Pressure transducer

I received an omega pressure sensor (model px309) with only three wires: Red (power supply +); Black (power supply)-; and white for the signal.

Should the output voltages of 0 to 5 and a voltage range from 9 to 30. It measures gage pressure.

I use a power supply that can output 12v to provide power to transduce and a Ni-9205 to collect the signal.

Red and black were connected to the power supply. The white one was hooked up to the ACH0 to collect the signal. In the test diifferential model, I used a wire to coneect to COM. ACH8 Also, I plugged the black wire pressure transduce to COM.

Is this the right settings?

I run my program and I found that the signal is not continuous, and it has a Variant. Please see attached picture.

Is this normal?

Hi Kang,

Depending on what beach you go the DAQmx task for the 9205, looks like expected behavior.

It seems that noise you see is on .1mV, which is quite close the specification of random noise for the NI 9205 module:http://www.ni.com/pdf/manuals/374188d.pdf#page=22

If you select a range of 0-5 v for example, DAQmx will put the module mode +-5V, which corresponds to a 116 uVrms noise, that seems pretty close to what you see.

Tags: NI Hardware

Similar Questions

  • How to wire a pressure transducer loop 2 wires for the NI 9949 RJ - 50

    I am new to the NC and data record.

    I want to connect a 4/20mA, pressure sensor loop 2 son of a Terminal screw OR 9949 RJ - 50... that will then connect to an entrance NI 9237 module.

    The NI 9237 module is installed in a chassis OR compact 9174 DAQ.

    I shall be measure the pressures of air brakes truck to compressed air system of zero kPa (gauge pressure) up to allowance of 1000 kPa (gauge pressure).

    The 2 wires on the transducer are identified as: food + and + signal

    I would appreciate it please advice on the appropriate terminals on the NI 9949 RJ - 50

    Unfortunately, it seems that you have the wrong module for this task.  The 9237 is used to measure the production of bridge-based sensors (for example, a strain gauge).  Signals that it is supposed to measure are fundamentally different from the 4-20mA signal pressure transducer outings.

    Take a look at the NI 9203, 9207 or 9208.  All have at least a few channels designed to take a current 4-20mA input.  Then, give the team nor a (877-387-0015) call to talk through your application and make sure that you select the right equipment for what you want to do.

  • How the pressure transducer Validyne system-> Validyne CD280-Dual-> SCB - 68-> PCI-6024E-> LABVIEW works and how we can calibrate the pressure sensor?

    Hello

    I am completely new to LABVIEW software.

    I learn a LABVIEW code existing my pressure (attached) acquisition system which has the path of data as follows: pressure transducer Validyne--> Validyne CD280 - Dual--> SCB - 68--> PCI - 6024E--> LABVIEW and I have a few question:

    1. How does the complete system of the transducer to the LABVIEW work? That is, if we apply the pressure of the transducer, it will change the resistance of the probe then...?

    2. How can I find the equation that expresses the relationship between the pressure and the tension to the keys of the Validyne double-CD280 in the LABVIEW?

    3. If I want to do the probe calibration, what are steps?

    Thank you

    You did not include most of the subVIs.

  • simple pressure transducer

    Hello

    Does anyone have a pressure transducer with offset null simple vi constructed in code. I'm relatively new to labview and most of the examples I have seen are way to complex for me to understand what is happening.  Is it better to use the DAQmx Wizard or data acquisition?  I have two sensor of pressure (excitation omegadyne-10V) hung on to with a module for 1121 SCXI-1327. I wrote a program to convert mV to pressure using a linear scale, but that's all I know how to do at this point! And I want to know how to read the two transducers.  Thank you!

    Don't forget that you will need to change your excitement of inside out and remove the part with the shunt calibration.

  • 6008 NI DAQ issues of reading pressure transducer

    Hello everyone, I am a new user of labeled and not one expert on electronics, so please bear with me.

    I'm trying to read a signal 4-20 my by a pressure transducer using a NI DAQ 6008 and Labview version 8.6. I am running Windows XP Professional. I'm the transducer, (http://www.omega.com/Pressure/pdf/PX01-I.pdf) providing a voltage of 10 v I have a 235 external resistance ohms between the terminals positive and negative data acquisition (although I am not sure that it is correct). At the start of my VI, I am able to measure a voltage of environ.9 data acquisition (verified by a voltmeter/ammeter) V and a current of environ.39 my.

    My problem is the fact that when I have a supply pressure of my sensor, I have not received any signal whatsoever by the transducer, despite the fact that it is important that the pressure should give me a signal. Is this a problem with my installation, connections, data acquisition (which has been reset in MAX number of times and pass the self-test) or the sensor itself?

    Thank you for your time and consideration

    Hi prenerk,

    You are right that resistance must be connected to the terminals positive and negative a way to HAVE to create a measured voltage to current.  Here are two troubleshooting steps you can try:

    1. check with an ammeter so you are detecting a change in the current when you apply pressure to the transducer.

    2. open the Explorer Measurement & Automation and use a Test Panel to measure the voltage of the device.  This will ensure that there is a mistake in your LabVIEW code.  Also check the MAX pinout diagram to see if your sons are correctly plugged into the channel of GOT it.

    If you have not discovered a problem with the above steps, we will have to learn more about your sensor pressure and how you connect it to the 6008.  We will need the information on the sons of the Omega instrument and how they are connected to the excitation voltage source and 6008.  Let us know how it goes!

    Brian

  • How to get mydaq recognize px309 pressure transducer % 3F

    It has only 3 wires instead of the usual 4. They are black, red and white and negative excitement, positive excitement and signal respectively. I plugged the black wires and red to an external power supply, the white wire to the myDAQ HAVE. Then I went to labview, opened DQA assisntant - acquisition - tension - ai1. Then I put samples of continuous get ot and drew a graphical indicator and press the run button. Yet, the only signal I get is noise and graph (amplitude time vs) shows no change when pressure is applied.

    Have you tried also runs the black wire from the transducer into one of the analog grounds on your MyDAQ? Measures data acquisition potential, without reference to ground the unit will not be able to measure anything.

    You can also use an output on the analog MyDAQ to produce the voltage instead of using an external power supply.

    I hope this helps.

    -Nick-

  • How to wire NI 9203 to Emerson PT5 pressure transducer

    Hi all

    I started studying with 9188 OR cDAQ and analog input NI 9203 module. I'm really new on the DAQ hardware and I confused about my wiring for the NI 9203 pressure sensors. I want to use 6 pressure transducers whose output signals are ampere (4-20 my). In the manual operation of NI9203 shows that there is only 1 negative port (COM). Could I wire 1 more NI 9203 pressure sensors? If it is possible, how would I do that? The pressure transducers that I wanted to use a 2-wire and its technical specifications is Emerson PT5.

    Hello

    You should be able to branch out of the output of the power supply 24 VDC 8 drivers. Does that help?

    -Carisa

  • Pressure transducer USB-6211 noise error

    Hi all

    I tried to connect a pressure sensor through my data acquisition by the voltage reading and when you use a multimeter to check the voltage measured by the son, I get exactly what I want, but when I look at reading through labview, I make so much noise that I can not even acquire a strong enough reading for the tests. When to play with filters and blocks measure, I'm sure that some of the ways to reduce the noise and get the reading of the amplitude of crete would, just do not know where to go next, as I have tried so many options like this already.

    Problem has been resolved. Simply a problem of ground that I threw.

  • To measure the pressure using a pressure transducer that provides the analog current output 4mA-20mA

    I wanted to acquire the current analog signal which varies from 4-20mA using NI 9207. I tried in 2 ways.

    method 1 - created an input channel current analog & used a reading Vi to acquire it. How can I give the channel connections in this...

    method 2 - using NOR-DAQ Assistant, I put the channel connections and I got Amplitude versus time graph. She also gave negative values. Can I do this way which is easier? How can I solve the problem

    First attachment belongs to the 1st method

    Second attachment belongs to the 2nd method


  • Problem with the sensor of pressure and NI 9205 with NI 9263

    I have a bridge pressure sensor based with a range of 0 - 5000 psi and requires a 10 V excitation. The output is 3 mV/V. I plugged on a cDAQ with a NI 9263 providing excitement V 10 and a NI 9205 to read the output values. I would expect a 0-30 mV output corresponding to 0 - 5000 psi, correct? (Incidentally, the NI 9205 module must be accurate to 0,1 mV or this should lead to a precision of about 16 psi, right?).

    However, when I tested it in fact the sensor on the run a test ranging from 0 to 1000 and back to 0 lb/po2 (now for quite awhile to States of pressure) the values I get from the NI 9205 are about 204.614 mV for 0 lb/po2 and then goes 204.867 mV to 1000 lb/po2 (actual results are noisy (, so these are averages). I enclose the SignalExpress file used for the test as well as the Excel file containing the resutls. Testing was performed with a number of other active sensors (load and displacement, LVDT and gagues strain gauges) in addition to the pressure that I'll eventually use them for my experience. In the SignalExpress file, the pressure transducer is the only sensor receives the excitement of the NI 9263 (ao_0) and ai_3 on the NI 9205 module.

    So this time I tried to run the test again, but changed the SignalExpress file (second .seproj attached) to include only the pressure transducer. But now I don't have an answer at all by the transducer (it remained simply flat).

    As to the connection of the transducer material, it is a connection 4-wire (excitement, - Exciation + Signal - Signal). I plugged the cables from signal to pins 4 and 23 (AI3 and AI11) for a differential pair. Then, I plugged the son of excitement to the pins 0 and 1 (AO0 and COM). Finally, I took a wire jumper and connected 23 on the NI 9205 module PIN to pin 1 on the NI 9263. I made this last step at the suggestion of support OR for a previous problem with a string potentiometer. This stage connected the ground between the two modules (the sensor did not answer with a pressure change when I don't have the wire).

    I appreciate any help as to why. I did not receive a 0-30 exit mV.

    his does not work like that.

    The probe should @30mA to 10VDC.  There is only available from the canal AO 9263 1mA.  you need another power supply.

  • DAQ USB-6008 will be able to power and record voltage for UMS T5 blood pressure at the same time?

    I would use my NI USB-6008 to power my blood pressure monitor UMS T5 (http://www.ums-muc.de/en/products/tensiometer/t5.html) but also to take readings of it, but I don't know if it's possible to do it properly. The power supply for the instrument can be as low as 5V, I can easily get the dedicated + 5V channel. I'm able to feed the instrument and connect it to an analog input on the 6008 and measure a voltage in differential mode. However, when you read the documentation of support for the instrument, I find the following:

    "Potential pitfalls of data acquisition: the pressure transducer is configured in a full Wheatstone bridge, the input voltage and mV signal output can be connected to the same reference (mass)." Therefore, the mV output signal can be measured using a differential voltage measurement. Therefore, do not make an asymmetric measure of pressure transducer mV output. "(http://www.decagon.com/assets/Uploads/MeasuringUMSTensiometerswithnon-UMSControlandDataAcquisitionSystems.pdf)

    My understanding is that the 6008 can take a differential measure if I attach the signal '+' and the signal "-" to the analog inputs of positive and negative terminals. However, it seems that all the ports of ground on the 6008 are grounded to the same reference, which would make my measure of invalid tension according to the above paragraph. So my real question is: if I try to record the voltage with one of the analog inputs on the 6008 in this way, is the valid measurement? Or I need to find a separate power supply, with a different reference field to ensure that the measure is accurate?

    The technical details of this device is very poor. The manual is not much better. Companies that want to sell scientific equipment should publish decent cards or get out of business.

    In section 3.4.3 General requirements the device is described as a "bridge not amplified circuit.  This information along with the impedance of the bridge should be in the specifications, because it is essential to apply the device under any circumstances other than the nominal behavior in 10.6 V.

    The answer to your question is:

    You can use it with the box USB-6008.  The 5 V supply will result in output voltages a little less than half (5/10.6) the voltage specified in nominal conditions. You can use the differential input mode on the box USB-6008.  The absolute input voltages will be approximately 2.5 V with the 5 V power supply. This voltage is in the range of the aircraft. The differences are likely to be less than 100 mV. The resolution of the USB-6008 on the + /-1 V is located about 0.5 mV so your resolution of pressure will be about 1% of full scale. The voltage input impedance and termination of the USB-6008 will present a few errors.  These can be in the order of 5 to 10%. I can't predict much better without the missing bridge impedance specification.  These errors should be relatively constant and systematic. A calibration of the whole system - sensor and together hardware DAQ should allow you to compensate for a large part of this error.

    Lynn

  • I can't Omega PX4200 pressure transducers where I can get a reading of Labview

    Hello

    I am a new user of LabVIEW, and I currently try to wire in a 4-20mA Omega PX4200 pressure sensor in an NI USB-6009 Data Acquisition device. I have sources of power like a P SOLA SDN 5-24-100, and I use a 3-wire system. I have the black wire going to SOLA a +, the ground wire to the GND on the NI USB-6009 case and the white wire going to the AI0. Currently, I am unable to generate a significant signal either in solution of measurement and Automation Explorer or LABVIEW Signal Express 2010.

    If anyone has an idea on what I'm doing wrong, just reply to this topic.

    Thank you

    Dev

    The USB-6009 measure the voltage.  The pressure transducer is a current signal.  You must add a kind of conditioning of signals.  At a minimum, you must use a resistor to the terminals so that you can convert the signal 4-20 mA, the transducer is a tension.  To convert a signal of 20 MA in 10V, you would need a 500 Ohm resistor.

  • Certificate of scaling of pressure sensor

    Hello

    My pressure sensor has the following calibration certificate:

    (lb/PO2) pressure unit of data (mVdc)

    ---------------------                   ---------------------------

    0                             -0.275

    100 49.824

    200 99.774

    100 49.844

    0                              -0.264

    Balance - 0.275s mVdc

    Sensitivity: 100.049 mVdc

    Calibration factors:

    Sensitivity = 20.01 mV/V

    the = 5 VDC voltage will give a result of 0 to 100 mV

    I'll use DAQ support to take the measure of pressure.

    the question is: what should be the two electric and two physical quantities (IN mV/V units) that I need to enter using DAQ Assistant when you descale the pressure transducer by the method of the colon. Once again, please specify the values in mV/V units.

    Thank you

    wisjaf12,

    You can choose one of the two points at your convenience.

    For example: First Value: 49.824 second value: 99.774. (Electricity). First value: 100 second value: 200. (Physics)

    CarmenC.

  • Read only the SPI using the box USB-8451

    I am using the NI USB-8451 box to read the SPI of a Honeywell digital pressure transducer data.  The difficulty comes from that part of Honeywell uses only the SPI in half-duplex mode, meaning that it only transmits data, but does not require anything beyond the appropriate chip select signal and SCLK to start transmitting its 32-bit sensor data.  To facilitate experimentation with the sensor, I bought a USB-8451 SPI Interface thinking I could easily configure the 8451 to read data from the sensor.

    However, if I understand the situation, there is a problem.  The 8451 considers full duplex data, i.e. a data word must first be sent to the SPI device in question before the unit will start to transmit back data of its own.  And since the MOSI data writing periods, since this particular device starts transmitting immediately once a clock is applied, synchronizing the unused address data will cause the unit will return its data before the 8451 begins listening for data in return.  The call used by the 8451 to write read action can be divided into the individual reading and actions of writing of any kind that I could discover.

    Can someone tell me if I missed something in this operation or if there is some quick tips, that I could use to be able to use the 8451 therefor?  Thank you!

    Doug G.

    Exactly correct. It's nice when things are easier than expected. Hope it works as easy as it sounds.

  • How can I use other companion DAQmax and data acquisition in labview code?

    Hello

    I'm new in labview and want to combine four different codes of labview and run them at the same time.

    I use my data acquisition PCI-6071E and BNC-2120. I want to send an analog output (flat DC signal) to control a blower, an analog volmetre to a (DC) output pressure transducer, also receive data from a pressure sensor and hot wire probe.

    I wrote four different VI for each of them (using DAQmax to output analog and DAQ assistant for analog inputs) and each of them works well, but when I want to use them together they do not work.

    Anyone know how to make them work together?

    Thank you

    Pooya

    To the DAQ Assistant, simply follow the instructions for the selection of multiple channels when you create (where he says quite clearly,' ' or to select multiple channels "). If already created, open the DAQ Assistant and click on "Add channels" at the top of the page.

    When you use a physical channel name, you simply click on browse instead of a single channel. Use the same or for multiple channels.