False values of voltage using NI 9225 and ELectrical Power Suite 2014

Hi all

I have a few problems regarding the use of a map of analog voltage NI 9225 with code exaple quality Applications of power (cRIO) (Delta-Sigma) of the Electrical Power after 2014. Running the example code I values of voltage wrong, they are exactly doubled! For exaple with a 230 V RMS input I get 460 V RMS. I think the problem is the module, because by default, the code example expects to use a module NI 9242. Before the upgrade to LabView 2014, everything worked perfectly with EPS 2013. Am I something missung? I configured the C module in the project manager in the same way that I used with EPS 2013. Can someone help me? Thank you

Gianluca

Hi Gianluca,

If you use other types of modules in the example, you must take care of the conversion of accuracy in the FPGA VI.

As FPGA VI resampling takes the signal of type fixed point +-24.1, you convert the +-24,10 and-24,5 to 24.1 in the Data Acquisition loop before feeding to the FPGA VI resampling.

And in the loop of data Communication, you restore the output of +-32,3 to 32,12 - signal + and +-32.7. Notice at the time the zoom in and zoom out are equal.

If you change other types of modules of different precision, you must change the precision of the given thread.

For example, 9244 accuracy is +-24.11, so when you convert it to +-24.1, you must restore it by variation of-32,3 to-32,13.

Please refer to the screenshot:

Tags: NI Products

Similar Questions

  • Access to the use of memory and processing power charge connect 6224 via SNMP.

    Hi support them

    Is it possible to access the use of memory and processing power charge connect 6224 via SNMP? If possible, should the mib I use?

    Thanks in advance

    Rogerio

    When looking through the available MIB found in the zip file for download of the latest firmware.  I see that the MIB fastpathswitching mentions 'agentSwitchCpuProcessGroup '.

    MIB fastpathlogging has 'agentLogInMemoryConfigGroup '.

    There is no reference document material about what offer different MIB.

    I hope this helps.

  • Setting a value to variable using an sql and javascript

    Hello

    I'm trying to be a value of variables using javascript and sql. The code is as below:


    < script >
    function getDueDate() {}
    function getVal (item) {}
    If ($x (point) .value! = "")
    Return parseFloat ($x (item) .value);
    on the other
    Returns a null value.
    }
    $x('P26_DUE_DATE').value = select sysdate + $x ('P26_PAY_TERM') of the double
    }
    < /script >

    Can this be achieved?

    Hello

    JS and SQL are 2 different languages, so you can't do what you want to do using this script. The good thing is that you can still do so using JS.

    $x('P26_DUE_DATE').value = select sysdate + $x ('P26_PAY_TERM') of the double

    the "Select Sysdate" try to use the date JS Object (http://www.w3schools.com/jS/js_obj_date.asp) to add dates together.

    Martin
    -----
    [http://apex-smb.blogspot.com/]
    [http://apex.stackexchange.com/]

  • Using After Effects and other Adobe CC 2014 on screen 4K.

    I use a new view 4K and I put across the user interface in Windows 7 to 150%, and all my other programs scaled very well. When I use my adobe programs the user interface is very low in the resolution of 4K, I have been able to scale Photoshop CC 2014 but am having problems scaling of other programs like AE, Pr and have etc. I need help with this because all these programs are unusable at this size, tools, numbers, etc. values are very small and hard to see. If anyone is having the same problem or has a solution please help, thank you!

    It is a known problem.

    We are doing the necessary work to make After Effects work well on Windows HiDPI monitors.

  • How to use the battery and the power cord in a smart way?

    Is it possible to do a power management, that use the battery as long as there is power in the battery and the first when there is no more power use the external power supply, which has been active all the time without loss of battery charge?

    Batteries wear out with use.  And they can wear out more when you do not use unless you physically remove them.  If you travel a lot, you will have to bite the bullet and buy new batteries every two years.  If you are usually connected to a power outlet, remove the battery and it will last much longer.  For more details, refer to the User Guide.  Different brands and models my vary in their recommendations.

    Good luck.

  • How "block of surge protection" works in NOR-9225 and NOR-9227

    Hi all

    Hi, I use NI 9225 and NI 9227 to collect data under high voltage.

    According to the data sheet, the NI 9225 has protection against overvoltages and NI 9227 overcurrent protection.

    I wonder what kind of protections, they, I mean, if high voltage or current applied, what will happen to the material? Fuse has popped up? Or trip relay? Can I re - opening after the failure to tension?

    Thank you

    Jenny

    Hello Jenny,.

    We have an IC isolation that will protect some of the internal circuits. There is no fuse or relay you can replace. Because of this, if you exceed the specifications (±450 VDC for the RMS A 9225 and 5 / 10 A RMS for 1 s max with 19s cool time 5A RMS for the 9227), so will most likely damage your card. If you open the module, you will lose the warranty of your device.

  • How to choose the maximum value in a column of a table by using the cursor and iteration

    Hello everyone
    I have a table loan_detail
    and a column inside loan_amount
    now I want to take the values in this table with the cursor, then using iteration that I want to take a max value of him using this slider
    Here is my table
    LOAN_AMOUNT
    
    100
    200
    300
    400
    500
    5600
    700
    I could make using the concepts of single loop, but when I tried to do by using the slider I wasn't able to do.
    Concerning
    Prost

    You can get the maximum value for a particular column in SQL, PLSQL needn't; Always avoid PLSQL so you can do it in SQL. But always as an example:

    declare
    number of Temp: = 0;
    Start
    for news in some loan_amt of loan_detail
    loop
    If (temp
    end if;
    end loop;
    dbms_output.put_line (temp);
    end;

    Edited by: Mahanam January 5, 2011 03:11

  • Caveat! A false Jobform company sends a questionnaire on the use of Firefox and promises a gift request credit card number to cover publication fees. I have inte

    They use your language and everything including the Firefox logos seem original and tht is why answered questions often asked about how I use Firefox etc... When you interrupt when they ask for your number of credit card they will send days nex sends you e-mail you asking to complete your order and providing new gifts in order to get the details of your credit card. Even if you do not respond they eventually send you an invoice from 49 euros in English language for engraving a membership to jobform.com as follows

    Thank you for your purchase on Jobform.com. Here's your bill:

    Invoice: Product: open the balance: Date: Options:

    1. 123296 subscription 5 days to Jobform.com premium 9 EUROS 13/02/2015
    2. 139622 premium subscription from 30 days to Jobform.com 40 EUROS 18/02/2015

    To settle your balance outstanding from 49 EUROS, you need to login to your account at Wakefield Finance...

    Hi peter, could you also post this information to the https://www.mozilla.org/en-US/about/l.../fraud-report/?
    Thank you

  • cDAQ-9174 with modules 9225 and 9239 for issue of power measurements...

    Hello to one and all.

    I do turn a cDAQ-9174 with modules 9225 and 9239 to measure the voltage (120, 208-240, 277 VAC) and current (via the clamp amp i400, Fluke - 10mA/exit A) respectively.

    I have the set scale scale Differential No. (1:1 I guess) for 100: 1 for current and voltage.  With the acquisition mode set to "N samples", 10 samples to read, at the rate of 500 m.  This feeds into 'EPM_Power.vi' (polymorphic VI of the palette of EMP... can be downloaded here, http://sine.ni.com/nips/cds/view/p/lang/en/nid/209826)

    I try to check my numbers with an Analyzer of power quality of 435 Fluke and supporting that upward is a Fluke 179 True RMS DMM with clamp amp i200 (for current playback)

    Now, I'm running is my numbers I take data look like they're following a curve of fishing... the averages are on the right, but when my blood pressure is expected to be relatively stable, it's over the map... same story with my current readings.  I tried several different filters without success.  I would like to have relatively stable values for voltage and current so my other values are correct.  Is there a signal conditioner or filter that will help with this?

    As always, a point in the right direction would be greatly appreciated!

    Thank you!

    Chad

    Hi, Will,

    Your request for a screenshot actuall lead me to the solution.  Not enough samples!  I took little data too slowly and only seeing part of the wave of fishing AC each iteration.  .. that became glaring when I tried to this graph.  So I increased my sampling of 100 to 5 Hz and things seem to have settled things down a bit.  Thank you for the good question!

    SH?

    C

  • 9225 and 9227 Module Constant phase shift (Drift)

    Hello

    I wrote a large VI and the am acquisition voltage and current with the 9225 and 9227. The phase of the voltage and current constantly derivatives. Its not just for phase a bit for all of the portion; the phase constantly cycles from 0 to 360 degrees. I have an excel file with locations showing what I mean. Series 1 = voltage 2 = current

    I wrote a small VI to do the same thing with nothing extra. I have not yet add the data record. I just use a graph to plot the voltage and phase. The voltage and current were still off!

    What do you think. Timers are bad on the modules? Slow computer? DAQ assistants are not good? IM using the internal timers in the modules.

    Synronizing at the beginning does not help because the signals would still go out of phase when even later. The load is a piezoelectric tube.

    I'm using Labview 2011 and NEITHER cDAQ-9178:
    9225 3CH 300 Vrms 24 BIT simultaneous AIN
    9227 4CH 24-BIT simultaneous AIN 5Arms

    See you soon

    I forgot to change to continuous. I do usually. The major difficulty that I did was remove the trigger "Start Digital Edge" completely. The trigger wasn't working properly with what I put it in, "PFI0"... Once I took it out, I got a sample invaded. I played with the numbers of the sample for the clock sampling and playback. I don't know why to use 25000 everywhere has not worked. But when I dropped the numbers to 6250 or something of this magnitude, the VI finally worked.

    The VI works with or without the clock I specified (Mod7/AI/sampleclock), which is great news! Thanks for this tip!

    The current and voltage are finally in sync. They will never out of sync! Finally I can calculate the POWER!

    This digital trigger was the worst headache... Thanks for the tips!

    I write the VI and the code snippet to my completed solution.

  • Authentication using the table and the existing database

    Very new to APEX here. In authentication methods and was curious to know if it is possible to link the Apex on an existing Oracle database and the table to get information for user name and password. Trying to use the employe_id and last 4 of their social.

    Any ideas or things I can read up to get a general idea?

    JosephPortello wrote:
    fac586, thank you. I didn't even know what I had done.

    Yet one thing that happens now after I corrected my incorrect syntax.

    CREATE OR REPLACE FUNCTION EmployeeIDAuth (
    p_username IN VARCHAR2,
    p_password IN VARCHAR2
    )
    RETURN BOOLEAN
    IS
    form l_count NUMBER;
    BEGIN
    SELECT COUNT(*) into l_count from table@db WHERE employee_no=p_username AND bdate=p_password;
    IF l_count > 0 THEN
    RETURN TRUE;
    ELSE
    RETURN FALSE;
    END IF;
    END; 
    

    Now returns an error of:

    Error on line 15: PLS-00103: encountered the symbol "end-of-file" when expects it one of the following values:

    not end up dominant static of final instantiable order pragma
    manufacturer membership card
    1. CREATE OR REPLACE FUNCTION EmployeeIDAuth)
    2 p_username IN VARCHAR2,
    3 p_password IN VARCHAR2

    Any other ideas?

    Remove the false token form.

  • Simple Script Execution: Days and only using 15% of the CPU power

    Dear community,

    I have a data file (.tdm) with 17 channels of waveform point each containing 13742499 data. I want to delete all the data points that are not actual load points. I have build a canal ("LoadTrigger") that is a trigger for points of interest. With the following script, I tried to erase the im not interested datapoints:

    Dim RowCount

    Number of lines = 1 to GlobalChnLength

    ChnVal if (RowCount, "/ LoadTrigger") = 0 THEN

    Call the DataBLDel ("1-17 ', RowCount, 1" ")

    End If

    Next

    The script is now running for more than 20 hours with no result. Can someone please tell me if I am wrong in my script, or if there is a way to make it faster?

    In addition, I wonder why DIADEM uses no more than my cpu power? As you can see in the photo attached DIADEM mainly use a carrot and never more than 15% of the CPU.

    Best regards and hoping answers of construcitve!

    If you try to delete rows based on conditions, you can use code like this:

    Option Explicit
    CreateExample

    "Make sure that the correct group is active.
    data. Root.ChannelGroups (1). Activate
    ' chncalculate allows you to set values that match a NOVALUE condition
    Call chncalculate ("Ch (1) = Ch (1) + CTNV (0 = Ch (1))")
    'Delete the lines that contain a NOVALUE in channels '.
    Call ChnNovHandle ("[1] / [1]", "" [1] / [2]'-' [1] / [3]'","Delete","XY", 1,0, 0)

    Sub CreateExample()
    data. Root.Clear
    Dim g: g = given game. Root.ChannelGroups.Add ("group")
    Dim c1: set c1 = g.Channels.Add ("c1", DataTypeFloat64)
    Dim c2: set c2 = g.Channels.Add ("c2", DataTypeFloat64)
    Dim c3: set c3 = g.Channels.Add ("c3", DataTypeFloat64)
    I have Sun: for i = 1 to 100
    C1 (i) = I mod 2
    C2 (i) = I have
    C3 (i) = i * 0.00001
    Next
    EndSub

    that will remove all lines that contain a 0 in the first round.

    will be converted to

  • How to change values to text using VBscript

    I have a model that contains the part of text-value 6 and a table. I figured out how to import the string table to the table, but I couldn't find how to change my text values. Below, you can see my table of text values.

    I'll send this table to DIAdem report function. Could you please help me how to change my text values in TDR using VBscript.


  • you want to send data using labVIEW to arduino using write visa and the process and to take action using arduino. A

    I want to send data using labVIEW to arduino using write visa and the process and to take action using arduino. After that, I want to arduino to send out necessary via a serial port to labVIEW which should be read using visa read and store in a chain. While I am able to write or read both individually, I can't do it consecutively. I used advanced read and write vi for checking my code, but nothing is helping. The wrong bed 'time delay before execution. " Please let me know where I can go wrong. Also is it possible to write code for hx711 using labVIEW

    1. you need not "\n" on your orders println().  This command adds an end of line character already in the message.

    2. you get the error because you have a loop around your reading.  After the first reading (well technically, the second because of you add an extra line end character), there is nothing left in the port.  As a result, you will get the timeout.

    3. you should really consider using a Structure of the event.  This way you just don't write and read when you press the Write button and you can also use the structure of the event to make the loop to stop.  I also go up to close the port inside the stop-> value Change event.

  • Connection of the PXI-4071 - can sound and LOs remain interconnected for the impedance of the cable 2, voltage DC/AC, frequency and waveform acquisition?

    Hello

    We use the DMM and SMU-6363 map to test a hardware device.  We will also use a PXI-2530 b switching matrix.  We will use the digital multimeter to perform the measurements of voltage, DC and AC, measurements of impedance (2-wire and 4-wire), frequency and waveform acquisition.  Can the PXI-4071 left be 4 wire connected (black jacks taken connected and red connected) mode and still be used to perform all other measures (including 2 impedance of the cable).  This would simplify the switch connections.

    Current measures use the son + and LO, but the HI and S-can remain connected.  The problem you are having is if you have an active device the digital multimeter and take you a 4-wire resistance and the measurement of voltage with all 4 wires connected and then change to a current... When you do this, short-circuit you the terminals of the DUT, on that you just take the measurement of the resistance.  If the terminal HAD, say, a power supply 10V, then you have just shorted out.  Of course, this isn't a problem if your Instrument is a passive device, or if you change just the unused two lead whenever there is an active device of low impedance.

    If you want to make voltage, current and 4-wire resistance, you need all 4 wires.  If you want to do the voltage and current, you will need 3 wires, but you could connect the s + Hi and then just do the two wires.  I vote running every 4 son to your DUT for maximum flexibility.

    2-wire resistance is a must if you are measuring resistance above 10 MOhm.  Alternatively, you can use 4-wire for all measures.