I need to control and communicate with a cell - Mecmesin Multitest 2.5 k - via USB/serial - ideas?

Help!

I am currently working on a project where I apply force on a hardware - software (non-associated labview) who is provided with the tool (Mecmesin Multitest 2.5 k) allows to control but (of course), I want to control it using LabView, as well as plot data using LabView.

This issue has developed several years on these forums, but it was never a good answer. I'm suitable for use of LabView for the simplicity of programming (loops, arrays, sorting data etc.), but am a newb in most other respects--unless companies have subVIs!

The tool is now using a USB-serial connection (PC - end is USB). I think I could use rs232-rs232, but someone in the dept said it still doesn't work, which could mean that the cable series is cheesy. I wish it works however, USB, as ideally something else is connected to the serial port currently.

Is this possible? If Yes, can anyone help?

The next step is to open the manual for programming the instrument. Determine the commands that you need to send to the device and to understand how the device sends information. You can use the signs VISA for that test or program such as Hyperterminal. To send something to the instrument, you use an entry VISA (tab ViWrite) and for the data, a VISA read (ViRead).

Tags: NI Hardware

Similar Questions

  • Updated at El Capitan and my mac doesn't see my camera Canon (connected via USB). Need help. Cannot import images.

    Updated at El Capitan and my mac doesn't see my camera Canon (connected via USB). Need help. Cannot import images.

    Try using using the Image Capture (in the Applications folder)?

  • HA Admission Control and DPM with clusters with 3 and 2 guests

    Hello world

    I have two clusters, with 3 hosts first and second with 2 hosts. And I have a few questions about the activation of DPM:

    1. 3 HA/DRS cluster hosts, I activated the admission control to the percentage of cluster resources reserved as failover alternative DART at 33%. Because I have no VM on this cluster, when I enabled DPM he put 2 hosts in standby mode. And now when I powered on any virtual machine on this cluster I get error "the operation is not allowed in the current state. It is possible that when I disable the admission control - but then I lose protection HA. So in Group 3 knots I would always have 2 guests under tension when strict admission control VMware HA is enabled? It is possible that if in DPM options I disabled power on 2 hosts management. Am I wrong? And DPM set power switch my third host (for example brings leaves host standby mode to keep the needs of failover)?
    2. Cluster of 2 guests when I activate CAC (50%) Should I disable DMP, because I still have 2 powered on the hosts?

    1 HA/DRS + DPM you 1 single host standby because HA need 2 host be powered.

    2. no DPM in two Cluster nodes if HA is enabled.

  • Blue screen (BSOD) and communicate with GPIB instruments

    I have a communication problem with a Pressuregenerator of PCS400 of Mory via NI-VISA. My simple application sends requests to the instrument and reads returned responses. These actions are performed in a while loop until you press the stop button on the front. The VISA-resource is initialized before the while loop and closed after leaving the loop.
    Now the problem is that the computer will work in a blue screen with the message "NMI: check / memory parity error parity." This blue screen can be caused by switching between the programwindow and another window (for example the file Explorer) - I know it sounds strange. Then the blue screen appears after about 20 seconds. The computer worked stable in the daily work with windows.

    I have already tried several things
    -Wait a while (up to 30 ms) after writing and reading of VISA-online blue screen
    -Communicate with the Agilent 34401 multimeter with the same program (but other requests of course)-online no crash!
    -Test memory on system for about 45 hours with memtest86 + -online no found error
    -Stresstest with Prime95 (product 100% of CPU load) for about 1.5 hours on windows => no crash

    Request sended:
    _PCS4\sREADING? \n
    _PCS4\sERR? \n
    _PCS4\sSTAT? \n
    _PCS4\sERR? \n

    The system is configured as follows:
    -Windows XP Professional SP3
    -LabVIEW 8.5 FDS
    -NI-VISA, NI-VISA run-time V4.2 V4.2
    OR-488. 2 V2.52
    -Adapt PCI-GPIB with 3 x Mory PCS400 Pressuregenerator (1.3 to addresses), Agilent E3641A (address 5) and Agilent 34401 connected (6 addresses)
    -Acquisition of data PCI-6052E card
    -Some RS232 (PCI) expansion cards

    Do you have any ideas of how this strange problem?

    Thank you.


  • I'm stuck on my account. This happens fairly regularly. It is almost impossible to find the number of apples and communicate with them.

    I'm constantly locked out of my account.  Why it can't get Apple as well as several accounts? Whatever you do, never having accounts families or several user names.  Information in addition to 60 Days (?) to connect with another Apple ID pending?

    Manage your devices in iTunes - http://support.apple.com/HT204074 - regarding, "When you turn on iTunes game, sharing family or automatic downloads, or when you download past purchases on your computer, iPhone, iPad, or iPod touch, this device becomes associated with your Apple ID. You can associate your Apple ID with up to 10 devices with up to 5 computers at any time. Each computer must also be authorized by using the same Apple ID. You don't need to authorize your iPhone, iPad or iPod touch.  "After that you associate a device in your Apple ID, you cannot combine this unit with another Apple ID for 90 days.  If you sit down and reason through you can do that Apple probably knows very well what she does and is likely to do to pressure from media companies who don't want a dormitory full of students turning on file sharing for all their media for an orgy of hacking.

    Don't put yourself in a position where you have to re-download. Make sure you have all your purchases backed up to a local disk. It's smart to do it anyway because if an item comes form the store that you are not able to re - download.

  • Variables shared on the CFP becomes zero and communicate with the PSC

    I use a PC with two network cards, one on one wired ethernet, the other on a private IP directly connected to a PSC-2120 (running 6.0.5 full). I use variables shared on the CFP and aliasing to static variable on the PC (initially LV 8.6, now 2009f2 with the same problem in both). PC variables are stored in a database of the Citadel. I have the CFP program built to run at startup, and it works fine when not connected to the PC.

    When connected to the computer all variables shared on the CFP will from time to time (almost 1/min for a few seconds) and randomly becomes zero. Even if they are resized only not to allow to become zero. I can say that what is happening because I use the variables shared as relay alarm thresholds module, and relays to suddenly slam on. I confirmed this by checking that the CFP has written to its internal flash memory card, and indeed variables suddenly are nil. On the PC it says that it cannot connect with the shared variables. This happens although LabView is not running on the PC!  Manager distributed system variables shared on the CFP are listed as (disconnected) during this time, but it continues to be able to read the correct values to live of the CFP entered the module.

    Interestingly, I have the same computer with the same configuration (and a PSC-2120 different) that does not have this problem. Both PC's are new from the factory. I tried to disable the second ethernet adapter, closing all wall-lights/virusscan and re - order network cards in windows networking/advanced settings without success. Simply disconnect the cable between the PCP and the PC does not cause this problem and shared variables become zero does not trigger the network on the CFP monitoring parameters.

    My questions are:

    (1) this problem can be fixed?

    (2) lost communication for a few seconds is OK, but given the PSC variables become zero is not. Can I stop the variables of the PSC to become zero?

    -David

    To update my previous post, it turns out that the re-installation was not the solution (the problem of variables start again occurring began). I finally realized that this only happens when I put the CFP time server IP (under Additional Configuration to the MAX). I was upgrading the server time on 192.168.1.1 IP (the IP address of the adapter of the CFP has been directly connected to). However, this compensation setting solves the problem.

    I do not think that the time on the CFP must need to be addressed too often (and perhaps it is synchronized if I never re - deploy files to the CFP?), so I guess that leave it empty. Curiously, I have the same setup time on an another PSC server IP and PC (running LV 8.6) without problem.

    -David

  • Can I create a page Web of Muse that it is child Pages could be open on different computers and communicate with each other?


    I want to create a Web site with 4 child pages that interact with each other when opened separately on 4 different computers.

    Is this possible?

    Example:

    Child 3-page Web site open on 3 different computers with triggers and the following objectives:

    PC #1 - child #1 - confirm button (trigger)

    PC #2 - Page #2 - accepted Confirmation Popup, resulting by clicking on 'confirm' page computer/child #1 (target)

    PC #3 - child Page #3 - calendar Popup changed, resulting by clicking on 'confirm' page computer/child #1 (target)

    Any ideas are much appreciated.

    THX.

    No this is not possible with Muse.

  • CS4 will not install on the new computer.  indicates the installer is corrupt and communicate with adobe.  now, what

    I had cs4 on my old computer which has since crashed.  Trying her my new compuater (with windows 8) and it is said that the installer is corrupt and to contact adobe for a fix.  So, how to install?

    Hi ip photo

    Please see the below Kb Doc: http://helpx.adobe.com/creative-suite/kb/error-installer-database-corrupt-install.html

  • need your input and help with menus

    http://i-Xray.BusinessCatalyst.com/index.html

    It all started yesterday. 1 project using muse

    mainly Visual Basic background, now I need to learn how to make menus.

    all feedback b4 I really this ruin would be greatly appreciated.

    you will need to add several page (s) under the poetry and try some sort of effect Lightbox.

    I'm still working my way through the various tutorials. you don't know any code at all.

    Thank you

    Randy

    I think you want to add a submenu to the menu of poetry to your site.

    If that's what you're looking for, all you need to do is add child pages to the page of poetry in outline view and create a structure of page to something as shown below.

    Then just open your master page in Design view (the one where you inserted the menu), click on the blue arrow icon to access the Menu options and change the type of Menu for all Pages.

    To create a viewer, you can use the built in Lightbox widget. Refer to the tutorial video below which will give you more idea on how to go about do - http://tv.adobe.com/watch/learn-muse/adobe-muse-creating-a-lightbox-slideshow/

    Thank you

    Vinayak

  • How my VISA control can communicate with an instrument via TCP/IP?

    VISA controls are my GPIB instrument when I select "Refresh", but none of the instruments TCPIP. I know that the TCPIP instruments are connected because I can their ping.

    If you are looking for how to build the resourcename, we find more here

    TCPIP SOCKET TCPIP [Council]: host address:ort:OCKET

    for example

    TCPIP0::1.2.3.4::999:OCKET RAW TCP/IP access to the port at the specified IP address 999.
    TCPIP::dev.company. com::InStr A TCP/IP using VXI - 11 device located at the specified address. This example uses the default value for the inst0 LAN device name.
  • Satellite A200 (PSAEOE) with Windows XP does not recognize the PSP via USB

    Hello!

    So I have problem with my laptop Satellite A 200 (PSAEOE). After Installing Win XP on it, I could download or download all the files to the PSP (Play Station Portable). To connect to her I use USB. I check the USB connection to all other devices works on detected well in all USB, also PSP on the same works of cable with the other PC with no problems. May be I have to install the USB driver? If Yes, please show me where I could download it. Or what it might be.

    Thank you for you answer.

    Hello

    If all other devices work while your machine is OK and your PSP is also correct. But to find the drivers who supposedly work, you´re definitely wrong here in this forum, I suggest you to check these site:

    http://forums.Gametrailers.com/forumdisplay.php?f=25

    You will find that several Playstation Portable related information.

    Good luck
    Welcome them

  • Dynamically loaded swf to communicate with the MovieClip on the stage

    I have a devil of a time here with a problem. I've got an xml document when it is clicked on a certain button on the stage, it loads its corresponding external swf into an empty movieclip on the stage. This empty clip is support for all external SWF.

    So what I'm trying to do, is when an external swf is loaded that has buttons on it, I need these buttons to communicate with the main timeline and delete an mc who is on the stage. I need the swf (s) currently loaded to be able to do. So whatever the currently loaded external swf is loaded, I need to talk to the main timeline. (I have external SWF files a certain need to do this) So I guess I have to somehow target the external swf currently loaded so he could talk to the main timeline.

    The code below in summary does not, but it gives no errors either. I don't pretend to know what I'm doing, and I did not find a solution to this particular problem. If someone could give me advice or direct me to a solution. I would be so happy.

    scenario focus

    var object: index = this;

    function loadComplete (e: Event) {}

    TweenMax.to (index.mcholder, 1, {alpha:. 5});

    Add the current module to the mcholder movieclip according to addChild

    index.mcholder.addChild (e.currentTarget.content);

    (e.currentTarget.content as MovieClip) .addEventListener ("eventTriggered", startListener);

    function startListener(e:Event):void {}

    var ext_swf:MovieClip;

    ext_swf = e.currentTarget.content as MovieClip;

    trace ("external swf");

    ext_swf. Button1.addEventListener (MouseEvent.CLICK, talktomainswf);

    function talktomainswf (): void {}

    TweenMax.to (index.mc_thatsonthestage, 1, {x: 1000});

    }

    }

    Now we have the first load we set to false firstLoad

    index.firstLoad = false;

    }

    function loadError (e: Event) {}

    trace ("Error");

    }

    }

    You can use the event to communicate with external SWF Dispatcher and a main timeline

    as follows:

    //in an external swf
    
    //Once loaded
    function onLoadComplete(event:Event):void
    {
         //dispatch an event in the form of a string
         dispatchEvent(new Event("Talk to Main Timeline"));
    }
    
    //On the Main timeline
    
    //listen for "Talk to Main Timeline"
    stage.addEventListener("Talk to Main Timeline", listenForCallsFromExternalFiles, true); 
    
    //if the event is heard, do this:
    function listenForCallsFromExternalFiles(e:Event):void
    {
         trace("I heard ya, now do stuff...");
    }
    

    That's the idea anyway, I use it all the time.

    hope that helps,

    ~ chipleh

  • Index and MATCH with several conditions

    Hi all! I hope I can explain what I'm trying to do here clearly. This is my first shot:

    In the table below "Office entry Worksheet - area 1" I am trying to insert a formula in column B which will result in an answer ' true / false ' and be used for conditional highlighting

    The first condition that must be met is that the checkbox in the table "Labor and material Checklist" column A or check, if it is not checked, then move to the next line. If the box is checked, then the following condition would need to index and match with column C of table "Office entry Worksheet - area 1" in column B. If there is a match in column C, then move to the next line of column B in the table 'labor and material Checklist '. This continues until there is a 'match', then highlight column B in the table "entry Office sheet - surface 1" in red (or something like "MISSING task")

    It is an estimate and the idea is to ensure that the Phases that are checked in the table "Labor and material Checklist" will not miss when you use the table "entry office map - zone 1".

    Thank you very much for your help!

    Tim

    Is it possible that I can download this spreadsheet file?

  • I can't use a website should I use with Firefox 5.0. So, I need to uninstall and go back to 3.5 or 3.6. Please notify. Also, my control panel blocks from Windows Explorer a PMP. That's why I use Firefox instead. Thanks for any help!

    I can't use a website should I use with Firefox 5.0. So, I need to uninstall and go back to 3.5 or 3.6. Please notify. Also, my control panel blocks from Windows Explorer a PMP. That's why I use Firefox instead. Thanks for any help!

    I thought it would be FAFSA is causing your problem.

    Install Portable Firefox 3.6.x on your hard drive to a Web site. This will not affect any of your current Firefox installation.

    http://PortableApps.com/apps/Internet/firefox_portable/localization#legacy36

  • Control valve tank with plots and twists of the water

    Hello everyone!

    This is my frist real project. And I am in love with this program so far!
    I have a little problem. Here it goes:

    The project of the University which has been given to us is to simulate a (entrance OPEN/CLOSED valve) water tank filling and emptying (output range of 0 to 100% valve), but with a few twists. It is done by the command to the movement by the following formula: flow rate [%] = position of the valve [%] * current water height [%]. (The output valve takes values in percentage from 0 to 100% giving the voltage digital output of 0 volt to closed(0%) or 5 volumes for open(100%), the strange part, is that the digital output can have a value of 0 or 1, so the position of the valve is "simulated".) And the supply valve (valve entrance) water can be open or closed (if 0 or 5 volts)

    -Sequences cannot be used (only one for initialization and cleanup can be used)

    -When the water level reaches 10% of the max height, LED lights, the output valve closes and the input valve opens. The LED is on until the button is pressed. The output valve cannot be opened until the water height is greater than 50%.

    -When the water level reaches 90% of the max height, LED lights, the entrance valve opens. The LED is on until the button is pressed. The output valve cannot be opened until water height is less than 50%.

    -There are two mode selectios:

    Control by hand: the person sets the position of the valve in percent and the flow is calculated by the formula and displays on the screen.

    Automated: Every 2 seconds the program assigns the place valve so that the current is constant, it is the value by the person.

    -(N'ont pas gottern ici encore) ESD (Emergency shutdown switch) the when is pressed both the entry and exit valves are closed. The entrance valve cannot be oppened up to the RESET_IN key. If the program is 'Control in manual mode' output valve cannot be opened until the RESET_OUT touches. In which case the output valve is in "Auto" mode, it does not start work until the RESET_OUT touches.

    My main problem here is the fact, I cannot use sequences and nested, then the loops do not work. The aqusition of data stops until the inner circle loop ends and I need both work. Especially for the 50% rule.

    I've been hitting my head for a few days and still couldn't come up with a solution.

    I have attached the VI.

    Tips are welcome!

    For the time being the outputs inputs/digital analogue are a simulation (when the time comes the USB-6008 material is used).

    There is some things that confuse me (as if the valves are percentage, from 0 to 100%, or digital, 0 or 1), but regardless, it's a detail.

    Review the task and consider what you want to do and that the "States" the system may be in.  [This with a pencil and paper, do not open LabVIEW yet...].  The description of your problem learn you something about the transition from one State to another.  Here is a thought exercise:

    Suppose you are in a "State".  Let the system to "do its thing" for a second, which could involve the water flow.  At the end of the second re - evaluate your state - it changed?  Continue to do this...

    This model of 'something' is called a State Machine and LabVIEW has some very good examples and tutorials on the creation and use.  The important things to do before you make your coding is (with the pencil and paper) decide on your statements, decide on the transition rules, and then you are ready to code.

    There are different ways to build a State Machine in LabVIEW, but the basic idea is quite awhile (to keep things moving along, a shift register which includes the current state, usually represented as an enumerated, Type whose values are simply the names of your States (made for a very mnemonic routine) and a Case statement whose power is the State of the shift register that contains the code to achieve this) particular State of loop, including the implementation of transitional rules (which can cause a new State on the registry to lag when it comes out of the investigation of the case).

    A great exercise.  Strive to keep your block diagram on a single computer screen by making use of the screw subgroup chore (as the management of changes in the water level based on the current water level and "States" (which here means 'values') two valves).  Once you understand the 'Idea' State Machine, you should be able to write the code for this problem quite easily.

    Bob Schor

Maybe you are looking for

  • Why search for Verizon continue to flock to the top?

    Whenever I do a search, he goes in search of Verizon. I have Google set as the default browser, but whenever I type something in the search bar I have to manually select the option of Google. If this isn't the case, it is by default in the search of

  • Gibberish, loading up on some Web sites

    So, for some time now, some Web sites have been appearing as pretty much gibberish. Just a series of things of brand black diamond-question thoes. I thought it was a problem of encoding, but I couldn't fix it, so I just used chrome for the weird site

  • Script error when you start using the ActiveX API

    Hello I try to start the TIARA of c# application and identify the data portal. This is the code I use in c#: tiara. CmdExecuteSync ("wndshow ('shell', 'normal')");tiara. CmdExecuteSync ("ScriptStart ('g:\\Build\\script2.)") (VBS') "); The first comma

  • Update of Silverlight for Vista Home Premium

    I installed the latest version of windows (silverlight) it shows that it is installed and whenever I go to shutdown or restart of the screen installation shows updated and hangs there. I left it on last night during overnight and this morning the scr

  • menu bars display in full screen, but body of window is off

    All the bars at the top of the page show all the way through the screen, but the body of the screen is smaller.  Even the picdures screen saver is smaller.  I hjave a few classics and severyting is 100% st This happened after I had the system cleaned