SCXI-1600 vs term of mass OR-6251 (SCXI-1000 chassis)

I just need a mental health check using an SCXI-1600 instead of a NOR-6251 module.

I have a SCXI-1000 chassis with 3 cards: (w/1300) module SCXI-1100, SCXI-1104 (w/1300) and SCXI-1126 (w/1327)

I'm sampling 62 channels of these cards to 1200 samples per second, using LabVIEW 2011 on Win 7 (32 bit).

We previously had a PCI-6040E to connect to the chassis, but we must move on to a USB interface.  I tried the 6251, but we have serious space and mounting issues since that a portable Setup.  I would use one SCXI-1600 module instead, but I'm afraid of all the severe warnings recommending a 6251 instead.  Are there problems I'll meet?  This seems to be such a basic configuration.

The 1600 allows me to save space the 6251, its power supply and cable 68 pin linking the two.

Hi CoryM,

Since you are using Win 7 (32 bit), I plan no compatibility problems. We strongly encourage our customers to use the 6251 because it is a new product with more features, but if you need the SCXI-1600, that should be good as well. I hope this helps!

Carisa - L.

Tags: NI Hardware

Similar Questions

  • SCXI-1102/SCXI-1303 undetected in SCXI-1000 chassis

    I use an installer with the SCXI-1303 module in an SCXI-1102 module in a chassis SCXI-1000 measures voltage and thermocouple.

    Until recently, everything worked well. but now it seems that I am unable to communicate with the interface. I've tried two different computers and hardware is not detected in MAX. The power light and fan on the SCXI-1000 chassis light again, it just seems to be no acknowledgement of the connection to the computer.

    A few more specific error codes:

    * When I double-click on the DAQ Assistant, in the block diagram window and click on 'Test' one of my entries, I received the following error message:
    "88705 error: the specified device is not present or is not active in the system."

    * I also got the error code "error-89130: not available in NOR-DAQmx device" in Control Panel test DAQ Assistant

    * During the closure of the DAQ Assistant test Panel, I see the error "error-89130: device not available for routing.

    * To the MAX, if I try to 'test' the chassis or the module, I've seen errors either saying that the device is not detected, or that the device must be reset or is being restored, or the traditional DAQ drivers must be reset (I tried the reset suggestions without result)

    At this point, I wonder if the Commission or the cable is pulled. Unfortunately, I don't have an other 68-pin cables to test at the moment.

    Thanks in advance...

    Thanks guys, I was able to solve the problem. To all those who might find this,

    * I noticed in the Windows Device Manager that two PCI devices have been listed as being not installed correctly.

    * After trying to update the drivers, I couldn't start the computer; got BSOD after the BIOS loading.

    * Some research led me to go out two PCI cards that I was not using, and I moved also to reinstall my data acquisition on an another slot card to force material.

    * Starts up, load drivers automatically and I'm gone.

    Doesn't explain not why it did not work in the other computer, but in the words of Spinal Tap, the authorities have said it was better left unresolved at this point.

  • Error 200320: SCXI-1001 Chassis - Mod 1502

    Hello.

    I'll put a SCXI-1001 chassis with several modules, 3 of them 1502 with their 1314 terminal blocks for the first time. I used this chassis before reading temperatures and worked fine, so I'm sure that the problem is me.

    I get this error 200320 on simultaneous sample and hold trying to acquire data of deformation, but I think that the configuration is fine. I'm still fairly new to LabView and have tried to check the installation by the Express data acquisition module and MAX; but I get the same result.

    Please help me to implement as soon as possible, I need your knowledge.

    Thank you

    usuario

    Hi user,

    You can learn the development of LabVIEW applications and data acquisition as shown in the following links:

    http://digital.NI.com/public.nsf/allkb/7D5F60ADBFD9CDC2862572BA005DD96A?OpenDocument

    http://zone.NI.com/DevZone/CDA/tut/p/ID/2835

    The latest links explains what a global virtual channel and a task.

    Regarding the shunt calibration, you can consult the following links:

    http://digital.NI.com/public.nsf/allkb/8D81F07FBCCB6DA186256BE6000691B3?OpenDocument

    http://digital.NI.com/public.nsf/allkb/286AC9A8DC0B689986256C2C0064B183?OpenDocument

    http://zone.NI.com/DevZone/CDA/tut/p/ID/3642

    You can open a new thread and forum to do not mix topics, that's my advice.

    Kind regards

    Jesus.

  • To put the module chassis SCXI 1000 with sinusoidal input name

    It is imperative to turn the module chassis SCXI 1000 with a pure sinusoidal nickname of entry sector rather than sinusoidal waveform?

    This will affect the performance and reliability of the DAQ module to provide power with a sinusoidal waveform of nickname, he entered input sector?

    Concerning

    Jamal

    Hi Jamal,.

    If you look at page A-2 of the 1000 users manual found here you can see that the power needed for the 1000 is not too strict. So I guess that if your hands are almost sinusoidal at all about these tensions ca then you won't have a problem.

  • SCXI-1600 is not communicating with MAX

    All-

    Maybe that someone else has come across this.  I have connected an SCXI-1600 module in an SCXI-1000 chassis to a computer with a USB cable.  I can not get NOR-MAX to recognize the device.  I ' v tried to restart and reconnect in different orders.  I traded with a well-known USB cable.  I even took it on another computer.  On the new computer, when I plugged it first of all, I have a message "new hardware found" Windows (XP) that the-1600 had been identified, but I still couldn't it appears in the tree, even if I tried to add devices manually.  I removed the other modules in the chassis and reconnect the - 1600, still no luck.  Any advice would be great!  Thank you.

    Reinstall the suite OR-DAQ solved the problem.  Didn't try it before posting because I had the same problem on several computers and so I thought of something else.  Please do not take into account post.  Thank you.

  • Error 1600 SCXI 200329

    Hello world

    I used my scxi-1000 chassis with the scxi-1600 daq card. Today, I get this error when I try to get data from several modules. Error number is 200329 and I have attached a screenshot for this error. Do you have an explanation for it?

    Thank you

    Bertrand

    Hey Berkayaga,

    It is possible that a single module, once damaged, which can affect three chassis.  There are a few steps that you can do to try to fix your equipment.  You might start by checking the fuses on the chassis.  Information on this may be found at Troubleshooting an SCXI chassis.  As well as specific information on the fuses themselves, there is also a documentation on where you can find them.

    If this does not work, you can consider the process of authorization to return goods (ARM) located in a National Commission of Instruments sent in for repair or calibration.

  • SCXI and acquisition with Windows 7 x 64

    Our just updated IT to update our phones and we have received Windows 7 x 64.

    Data acquisition, we use the SCXI hardware in the SCXI-1000DC chassis and an acquisition of data SCXI-1600 USB card.  Doing my gear OR set up on the new laptop, I discovered that neither does not support Windows 7 x 64 for the SCXI-1600. The other options to connect the SCXI-1000 chassis to a laptop are the DAQcard-6062E and DAQcard-6036E, nor which supports Windows 7 x 64.  According to the information below, there are no plans to support x 64.

    What are my options to continue the use of the material, we have invested with the laptop running Windows 7 x 64?

    You will need the NI SCXI-1349 connection block and SH68-68-EP cable.  The SCXI-1349 plug at the rear of your SCXI module and cable connects the 1349 the USB device series M mass termination.  The rest of the configuration is identical to any PCI/PXI E Series device.  The USB M Series device will be the analog-to-digital converter for the system instead of the SCXI-1600.

  • SCXI and USB 6210? Getting started...

    Hello

    I'm looking to get an online data collection system and had a few quick questions of SCXI signal conditioning high-performance...  First of all I have is 1 chassis SCXI-1000, SCXI-1121, SCXI 1180 and 1100 SCXI modules assorted.  I also have a few blocks of different endings, SCXI-1300 block and a 1320.

    (1) by section "what to start' in"the SCXI chassis user manual"everything only I need is hardware of SCXI chassis (in my case the SCXI-1000 chassis), a 120 VAC power chord and a computer.  However in the ' what you need to get started section "of the manual"Getting started with the SCXI"I need also a SCXI-1200 or"Hardware DAQ"module  So my question is this, can I set the SCXI-1000 chassis directly to my laptop without any additional hardware?  I feel, I need something else, but want to make sure...

    1 (b) assuming I need a material additional DAQ to start, I can use a NI USB 6210 somehow?  I checked the manuals and little is said about SCXI.

    (2) neither manual mentions a controller, but the SCXI Advisor shoot one.  They should say if all I do is collect data, not active control?

    Thank you

    Mike

    Mike,

    There are two options for you to get your SCXI system operational. You will need either an M series or a card of E-Series you connect to your SCXI system or you could focus on the NI SCXI-1600, which is a controller for your SCXI chassis and allows you to connect your laptop to the controller via a USB cable. This controller will operate similar to the E-Series and M-series cards.

    We also offer a couple of PCMCIA cards which allows you to integrate the SCXI system with your laptop. It comes from the NI DAQCard-6036E, and NI DAQCard-6062E. Unfortunately, your NI USB-6210 is unable to be used as a device for the Acquisition of data for your SCXI chassis. The list DAQ cards which are possible to use with your system can be found here on the tab 'controller '.

    http://ohm.NI.com/advisors/SCXI/pages/SCXI/modules.XHTML?conversationContext=1

    In addition, we have an old manual SCXI of starting you can find here:

    http://digital.NI.com/manuals.nsf/WebSearch/C195895E8C9CAE3D8625691F004DB285

  • pilots and SCXI?

    Hello

    I have a SCXI-1000 chassis, with a 1100 unit, a unit of 1520 scxi and a 1600 scxi. Now when I connect, I get only a "USB not recognize unit."

    I installed 9.1 OR DAQmx and LabVIEW 9.01 32bits, need more? I have never tried to use before SCXI.

    I got Windows7 64 bit.

    Thanks in advance.

    For these devices, it is not currently planned to add support for 64-bit operating systems.  This is due to incompatibilities between these older devices USB design and implementation of USB in 64-bit version of Windows.

    Kind regards

  • Calibration of the scxi-1520 and scxi-1314

    I have a SCXI-1000 chassis with the scxi-1520 module and the scxi-1314 terminal block and I was wondering how to do a calibration on it?  What I have to send back them to OR for calibration or I can send it to a local company to perform the calibration?  If I can send it to a local provider, what the procedure would be to calibrate?  According to the manual of the module scxi-1314, everything we do is check the resistance and for the scxi-1520, it performs internal calibration.  Is this correct?

    Ron

    Hi Ron,

    You can send them to OR get these calibrated, or you may have a local company. Manual calibration procedures are available herefor many of our products online. Here are the procedures of calibration of 1314and 1520.

    Other issues on the calibration should be directed to ni.com/calibration.

    Chris W

  • Information SCXI

    This is the first time I work with platform SCXI. We have a SCXI-1000 chassis, a 64 mio pc HAVE, a 32 entry SCXI 1102 and a block of endings 1303 for reading signals from thermocouples 32. Now we have the opportunity to purchase a module SCXI-1100 and 1300, one block of endings. Could they to Thermocuple read? What is the difference with 1303 1102 and 1100, and 1300 respectively?

    Francesco

    Thank you very much for your answer. I should say: "that's all".

    Thank you once again

    Francesco

  • The measurement and Automation Explorer is completely empty. No idea why?

    Hello

    When I open Measurement & Automation Explorer, there is nothing in the main boxes, boxes are completely empty. The top toolbar with file/Edit/tools is there, but all that is empty.

    I had good success with the SCXI-1000 chassis in a few years when I did further research at another University. At my new University, it gave me access to an SCXI-1000 chassis with an SCXI-1600 module to connect to the computer via USB. I reinstalled windows on an old computer to 64-bit, had the hand of the entire system, the Measurment and Autiomation Explorer open as expected, then was that NEITHER ceased support for the SCXI-1600 on a Win 7 Pro 64-bit modules. So, I've divided the hard drive into two partitions, installed Win 7 Pro 32 bit and reinstalled NOR then installed the updates for a few hours.

    When I try to turn on measurement and Automation Explorer, there is nothing in the window. What can I do to get the various boxes? I went to tools > reset Configuration data.

    Nicholas,

    I recently saw a similar problem on another post. This person has received the exact same blank screen because Windows Firewall has blocked MAX (nimax.exe). Try turning off your firewall temporarily and then see if MAX runs. If so, you will need to add an exception in your firewall settings. If this isn't the case, you could have operated in a corrupted installation and the need to do a repair.

    Ryan

  • application is not running on the target system

    Hello, this is my first attempt using this forum & creating applications LV and I met a brick wall.

    I have created a simple VI to display some readings of T/C, which works well on the development system (LV teacher 8.5), but not on another PC target.

    The target computer is not installed, just software LV DAQ - MX for Windows V8.6

    The DAQ hardware is a SCXI-1000 chassis with SCXI-1600 USB interface, taking data with SCXI 1102.  I use the same material on both

    PC development and the target computer.  Data acquisition equipment works correctly on computer target (using MAX).  My application of load, but simply

    work on the target computer.

    Hi Bob,

    1. Ensure that you have included drivers OR DAQmx and measurement and Automation Explorer (MAX) with your Installer.
    2. Use target MAX on the computer to check if you are able to see the device.
    3. Ensure that the name of the card MAX device corresponds to its device name in the program i.e. Dev1 MAX = Dev1 in the program
    4. Also, check the names of device between the development system and the target system. IE Dev1 on dev = Dev1 on the target system.

    I hope this helps!

  • use the cyclic data record

    Hello

    Before you begin, let me briefly introduce what I the material and what I want to achieve.

    I got the chassis SCXI-1000 (chassis), SCXI - 1102C (module), SCXI-1303 (terminal block,

    mainly for the entry of temperature sensor); A few months ago I bought crossed SCXI-1180 and SCXI-1302 terminal to my digital input signal (see attached PNG file). In fact, I have two digital signals at the entrance. The first digital signal (enter as PFI0) Injection that will trigger the start of the recording is my

    Temperature profile; The second is entry signal (such as PFI1), that I want to use as a trigger of reference to stop recording and save the wrapping cycle.

    And then the injection cycle keeps going. Please see the attached VI for more details.

    I get my digital triggering working to start recording my profile of temperature sensor. However, I can not get the second part trigger to stop recording.

    The error of the execution of my VI, is a trigger of reference applies to finishes of sampling. It seems I need continuously to record my temperature profile sampling! Thus, by browsing my VI, please give some suggestions for the task and solutions to solve the conflictions for sampling.

    Your help is highly appreciated.

    HW


  • How to use HX85BA sensors with NI DAQ system?

    Hello

    I'm trying to connect two sensors of Omega HX85BA moisture to my NI DAQ system. The sensors detect moisture, barometric pressure, and temperature, using three 0 to 10 VDC positive outputs @ 10 mA or less, with a common thread of output. I have a SCXI-1000 chassis with two modules SCXI-1100 cards, 1102 b and a single 1102. Map of 1102 uses a block to connect SCXI-1303, while the other three cards use some 1300 years. The system will mainly use thermocouples for temperature measurements, but there will be other types of sensors, such as the two HX85BAs.

    The manual of the product for the HX85BAs is not very clear on how to connect the sensor to the system. Terminal blocks each have 32 channels with the positive and the negative screw terminals, it doesn't connect three positive sensor outputs to individual channels, with the municipality of output connected to each channel? It will work installation with SCXI cards I have, or I will have to purchase different cards? The system will be expanded to include at least 4 cards in a separate chassis SCXI 1000 more, is not a problem for the purchase of different components to adapt to the HX85BAs.

    Thanks in advance!

    Hi BBalmforth,

    You can connect the sensor to the SCXI-1100 1300 TB. Connect the positive output of CH + and CH - CHSGND. Common output wire must be connected to GND.