Configure a load cell connected to multifunction DAQ

I'm trying to connect a LCDH Omega 10 K scale by a module of constraint Dataforth SCM5B38-37 then AI0 gauge on a USB-6251 multifunction data acquisition.

In addition to the load cell, the application will use a pressure sensor with a strain gauge bridge module plugged into AI1 to read the pressure of hydrulic.  The pressure transducer using a measure of the force will be calculated and compared with the reading of load cell.

Currently, the USB-6251 box is configured to read analog voltage on AI0.  The scale is then performed using linear Fit.VI.

Review the example Bridge - entry continues shows some additional controls may increase the accuracy of the reading.

Unfortunately trying to use the DAQ Assistant only added to my confusion of the best practice in the acquisition of a signal from a load cell.

__________________________________________________________

Here are some of the questions I have...

What is the best practice implement a load cell using a full bridge strain gauge conditioning module labview?

What is shunt calibration?  Is this similar to the USB-9237 DAQ modules-specific?

How do you know if your sensor is set up like a full bridge, half bridge, etc, the document load cell does not it?

You can use the VI delete DAQmx perform bridge?

_________________________________________________________________

I don't know, I understand all the information of the map calibration either.

Why is there resistance to entry and exit values?

What is the value of the zero balance signafiance?

Specifications of load cell

Lbs mV/V             

0.000 0

2000 0.598

4000 1,201

6000 1,804

8000 2,404

10000 3.002

Excitement: 10 VDC

Input resistance: 350,41 Ohms

Output resistance: 352,08 Ohms

Balance zero: 0.01 mV/V

Hello!

I was looking into your question and found a few links that might help you configure your system. The first is a white paper that discusses some of the basic concepts of the gauges of constraint and their use with LabVIEW, shunt calibration and also covers some of the resistance values you see:

http://www.NI.com/white-paper/3642/en/

As you can see in this article, we have some specific hardware, built for the kind of measures you take. Because these devices are manufactured specifically for this application, they would provide better and more reliable results. That being said, you can configure your system with your USB-6521, it might not be as accurate as one of these systems. To set up your measurement system, rather than defining the analog input as a measure of deformation, you want to use the voltage custom with the option of excitement. This article treats this yet:

http://digital.NI.com/public.nsf/allkb/C66F92BDE229F45A86257B6D004D6033

We get you your data in the form of Volts/Volts (this document addresses more info on bridge probe scaling: http://zone.ni.com/reference/en-XX/help/370466W-01/measfunds/bridgescaling/), , but you should make sure you standardize this value to the excitement. Devices such as the NI 9237 provide the voltage and can take this known value in to account within the program, but you may need to do this manually for your application. Here is additional information on the NI 9237 and how it is configured to read information from strain gauge:

http://digital.NI.com/public.nsf/allkb/892C84122A6501AE86257547007E5C53?OpenDocument

Regarding the configuration of unit load and the information, you can try to contact the manufacturer for more information!

Thank you!

Tags: NI Hardware

Similar Questions

  • Load cell, compatibility linear actuator

    I am putting together a new test configuration that will operate a linear actuator, in charge of sampling. Data will be graphically represented on a graph of load vs displacement. I am new to Data Acquisition, so I just wanted to run my installation by a person who has a little more experience before you start spending money.

    Motor and controller are ZAGARELLA. The engine is a NEMA size 23, 60mm travel (model #NA23C60), the force of 175 lbs; controller is a phase 2, up to 2. 5A by the motor controller step-by-step phase (Model #T - MCA). Engine connects to the controller with a Minidin 8. Data cable to the controller is a controller's connector Minidin 6.

    Load cell is a cell of weighing Subminiature Honeywell Model 11 with a capacity of 100 lbs (range BR code). I also use a Honeywell model UV-10 online amplifier to convert the millivolt signal ±10Vdc signal.

    Input module is a NI 9219, 4 channels, with connector 6-pos or blocks.

    Output module is a NI 9265, 4 channels, with connector 10-pos or blocks.

    My question is, what is the best way to connect to the motor controller for input/output modules? Also, same question for the Load Cell/amplifier and the input module. My setup looks, as it will work. I'm off little blind here and would be greatly HERE can anyone shed any light. Thanks in advance.

    Jeremy robots

    Hi Jeremy,.

    I'm not too familiar with the ZAGARELLA products but you were seeking to use the modules OR to send commands to the controller or you were looking to them for use as drive for the engine? Whatever it is, these modules will not be able to do it either. A quick search on google shows that the motor controller communicates via serial and or one of the modules that you have listed has a connector series. If you plan put these modules in a cRIO, you could use a NI 987 x to communicate through series.

    If you thought using modules OR as a drive for the engine, it is not possible either. You say your Motors require 2. 5, but most NI 9265 will be able to output is 20 m.

    In regard to the load cell, it seems you're going to need power external amplifier, which make no incremental signal you're trying to read. 9219 read only of differential signals, then you probably want to consider using a NI 920 x or a 9221. To connect the signal of the amplifier to one of these modules, the signal wire should go to one of the channels I and the Earth to the Terminal COM.

  • How can I convert a 3 mV/V to a load cell signal to something I can read using a field point AIO-600?

    I use an external powere supply 12V to supply voltage to the load cell, and I'm reading data using AIO-600. There is a separate module that can do this, but is it possible to do it with the current material?

    Russell,

    The lowest input for the AIO-600 range is 0 - 6V.  This means that you will have a resolution of 12 bits (11.3 effective) broadcast more than 6 volts.  I guess that your full scale signal max will be 36mV.  Without amplification, you will not have a very high quality signal.  It would be the best bet to amplify your output signal mV as close to the load as possible cell and then connect this signal amplified to your AIO-600.

    Best,

    Chris LS

  • Load cell amplifier signals

    Hi all

    I am building an app to read a signal in a cell in charge of RS & amplifier using a USB6210 data acquisition card. I ' v tried SE & Diff connections, and I see a change in signal when I apply a load, but it is not quite what I expected (see attached screenshots of MAX).

    I am providing an external 5V to the amplifier, so no load can expect approximately 3V from the amp - it works OK. With a load applied, I expected to see a constant voltage drop, however I seem to have become unstable, with decrease in voltage.

    No idea why this is happening, any help would be appreciated.

    Thank you

    Martin

    Link to RS load cell & amp data sheet:

    http://docs-Europe.origin.Electrocomponents.com/WebDocs/0099/0900766b80099349.PDF

    It looks like a 100 Hz hum... setting ground and shielding and clean power for your amplifier?

    Everything changed power devices (power, converter,...) nearby (shure your PC ;-)?) Electronic ballasts modern economy energy light or dimmer?

  • 9237 with deck full load cell: support cell_null_off_shuntcal.vi survey error 200077

    Normal
    0

    fake
    fake
    fake

    EN-US
    X NONE
    X NONE

    / * Style definitions * /.
    table. MsoNormalTable
    {mso-style-name: "Table Normal";}
    MSO-knew-rowband-size: 0;
    MSO-knew-colband-size: 0;
    MSO-style - noshow:yes;
    MSO-style-priority: 99;
    MSO-style - qformat:yes;
    "mso-style-parent:" ";" "
    MSO-padding-alt: 0 cm 0 cm 5.4pt 5.4pt;
    MSO-para-margin-top: 0 cm;
    MSO-para-margin-right: 0 cm;
    MSO-para-margin-bottom: 10.0pt;
    MSO-para-margin-left: 0 cm;
    line-height: 115%;
    MSO-pagination: widow-orphan;
    font-size: 11.0pt;
    font family: 'Calibri', 'sans-serif ';
    MSO-ascii-font-family: Calibri;
    MSO-ascii-theme-make: minor-latin;
    MSO-hansi-font-family: Calibri;
    MSO-hansi-theme-make: minor-latin ;}

    Hello

    I'm trying to use the example
    load_cell_null_off_shuntcal.VI with a scale of full-bridge (Honeywell
    Model 31, not amplified). I'm using LabView 8.6, cDAQ-9172 and NI9237. The
    load cell is connected to the pins 2,3,6 and 7.

    Entries for the front side of the VI
    are: excitation10V internal; mV/V 2.1492 (calibration sheet); weight max 10
    lbs; resistance bridge 350 ohms (Honeywell specifications); 9237 shunt internal
    100 kohm resistance; map of shunt R4 (default setting). I chose
    "Do not offset null" and "shunt cal.

    This is the error I get:

    Error-200077 occurred at DAQmx
    Do a calibration Shunt (bridge) .vi:1 or the possible reasons:

    Measurements: Requested value is not
    support for this property value.

    Property:
    AI. Bridge.ShuntCal.GainAdjust

    You asked:-61.980405e3

    Valid values begin with: 500.0e - 3

    Valid values ending with: 1.500000

    If the "shunt cal.
    green button is not selected, there is no error. I understand that the Gain
    Change value should be approximately 1, whereas I get is much larger. The Subvi DAQmx PerformShuntCalibration
    .VI (bridge) contains a "Call library function node" which I did not
    find out how interrogate.

    Someone else has experience
    with this error? Do you have any advice on:

    1)
    How to 'see' the calculations being
    carried out inside the "call library function node"?

    2)
    What the correct shunt element
    a full-bridge load cell location is? (although changing this location only)
    does not eliminate the error, I can't find this info).

    3)
    What can I do wrong with
    my entries to cause this error?

    Thank you

    Claire.

    Hi Claire,

    You must physically connect the SC of arm of the bridge terminals (normally R3). The terminal is not provided for the connection of external resistors.

    See the example

    C:\Program NIUninstaller Instruments\LabVIEW 8.6\examples\DAQmx\Analog In\Measure Strain.llb\Cont Acq strain samples (with calibration) - OR 9237.vi

  • load cell

    Hello

    I use NI 9219 with model Honeywell scale 11. The problem is that the voltage for 9219 is 2, 7V while the calibrated for the load cell voltage is 5V. When I connect the two device, I always get the release of Labview, but how can I calibrate the sensor?

    Load cells operate on a principle of millivolts per Volt.  You should not need to calibrate the load cell just because you use a different voltage.  You just need to use the value of mV/V provided with certificate of calibration of the load cell in the setup of the scale of the load cell chain.

  • load cell 9237 + full-bridge: load cell_null_off_shuntcal.vi - error 200077

    I try to use load_cell_null_off_shuntcal.vi with load cell (Honeywell model 31, not amplified). I'm using LabView 8.6, cDAQ-9172 and NI9237. Entries: excitation10V internal; mV/V 2.1492 (calib. bin); weight 10 lbs max. resistance bridge 350 ohms (Honeywell specifications); 9237 internal shunt resistance 100 kohm; map of shunt R4 (default setting). Selected "offset null" and "shunt cal.

    Error-200077 occurred to Shunt calibration perform DAQmx
    . VI:1 (bridge) or the possible reasons:

    Measurements: Requested value is not a supported value for
    This property.

    Property: I. Bridge.ShuntCal.GainAdjust

    You asked:-61.980405e3

    Valid values begin with: 500.0e - 3

    Valid values ending with: 1.500000

    If "shunt cal' green button not selected, no error. Setting the gain should be about 1. Subvi DAQmx PerformShuntCalibration (bridge) .vi contains "Call library function node" which is locked (?).

    Any ideas?

    What is the location of item correct shunt for a full-bridge load cell? Change this location does not eliminate the error.

    Hello, YTC,.

    The problem is most likely in your external connections of the NI 9237 and the load cell.  As mentioned in NI 9237 Operating Instructions and specifications, page 9, SC + SC - pins must be connected to the terminals of the resistance specified in the .vi of Shunt calibration perform DAQmx (bridge) (in the case of a full bridge, it would be R3).

    Let me know if you still have problems with your calibration.

  • With the help of external excitation of load cell 9219

    I want to use 9219 to assess the load cells.

    Internal excitement for 9219 is not sufficient for my application, this device can be used with an external excitation source?

    If Yes is it as simple as cross the bridge of load cell using the external voltage and the output of the bridge on the 9219?

    Power max would be approximately 20 mV.

    Then the 9237 is probably your best choice. For this voltage, you will need to use an external source.

  • calibration of the load cell

    I have a new 5 lb load cell, I want to replace the old one with.  The former load cell had previously been calibrated to the program and was located in the drop down menu, as well as several other cells.  So far all I've done is went to edit items in the menu properties ring and add in the name of load cell - I did in all references to the program.  There are several references to the cell of support throughout the program, but most is a case of event "loadcell scale" that passes across the cell to the variable global-loadcell scale and a cell nationally for initialization.  I have attached a manual for the program I work with (the cell for initialization is not displayed).    I have never calibrated a scale before and do not know the appropriate measures, so any advice will be appreciated.  I have to do something to the MAX?

    as a side question what are and when do you use the stringsand values of the nodes property.

    Thank you

    Hello

    Here are two links about calibration with our software. It is usually done through MAX and step by step instructions can be found in the links.

    http://digital.NI.com/public.nsf/allkb/D22B70F72C3FB23286256F15005C2575?OpenDocument

    http://zone.NI.com/DevZone/CDA/tut/p/ID/4845

    As to when to use property nodes, they are used whenever you want to programmatically set some properties or attributes. For example, if you want to programmatically change the color of the text in a control that is based on what is displayed or channel indicator or resize the control or the indicator.

  • How can I configure my Internet XP connection so that it will automatically connect to my WiFi (wireless) without asking me the password whenever I connect?

    How can I configure my Internet XP connection so that it will automatically connect to my WiFi (wireless) without asking me the password whenever I connect?

    original title: automatic Wifi connection

    Hello

    By default, Windows XP attempts to connect to networks without wires in the following order:

    1. Available networks that are in the list of network preferred (by order of registration)
    2. Preferred networks in the list proposed (by order of registration)
    3. Other networks according to selected advanced settings

    Check out these articles and check if that helps:

    http://www.Microsoft.com/resources/documentation/Windows/XP/all/proddocs/en-us/wireless_networking_overview.mspx?mfr=true

    To connect to an available wireless network

    http://www.Microsoft.com/resources/documentation/Windows/XP/all/proddocs/en-us/wlan_client_connect.mspx?mfr=true

    To set up automatic wireless network configuration

    http://www.Microsoft.com/resources/documentation/Windows/XP/all/proddocs/en-us/wlan_client_configure.mspx?mfr=true

     

    5 steps: how to set up your network wireless home

    http://www.Microsoft.com/athome/organization/wirelesssetup.aspx

  • Error 0 x 80070035 "[name of the PC] is not configured to establish a connection on Port"File and Printer sharing (SMB)"on this computer".

    Original title: "home network".

    0 x 80070035 "[name of the PC] is not configured to establish a connection on Port"File and Printer sharing (SMB)"on this computer".

    That's what I get on my ThinkPad when I try to do the networking. I placed all the parameters correctly: all 3 of my computers are set the same. 2 workstations work perfectly. They can even see the laptop but not to open all files. The laptop can, open same network t itself file. But he can see the whole of the network (including a printer, is told by the way). I, ve gone through all parameters twice. I've even set the C drives to share via the menus of context on each computer. I have visited many forums with none even come close to this problem (they all will tell you how to set the network parameters or they get very technical and over my head).

    I tried with & without a firewall.

    No password.

    I use a wired network.

    Laptop & 1 desktop computer running vista - 1 desktop computer running xp.

    They have all the latest service packs ms & udates.

    Any help will be appreciated.

    These files can be opened by norton or lenovo network setting apps. The settings do not work only in the application of network of vista. Nor can I work norton or lenovo from other computers.

    PS how can I ensure these settings effect may just lan and do not extend to the internet?

    Thank you bb

    I tried new drivers & returned through sharing & permissions, according to Jack, but he was not good. At the level of experience of my computer I can miss something obvious, but I doubt it because I've been very careful in following the instructions. I am experencing other problems with this laptop as well and I'll just get all. I am planing to get back anyway, that's why I needed network. I wanted to Exchange files over the network to avoid having to ' wait & swap "usb drives. Thanks anyway. Thank you BB bb

  • Loading cells text in HFM using FDQM

    Version FDM - 9.3.1
    Target - HFM 9.3.1
    Environment - Windows Server 2003 32-bit.

    Integration settings - Enable cell loading set to text.



    We try load cell textual information in HFM using FDQM. We checked LoadB action that allows to load cell text information. But not able to understand where it's source for the text in the cell. He reads from the Recordset, but cannot find "MemoText" field in the FDM Tables, but all other fields are available.
    As shown below, in LoadB

    "Open the recordset containing the memorandum items so that we can load in the form of text in the cell
    Set rsCellText = API. SQLMgr.fTrialBalanceConvertedMemos (CStr (strLoc), CStr (strCat), CStr (strPer (0)), True, 1).

    It reads the information of TxtCase of

    "Memo for cell text.
    strCleanMemo = rsCellText.Fields("MemoText").value
    strCleanMemo = Replace (strCleanMemo, "[", + "\n]" & "[" "]")
    strCleanMemo = Replace (strCleanMemo,"]","]" & vbTab ")
    strCleanMemo = Replace (strCleanMemo, ' * ', vbcrlf & vbcrlf)
    If Len (strCleanMemo) > 1950 then
    strCleanMemo = Left (strCleanMemo, 1950)
    End If

    I would like to know which table it reads the MemoText field, as it does not exist in tDataSeg1-tDataSeg50 (view - vData). Not sure where the information in the memo are provided and gets load in HFM. We see xxxx - B.dat file from Outbox to folder view - "TxtCase Loaded by direct intervention of the API."
    But I do not see any text of cell in HFM in the specific intersection.

    Can you please let us know how the cell text should provide information in the source file. We strive to import simple text file containing the text of the cell for each row. Any help would be much appreciated.

    Thanks in advance.

    No problem. If you have a chance, please mark the post as helpful or appropriate. Thank you.

  • How to configure a PXI controller - 8106 RT with DAQ traditional channels?

    I have a few tips that are not compatible DAQmx and do not appear in MAX (4.6). I want to configure them to traditional DAQ, but if following the instructions (Tools / traditional DAQ/Remote Configuration) the error "connection to the remote computer XXX error. Make sure that the GDR server is running on the remote computer, and that both computers are connected to the internet. "They are connected with a crossover cable, and the machine rises to the MAX. I can run the server of the GDR on the local computer, but don't see no way to start it on the remote control of the RT. Right click on the machine of MAX and say 'Create New' do not have the traditional DAQ option, only items DAQmx. And MAX takes forever with this controller and often crashes... I made sure the software on the controller corresponds to that of the local computer (LV 2009 with RT).

    Lauréla,

    Please post on the Forums of NOR. It seems as if you do not see MAX devices, then you do not have the driver installed on your computer. I recommend to install the driver traditional DAQ 7.4.4 on your Windows machine, then you can install the T-data acquisition driver on your RT system. From there, you should be able to see your T-materials under your remote systems.

  • Handling of load-balanced connections Sonic MQ

    Hello

    We have successfully connected to a single broker Sonic WebLogic, but we now plan to connect using a factory of (foreign) connections which is load-balanced, i.e. Specifies several brokers.

    I have read the Sonic MQ Application Programming Guide and noted that...

    "Any broker in a cluster can redirect incoming client connections to another broker in the same cluster for load balancing purposes. Load balancer must be configured on the broker. The client must also be configured to indicate that it is ready to have a connection forwarded to another broker request. »

    It indicates a client should call:

    ConnectionFactory.setLoadBalancing (true) before calling the createConnection() method.

    Given that I use JMS wrappers I do not have the ability to do that, can you confirm if this is the default?

    Also if you know traps around this scenario I would be grateful for pointers until we test this out for good.

    Thank you very much

    Mandy

    I think that the res-Ref wrappers can allow invoking extensions, but they must be exposed via an interface (cast to the interface from abroad). Sonic MQ can provide configurable load balancing behavior as well.

  • How can I configure Mozilla to accept connections on port "The World Wide Web (HTTP) service"?

    When you try to embed a video in PowerPoint 2013, the "Sorry we are having some problems of temporary service" message.

    In research on the problem to Microsoft. com, the following message is received: Mozilla is not configured to accept connections on port "The World Wide Web (HTTP) service."

    Hmm... PowerPoint must use Windows to connect to the internet and not Firefox.

    What method do you use?

    (1) insert > video > video online

    Apparently, there was an update this month that allows this, as described in this message of support (see the top of page answer right after the question): embedding video in powerpoint 2013 - Microsoft YouTube community online

    Here's how it should work after the update: Insert or link to a video on YouTube - PowerPoint

    (2) developer > control > Shockwave Flash Object

    As documented on other sites, such as: https://www2.palomar.edu/pages/atrc/2013/09/17/how-to-embed-youtube-video-in-powerpoint-2013/

Maybe you are looking for

  • Firefox window blocks everything else on the computer screen

    for some reason, the FF window now blocks EVERYTHING else on my computer screen. for example, if I open a PDF file, it opens BEHIND the screen of FireFox. normally the PDF file would be open, and in the new window on my screen, hide at least a portio

  • HP DV6 Question Quad (RAM upgrade)

    Hey guys,. Here's my laptop. I get conflicting documents on what exactly I can move the RAM to. On these forums, poster says he can be upgraded to 16 GB. However, Crucial and the documentation just says HP 8 GB I think. Anyone has any experience with

  • Update, Qosmio F - SQL Server 2005 Express Edition Service Pack 3 fails

    This update continues to be marked by the automatic update from microsoft, but fails to update no matter how many times I try to treat it... Why is it so? What is the aim of SQL servers and do I need? Deleting it will solve my problem?

  • delete the printer driver by political group or batch script

    I am migrating our server F & P from 2003 to 2008.  I created all of the printers in server 2008 and tested, and they print correctly.  Test the issue I am running is this: I can add the printer Server 2008 for fine local workstation, but users are u

  • automatic device not installed

    I installed driver cure to repair some drivers and it says no audio installed now