground pressure sensor.

    I attached how I got the system wired.  How to connect up to my analog input module pressure sensor OR 9215 to break the ground loop?

    Thank you

    Dan

    Ahhhh Yes,

    I have solved this problem.  It is a ground loop.  I had the field related to 2 places.

    We have removed one of the references on the ground and poof everything worked fine.

    One of the problems I've found is a reference of external power supplies and others do not.  If you provide an additional reference to the Earth... Ground loop!

    Ground loops can be mean terrible things.  According to my experience, they are the number one electric goofiness and the noise causes.

    I have compiled some resources on my website to help people

    http://www.autosofttech.NET/resources

    The best is field wiring and considerations of noise for analog signals

    If you want the final word on the subject see this book

    Techniques of reduction of noise in electronic systems

    Hope that helps!

  • PXI-6224, SCB-68, power supply for pressure sensor 9-28Vcc

    I use a block of SCB-68 connection to connect a Honeywell FP2000 series pressure sensor to a PXI-6224 DAQ module.  The transducer requires 9-28Vcc, but the connection block seems to have only 5v pinout.  Do I need to use an additional power supply to the transducer?

    Yes.

    Don't forget to connect the ground from the power to the Gnd of the AI of the 6224.

    Lynn

  • Out of reading 2-wire 4-20mA pressure sensors using the NI9203 module

    Hi all, I'm sure that a lot of people can give a response would we apply for this.

    While looking at the wiring diagram on some pressure transmitters (2 wire 4-20 mA output, direct wire), that I'm about to buy, I noticed a slight inconsistency with the wiring in the NI9203 manual. Here are links to manuals for sensors:

    Swagelok pressure transducers

    Following the manual of NI9203, the + ve terminal of power supply (24Vdc in my case) must be connected to the Brown Terminal in the transducer and - ve supply ground COM on the NI9203. This leaves the Green Terminal in the transducer must be connected to AI0... AI7 on the NI9203 module. This provision makes perfect sense to me. However, the user for the transducer manual suggests that the Terminal green must be wired to zero volts land i.e. COM. This configuration makes less sense to me.

    My question is, the internal electronics of the transducer released still the same my current for the same measured pressure given by the curves of calibration for sensor, even if the Green teminal is not connected to zero volts?

    I have a hunch it's okay, but I wanted to be sure before that I lost many years spent money / bad record pressure. This should be a matter for manufacturers of sensors I know, but I found this forum is * much * more useful!

    See you soon,.

    Chris

    Hi Chris,

    The 9203 is a device for current descendant be plugged on one side common or 0 volt.  You must connect your pressure in 2 wire mode sensor.

    + 24 v DC to the positive terminal on your sensor pressure 1 terminal or brown flying lead, Terminal 2 green flying lead is connected to AIx on your 9203 and the commune of the 9203 connected to 0 Volts on your power supply.

    All 2-wire device used in a current input down is still as cable + volts then 2-wire device + then then 2-wire device - (or 0) then HAVE + then common then 0 volt.

    In my experience the 9203 can be rather prone to noise pick up devices with 2 sons and need a suitable clean power supply.  For 2-wire devices, I take 1024 readings and then averaged to reduce noise.

    See you soon

    Stephen

  • Barometric pressure sensor

    Hello

    I wonder if anyone has used a sensor of barometric pressure with myRIO Council. If Yes can you suggest any probe good accuracy?

    What of it? http://www.embeddedadventures.com/barometric_pressure_sensor_module_mod-1009.html

    It will work properly with myRIO?

    Thank you

    Christos

    Hey Christos,.

    Looking at the sensor as you linked, which should work perfectly with the myRIO. It uses the I2C/SPI protocols to transmit his readings to a controller, and the myRIO dedicated pins for I2C/SPI communication. Therefore, you should have no problem pulling measures the pressure of the probe.

    The myRIO has also 3, 3V output for external sensors power PIN, so you should be able to feed the sensor directly from the myRIO as well.

    I hope this helps!

    Ryan

  • How the pressure transducer Validyne system-> Validyne CD280-Dual-> SCB - 68-> PCI-6024E-> LABVIEW works and how we can calibrate the pressure sensor?

    Hello

    I am completely new to LABVIEW software.

    I learn a LABVIEW code existing my pressure (attached) acquisition system which has the path of data as follows: pressure transducer Validyne--> Validyne CD280 - Dual--> SCB - 68--> PCI - 6024E--> LABVIEW and I have a few question:

    1. How does the complete system of the transducer to the LABVIEW work? That is, if we apply the pressure of the transducer, it will change the resistance of the probe then...?

    2. How can I find the equation that expresses the relationship between the pressure and the tension to the keys of the Validyne double-CD280 in the LABVIEW?

    3. If I want to do the probe calibration, what are steps?

    Thank you

    You did not include most of the subVIs.

  • Suggestion of programming pressure sensor

    Hello

    I'm reading 19 values of pressure transmitter.

    2 PT are 0-350 bar

    another 17 are 0-200 bar.

    in accordance with the equation y - mx + c

    I want to generalize and define values for each PT.

    using some code, the node farmula... If I consider every PT that takes a lot of space in the main VI.

    Please see the attachment and give some suggestions.


  • sensor pressure and USB-6210

    I am installing a USB-6210 to read pressure sensors (specifically: http://www.omega.com/pptst/PX309-100mv.html), with labview2013.

    Its one task relatively simple and should not be this hard...

    I used a similar system to take similar measures in the past, but its been awhile.

    I don't know if it's just a wiring problem or incorrect installation of labview (program), but I can't measure the tensions that are remotely close to the sheet of calibration of the probe.

    Thoughts?

    It turns out that my diet was not come out correctly. Passed to a diet different and everything works as it should.

  • Change the type of sensor pressure strain for SCXI-1314

    Hello!

    I would like to ask how to change the type of strain sensor.

    I use SCXI-1314 and one of my pressure sensor is a type of 4-wire. (Output 100mV, excitement: 10vdc)

    , but I want to replace the sensor by another type of current output type. (Output 4 ~ 20mV, excitement: 12-30Vdc)

    I checked the output voltage cable 4. (Defined as full-bridge OR max)

    Power + = Ex +.

    Food - = Ex-

    CH+        =  S+

    CH-         =  S-

    However, I don't know how to connect the type 2-wire with SCXI-1314.

    In addition, I don't have enough experience on the types of bridge (bridges and whole, quarter half)

    Can you explain what I need to do?

    One thing you migth try is to put a 5 Ohm resistor in the current loop. causing a drop in voltage 100mV on resistance, however check the soil conditions.  Since you will need a new calculation of sensitivty sensor anyway you can use a resistor 4.7 Ohm or 2 10 ohms in parallel... just find a good match between your 1314 input voltage range and the 4-20mA.

    And you will have an external power supply for the sensor.

  • Problem with the sensor of pressure and NI 9205 with NI 9263

    I have a bridge pressure sensor based with a range of 0 - 5000 psi and requires a 10 V excitation. The output is 3 mV/V. I plugged on a cDAQ with a NI 9263 providing excitement V 10 and a NI 9205 to read the output values. I would expect a 0-30 mV output corresponding to 0 - 5000 psi, correct? (Incidentally, the NI 9205 module must be accurate to 0,1 mV or this should lead to a precision of about 16 psi, right?).

    However, when I tested it in fact the sensor on the run a test ranging from 0 to 1000 and back to 0 lb/po2 (now for quite awhile to States of pressure) the values I get from the NI 9205 are about 204.614 mV for 0 lb/po2 and then goes 204.867 mV to 1000 lb/po2 (actual results are noisy (, so these are averages). I enclose the SignalExpress file used for the test as well as the Excel file containing the resutls. Testing was performed with a number of other active sensors (load and displacement, LVDT and gagues strain gauges) in addition to the pressure that I'll eventually use them for my experience. In the SignalExpress file, the pressure transducer is the only sensor receives the excitement of the NI 9263 (ao_0) and ai_3 on the NI 9205 module.

    So this time I tried to run the test again, but changed the SignalExpress file (second .seproj attached) to include only the pressure transducer. But now I don't have an answer at all by the transducer (it remained simply flat).

    As to the connection of the transducer material, it is a connection 4-wire (excitement, - Exciation + Signal - Signal). I plugged the cables from signal to pins 4 and 23 (AI3 and AI11) for a differential pair. Then, I plugged the son of excitement to the pins 0 and 1 (AO0 and COM). Finally, I took a wire jumper and connected 23 on the NI 9205 module PIN to pin 1 on the NI 9263. I made this last step at the suggestion of support OR for a previous problem with a string potentiometer. This stage connected the ground between the two modules (the sensor did not answer with a pressure change when I don't have the wire).

    I appreciate any help as to why. I did not receive a 0-30 exit mV.

    his does not work like that.

    The probe should @30mA to 10VDC.  There is only available from the canal AO 9263 1mA.  you need another power supply.

  • cDAQ 9184 with module NI9232 for sensor accelerometer and pressure

    I'm currently building a CompactDAQ system for measuring two signals: vibration and pressure in the duct system. The system will consist of the following elements: chassis, NI 9232 Module cDAQ9184. Accelerometer IEPE to measure acceleration (vibration) and pressure for the measurement of the pressure sensor. My question is: can I connect the pressure sensor in the NI 9232 module? I have found on the site OR that this module is designed for acceleration and offers a conditioning of signals and excitation for accelerometers but nothing mentioned about pressure sensors. On the other hand the site mentions NI9237 module as the best choice for pressure sensors. So can I just use 9232 OR both sensors or not.  Thank you

    Here's an IEPE (source of ICP (r)) of transducors pressure: http://www.pcb.com/products/browse_productlist.asp

    Kistler is another provider.

    To help choose - can you tell us more about what you are going to connect the sensors - what is the unit under test?

    What do you do in the hope of discovering or surveillance of the unit?

  • Graph of pressure NI 9237 sensor error

    I use my NI 9237 module with four pressure sensors that works very well as long as all the readings from all sensors in pressure four are on the same graph. However, what I need them because I would need each of pressure sensors to be on their own individual chart but when I try this I get an error that the device is already in use.

    So for the moment I can't get a graph either showing the four pressure sensors or just a chart showing a pressure sensor, but I need four graphs showing their own pressure sensor. Any suggestions?

    Thanks for any help

    As you use the screw Express and dynamic data type, use the 'Split signal' function of the range of Manipulation of the Signal, the slide down so that it got 4 outputs and son each in a graph.

  • Pressure.VI help

    Hello

    Im a student working to the test with the polymer PDMS under pressure. I hope you can help me with a little problem I have met. I use a Honeywell P40 series pressure sensor connected in a 6218 USB of NOR. It is a program to measure the pressure with an inside diameter of the hose included, average of 1.2 mm is the for the 40PC honeywell series user manual:

    http://docs.Google.com/viewer?a=v&q=cache:rLeZ4BjNn6EJ:content.Honeywell.com/sensing/prodinfo/pressu...

    I have connected 1 pin 2 on the pressure sensor to pin 19 and 20 respectively pin on the NI USB 6218 case. I have not connected pin 3 No PIN pressure probe of the OR usb-6218. This will effect? Should I connect it on a spit of land and if so is it important that we?

    I have the program VI in labview 9 but still can't get used to it as I'm accustomed to Labview 7. I will include the program in the email and you would appreciate it if you took a glance and could let me know how I would use it or if the program is terminated.
    Thank you very much
    John Stewart 3rd year mechanical engineering.

    Hiya John,

    The link you sent me on the pressure sensor, in "General information", it is said;

    "The analog output of 0.5 v to 4.5 v is linearly proportional to the inlet pressure.

    In other words if the voltage measured at the output that you read is from 0.5 v to atmospheric pressure (101KPa), the measurable maximum output pressure will report to a pressure of 909KPa, for example if the boost value K is equal to 1. (unless it is inversely proportional, where when the pressure drops the voltage increases). But you need to calibrate your sensor to determine the boost value K.

    For calibration what is the measured voltage to output for a known input pressure is proven and you can then use some simple mathematics, know the boost value K.

    IE (0.5V / (101KPa)) = K (Vout / (known to pressure))

    But if you have any additional doubts on this I do not know that a call to specialists of Honeywell would not go a miss...

    Concerning

    Ben Hartrick