the variable entry in the time

Hey,.

I'm a student in a lab, and we have a DC power supply "Agilent E3646A". I have the driver for this installation. The front panel, I can change the voltage, but what I need is to simulate the power of a photovoltaic panel of Satellite.

So, I need to have a table with a different value input voltage changing every minute.

IM new in labview and know him well.

I have attached the VI of our supply.

Thank you in advance. !

Hello sarri,.

The code I give you is a little advanced, but hope that you will understand.

It has two parts.

First to use another while loop as shown in the figure.

The loop has two shift registers

(1) time loop

(2) the index of the table of values of voltage.

Verifying part timer elapsed 60 seconds. 60 seconds of expiry, increase the array index to select the new value of the voltage. Also replace the timer to the new value shift register.

Tags: NI Software

Similar Questions

  • OTL entry sets the time - fast formula not triggered

    Requirement: Throw an error during the registration of the time, if time is < 40 and overtime are > 0

    Approach:

    (1) created PLSQL function and defined in HRMS

    (2) created a quick formula, using the above function

    (3) created a new TER

    (4) fixed to the TERG existing (operation TERG prod)

    Details:

    My function (only for a CEP). Returns 1 always (error throw)

    SET DEFINE OFF

    CREATE OR REPLACE

    APPS PACKAGE. XXXXX_HXC_CUST_RULES as

    -Description:

    -History

    FUNCTION XXXXX_CHK_OT_ELIGIBLE return number (p_resource_id in NUMBER);

    END XXXXX_HXC_CUST_RULES;

    /

    CREATE OR REPLACE

    PACKAGE BODY XXxxx_HXC_CUST_RULES AS

    FUNCTION XXxxx_CHK_OT_ELIGIBLE (p_resource_id NUMBER)

    RETURN number

    IS

    l_return_rule_status number: = 0;

    l_timecard_info hxc_self_service_time_deposit.timecard_info;

    l_tbb_array HXC_BLOCK_TABLE_TYPE;

    BEGIN

    l_timecard_info: = hxc_self_service_time_deposit.get_building_blocks;

    l_tbb_array: = hxc_deposit_wrapper_utilities.blocks_to_array (p_blocks = > l_timecard_info);

    RETURN 1;

    -0 is not an error

    -1 is error

    EXCEPTION

    WHILE OTHERS THEN

    null;

    END;

    END XXxx_HXC_CUST_RULES;

    /

    DEFINITION FORMULA FAST

    /* *****************************************************************

    Name of the form: XXXX_CHK_OT_ELIGIBLE

    Description: Prevent OT when not eligible (40 hrs of ST)

    Contexts: no

    *****************************************************************

    */

    /*

    Initialize Variables that may be null

    *****************************************************************

    */

    the default for db_pre_period_start is ' '

    the default for db_pre_period_end is ' '

    the default for db_post_period_start is ' '

    the default for db_post_period_end is ' '

    the default for db_ref_period_start is ' '

    the default for db_ref_period_end is ' '

    default value of timecard_hrs is 0

    /* ******************************************************************

    READ IN THE INPUT VARIABLES

    ****************************************************************** */

    The ENTRIES ARE id_ressource (number)

    submission_date (text)

    db_pre_period_start (text)

    db_pre_period_end (text)

    db_post_period_start (text)

    db_post_period_end (text)

    db_ref_period_start (text)

    db_ref_period_end (text)

    timecard_hrs (number)

    IF (XXxx_HXC_CHK_OT_ELIG_FN (resource_id) = 1)

    THEN

    (rule_status = 'E'

    Message1 = "XXxxx_CHK_OT_ELIGIBLE."

    return rule_status, message1)

    ON THE OTHER

    (s = rule_status"

    return rule_status)

    The above formula is verified successfully. I can see the FFP plsql pkg also created in the DB.

    The RULES of SEIZURE of TIME are created to record and SEND/QUOTE

    Then, they are attached to an existing rule GROUP on the TIME ENTRY. Yes, at least it is not the question, because it's a group of working time entry rules.

    MY QUESTIONS ARE BELOW

    (1) the function accepts id_ressource as an entry, should I set this as a FDF segment or is not not necessary?

    (2) even if I attach a formula seeded to my TER, it is not triggered. I suppose, then, that there is nothing wrong with the quick formula atleast.

    Pointers would be a great help.

    Thank you

    Germain

    I found the reason why it was not a device.

    The quick formula was an effective start date of November 7, 2013, the TER also had effective start date of November 7, 2013.

    The score card I used was in October.

    Silly, but he did.

    Thanks for your help.

  • Calendar entry for flights crossing the time zones

    I fight to find the right method to enter a calendar event that starts in a time zone and ends in another.  It is often the case for the flights of the airline.  Example: My departure is from New York (JFK) at 12:00 noon Monday and arriving in Seoul, Korea is at 16:24 (local) Tuesday.  I cross the road to the International Dateline. With Support for time zones "lit", how to capture these moments so that they are accurate, once arrived in Seoul?

    Even more difficult: due to cross the International Date line on my return, I arrived in New York earlier that my time of departure the same day.

    Any suggestions on how to deal with this would be appreciated!

    Paul

    If you enable support for time zones, each entry includes the time zone setting. Select the correct time zone for each. Move real time you you include an entry in the "travel time" drop-down menu. Otherwise enter time manually and identify the zone in the title.

  • BPEL variable, take the time to 1 min just for the assignment of value in the Assign activity

    Hi all

    We get a strange question in BPEL production environment.
    In the assign activity, there are 3 variable hardcoded with some default values.
    At runtime, a variable is to take more time from 1 minute to initialize just this default value.
    We do not get this problem of time in Dev and Test environments, so we are not able to test this.
    Please look in the Audit trail for the BPEL instance in the Production environment.
    Request to let me know if them work around or fix for this issue.



    Thanks in advance,
    Kumar

    Hello

    Yes, you can get to know the time for the verification that frm trail... u can also make use of activity sensors to calculate the exact time.

  • How to pass a value to a variable at the time of the execution of the script

    Hi all

    I need to establish an ODI scenario, with below requriments.

    I have to move the data from flat file to target DB.

    flat file name is not consistent, it must pass the file at the time of the excution of scenario name.

    All could help me how to build this ODI scenario.

    Appreciate your help.

    In the data store where the filename in the "name of the resource', you enter a variable #myVar

    You must recompile your script after that this change was made.

    And when you run the script, follow the «-Project.myVar=c:\filename.txt settings.

    Probably, this link may help you http://blogs.oracle.com/dataintegration/2009/04/using_parameters_in_odi_the_dy_1.html

  • How a class can access a variable on the time line of MovieClip?

    in my class, I declare a linked clip:

    a_mc = new LinkedMC ();

    trace (a_mc.testString);

    and on the time line of LinkedMC:

    testString var = "test string";

    but I can't access test string.

    clues?

    Thank you!

    you try to access the variable before it is defined.  have your LinkedMC send a rental event your knowledge the variable has been set and assign a listener to a_mc to detect this event.  in the listener function, use your trace() function.

  • How to find the time between two channels of entry in the data acquisition card or pci 6036

    Hello

    I read a lot-related posts on the simultaneous measurement of two input voltage of similar channels in map data acquisition. I know that the best material is "simultaneous measurments of the Series DAQ cards" but I only pci data acquisition card 6036 and I try to understand what is the time between the reading of the two channels . This period is always constant? (must it rely on a voltage (amplitude, frequency, waveform..). I send the sine wave (s) to the two channels and read the values of V, if they read the same value, the difference should always be zero but I get-0,002 to 0.002 Volt difference (I must find a way to convert it in time). A screenshot of my VI is attached. I wonder how I can accurately measure the time delay between the channel.

    I am open to any suggestion, my final goal to read exactly two channels at the same time ((ou connaître le délai exact donc je peux correspondre les données correspondantes étant donné le temps de retard))

    Hi spinup,

    better you should post your question in the forum of LabVIEW, LabWindows/CVI is used

    Good luck.

  • vertical scale 5152, set up the time

    I use a pxi-5152 in a pxi-1033 chassis.

    I want to play a trick and change the vertical scale between my measurements to increase the dynamic range of a signal that I'm scanning.  I made a loop and changed the vertical scale property node after each measurement.  It works, but when I timed it the loop, she seems to take 80 ms.  I need to do in less than 10 milliseconds.

    Is it still possible?  I see the real time it takes for the scale change or there at - it something to speed up?

    I also have access to a PXI real-time with the same card.  Could only accelerate things or is the digitized the limiting on the change of scale factor?

    Thanks for any help

    -Rich

    Hi rich,

    I see the real time it takes for the scale change or there at - it something to speed up?

    You probably see the effect of the two.  There may be something of software that you can do to improve it, but there's always a material limitation when you do this kind of operation.  Switching of the means of vertical lines change from one type of mitigation to another path.  This means there is sedimentation time involved (electronic and mechanical, with obviously dominant mechanics).  The main element in the run-in will probably be the relays that are reversed.  There must be a minimum period after the passage of a relay to ensure the data read by the digitizer are correct / valid.

    Since you know your signal is going in a direction (descending), then you may be able to improve somewhat by calling the property to write the new range, and then commit node & initiating in a loop which contains a minimum transformation (possibly using queues in LabVIEW to transfer the data in another loop of treatment).  If you need treatment in the loop, you must perform the treatment after niScope commit but before launch niScope.  This will allow the digitizer begin to settle in the new configuration, and you have to wait less time when you're ready to call to initiate.

    I also have access to a PXI real-time with the same card.  Could only accelerate things or is the digitized the limiting on the change of scale factor?

    Real-time processing ensures determinism, it does not necessarily speed up the application.  Determinism and defining a specific priority for calculations of software can help speed upward the software that you are running, the interface of low-level PCI/PXI bus driver with the instrument probably will not change radically.  You can certainly try though and let us know what you measure.

    I can understand if it is a fundamental issue, because normally people are not change the range of vertical entry on the fly, but if there is a way around it, it will avoid the need for me to put an amplifier with variable gain on the front-end server or buy a digitizer with more bits.

    This isn't a perfect suggestion, but just a thought: If you have several scanners at your disposal, you might try setting each to a fixed vertical range and split the signal among them.  TClk, you can synchronize these devices together to make sure your data points are sampled at the same time, and then introduce delays in the initial on each digitizer trigger so that they take account of the waveform in a cascade mode.  For example,.

    Digitizer #1, attached to the greater vertical reach, would have no time limit from the time wherever the trigger is received.

    #2 digitizer, set at a smaller vertical range has a fixed period of (10ms?)

    Digitizer #3, value the smaller Beach, has a fixed time (20ms?).

    Otherwise, if you have assistance with the OR, you could try to call a technical sales engineer to discuss other possible options.

    -Andrew

  • Detect if the time/date browse button has been used

    Hello

    I have a checkpoint vi for the indexing of the elements inside the table and shows the date and the corresponding time. I have the large number of control inside my block signals, but I removed all the blocks out of words and signal for this discussion. I replaced all the unnecessary controls with constant values, so don't waste time analyzing.

    If you take a look at my vi, you'll see Star/Stop control, control of Position and an indicator of time stamp (time waveform) with a browse button date/time of another control of time stamp. I have all the features you want, but I have the problem when you use the button browse to bring up the time. I can't put my time twice in a row to the same value as I compare value "Last entry of time" with a current comes to enter. I wouldn't completely change my blocks because I use a lot of variables and property inside my project nodes. I'm looking for a way to figure out if the time/date browse button was pressed so I could jump to the desired in my registered wave point.

    I'll skip the actual implementation of what you're doing because it's not all that clear to me and answer the question about to detect if the button has been pressed. This general question came several times before in this forum. There is no built-in mechanism to detect that the Browse in a timestamp control button was pressed. When you click this button, it actually opens a VI with the title 'Set Time and Date'. For example, you can monitor to see if this window opens.

  • Ticker display the time spent on a page

    I want to show the user the time elapsed since their entry to the page in a ticker that updates every second. So I would say...


    00:00

    00:01

    00:02

    00:03

    ...


    I look at some different methods to do it, but I'm not an expert in Javascript and I am very new to animate dashboard. While I try to figure it out, I wanted to see if I could get feedback on the good way (s) to do so.


    I've been playing with creating a symbol that has a 1 second loop the timeline, and having an action triggered in this timeline. But that requires you to get time at the beginning and then access and manipulate, and who gets easily spoil/do evil.


    Or you can create a loop that increments every second, probably using setInterval() , and just ++ a counter variable. But I don't know if that would be executed asynchronously of everything.

    So if anyone has treaded similar waters, I would be very happy comments!

    Thank you very much!

    Here's an improved version.

    I wanted to see a 0 in the 10s place if the meter has been less than 10. Then I wanted to 00:00:00, not 0:0:0.

    And before I had done all of the globally accessible variables, which is not necessary. The only one that must be accessible in the world (in my case) is timerCombined, which you can access at:

    sym.getComposition () .getStage () .getVariable ('timerCombined');

    sym.getComposition () .getStage () .setVariable ('timerCombined', 'value');

    ----------------------------------------------

    Create variables

    //"... Variables in the format' should say '00:00:00 ' instead of ' 0:0:0 ".

    var secondCounter = 0;

    var secondCounterFormatted = '00';

    var minuteCounter = 0;

    var minuteCounterFormatted = '00';

    var hour meter = 0;

    var hourCounterFormatted = '00';

    Combining everything into one

    sym.setVariable ('timerCombined', ' 00:00:00 ');

    {setInterval (function ()}

    Number of seconds

    secondCounter ++;

    If (secondCounter< 10)="">

    secondCounterFormatted = "0" + secondCounter;

    } else {}

    secondCounterFormatted = secondCounter

    }

    Convert seconds to minutes

    if(secondCounter == 60) {}

    secondCounter = 0;

    minuteCounter ++;

    If (minuteCounter< 10)="">

    minuteCounterFormatted = "0" + minuteCounter;

    } else {}

    minuteCounterFormatted = minuteCounter

    }

    }

    Convert minutes to hours

    if(minuteCounter == 60) {}

    minuteCounter = 0;

    hours counter ++;

    If (hour meter< 10)="">

    hourCounterFormatted = '0' + hour meter;

    } else {}

    hourCounterFormatted = hour meter;

    }

    }

    sym.getComposition () .getStage () .setVariable ('timerCombined', (hourCounterFormatted + ':' + minuteCounterFormatted + ':' + secondCounterFormatted));

    (}, 1000);

    ^ Here set the time in milliseconds

  • Where are the time Machine connects to Sierra?

    Running Sierra on a late 2009 iMac. Time Machine works well and updated regularly. However the Time Machine buddy Widget shows all the information. Study using the Console indicates that the TM log entries are to be absent in Syslog. Nothing shows using the "backupd" as a filter on the system log. Under Diagnostics and usage data, I see information for the type of backup and "NOERROR" for completion.

    Does anyone know what has changed in the Sierra? Or is it my unit?

    Boyfriend of Time Machine does not work with Sierra

    Console is 'redesigned', extraction of newspapers can be difficult...

    But!

    You can copy the lines of command in Terminal to see the logs of Time Machine ("TM on the fly" and "History of TM")

    Since this discussion started I see how activity time Machine Console with macOS Sierra;

    By copying this command (from John Galt) line in the Terminal I can view TM-history-logs for a few days back;

    Newspaper show - predicate "subsystem =="com.apple.TimeMachine"'-info | grep ' upd: (' | cut-c - 1-19 140-999)

    This command line also gives information of a TM running 'on the fly', but after the TM copy I have to put in the line of command again TM start-up for a new copy (I can live with that...)

    By copying this command (from chockosmund) line in the Terminal I see what TM is ding on the fly, but did not get the newspaper story.

    the log stream - style syslog - predicate 'contains a [cd] 'Time Machine' senderImagePath' - info

    This command line is that Terminal shows constantly new TM-logs.

    Thank you John Galt and chuckosmund!

    I have a separate Terminal office-space computer were I can watch how TM is progressing.

    Hope that answers your question.

    Jur Kuipers, Netherlands

  • WatchOS3 - Mickey will not tell the time

    I read that with the new OS Watch, tapping on Mickey or Minnie will launch a saying of the time in their voices.  So far, I have not been able to get them to tell the time by pressing.  I have a first generation Watch (pre-series 1), who had no prior problems with operating system updates.  To resolve this problem, I have:

    • Removed and re added the dial of the watch and attempted some Mickey and Minnie.
    • Ensure that the 'Tap to talk Time' is enabled.
    • Made sure the silent mode on the watch.
    • Turned the alert Volume all the way up to maximum.
    • Ensure that other alerts work - who they are.  Mail alerts, messages, calendar entries and Facebook all sounds off.
    • Operated as a Single, double-tapped and triple-operated by trying to make them say both.
    • Restarted the watch and tried all the steps again.

    I'm puzzled.  Anyone else having this problem?  What is only an option for the 2 series watch?  All of the suggestions that I have not yet tried?

    Next step may be to restore your watch.

  • Why the clock on my calendar seems to be working at the time of the West Coast, while my ThunderBird is defined and works well at the time of the coast is?

    Ladies and gentlemen:

       I recently started using Calendar add-on to Thunderbird. My test event entered okay. But, my invite showed up on my colleague's PC with meeting time 3 hours late. That is, my meeting time is 11:00 Eastern time. But, he got the time as 14:00 which was 11:00 West coast time. I checked my "Calendar" tab view. There is a tell-tale sign that a red horizontal line that appears to follow my PC's real time clock but lagging 3 hours behind. So far, I could not find how to synchronize them.
    
       My PC was used on the West coast for several years before I brought it to the East coast a few months ago. Since then, my Thunderbird appears to run fine with the clock setting changed to the East coast time zone. Could there be any kind of memory somewhere that took my numerical 11:00, but recorded it by tagging as a West coast time zone entry?
    

    Abe (2015-10-11 11:54)

    Make sure that you do not have the correct time zone set lightning (calendar).

    In the top right of the Thunderbird window, click the menu button > Options > Options > calendar > general tab > zone

  • How to use the time of 80 MHz with the example of the soul base angular Position-buffer-Cont-Ext Clk.vi

    Hello:

    I use the card PCI-6602 and am a bit new to the use of timer/counters.  I use the Commission to measure the position of a quadrature encoder.

    I would use the time base internal 80 Mhz the map of 6602, but examples of LabVIEW will not allow that.  In the example, he States:

    2. call the DAQmx Schedule VI (sample clock) to configure settings of clock synchronization external sample Mode Sample and sample clock Source.  The sample clock Source will determine when a sample will be inserted into the buffer.  The 100 kHz, 20 MHz and 80 MHz time bases can serve as the sample clock Source.  The Edge parameter can be used to determine when a sample is taken.

    Is it possible to use the database internal time for my application?

    I have attached the example vi that I use.

    Thank you.

    Pat

    Hi, Pat.

    The rate determines how fast the samples are acquired and put on the material buffer.  This value depends on the basis of time, which is specified by the source of the VI DAQmx calendar entry.  The default source on the PCI-6602 map is the on-board clock (80 MHz).  The rate must be a divisor of the source.  For example, 80 MHz, 40 MHz, 20 MHz, 10 MHz,..., would be acceptable values for the entry rate.

    So the answer to your question would be to remove the entry from the source and ensure that the rate is a factor of the source.  Leaving this entry unwired will use the clock shipped by default of the unit, as shown in the screenshot below:

    I hope this helps.

  • Is there a way to SHORTEN the time of compilation

    Hey guys,.

    It is extremely laborious when you book only little change in the labview project and the overall program needs to be recompiled all over again!

    Need me often 30 minutes or more. He lead me almost crazy...

    Is there anyone who have samilar experence and find a way to solve it, or even just shorten this boring waiting?

    Thank you very much!

    Nick

    You can't really affect time itself, but you can affect how much/often it RECOMPILES.

    For example this type of variable, is it used in a lot of vi? If this is some common AE, perhaps it can be divided in a couple different ones?

    You can write to protect a folder which shouldn't really change and adjust the LV of "Treat read only VI as locked"

    There is also a framework of LV in the .ini file that can facilitate the optimization that will reduce the time of compile/save.

    /Y

Maybe you are looking for

  • Why repeat the updates?

    Current update is 38.1. He has developed twice this morning for an update. I did it once already. This isn't the first time. Updates several did so.

  • 3.3 - sorry we could not connect to Skype

    reasantly I updeted my computer wendows 10 and my skyp has stoped working, my internet conection is fne, I have reinstaled the app and rebooted the Pc and it still isn't working, I have check internet explorrer and it is updeted,. What should I do?

  • laptop Envy 17 17-j034ca: can not detect the usb flash drives

    Hi, I recently installed windows 7 64 bit on my computer laptop envy 17 17-j034ca. Then, I managed to install all the drivers but it still doesn't detect usb flash drives and I tried to install from the web site of HP and my config but it still does

  • Svchost 100% Windows Update the cause

    From my own post about my problem after taking the post similar here. I have a Panasonic Toughbook, running Microsoft Windows XP Tablet PC Edition 2005, Version 2002, Service Pack 3. I use Microsoft Security Essentials antivirus. This toughbook gets

  • Windows Vista 64-bit Home Premium operating systems disks

    I bought my laptop with Windows Vista 64-bit Home Premium installed on it.  I've since upgraded to Windows 7.  Unfortunately, there are a few programs that will not work under Windows 7 little matter what I do. So, I would try to run a virtual Window