Camera USB/DirectShow is no projection in NOR-MAX

I have a gray Chameleon Point CMLN-13S2M USB 2.0 camera and it does not appear in the NOT-MAX.

I am well aware of Directshow registration with cameras RMP problems and I can tell the camera is correctly recorded.  It comes without any problem in AMCAP, a simple viewer of Directshow and also in Skype.

I just uninstalled the drivers PMP and spent some time going through the registry to remove all references that I could, at Point Grey and PMP.  Then I reinstalled the 32-bit version of the newer drivers, v2.5 beta 2.  I've also upgraded the firmware of the device later.

I am running

LabVIEW 2012 SP1 f2, 12.0.1

Vision development 12.1.0 module

MAX 5.4.0f0

NOR-IMAQ 4.7.2

NOR-IMAQdx 4.1.0

Any ideas?  It is a side project and I worked on it market now for over a year and I've learned to get this camera working.

Hello

The way list us the cameras is quite simple using the only mechanism of API Miscrosoft has for him. One thing I noticed, is that we need the "DevicePath" registry key to be filled (in addition to the key "FriendlyName"). Can you check and see if the entrance of advanced camera Grey was who? If this is not the case, try adding a string with a text for her.

Eric

Tags: NI Hardware

Similar Questions

  • Imaq is for cameras USB driver always available? He can not find on the Web site of NOR

    Cannot find or-Imaq for cameras USB Driver.  Is it still available on the Web site of nor?

    Thank you

    Or here.

    Andrey.

  • Camera USB LabVIEW 8.6 problem

    I'm new kind of LabView.  I'm working on a project of image editing, but I have problems with a camera USB and LV.

    I installed the NI_IMAQ file USB, I the NI Vision module.  I read the "NOR-IMAQ_for_USB_Camera_User_Guide", but I can't find that they use the numbers (fig. 6).  I imagine that the writer may have used a slightly different version of the LV, but I can't find anything in my icon palette that has a "USB Grab of acquisition".

    When work to aid the Vision Assistant, I am able to capture images of my USB camera so I know it works, I can not set up as a VI.

    Y at - he one day User Guide or matter who has a direction?

    Thank you for your time,

    -Rocky

    Hey, Rocky,

    Attached the files you're looking for...

    Andrey.

  • Camera USB IMAQdx low resolution-'for internet use"Option?

    Hello

    I have a USB camera compatible directshow. The resolution of the camera should be close to 2600 x 2000 (not sure on exact). When I see 'Available video Modes', the highest resolution is 352 x 288 resolution.

    The company that provides the camera told me to uncheck "for internet use" in the property of the Group of experts to help solve the problem of resolution.

    I don't see that option anywhere. I looked into the property node IMAQdx as well as for any "attribute of the internet."

    Any ideas on where this option is or how I can fix this low resolution?

    Thank you

    Tony

    The option 'Use Internet' apparently does not exist in LabVIEW (the less c. 2010).

    For the directshow driver, the seller of the camera (Motic) created a new registry file that I imported into the registry and cela solved the problem of resolution.

  • How can I change the default actions of what made the PC when I connect a camera, USB key, etc.?

    How can I change the default actions of what made the PC when I connect a camera, USB key, etc.. My Samsung intensity II is what I need it for; I want to shoot to the top of this window asking me what to do when I plug in, but one day, I was not thought and press on 'No Action' with the "always perform this action for this device" (or said something like that) checkbox checked.

    Once it is connected, go to my computer, and your camera photo or anything else that should be there.  If is isn't, it is a completely different issue.  It's done, right-click on it and select Properties.  Click the AutoPlay tab.  There is a version of dialogue enough of this "window asking me what to do."  You can figure out what to do from there.

  • The camera USB driver

    My guard computer crashing, either go to the blue screen or gel so bad that I have to press the power button. Then windows troubleshooting tells me that the problem is due to a camera USB driver, but she doesn't tell me how to remove it. When I try to find this driver, I can't. Can someone give me some ideas on removing this driver? Please keep in mind that I am barely computer literate.

    Hello

    You will probably need to reload the drivers of the device and any camera control software.

    Login as an administrator.

    Double-click Control Panel / Device Manager - Imaging - writing down of the brand and model of camera.
    on this subject and on the tab of the driver is version. Now, click on update drivers (who are unable to do anything as MS
    is far behind the pilots of certification). RIGHT click on the camera - UNINSTALL - REBOOT - it
    will update the driver stack.

    Now, go to the system manufacturer's website and download the latest driver for the camera and the other related camera
    software (if not more recent get the same).

    Download - SAVE - go to them and RIGHT CLICK - RUN AS ADMIN - reboot after each driver.

    Look at the sites of the manufacturer for drivers - and the manufacturer of the device manually.
    http://pcsupport.about.com/od/driverssupport/HT/driverdlmfgr.htm

    How to install a device driver in Vista Device Manager
    http://www.Vistax64.com/tutorials/193584-Device-Manager-install-driver.html

    Then let windows updates on however prevent loading of drivers who are often older than the
    those that you have installed. If updates suggests a pilot and then HIDE it and watch manually to see if their
    really is a more recent version (at the time system manufacturer and the sites of the manufacturer of the device).

    How to disable automatic driver Installation in Windows Vista - drivers
    http://www.AddictiveTips.com/Windows-Tips/how-to-disable-automatic-driver-installation-in-Windows-Vista/
    http://TechNet.Microsoft.com/en-us/library/cc730606 (WS.10) .aspx

    ===========================================

    If you still bluescreens then so tell us BCCode or Code numbers to stop.

    I hope this helps.
    Rob - bicycle - Mark Twain said it is good.

  • Pilot the camera USB 2.0 Video 8 64-bit Windows

    Hello
    I'm having a problem with my webcam. I recently installed gesture control which darkens my camera, so I uninstalled the driver now I can't go back there. Even when I install a pilot Imaging it does not work and video/camera USB 2.0 driver is not yet on the list. Please help me to get the video/usb camera driver.

    Hi Sunny,

    I would need more information to help better understand the issue.  Please answer these questions:
    1. What is the brand and model of the webcam?
    2 do you have uninstalled the software of gesture?
    3. What is the brand and model of the PC?
    4. do you get an error message when you try to install the driver for the webcam?
    I suggest to install the drivers for the chipset of the website of the manufacturer of the computer.  Then, try to install the driver for the webcam and check.
    Please come back with information to help you solve the problem.
  • NOR-MAX noncontinuable exception

    Hi all

    See the attached screen shot. When I try to change the NOT-Max camera file, I get this "noncontinuable exception." The file in question worked last week, and in fact NEITHER-MAX won't let me change any camera file that should work with this camera.  The camera is an evaluation from a supplier of the probe card and the file .icd I'm moving them here - it is currently working with my file .icd updated the. For me, it's the strangest part: the equipment works, everything is speaking to each other, I just can't change the camera files.

    I tried to run OR-MAX as administrator, thinking business HE might have eliminated a change of permissions. When I run as administrator and select the new camera file, NOR MAX just goes down, no pop-up window and the file still does not change.

    Windows 7; OR MAX 14.5, LV 2014

    PCIe-1433 framegrabber

    I'll try to save a file in Camera File Generator. I'll post more information, if I learned something.

    Thank you!

    Crazy

    Installation LV2015 has indeed fixed the problem OR MAX. If a part of my previous installation has been corrupted. When I tried to repair the installation earlier today, I had not the installation media going TO practice (or I could have deleted the download like a fool), so I jumped this repair. Now it seems that the problem lies in him GOING somewhere.

    See you soon!

    Crazy

  • NOR-MAX for VISA DURATION 4.2

    Hello

    I posted this on the Narcotics Control Instrument, but maybe it's not in the right place since I didn't get all the answers. Here's my question:

    I have been looking for more than an hour and have not led to what. Is there a version of the NOT-MAX runtime? I have deployed executables LabVIEW on many other PCs and I encounter a problem on one of them where the resource VISA vi does not return the exact list of the COM ports even as drop-down list of the vi box VISA set up a Serial Port. For example, the list box might show COM1 and COM4 COM10, but VISA resource find vi shows ASRL1::INSTR, ASRL4::INSTR, and ASRL12::INSTR (should be ASRL10::INSTR). So I need a program such as NOR-MAX to change the alias on the PC with the execution of VISA 4.2 engine so that it corresponds to what shows the VISA configure Serial Port vi, incidentially, that matches what is shown in Device Manager.

    The PC in question does NOT the LabVIEW development environment installed, so it has NEITHER-MAX. is there a stand-alone exe OR-MAX I can install it on other PCs without the IDE LV? Manual change of visaconf.ini will do the trick? If so, is that what I should remove all aliases, uninstall all external (VCP) ports in the Device Manager, restart the PC (or not) and start again, reconnect the devices?

    Thank you for any help or advice.

    Ed

    James,

    Thank you for your response. Yes, I use the Application Builder and no, I chose not install NOR-MAX, but I see there in my file of project under "additional installers. I will consult the installer to run from the Web site of NOR.

    After that I posted this question, I have experienced yet. I have manually edited visaconf.ini so that numbers ASRL agreed with Device Manager of / ' Windows port assignment and low and behold, it worked! If this happens on the PC client, installation of MAX would probably be best solultion, however.

    Now I'm wondering if the VISA resource find vi is just read this file? I guess that Yes.

    Thanks again,

    Ed

  • NOR-MAX 5: analysis of the instruments?

    I am currently developing my GPIB-USB-HS adapter using NOR-MAX version 5. Where is the button "Scan for Instruments' past? All my other systems are on the NOT-MAX 4.4.1.

    I was still unable to find it, so I went back to version 4.7.7 (who has the button "Scan for Instruments" in the usual place) and went from there.

  • Attributes in NOR-Max not found in Labview

    I'm currently developing a LabView tool (2011) with a camera from Manta G504B Ethernet. It is NOR-compatible IMAQdx but in NOR-MAX attributes are not the same as those found in LabView. I would like as my tool developed to be able to change the exposure and acquire values and lifestyles. However, these attributes do not appear in LabView but appear in NOR-MAX. is it possible to get these attributes in LabView, since these values will change very frequently?

    My Manta firmware is 00.01.44.00

    NOR-IMAQ is 4.6.1

    E/s-OR-IMAQ is 2.5

    NOR-IMAQdx 3.8.1

    NOR-MAX 5.0.0f1

    Answer was found with the example OR grab and Attributes.vi base.

  • Lost the remote TCP/IP connection through NOR-MAX

    "" Hi, I try remotely LeCroy WaveRunner 104Xi oscilloscope through "resources VISA TCP/IP" option in Control Panel OR MAX, but the VISA session is lost (0xBFFF00A6) right after I viWrite the '* IDN?" order several times.

    Here's my full configuration process:
    1. install the Passport VISA LeCroyVICPPassportInstaller_1.0.8.exe , LeCroy oscilloscope

    2. setting of the session to follow the instructions (as per joint, LAB_WM827.pdf)

    [by the way, I'm not able to configure the session as "VCIP:'type', I only be able to configure the session asTCPIP0:: 1861:OCKET" ""]
       
        TCP/IP resources VISA-> Create TCP/IP new resource...-> Raw Socket
       
    Set the IP address: 137.57.96.8 (this IP address is successful to be ping by PC and this address is of course belong to the 104Xi WaveRunner oscilloscope)
    Set the port number: 1861

    3. then, there will be a new resource under "resources VISA TCP/IP". Select this option and 'VISA Open Session'. A dialog box will be retractable to the communication of the instrument.

    4. go to the tab "viSetAttribute", the value VI_ATTR_TCP_KEEPALIVE to VI_TRUE, VI_ATTR_TMO_VALUE the value of 15000.

    * NOR-MAX Version: 3.1.1.3004
    * Connection: The ethernet cable is used through the server (No 1 to 1 connection) of the company,

    "After all this, I send '* IDN? \n"(dans l'onglet viWrite ), it returns the value 0 (which means no error occur) and return 6 counts. " So I try to read a buffer (in the viRead tab), but it send me error timeout (VI_ERROR_TMO). After that, I erase the buffer in viClear. Then, I repeat same steps, this time he (viRead) return me connection lost error (VI_ERROR_CONN_LOST).

    This situation is repeated every time that I reopen it new session. In the meantime, I also doubt that the viWrite did send the command to the oscilloscope or not. The oscilloscope was no response after that I have send "ASET" (WaveRunner 104Xi order for autosetup oscilloscope) through the viWrite tab.

    I am out of ideas now.

    What's my configuration OR-MAX problem? or hardware problem? or a server problem?

    Please advice... need help badly.

    Thanks in advance.

    Hi wcsoh,

    Alan and LeCroy Support here...

    Use the type of connection (LXI).  The problem is that when you use the connection TCP/IP (ACIP) type, you cannot configure the WaveRunner 104MXi as a resource of VISA's TCP/IP in MAX.   Note that MAX is not 'hip' to recognize any instrument type TCPIP, including extended via ACIP LeCroy.  To use LXI, go to utilities > utility configuration... and then to the 'Remote' tab, where you will find the button on the left side of the screen of config to use 'LXI '.  If you don't see this button, you will probably need to update your firmware.

    With the type of connection (LXI), you can automatically find the scope through MAX, enter this amount as a VISA TCP/IP resource and configure an alias as well...  You can also point your web browser to the IP address of the bezel, and you will see a status page with information, including the name VISA always useful resource to use when you configure the scope in MAX.  (Note, however, there is no need to use MAX at all when the connections of VISA to LeCroy oscillloscopes!)

    Here is a link to an application note written by me which deals with the LXI interface, including information about the configuration of MAX to show scope: http://www.lecroy.com/tm/library/AppNotes/LXI/LXI_Interfacing_AppNote.pdf

    Do not hesitate to contact me for any question!

    Best regards

    Alan Blankman

  • Compatibility of 64-bit of Windows 10 NOR-MAX,

    I plan my (forced) migration to Windows 10 this summer for my PXI chassis and noted that NEITHER-MAX 15.0 is not currently compatible with Windows 10, 64-bit (http://www.ni.com/white-paper/52818/en/) operating system. Has anyone heard any news on when / if will produce a compatible NOR-MAX always remember how much time OR took based on the latest versions of Windows?

    National Instruments for Microsoft Windows 10 product compatibility

    MAX 15.0

    It is compatible with Windows 10 64-bit as long as you are using the 32-bit version of the software.

    Event take a look at National Instruments for Microsoft Windows product compatibility 8

    You will see as very few packets NOR is designed to run natively as a 64-bit software on 64-bit Windows version.

    Read also

    LabVIEW 64-bit vs 32-Bit Applications FAQ

    White paper: the use of products from National Instruments on 32-bit and 64-bit Windows
    White paper: National Instruments for Microsoft Windows Vista product compatibility
    White paper: National Instruments for Microsoft Windows 7 product compatibility

  • NOR-MAX recorded do not channel setting correctly

    I'm developing a system using Modules of field Point. I use 7 LabVIEW and NOR-MAX 3.0 to create commands and attributes for the FP-PG-522 module.

    To create an attribute of channel using NOR-MAX 3.0, I selects the FP-PG-522 module in the section of remote systems, select the desired channel, select the string attribute, then select Create point.

    A new item now appears in the data area. When I select this element, there is a description in the section 'Channels used on this point', which describes the attribute. For example, I chose the market time attribute. For this article, it says "this element represents the 'On Time' of the channels selected attribute. This attribute can be set to any integer between
    1 to 65535. "

    Then, I hit the SAVE button, which shows how to save the IAK file locally, so to save it on the module of ps2000.

    If I close OR-MAX 3 and restart, when I select the same new element in the data area, the attribute has changed. It doesn't seem to be any pattern, but I had such combinations as the "control" attributes, instead of the attribute "Generate impulses", etc..

    Then I have to delete all these channels and start again, however, data are still once, not saved correctly.

    Someone else knows this and knows a way around it?

    Chris,

    This problem has been identified and the fix will be included in the next version of FieldPoint software.

    I just wanted to let you all know.

  • free software of the camera USB

    Hello

    How can I save videos in Labview with my usb camera? I would like to use some
    free toolkit and driver for it.

    Are there? I tried a few old IMAQ driver, but it does not work.

    I have to use the expensive Module Vision Development to record videos?

    The cheapest 'NI Vision Acquisition software' will do the job?

    I use Labview 2015.


Maybe you are looking for

  • Looking for a way to open files .avr

    Police of XXX service gave us a .arv file to watch a defendant under interrogation. Nothing seems to open it, not even A / V VLC plant. And we do not have Final Cut Pro to recognize. According to this source, the format was somehow the child to Apple

  • HP Officejet Pro 8610 webization Service

    I can't get the Web Services enabled PC or printer which Scans/Prints fine on the wireless network. The only advice I can see relates to the 8600.

  • Tecra S1 - pilot and DVD issues

    Hello I recently bought this laptop which had a few software bugs from hangover of the previous user. He advised the DVD player was not working, but at the beginning it worked for me very well so I reinstalled windows to have a fresh start. As a reas

  • 2610 HP printer does not print

    Some print jobs print and others don't. What should I do? HP support computer-generated test says that all tests are normal.

  • hp6520: I use all four cartridges on my hp printer 6520 or I can use black

    Let me use my black cartridge on my printer hp 6520. Do I have to use color those also to print