Visa timeout error-1073807339 reading
Hi all
I'm doing a command autoscale on a DSO-A4034A Agilent (Keysight) connected by ethernet.
But for some reason, I get a time-out error. When cheking to find system errors.
I have set up the session visa in 'MAX' and use the name of the session of the instrument of the call in the VI.
I founded these screws on the Agilent instrument driver. and I can do all the other stuff I have to do without errors. EMPLOYMENT INSURANCE. the installation type measures Aqusition read data and so on without error.
The error is caused by reading order just after the ": SYSTEM: ERR?" command under the request of error VI.
I have updated the firmware of the scope and upgrated 15.0.1 VISA but so far without success.
There is no problem when I turn the "climax of the performance"...
Is there someone who can give me some advice about where to look this anomaly? Any help would be gratefully recived.
/ Best regards Rasmus
Rasmus.Olesen wrote:
There is no problem when I turn the "climax of the performance"...
Crying just "timing issue." Add a little wait between ordering autoscale and error checking. It is likely that the oscilloscope does not resemble incoming orders when he assures the autoscale. If you wait for it to do the job.
Tags: NI Hardware
Similar Questions
-
More information:
- It is connected to the PC via a RS232 - RS232 cable
- I downloaded compliance IVI, IVI drivers and layer Keithly e/s
- Measurement and Automation Explorer:
- Parameters: baud rate 9600, data bits 8, zero parity, no flow. The baud and flow are certainly correct for the instrument, but I don't know what are the other parameters.
- Panel test VISA: write * IDN? \n returns 6 characters without error. When I try to read, even only 1 byte, I get a time-out error.
- The port is never exceeded the query and analysis stage when checking this. I tried termination characters \n, \t, \r and no (all that were available) and yet it returns a time-out error
- LabVIEW (v 8.2): I use ke6485 drivers have been installated opening initialization of ke6485 vi and then others to read my measure. Ever it gets past the initialization phase and will return the error: 1073807339 "timeout expired before full operation".
- Crescent thetimeout period is not correct the problem.
Any help would be much appreciated!
Check what type of cable you are using. It requires a null-modem cable, but if you use it, try a straight through cable. The manual should specify the type of cable.
-
VISA read timeout error - Keithley 2400
I am using LabView 2010 to control a Keithley 2400 Sourcemeter on an RS-232 Interface. I want to set up a measure and take a reading (e.g. ground voltage / time)
I took an example of "Keithley ReadOnly" library code, but the 1073807346 error to retrieve box "initialize". Then I looked in 'Initialize. VI"and found there is a VISA (timeout error?) reading error. I have attached two screenshots that shows errors. Can someone tell me how to solve the problem? Moreover, I am very new to labview, if someone could suggest a solution that is not too 'Advanced '? Thank you very much.
-
I have connected the Ocean Optics USB4000 detector and the program worked well.
But now it displays a Visa Write error-1073807339 of the 'Initialize.vi' & other screws that I used as it is also for driver package.
I read the previous discussions on this error but cannot get my program to run.
All solutions?
Thank you
Have you unplugged the USB device and, after a few seconds, plugged in? I have USB devices work for weeks and lose their sanity. The only way to reset have disconnected them and reconnected them.
-
VISA READ timeout error - several GPIB resources
Hello
I'm working on a network 3 GPIB instruments (Optical Attenuator, amplifier in fiber, spectrum analyser), controlled using VISA sessions in Labview. When it is run separately, the three corresponding screws (which are located in three different Labview projects) works as expected. However, when they ran at the same time, one gives them timeout-1073807339 VISA READ errors. These errors seem to occur when another instrument is sending / receiving data / instructions at the same time that she is.
The exact context of the error is either:
-an another VI is running, which includes sending several requests and responses every 100 ms, reading
-by failing VI starts, I get an error of waiting time among the first Subvi containing a VISA READ operation to perform (sometimes initialize.vi (in situation 1), sometimes one of the subVIs connected the rest State (State 0) on the time-out of the structure of the event).
or:
-the default VI is running,
-by starting another VI, which includes repeatedly sending requests and reading the responses, the failing VI up a mistake among the first Subvi containing a VISA READ operation to run (one of the subVIs linked the State of rest (0) on the time-out of the structure of the event).
What I tried:
-gradually increasing the delay between the VISA WRITE and READ operations to the relevant instrument (from 10 ms to 10 s), without success. More surprising are my observations which, by operating this single VI, increasing WRITE it / READ delay results in the same timeout errors. I have found no mention of such behavior through research google and forum. I hope that this may point to a solution to the main issue,
-switching between synchronous and asynchronous VISA WRITE / READ operations.
-reorganization of the network GPIB to a star to a linear topology topology (three instruments have different GPIB addresses in case anyone is wondering).
My thoughts:
It seems to me that the error is related to a delay introduced between a VISA request and its read operation by the transmission of another query associated another instrument in the same network GPIB. But I have no idea why pass a query to another instrument would introduce such a delay, or why this delay would result in a timeout error (and only an instrument, while writing / reading in each pilot screws are basically the same). Hopefully a more experienced Labview-er will be able to shed some light on my problem.
Included is the project that contains the default VI (main.vi) and the driver custom made use of.
There are a number of ways. If you want to keep separate projects for instruments, you can use semaphores often slandered around calls for screws in each project. Locking is not expected or predictable, but it would give controlled access to the GPIB resource.
-
VISA Read Timeout error when using stream protocol BUS
Hello
I currently use the BUS Flow for a MFC Protocol and I get a time-out error VISA Read:
Error 1073807339 occurred at VISA read in Send Setpoint.VI-> test.vi
Possible reasons:
VISA: (Hex 0xBFFF0015) timeout expired before the operation is complete.
I extracted the part of the code and it attached to this message. Basically, whenever I hit OK to send a VISA just read set value times out. I tried to use the sub VI who is responsible to send the set value and it was fine (when manually, I copied and pasted the ASCII codes) but when I run the global function it times out.
Any help is appreciated
Thank you
If you have a command string or a constant you enter a \r\n in, you MUST have '-' Codes view turned on in order to enter the actual control characters. If you enable "\" Codes to display and see \\r\\n, then you have used Normal display to enter the characters.
-
Error 1074000000 has occurred in a place not identified and error 1073807339 occurred at VISA read
I worked with it for a few weeks. I am a student doing research for a professor and I'm new to LabVIEW (about a month).
We calculate impedance in an object with a pulsar. The pulsar will pass through an area of resistance before and after the charge for the calculation of v1 and v2. This will help us to find the current which will help us to find the impedance of the load. In this case, the charge will be a potato. The pulsar will send piulses so we want the program to record the waveform on the trigger. We have v2 and v1 of two channels. We must record the results of v1 and v2, more than an hour. In addition, if v1, v2, and timestamp are in the same excel file then our lives would be easier. A waveform by file.
-TL; DR Two channels to be recorded with a timestamp
The oscilloscope is the WaveRunner LeCroy_64xi.
We use LabVIEW 2014 SP1.
I connect to an oscilloscope via an ethernet connection. The connection should not be the problem, but I connected the oscilloscope using NOR-MAX. The oscilloscope is an ACIP device with a remote system, passport and alias. The instructions step by step are http://forums.ni.com/ni/attachments/ni/140/43120/1/VICP-NI-MAX.doc.
After you connect the oscilloscope, we took a sample program from the driver and changed that to do what we want. We got a few results. FetchWaveform strives to get the shape of wave, but after four or so waveforms, it gives two errors. We tried to add delays, loops and run the program once at a time. I'm willing to start one of these well. This worked well enough at some point, but I think it was because we had a few labVIEW programs open and he was capture the waveform at every second. I think it was because the LABVIEW program has been slowed by the natural delay. It was exactly what we were looking for, but mistakes don't happen often. After the computer restarts and the program, the waveforms are extremely fast registered and errors occur. It is not consistent and we do not know how to register only when the waveform is triggered. Trouble downloading my code here.
Moving this, we tried ReadWaveform. For some reason, we cannot get readWaveform to work. It does not display the waveform and gives the second error or timeout error. We understand why he will not really connect with the oscilloscope. I do not change one of the examples that use readWaveform for this problem.
I'll be here tomorrow, but I'm almost done for today. If there is anything I missed, I'm sorry and I'll it download tomorrow. I'm out of ideas. To message me please and thank you very much for your effort.
Hi Tnorm007,
I appreciate the level of detail you have included; It seems that there are a lot of things here. However, it still seems to be inconsistencies and ambiguities in the errors that you receive. You are able to get a consistent result by running the code you provided?
For example, if you run the code once, it works as you want?
If you press run is once again, it works the way you want? (I assume that you configure the writing on a measure VI file to add the data to the current file?)
Does it fail after a number of iterations, or it does not depend on the speed at which it runs, back to back? Which two errors are thrown into this case?
Also, just a general tip, if you want to receive signals continuously, you want only portions of reading/writing the code of the loop (and initialize/close the task only once), rather than put all the code in a large loop or by pressing the button several times run
Best,
David Randolph
Engineering applications
National Instruments
-
Hi, I'm getting "the resource is valid, but the VISA cannot currently access" error. " I use a 4 x serial usb adapter.
Is it because I use modbus I get this error of visa. All of my com ports work correctly when it is connected to my other devices. Multimeter, power supply etc..
In the interactive control of VISA my devices are labeled under
It's the ony com port connected right now. I made a loopback test in labview using Scripture series and read vi and also to the MAX I can write/request to the device but there is no answer or a timeout error. I connect the RS232 to RS-485 to connect to a device of ADAM. All reviews are very helpful.
Thank you
Damien
Hi Damien
I confirm that you are experiencing an error of 1073807246 code? It is possible that you did not close the reference to the VISA session in a previous program and as such are not allowed to communicate with him in another.
The following links explain this more in detail and show that the VISA close function required to close the references you can do during the execution of your code. Don't forget to close the reference every time you want another program to take control of the serial port, as labVIEW does not release from default serial ports.
-
I am not able to connect to my facebook account and the error message reads "this website does not provide information on property." Help, please
only title bar opens and nothing else... Please help
Clear the cache and cookies from sites that cause problems.
"Clear the Cache":
- Tools > Options > advanced > network > storage (Cache) offline: 'clear now '.
'Delete Cookies' sites causing problems:
- Tools > Options > privacy > Cookies: "show the Cookies".
Also check the cookie exceptions.
-
I use VISA 5.3 for reading my serial device... my standard equipment is WELL this device will send four messages of single command given to this instrument... and I read messages through VISA series READ... during the race, the VI buffer displays the messages individually... I need all four messages to display at the same time in VISA SERIES READ BUFFER, in order to find more than this block, I should add from the palette of function.
Disable character of endpoint detection or make four readings where you concantanate simply the strings together.
-
Hello
I recently tried to install Cool Edit Pro 2.0 on my Dell laptop. After that it's over, I went to open the application, but the error which read "wmvcore2.dll is not found" message popped up. He suggested I have reinstall the application, so I did, but nothing helped.
I wonder what I can do to solve this problem and can you give me any help will be greatly appreciated.
Thank you.
http://www.Microsoft.com/Windows/compatibility/Windows-Vista/default.aspx
Windows Vista Compatibility Center
First thing to do is to check its Vista compatibility at the link above, and if not to see what patches/solutions are available from its manufacturer...
http://www.howtogeek.com/HOWTO/Windows-Vista/using-Windows-Vista-compatibility-mode/
If this is not Vista compatible, read the info on the link above.
It works for some programs, but not all.This applies to software programs, NOT hardware drivers.
If it's Vista compatible > uninstall it > Re-download/save to your desktop > right click on the setup.exe > run as admin.
If the above does not help, try to repost in Forum Vista programs:
http://social.answers.Microsoft.com/forums/en-us/vistaprograms/threads
See you soon.
Mick Murphy - Microsoft partner
-
semaphore timeout error 0 x 80070079
Backups continue to fail. I tried "autobackup" and it stops. Used backup vista and has received 0 x 80070079 semaphore timeout error. How can I remove or change this feature.
Hello
Method 1:
Temporarily disable the antivirus on the computer running software and check if you are able to install the software.
Disable the anti-virus software
http://Windows.Microsoft.com/en-us/Windows-Vista/disable-antivirus-software
WARNING: Antivirus software can help protect your computer against viruses and other security threats. In most cases, you should not disable your antivirus software. If you need to disable temporarily to install other software, you must reactivate as soon as you are finished. If you are connected to the Internet or a network, while your antivirus software is disabled, your computer is vulnerable to attacks.
Method 2:
I suggest you perform the clean boot and check.
How to troubleshoot a problem by performing a clean boot in Windows Vista or in Windows 7
http://support.Microsoft.com/kb/929135
Note: After a repair, be sure to set the computer to start as usual as mentioned in step 7 in the above article.
See also:
Error 0 x 80070079: the semaphore timeout period has expired
-
The error message reads "the user profile Service service has no logon.
User profile cannot be loaded.
Thanks have tried but can't boot from a fix it for me or set me.
Go to Safe Mode with networking
Shut down your computer > turn it back on and immediately and repeatedly tab key F8 until you see a black and white screen. The top/down arrow and select Mode safe mode with networkingYou can log into your own account in safe mode?
If so, you remember the date when you did NOT have this problem?
If so, this date no more than 10 to 15 days ago?
If so, again, please do a restore of the system in safe mode.How to restore the system:
Start button > Search box, type system restore > press the Enter key > uac prompt > click on choose a different restore point > next > select dates as your restore point, until the click > next > finish
To sit and wait. The machine restarts when it's done.Once completed, restart your machine and see if you can log in to your account in NORMAL mode.
For the benefits of others looking for answers, please mark as answer suggestion if it solves your problem.
-
DRAC5 "Error when reading the SSL socket connection"?
Hello
I have a BIOS of PowerEdge 1950 2.7.0 with DRAC 5 Firmware Version 1.65 (12.08.16).
When you try to connect to 'Console' I get the error "Error when reading the SSL socket connection".
I spoke the line "# jdk.tls.disabledAlgorithms = SSLv3, RC4, MD5withRSA, DH keySize.< 768"="" at java.security="" file="" but="" still="" the="">
I use WIndows 10 with the latest FireFox.
Thanks in advance,
Asaf
I found that these changes were necessary to connect to the DRAC 5 using Chrome and Java 8, rev card. 91:
In Control Panel > Java > Security:
1. check "Enable content of java in the browser.
2. set the security level to "High".
3. Add each DRAC IP to the list of exceptions
In C:\Program Files (x86)\Java\jre1.8.0_91\lib\security\java.security):
Remove MD5:
jdk.certpath.disabledAlgorithms=MD2, RSA keySize < 1024
Remove SSLv3, RC4 and MD5withRSA:
jdk.tls.disabledAlgorithms=DH keySize < 768
-
Action Center reports an error that reads "reconnect your drive, it is important."
Action Center reports an error that reads "Reconnect your drive, it is important," the history of your files was disconnected, too long. to keep save copies of your files, reconnect your player an n, then run a backup. Any help?
Thank you
I noticed that my D drive (sd card) drive back unknow question under him? I can put it is not to back up files from the file history and I don't get an error, but I prefer to have connected and backup my files... I hope someone can help.
Maybe you are looking for
-
Pavilion g6 2320tx: ACPI\VEN_HPQ &; unknown device DEV_0004
I have recently installed windows on my g6 2320tx pavilion 10, but after the installation of windows, two unknown devices are presented in the other devices section, Device Manager 1 ACPI\VEN_HPQ & DEV_0004 and 2 root\iwdbus Please help me solve this
-
Hello I'm trying to delvelop software (Labview) that is installed on the student laptops (12-20 max) who reads data continusly (max 1 to 10 hz) of the computer laptop instructors. It tells me that the use of a text file and have the students to read
-
Pavilion dv6 7010tx backlit keyboard only in a few countries?
Hello I own a HP Pavilion dv6-7010tx notebook and I wish to Exchange my no backlit keyboard for a backlit. I saw in the manual (http://h10032.www1.hp.com/ctg/Manual/c03582006.pdf) that the laptop only supports backlit keyboards, however the models in
-
Resale of Windows 7 Home Premium upgrade.
Is it legal to sell my copy of Windows 7 Home Premium Upgrade. It has been installed on a computer at a given time, but this computer is no longer in use due to a complete failure then the replacement. If it's legal, what information should I give th
-
Why windows live mail doesn't stop me to set up my gmail account using pop
I am trying to access gmail via Windows Live mail 2011. However, I am not able to set up Gmail account in windows live mail using pop. A message prompts me to use the IMAP protocol. Why?