Ehternet Instrument IP address using MAX

I have several instrument (PS, freq counter etc) which are the instrument of ethernet. I'm trying to find:

1. If the IP address can be changed in the form of LabVIEW for each instrument? It is after that the aircraft is initially configured with an IP address and not it is moved to a different subnet. Also, when the IP address is changed to the MAX instrument automatically identifies the IP has changed to this instrument or MAX needs to be updated?

2. I would like to read the IP address of each instrument by programming in LabVIEW, during initialization. Once the obtained IP address I have a ping of the instrument using LabVIEW himself.

Any help! TIA.

lvrat wrote:

Thanks, so if the IP address is changed physically on the instrument then MAX automatically is updated with the new address?

N °

If this is not the case, is there any file that can be changed and uploaded in MAX. My guess is that this will avoid going to each instrument and configure it again.

No. (it depends on what you mean. There is a MAX configuration file where this information is kept, but I don't know what you mean by "send to MAX. This is a file on the computer, not the instrument, and the instrument is not able to change).

Also it is necessary to configure all instruments of beofre MAX screw LabVIEW (visa) can communicate with instruments?

N °

It was easy.

Tags: NI Hardware

Similar Questions

  • I can not format my sbRIO 09632 using MAX?

    Hi, I have NEITHER Robotics starter kit 2.0 with sbRIO 09632. At the beginning it works perfectly, now I have the status light will blink twice. The State Max system is: not connected - no software installation. I tried to reinstall the software in the target but the LabVIEW Real time hardware wizared does not start, it shows me "Time out while communicating with the remote server Make sure that the system is online. "I also tried to format the target but when I swich the SAFE MODE ON and reset the target do disconnect the remote systems to MAX.

    Please is there anyone who can help me to solve this problem, thank you.

    Thank God,

    I solved the problem,

    1. when run in safe mode, the address must be static

    2 - after formatting with MAX, reset to start in normal mode

    3-to get a new IP address, use the hardware configuration wizard

  • To configure a serial port using Max

    Hello

    I inherited the code that configures the serial port using the old serial code that does not allow the user to specify the term tank and turn on the tank of the term. Any ideas on how this is done with the old serial code? I also tried to use MAX to set the term tank and allow it, I am able to communicate with my device series with Max and I saved the changes when I was. So when I run Labview code, my serial code fails because the term tank is wrong and it is also not enabled.

    Everything that happens with the definition the term tank and allows ideas?

    What you do to the MAX to the extent where the configuration of the serial port have no impact on the LabVIEW program. (EDIT: which means that you can change the configuration in the LabVIEW code.)

    If you load the old code in a version of LabVIEW using VISAS for all the stuff of series then the screw would have been automatically replaced by shell screw that NEITHER created for the old code. Inside, they use just VISA. More than probably the old code use via digital port numbers. For example, in the library of \vi.lib\Instr\serial.llb of , you will see these shell screw Serial Port' writing' for example, calls the 'Open Serial Driver' VI which simply returns a VISA session. It then uses the Write VISA. So you can define characters of termination by obtaining the VISA of "Serial Driver Open" resource.

    Of course, the best solution is to rewrite the code using VISA directly. It will be much cleaner.

  • IP address used by another computer

    I got a message saying the IP address used on a different computer on people. What can I do?

    If it is the public ip address witch, I hope it's

    Click on the button search for cmd right click Run as administrator

    Type ipconfig/release

    Press enter

    Type ipconfig/renew

    Press enter

    weight for all processes to complete and then type exit

    Restart your computer

    now, I would recommend hide your ip address, once you have renewed

    Tor navigation Pack works well for hide ip address

  • Email address used for the deleted record

    A couple of years ago, we bought adobe Acrobat pro 10. My Secretary has registered the software from his email address. She has since left the company and its address has been deleted. I have a new computer and cannot download the software because the email address used for registration is no longer valid! I have the serial numbers and the license certificate. Can someone help me.

    Thank you

    Peter

    I don't know what are your terms of organization, but I think that you could recreate the e-mail address so that it can be used.  I know that I can create sub-accounts for my own e-mail service.

    Aside from that, you might be able to download the demo version of the software directly via the site linked below and then use your current serial number to activate it.

    Don't forget to follow the steps described in the Note: very important Instructions in the section on the pages of this site download and have cookies turned on in your browser, otherwise the download will not work correctly.

    http://prodesigntools.com/tag/DDL

  • Syntax using MAX problem

    Hello members,

    A problem of ridiculous syntax this time, or perhaps my lack of understanding of the use (MAX)
    I get a group by error / missing, according to how I change the construction of the table.

    create the main as table)
    Select
    T1.cust_id,
    T1.invoice_amt
    T1.acc_no,
    T2.cust_id
    T2. Max (pay_ack) - number of float that increments each time a payment is made.
    Of
    Table 1 t1,
    table2 T2
    where
    T1.cust_id = t2.cust_id
    )


    Out of the final table should have the following fields:

    Cust_id, acc_no, invoice_amt, the max value of pay_ack for each cust_id. There is a 1:1 relationship between cust_id and acc_no!

    All entries will be very useful.

    Thank you!
    novice

    user8635888 wrote:
    Hello

    The error I see is:

    Max (T2.pay_ack)

    Shall appoint this expression with a column alias...

    Use max (t2.pay_ack) max_pay_ack or any alias that you love.

    Kind regards
    Avinash

  • Unable to connect/configure PSC 2020 using MAX

    Hello world

    It's something like 'part II' to try to use the serial ports on our PSC 2020.

    I received a lot of help and valuable advice from Ravens Fan and others of the Spanish community in regards to the series, and I really appreciate their help.

    However, just at the point where I was about to make the first test, after you advise them, someone in my Department decided, there are a few days to clear (or format) all the contents of the memory of one of our CFP-2020. In addition, there is an another PSC 2020 which was unused for a long time. Well, the point is that none of the two controllers is seen by MAX, and therefore they cannot be configured by MAX either. Both are simply a blank space (i.e. nothing appears) under remote systems or devices and interfaces or another area in the configuration tree.

    Another important detail, we used an earlier version (4.xx, I guess) and also a newer version (7.x.x, I guess) of MAX trying to win again the communications. None of these versions could not communicate with one of the CFP-2020. In other words, none of these versions could establish communication or to show a sign of life of one of the two controllers. We test the wiring and it worked perfectly with the other PSC 2020 we have in existence.

    I guess that LabVIEW 7.1 or 8.0 (versions that we have) will not be able to communicate with controllers or if MAX has failed before. Am I right?

    Is there another way to see (contact) the CFP-2020?

    Is there a procedure for 'must - after setting shaped' that we are missing?

    This behavior has something to do with the DIP switches in the cFP2020?

    Is it possible to say if controllers are broken (fatal hardware failure) or if it's just a matter of software configuration?

    The deadline to begin a period of testing laboratory with the cfps 2020 is very narrow, and I really need to retrieve both controllers, so I would really appreciate help for this problem.

    Best regards

    I am pleased to have been able to help so far.  Let me see if I can help even more.

    First of all, to find a big stick and go fight anyone who formatted your controller overhead.

    Second, if you don't see anything under remote devices, I'll guess that the controller has been the installation of another IP address on a different subnet than where you are now.  I would try the instructions on page 3-6 of this Manual to reset the IP address.  That set it back on a 0.0.0.0 address who should make them findable by MAX again.

    Thirdly, MAX is up to version 4.6.2 (perhaps under minor version or two later.)  So, I'm not sure "newever version 7.x.x" unless you have a version number of LabVIEW.  Yes, you will certainly get the controller found in MAX before LV can work with it.  Definitely check the DIP switch as safe mode and reset the IP.  It is possible, that there is an incompatibility of drivers fieldpoint between the remote control and your LV version but I think that the two would have to endure somehow offset, otherwise you wouldn't be able to update the fieldpoint drivers in the controller.

    If there is a problem with the controller, I don't know how to determine that.  A person of OR may be able to help.  I think that the most likely scenario is that the last IP address defined for the controller corresponds to the setting for your PC or network, you are now, in which case the IP reset should help that.

  • Instrument to appear in MAX, but not in the executable file

    Hello

    I use two HP34401A make some measurements, and I transfer the program to a laptop with complete labview at your desktop with only the runtime engine.  Instruments made their appearance in the Office MAX, but when I run the executable, they cannot be detected.  I expect some suggestions on how to possibly resolve this problem.

    Thank you

    Not to mention that MAX, what other driver is installed and necessary? Did you install NI-VISA, NI GPIB? At a minimum, you should include NI-VISA in the installer that you built.

  • Can I get the IP address used by other devices for synchronization?

    One of my machines was stolen. I wanted to know if I could find it using the synchronized information. I hope that there is a way to determine the IP address, and then contact the owner of this IP block to see if they can help me more. Is this is possible, and if so, what steps should I take? Thank you.

    It is a question that we receive from time to time, and no, this is not possible. Please contact the police for your inquiry.

  • How to 'book' an Ethernet cDAQ chassis based in software without using MAX?

    I am using a cDAQ 9188 with Ethernet interface. It works fine, but the Setup is a little more than I like for end-users. Users are all first go to MAX and do make a connection. After that, users must 'reserve' the chassis to ensure access to it. This isn't a problem except that I still have to find a way to make the code G. I have rather not in my documentation would include a series of steps require that the user enter MAX whenever they want to turn on the system. I would have preferred it is comparable to a USB DAQ experience where it works.

    All the examples there or suggest on properties, methods or screws to use for my situation?

    You'll want to use the DAQmx reserve Network Device.vi, under measures of e / s-> DAQmx--> DAQmx advanced-> configuration of the data acquisition system.

  • I can't install the driver OR RIO in my cRIO use MAX 5.5

    I can connect it assign a static IP address. Whenever I try to install the software in the FPGA target, appears the following error message:

    "Timed out while communicating with the remote server. Ensure that the system is online.

    What can I do?

    I have LabVIEW 2013 installed in WINDOWS 8

    Hi Vitrion,

    As seen in the Service request, this problem was resolved when you demoted to LabView 2012 and reinstalling Max5.5, so we can properly install the Driver NOR-Rio and sees the device to the Max apparently, it was a compatibility issue in versions...

    Just a recommendation for a few cases like this. Please check the compatibility chart before installing an update of LabView in order to have the right version of Driver OR-Rio.

    http://digital.NI.com/public.nsf/allkb/577CC9A7DCFC73DF8625738400116CC3

    Kind regards

  • 2011 Streets and Trips will allow me to enter & register names of specific locations associated with the addresses using the route planner?

    I'm a traveling salesman.  I use the trip planner feature.  I want to enter and save the name of the company and address to create slaes itinerary in a State or a specific city.  A previous version of streets & trips were only allowed to enter and save the address, city, State or zip... I want to record the name of every company too!  Is this possible with the 2011 version of Microsoft Streets and Trips?  Thank you.

    Hello

     
    The question you have posted is related to Streets and Trips and would be better suited in the forum of Streets & Trips. Please visit the below mentioned link to find a community that will provide the support that you have requested.
  • Automing instruments LabVIEW without using specific and coding drivers for each instrument?

    Hello

    I'm new to programming in labVIEW, but I had a few questions about its use in automation and remote controlling several instruments.

    It is an ideal software for automation of various instruments remote control/test. Assuming that I have several different companies spectrum analyzers, I understand that labVIEW drivers of instruments or special VI who would control these instruments. But if I wanted that all these spectrum analyzers to say, to display on another computer using the remote control, but I wanted to use the same program for effeciency, is it possible to do so universally for all instruments using labVIEW? Or do I have to adapt this program for each instrument to instrument drivers cause?

    If labVIEW can do that, is - anyone had good success for such scenarios in other programming languages?

    Thanks in advance for your help.

    Look for the Abstraction Layer material (that you can implement with LabVIEW, even if it's a slightly advanced topic).

    The idea of having a specific Test and be able to enforce it against different Instruments, decided at run time, is the kind of situation that LabVIEW and use of HALs was designed to facilitate.

    The concept is to 'Levels of Abstraction'.  When you create a test, you can design it for an abstract Framistan, since all the Framistans are supposed to be able to measure Vorbels in the range of 0 to 100, so you just plug a generator of Vorbel to generate Vorbels in a certain sequence of Test (linear, random, quickly, slowly, pulsatile, continuous, you decide), measure your Abstract Framistan records readings and compare with the results you expect write the results in a nice report.

    Of course, Framistat of OR uses Ethernet as the means of communication, while HP uses a serial port (depending on the model, series settings can change) and uses of the Intel one. DLL to communicate with the low-level API.  So you also need to develop, for each specific instrument, an "interface" between its API and the Framistan abstract (not everyone uses Vorbels as the input unit, for example).

    So the good news is that it can be done, the bad news is that one of the best ways of handling this type of question on the programming is to use OOP (OOP).  A number of languages (LabVIEW, C++, JAVA) support OOP, but LabVIEW might have the advantage when it comes to interact with the material.

    I you are looking for Hardware Abstraction Layer?  Add LabVIEW to the search query and you should find some nice presentations by Elijah Kerry...

    Bob Schor

  • ResetSys on the GPIB instruments with addresses


    Style calls (alias calls NOR-488. 2 multi-device) IEEE 488.2 use a two-byte addressing scheme where the subaddress is in the high byte.

    Help OR-488 online. 2:
    "Some of the calls to several NOR-488 devices. 2 have an address or address list setting. An address is a representation of 16 bits of the address of the GPIB device. The primary address is stored in the low byte and secondary address, if any, is stored in the high byte. For example, a device to the main address 6 and secondary address 0 x 67 has address 0 x 6706. A NULL address is represented as 0xffff. An address list is represented by a list separated by commas in addresses, such as 1,2,3.

    So, your channel should be:

  • WRT320N cannot not get IP address - using only 1 - no internet access - wired port hang web configuration pages

    You just bought Linksys WRT320N as a replacement for Netgear MR814.

    Cannot connect to internet using Linksys WRT320N.

    Installation program:

    ISP: Cox Communications (cable)

    Firmware: v1.0.03 build 010 July 24, 2009

    1 Wired - port 1 - for Windows XP SP 3 Dell Desktop

    Attempts to install:

    1 tried to set up using the installation CD.  Spread the slightest error with error 302: internal error. tried this 3 times

    2A through the reset, power off, procedures.

    3 Ping 192.168.1.1 = success

    4. connected to 192.168.1.1 via web browser

    5. set the MAC Address = MAC address of the computer

    6. the MTU value = Auto (1500).  Also tried manual 1492

    Could not get the IP address by using the version of IP address and IP address renew.  Address Internet IP still = 0.0.0.0

    Note that the web page will be still stuck after perfoming some actions (such as after switching between the tabs).  Message at the bottom of the browser is "Wating for http://192.168.1.1/Status_Router.asp.  And the hourglass on the screen away.

    Note that the old Netgear router still works fine and I am currently connected to the internet via the Netgear router.   I'd appreciate any help getting the new Linksys router connection/operation/obtaining of internet and IP address.

    Thanks for your help.

    I do not know where I was unclear: you wrote before you used the function of the cloning mac address on the WRT to clone the MAC address of the PC. In this way the WRT "pretended" to be the PC if you connect to the internet port WRT to your modem.

    However, the PC is not the device you used before you connect to the internet. The PC was not the device that was connected to the modem before. If you connect the computer directly to your modem it will either be an IP address. There will be no internet.

    The problem is that your ISP only allows you to have an active internet connection single at some point. If you plug in a different device with a different MAC address to your modem this device won't get an IP address. That you have connected the Netgear router to your modem and you have internet, then you need to clone the MAC address of the Netgear router. This is the device that is connected to your internet connection and it is the MAC address that your ISP has locked your internet connection.

    So, once again: clone the MAC address of the router netgear (internet or WAN port MAC address) of your Linksys router. Then the Linksys's router Netgear and your ISP should accept the Linksys router. The netgear router should have a status page in the web interface, showing you the right MAC address to be cloned.

Maybe you are looking for

  • Remove firebug

    to remove the bug next to the Home icon icon

  • Cover is correctly displayed in iTunes

    Hello My clutch is not displayed correctly in iTunes, it is display with a black background and jagged edges - https://itunes.apple.com/au/podcast/born-to-kick-arse-podcast/id1097445993?mt=2 It should look like this (white background, clearer, crispe

  • Second installed hard drive is not seen by BIOS

    I have a computer "AMD A8 - 5500 3.2 GHz APU" - A8 Vision HP Pavilion BIOS v.7.10 I installed a second SATA drive (a 3 to Seagate) and booted up the factory installed OS on the first SATA hard drive. The operating system has no problem see the car, s

  • Impossible to uninstall the program

    I tried over and over to unstall program.   What is my next step without having to buy one of these services to uninstall?

  • HELP PLEASE WITH MY WIRELESS NETWORK CARD CONSTANTLY HAVE TO BE SET

    I have an Acer Aspire 5552-3691 with Atheros AR5B97 wireless adapter. It shows that I have an internet connection, but I have to constantly run network diagnostics it resets then the wireless network adapter and which solves the problem. However, the