median filter

Hi, I am a student on filtering

My basic test comes from the txt file with table 1 d, with the help of tool median filter, where the signal is filtered median

But it does not work which means it does not indicate the field. I don't know why even there no error that occurred.

Please let me know the reason.

Try something like that. Filter seems to work pretty well:

(Note that read the spreadsheet file has changed in LabVIEW 2015, so the red X on the icon. Just ignore that. Be sure to select "convert" so that you get the first column instead of the first row)

Tags: NI Software

Similar Questions

  • Median filter isn't removing the spikes at the bottom

    Hi all

    I use a median filter to spikes of deletes for the acquisition of data in real time of thickness of an ultrasonic sensor. This filter works great to remove the upper tips, unfortunately the bottom spikes remains unchanged. I tried changing different grade in the median filter with no improvement.

    Attached images are code LabView and graphs showing data before and after filtering. How can I remove these spikes at bottom also?

    Thanks in advance

    Apple

    I think that your problem is your acquisition code, not the filter. The data set, you have posted has 969 items with values of 2.681 or 2.68. The other 54 elements have values as 2.6, 681, 1, 2, 8 and 68. An element has the value 5, and there the value 9.

    What this suggests to me, it's that your VISA reading is not correctly synchronized with the data transmission device. Either you get partial messages or an occasional character that does not match your regular expression.  Look at the default value stored in the reading string in your VI. Unplug the display of ' \' character mode. It contains '\D5U=\F32.681\r\n2.681\r\n2.681\r\n' it looks like there two characters special \D5 and \F3 and then three values separated by carriage return, line feed. With this kind of protocol you should probably use the jump as a default line end character. Then search for the string \D2...\F2 which can be a header or channel ID. The rest is digital data.

    Check with the creator of the device that sends data to determine the complete communication protocol.

    Lynn

  • application 2D median filter for images

    I want to use the function of median filter 1 d LabVIEW for application table 2d (image).  Vision doesn't have this feature (or I can't find not [using Vision 8.5]).  I guess I can apply the median filter 1 d through the lines, then transpose the table and apply again to the lines (really columns but now the lines because of the conversion).   Just like the use of the 1-d FFT for 2d FFT.  Is this strategy the correct?

    Thank you

    Don

    That's where - IMAQ NthOrder under treatment: Image filters.

    By default, it takes the middle number of a 3 x 3 around each pixel region, but can be set to be the biggest, smallest, or any place in between.  Also the ability to set the size of the filter, although it is a non-linear filter (so why it is not separable as you suggest) the time increases significantly.

  • Cannot use median filter inside the While loop

    Hello

    I try to use the filter median insade loop some time. This seems to be a problem. I have attached the VI. Help, please.

    This entry is a number of sample as 1,1.1,1.05. This output is zero.

    Thank you

    Vincent

    Since you get only one value per iteration, you can use the point-by-point version of the median filter.

  • How to use the median filter?

    I need a median filter that ignores the pixels with color difference above a certain threshold. So, if the pixels are relatively narrow in the value of the color, apply median as usual. If not, ignore.

    Hi bluefalcon,.

    Please check the following link and let me know if you be useful: How to use the median filter in Photoshop CS6 - for Dummies

    Concerning

    Tanuj

  • Apply a median filter in video?

    Hello.

    I just wanted to start to say I'm a complete Newbie using Photoshop for purposes of video editing, so patient with me to work, I work with Photoshop and we recently received a promotional video clip about a new Middle East project we collaborate, we would like to publish it on our intranet home page. However, they have chosen print information in the video, which is in Asia and type of a gene for eyes.

    I learned that I can use the CS6 video editing option and place a median filter on the denunciation without removing completely the bottom. I tried to use the Rectangle tool of selection (with the video clip selected in the layers) to join the information and then apply the median filter. It works fine, but as soon as I start the video, he is gone again... So, how can I make sure that the effect of median filter is applied throughout the whole sequence?

    I'm sorry if this thread is in the wrong forum, I was not really sure... Thank you very much for your help!

    Best regards

    Jesper Pedersen

    Hello, I beg to differ. Convert the video layer to a smart object, select an area, apply your blur. IT can even be animated

    See:

    and:

  • Use a table file bmp for future median filter 2d?

    Hello

    I want to try to read a bmp file, it unflatten so I can use the 2D for a signal processing table.

    The signal processing be especially will be nodes of formula.

    My problem is that I can't just understand how to divide the table 2D of Unflatten Pixmap VI in a component Width and Height.

    The component Width and Height would then serve as input for my formula node.

    Did I miss something or I forgot a simple VI?

    Thanks in advance!

    I think that you just need the function of the size of array with an array of Index.  This will give you the number of rows and columns, which is the same bone width and height in number of pixels.  Or you can simply use the function of the size of matrix (will be slightly faster)

  • Bilateral Image filter

    Hello

    anyone has any idea if a bilateral filter for smoothing the image is available as part of the vision? I have not found anything similar. For the moment, I use a median filter for image smoothing (noise remove texture) but the results produced by ImageJ / Fiji with a bilateral filter are much more impressive.

    Best regards

    Paul

    Hello

    Please see the image (diagram) below:

    Try allocation of memory for the new image. Tell me how it goes.

    Best regards

    K

  • TC data show the HF peak

    I have a 1102 I read in 32 TC data channels with.  Attached is a screenshot of a band chart showing data and as we see there are HF narrow peaks in the data with a dT of 2 degrees.  I say HF only with regard to the acquisition of 1 Hz rate.  Any suggestions on how to calm up to 0.5 degrees of noise dT?

    two options:

    1 - Use a filter.  A median filter is known as the peak 'killer '...

    2 - find the root cause. analyze the configuration & ground (ground loops?), increase the sample rate, seek sources of EMC noise (power devices: pumps, heaters, room ligth;) RF devices:,...), connect a scope, look at correlations.

  • With differential measurement noise problems

    Hey everybody.

    I use an NI USB-6211, LABView 2011 and Win7.

    I'm trying to measure a voltage through a resistor using differential mode. But unfortunately im getting a lot of noise (on + - 25%).

    The voltage source I use is variable and can go up to 600V. With my diet I am essentially heat a metal plate.

    A parallel voltage divider is used to reduce the voltage by one hundredth (1 MOhm and 10 kOhm).

    Two wires attached to the lower resistance then go directly into analog input 0 + 8 of data acquisition.

    So if I'm trained 600V to the plate, data acquisition should get about 6V... and that's what I measure with my voltmeter attached to the acquisition of data input pins.

    I also tried these resistances of POLARIZATION and connected the + and - leads to the analogous to the ground like a resistance I used 10 k, 100 k and 1 MOhm and 2 MOhm and

    continue to receive a bad signal.

    Two sketches of the wire as a signal (100V, supposed to measure 1V, no CORRECTION) are attached to the post.

    Concerning

    EDIT: I forgot to write that I even tried an another NI DAQ and still get this noise problem.

    Also, I measured the voltage source signal using an oscilloscope and I see that noise. But the differential mode isn't supposed to

    reduce noise to a minimum?

    Hey everybody,

    come to understand that the voltage divider resistors are medium to high and they were the main reason for the noise

    problem.

    Before I used 1 MOhm and 10 kOhm, now I use 100 ohm and 100 kOhm and with a median filter in this regard, it works very well!

    But still, if I use the resistance of these BIASES they do not change.

    Concerning

  • How can I remove the noise of a signal?

    Hello world

    I need to extract the phase of a signal of 1.5 Hz which has little noise (see attachment files) and MAX I was recording the signal to 500 Hz. I use the filter of Labview tool value 'smoothing' with a factor equal to 3. Could someone tell me if using the right tool for what I want to do? If so, is-3 a correct number of retativaly? Looking at the result of the signal is much better.

    Thank you

    User

    here still to play

    I did not use the ordinary filter, but also more you know about the original signal, the better your filter or fitting can be.

    If you know it should be a sinus (maybe even with a constant frequency) and you want the phase, go with detection of tone or a linear adjustment...

    the median filter is good against the spikes, the filter of SG will make an adjustment of polynomial (original designed for adjustment of Spectra but it's also nice noise with propper settings)

    Uups, the pic was taken without median filter...

    If you tell us more about the signal, we could offer best filter.  And keep in mind that most of the filters have an influence on the phase of the signal (late phase/group).

  • BMP bmp binary & write on MacOSX

    Hello world

    I have two questions for you!

    First of all, I need to convert a bmp to binary file.

    What I'm supposed to do is: open the thread bmp with a file, unflatten dialog, to use it for other animals (convertir convert in grayscale, using a median filter, etc.), and after changing the it, I would like to save it. So I used the flatten pixmap vi and vi of the bmp file writing. But here's my problem: he always write the new txt file, and not bmp! :/

    Second question: I would like to convert the same bmp in binary file. I know how to get from binary to bmp, but not vice versa.

    Any help ?

    Thanks for all!

    Valentine

    (If you need more information, just tell me!)

    Convert just a file that is located in a specific location means that you can follow the code below

    But I'm not sure the MacOSX.

  • How to get the peak value

    Can I know how to get the 2 peak values v1 and v2 and 2 duration times t1 t2.

    Since there is a small amplitude noise, difficult to use the Max simplely.

    May need to use a filter such as wavelets or TREE to Denoise it first. Can I know how to use, any Toolbox in labview. How do the curve smooth first.

    Thank you.

    I can't watch your vi now (I'm at work), but if your signals are long enough, you might consider a median filter.  You can set the number of points before and after use.  I found that it is useful for some smoothing problems (not all).

  • Lets read file as create table

    Hello!

    I have some data saved in the file .lvm format.  I would now like to organize these data in a sequential table so I can apply a median filter to it.

    When I run the VI I noticed that the first component of the indicator seems to have crashed on each loop (instead of going to the next item), I think that the table should be built outside the loop, but I'm not sure how this code.

    Please find my current VI attached.

    Peter

    use the shift register... See the attached excerpt

  • How get index of the Terminal to reduce

    You need to get the index of the minimize point, marked as cycle of red in the attached word file.

    It's a data from different sample devices, so the curve may change a little each time. But the general pattern is the same, just 2 main peaks.

    I tried to detect the difference is to see when the difference values change from negative to positive.

    But, there is noise pic lowercase when zoom, disclosed being in the word file. Then tried to use the median filter to make it smooth. (Perhaps, best method should be used.)

    But there are still some larger regional summit. Any suggestion?

    Is there a function to labview or best algorithm can do. Thanks a lot for your help.

    How can u say the regional summit broad peak, you want to remove only the minimum index data, or you want to smooth out all the data that I tried with moving to find a way to filter the attached vi.