Measurement of voltage USB 6008 ranging from 1mV

Hi, I'm doing a supply at an angle for an amplifier using the USB-6008. The ranges I look are - 0.5 to 0.5 V and 0.5 to 2.5 V. To generate negative tensions, I use the + 2, 5V for a reading of differential voltage output as a 'ground '. Voltage measures have a delay that varies with the voltage, given this by drawing these variations and set the output to data acquisition there is still a 'noise' of +/-1 mv so that it is clear from this parcel of tension against the sample number:

It seems that at some point the values are being rounded up to the nearest millivolt. I need to get to a resolution of 0.5 mV for my device, it will be possible with the USB-6008?

The AO USB-600 x has a range from 0 to + 5 V and the 12-bit resolution. 5/4096 = 1.22 mV. Absolute precision is 7 mV typical and 36.4 mV maximum full scale.  The noise of the AO is not specified.

If you measure the results with THE USB-6008, you have at least 0.5 mV, similar resolution system noise and absolute accuracy of 2.5 mV or more.

It's probably as good as you will get with the box USB-6008.

Lynn

Tags: NI Hardware

Similar Questions

  • Get incremental counter/sound to work along side with action with usb-6008 with labview tia sal22

    Get incremental counter/sound to work along the coast with usb-6008 with labview tia sal22

    Hi all

    I can get this vi to work if they are distinct from the vi but I can't join them together

    Example of my error:

    If buffers are set to 0 the freq counter increment works, but no sound
    If the buffers are set to 1 the audio works fine but is not increment the Freq counter
    If the buffers are attached to more 1 clicks and pops are coming

    That's what I'm doing:
    (1) have the frequency of increment of my internal sound card to a certain level as .01hz a second until he gets to 20 000 hz

    (2) use my device usb-6008 daq, which is connected to the same machine to measure the voltage at the same time. (I am in a position very low voltages between 1-5volts)

    (3) output to a worksheet text file which will show you:

    time in seconds, frequency, voltage
    0,400.01,2

    1,400.02,2.5

    2,400.03,1

    I'm a bit confused about how connect the increment and the audio during the measurements with the usb-6008 housing on the same machine
    at the same time and in the same VI.

    Anyone have any ideas?  I'm using labview 8.5
    TIA sal22

    Ha ha you have been deceived by a dynamic thread. Insert a convert from Dynamic Data Express VI (Palette to own: Signal handling screw Express) between the daq read and build the array function. Then it won't work. Now the value in the dynamic data is only converted to a numeric value

  • Want a ramp of output voltage over time and measure input 2 analog USB-6008

    Hello

    I want to produce an analog voltage output signal that increases over time with a certain slope, which I'll send in a potentiostat and at the same time I want to read voltage and current (both are represented by a voltage signal) that I want to open a session and ultimately draw from each other. To do this, I have a DAQ USB-6008 system at my disposal.

    Creation of the analogue output with a linear ramp signal I was possible using a while loop and a delay time (see attachment). Important here is that I can put the slope of the linear ramp (for example, 10mV/s) and size level to make a smooth inclement. However when I want to measure an analog input signal he's going poorly.

    To reduce noise from the influences I want for example to measure 10 values for example within 0.1 second and he averaged (this gives reading should be equal or faster then the wrong caused by the slope and the linear ramp step size.) Example: a slope of 10 mV/s is set with a 10 step size. Each 0.1 s analog output signal amounts to 1 mV. Then I want to read the analog input in this 0.1 s 10 values)

    Because I use a timer to create the linear ramp and the analog input is in the same loop, the delay time also affects the analog input and I get an error every time. Separately, in different VI-programs (analog input and output) they work fine but not combined. I searched this forum to find a way to create the ramp in a different way, but because I'm not an experienced labview user I can't find another way.

    To book it now a bit more complicated I said I want to measure 2 input analog (one for the voltage of the potentiostat) signals and one for the current (also represented by a voltage signal) and they should be measured more quickly then the bad of the analog signal. I have not yet started with because I couldn't read on channel work.

    I hope someone can help me with this problem

    An array of index. You want to index the columns for a single channel.

  • NEITHER USB-6008 can measure an AC voltage

    Now I am doing a project, but don't know if DAQ can measure an AC voltage or not, the acquisition of data that we use is USB-6008.

    Yes, it can measure AC in the range of +/-10 volts and have a sampling rate of 10kS/s.

    Application engineer OR

  • NEITHER USB 6008 voltage offset using CSR and measurement of diff.

    Hi all

    I am currently trying the NI USB 6008 housing and I'm getting problems when reading voltage analog using CSR or differential.

    So basically, what I want to measure is a PWM signal (0 to 12V), which is divided by a divisor of tension (by two). But instead of measured 0V and 6V

    I am in a position a constant 0.8V and approximately 3V.

    On the side of digital data acquisition, I give you on impulses for the SSR... and it works fine.

    I connect it that way: http://digital.ni.com/public.nsf/allkb/95CC0CB11D7DF3D18625712E000C4ABD?OpenDocument

    Would apreciate any help

    Best regards

    EDIT: Attached graphics acquired are

    What is the impedance of your voltage divider?  The input of the USB-6008 impedance is not very high. If the impedance of the partition is large, it could cause the effect you see.

    Lynn

  • DAQ USB-6008 will be able to power and record voltage for UMS T5 blood pressure at the same time?

    I would use my NI USB-6008 to power my blood pressure monitor UMS T5 (http://www.ums-muc.de/en/products/tensiometer/t5.html) but also to take readings of it, but I don't know if it's possible to do it properly. The power supply for the instrument can be as low as 5V, I can easily get the dedicated + 5V channel. I'm able to feed the instrument and connect it to an analog input on the 6008 and measure a voltage in differential mode. However, when you read the documentation of support for the instrument, I find the following:

    "Potential pitfalls of data acquisition: the pressure transducer is configured in a full Wheatstone bridge, the input voltage and mV signal output can be connected to the same reference (mass)." Therefore, the mV output signal can be measured using a differential voltage measurement. Therefore, do not make an asymmetric measure of pressure transducer mV output. "(http://www.decagon.com/assets/Uploads/MeasuringUMSTensiometerswithnon-UMSControlandDataAcquisitionSystems.pdf)

    My understanding is that the 6008 can take a differential measure if I attach the signal '+' and the signal "-" to the analog inputs of positive and negative terminals. However, it seems that all the ports of ground on the 6008 are grounded to the same reference, which would make my measure of invalid tension according to the above paragraph. So my real question is: if I try to record the voltage with one of the analog inputs on the 6008 in this way, is the valid measurement? Or I need to find a separate power supply, with a different reference field to ensure that the measure is accurate?

    The technical details of this device is very poor. The manual is not much better. Companies that want to sell scientific equipment should publish decent cards or get out of business.

    In section 3.4.3 General requirements the device is described as a "bridge not amplified circuit.  This information along with the impedance of the bridge should be in the specifications, because it is essential to apply the device under any circumstances other than the nominal behavior in 10.6 V.

    The answer to your question is:

    You can use it with the box USB-6008.  The 5 V supply will result in output voltages a little less than half (5/10.6) the voltage specified in nominal conditions. You can use the differential input mode on the box USB-6008.  The absolute input voltages will be approximately 2.5 V with the 5 V power supply. This voltage is in the range of the aircraft. The differences are likely to be less than 100 mV. The resolution of the USB-6008 on the + /-1 V is located about 0.5 mV so your resolution of pressure will be about 1% of full scale. The voltage input impedance and termination of the USB-6008 will present a few errors.  These can be in the order of 5 to 10%. I can't predict much better without the missing bridge impedance specification.  These errors should be relatively constant and systematic. A calibration of the whole system - sensor and together hardware DAQ should allow you to compensate for a large part of this error.

    Lynn

  • Beyond the limits of voltage on USB 6008?

    I use a USB-6008 to measure analog differential in the range of 3 to 5.3 V. I chose this range, because outside this range, I'm not interested in what the tension is, knowing that his "on the rail", but to aid resolution of the ADC in this range.

    So for a signal of ~0.05 supply V, I expected to read 3V, telling me that the voltage is lower than 3V, but again it returns 0.05v?

    As it was unexpected for me, could someone please explain what can / should I expect of my USB6008 of the responsed to the signals that are the limits? Is there an effect of "rollover"? Should you return the value to the limit? Depends on how far the limit is?

    Thanks for your help!

    Entry level do not think that way. You specify the range is that you wait for the signal and the DAQmx driver will set the most appropriate range that the device supports. The actual ranges are indicated in the guide. In your case, because you specify a max of 5.3, the device would be defined the +/-10 volts range.

  • No output from USB 6008

    Hello

    I have connected a USB-6008 to my computer, but can not read all the output using a multimeter when I run my program.

    I went in the measurement and automation explorer, disappeared to the USB6008 device, click on test panels and selected the output voltage. From there, I gave different output voltages and I read all this tension using a multimeter to the USB-6008.

    To check if there is something wrong with the pattern-block etc that I install, I connected the 6008 to another computer that is running labview and used the same program. It gives a very good result.

    Does anyone have any suggestions as to what can go wrong on the first computer?

    Much appreciated,

    Paul

    If I were in your shoes I download DAQ - mx 8.6.1 (here) on the PC that you are having problems with and give it a go.  I don't know why the 6008 would work on a PC and not the other when they have the same configuration.  Do you have other devices working on the troubled PC?

  • USB 6008 weird analog voltage reading

    Hello

    I use the USB-6008 to measure a voltage of a Lithium battery, 3.66V.

    the battery come with a blocking diode (in series with the battery) of 1N5820, who have a fall of voltage drop of 0, 1V.

    battery with diodes in series (this is the way in which the battery is shipped with)

    -measure with DDM yield 3.66V without you connect to usb6008

    -measure with DDM yield connected to usb6008 (putting OUT VOLTAGE USB6008) 3.59V

    -able with USB-6008 performance 3.59V connected to usb6008

    battery with diode removed

    -measure with DDM yield 3.66V without you connect to usb6008

    -measure with DDM yield connected to usb6008 (putting OUT VOLTAGE USB6008) 3.66V

    -able with USB-6008 performance 3.66V connected to usb6008

    Thus, it seems that the problem is in the led in the series.  This is why the battery voltage has fallen to 0.07V? the series diode will hurt the USB-6008?

    Maybe people who know the circuits inside the USB-6008 can give me an answer.

    Thanks in advance.

    Hi learnerd,.

    The fall that you see is falling forward in the diode.  As an entry class device, the USB-6008 case has a relatively low input impedance (144kOhm) and thus draws a little current of the device.  Looking at the datasheet of the 1N5820 (http://www.onsemi.com/pub_link/Collateral/1N5820-D.PDF), Figure 7 shows that at 25 ° C, a draw of 50mA will cause a fall front of 200mV.  While the figure does not extend the curve below 50mA, extrapolating the given curve would indicate that a drop of 70mV would cause only a few current microamperes.

    A DMM will have a much greater input impedance (GOhms instead of kohm) and won't draw enough current to influence the measure, that's why you wouldn't see the decline with only the DMM.

    The diode will not harm the USB-6008 somehow.

    Good luck with your application,

    The f

  • USB 6008 angle measurement?

    Hello friends

    I need to measure the angle of rotation of the shaft with a precision of 1 degree... is there any simple device can I connect with usb 6008 custom angle...

    Thank you

    180 degrees is very different from continuous rotation. Just wire a potentiometer to one of the analog inputs and try to calibrate it.

    You don't need the LM317.  It will not help you with what you have described so far.

    Lynn

  • AO. MaxRate, AO. MinRate, AO. Properties of voltage. RNGs for hardware DAQ USB-6008

    Hello

    in one of my report, I use the AO. Property of Voltage.Rngs to see if the selected DAQ card takes in charge the application voltage range. This works very well for my PCMCIA card as well as a PCI card. Now run the same VI with a USB-6008 device, this property gives all the return values. In addition, the report of AO.max.rate and AO.min.rate of the '0', the output is-200197 error properties. I use DAQmx as it is supposed to support the same functions for all DAQmx devices. Can someone please tell me what wrong here and how can I get around this?

    Best regards

    Gabs

    AO.min.rate and AO.max.rate are 0 and error-200197 back because the USB-6008 case supports the outputs analog hardware timed.  The description of error is "device does not support this property."  There is an entrance to the knowledge base for this question.

    By selecting 'Use Waveform' uses the synchronization of the sample clock.  The waveform data type specifies a delta t, which is used to set the sample clock frequency.  It is not supported on the box USB-6008.  You shouldn't set your calendar of sample type or explicitly assign the "On Demand".

    The DAQmx driver supports hundreds of different devices.  Not all combinations of properties are valid for all devices.

  • USB-6008 LABVIEW 8.2. SINGLE CHANNEL WITH DBL INPUT VOLTAGE OUTPUT COMPARISON

    I AM WRITING A PROGRAM THAT USES A SIMPLE USB-6008 ANALOG INPUT CHANNEL. I WANT TO READ CONTINUOUSLY THE VOLTAGE FOR 60 SECONDS. I WANT TO COMPARE A TENSION FOR THE PREVIOUS OF THIS SAME CHANNEL VOLTAGE, MAINLY FOR THE PERIOD OF TIME MAX VOLTAGE GIVEN, THEN GET A FINAL VOLTAGE READING. THE OUTPUT OF THE VI IS A DBL. I WANT ONLY TWO TENSIONS OF EXPORT TO EXCEL. TO SAVE TIME, I KNOW HOW TO EXPORT. CAN SOMEONE HELP ME WITH THIS ONE.

    VI needs an register shift related to the Max & Min function. The current value would be the entrance is and the entrance of x is the left shift register. The max value gets wired for the shift register to the right. Don't forget to initialize it. The output of the shift register is the max you would write and the value of the DAQmx Read out of the loop of wire will give you the last reading.

    Your waiting for 45 seconds makes no sense since you said that you wanted to read continuously. You also said that you wanted to read 60 seconds and all this logic is missing. A simple function of time elapsed, it's all you need.

  • How to measure high voltage (60-70 v) and current (75-80 a) using a DAQ PCI or USB DAQ

    Hello

    I work with a system that works on about 5kW. The output of the system voltage can go maximum up to 60-70 v and thus the corresponding current around 75-80 a. I have 10 these systems that I want to read one by one continuously for long periods.

    I am designing the automated system best suited for this and looking for the best material that would be appropriate for this purpose. Looking for options, I found that an SCC - A10 attenuator may be used to get the tension down by a factor of 10. But I'm confused, if the high current will pose a problem and also how to measure this high current.

    I need to measure the voltage and current at the same time. Please suggest what would be the most appropriate fitting for the same (preferably PCI or USB)

    The hope of a quick response. Thanks and greetings

    Reena Sharma

    Facilitated learning

    Reena says:

    Hi all

    There is good news that the idea of using a compact data acquisition has been accepted by the authorities of the society. I'll be very grateful, if you could suggest me with some hardware modules suitable for my application and how I can use them best.

    Thank you very much

    Reena

    I was able to make a few suggestions, but do not have the time to understand your needs and the forums are not the best solution.

    Your Local OR representitive actully gets paid to do this kind of thing.  a google search suggests THAT LME is in Pasadena. Zack Collins would be the contact rep

  • During playback from the USB-6008 on the OID entry, he reads the top floating. How to configure to read low when it is anything but 5V

    I have a USB-6008, I read an entry on the OID.

    I want this 1 report when the line has a present 5v and 0 for something else.

    When I don't have anything on the lines.  It reads 1

    How configure it to read 0 when nothing is connected?

    Also how I re this in c#

    Thank you

    Hello ashitakaLax,

    The USB-6008 housing has an internal pull-up resistance to 5V (according to page 22 of the User Guide and specifications) that pulls the line to 5V when nothing is connected.

    In order to change this, you can add an external resistance of menu drop down to make the output to the earth when it is disconnected, if your device should get power to fuel the high line.

    Kind regards

  • USB 6008 DAQ scaling

    Hello

    I use a USB-6008 to read signals from analog voltage to a device I built. The task is configured with a single reading finished with a range of + -10 volts. There is no scaling in the task or in the block diagram, but when I acquire data in a PDM file and view it, I get a load of PieceWisePolynomial scaling properties. I have attached the TDMS spectator to this post.

    Does anyone know where it comes from because I can't explain it.

    Thank you

    The data goes back as an unsigned integer between 0 and 32 k of the ADC on the 6008.

    The scale is to convert this unsignedInt to floating point value between-10 and 10.