Hello!

We just bought a USB-6210 hardware DAQ, which we could successfully install and it appears as well in the device OR monitor.

So, we come the 6210 connected via USB to our Win7 Pro 64 bit machine and wanted to see if his works.

We opened NI Max (ver. 14.5), our device is recognized and we have opened a "Panel Test". We don't see anything, but the noise that we didn t connect anyhing to any unit. But we see a lot of crazy things. Our analog inputs are beginning to increase and saturate all about 11V. Some other entries are reduced to 11V slowly to 0...

However, we connected a photoreceptor (thorlabs) to acquire certain signals... nothing is showing just all the same strange signals.

We went on our device USB-6000 and everything works. We don't get anything with USB-6210.

Any ideas?

Thank you

The readings that you see with nothing plugged are expected. When you do not connect a signal, what is your configuration (diff, CSR, etc.) and what is your wiring? Looks like you selected a configuration but your wiring does not match then you still have a nice.

Tags: NI Hardware

Similar Questions

  • Amnesty International and counter sync + USB signal stream (USB-6210 vs USB-6341)

    Hi all

    I'm at a stage of identification of a material suitable for the following tasks:

    • 5 analog inputs (AIs) of reading at the same time, tensions at a rate of kSps (at least) 10,
    • application captures 2 inputs using timers (detection of contours with timestamps), square wave entry with duty ratio of 50 percent and about 1.5 kHz frequency and variable pulse width / frequency (from 2 sensors hall, representative of the DC motor rotation speed and direction, quadrature signals), resolution of timestamps should be (at least) 50 ns,
    • AIs and counters should behave in a deterministic way, and must be synchronized in a way,
    • data to be transferred via the USB port of a host computer with Matlab Data Acquisition Toolbox (unfortunately not LabVIEW).

    I've identified the long USB-6210 USB-6341 and potential candidates of material to accomplish the above tasks, but after reviewing several documentation and the topics of the forum, I'm still a bit confused, if both are fully working and my approach described below is not working properly.

    Counters: I intend to use the internal time base available 20 MHz as being the source of meter to get into account the resolution of timestamp 50 ns. External impulses hall are used as sample clock (about 1.5 kHz, see above). As the pulse width varies, the sample clock is not constant.

    AIs: Using a 10 kHz internal clock signal derived from the time base of 20 MHz for timing and analog inputs (trigger) start-up and counters simultaneously material should translate into the required synchronization and deterministic behavior.

    It work? Other recommendations?

    Next is the USB data transfer: all HAVE 5 and 2 data entry of the meter must be correctly transferred to the host computer (the corresponding rates are shown above). USB-6210 is capable of 4 USB signal flow, device USB X range (6341) offers 8 of them. Unfortunately, I could not understand the exact meaning of the expression "signal flow" still. Do I need 1 flow of input signals (would be 7 for my application described) or 1 stream for all analog inputs and 1 for counter inputs (lead 2 streams for my request). Is there no further details on this approach (more than Streaming of signals of NOR) USB signal flow?

    Any challenge to the described application that I might have forgotten? 6210 USB seems to a very limited number of entry PFI, maybe even too low for my meter participate application?

    Looking forward to your comments and advice.

    Concerning

    jAwA

    1. I recommend the X-6341 series on the M-series 6210 sake of counters/timers.  It is more of them, and each of them is more capable.   It can also have a great FIFO embarked for meters that may be important in certain tasks, although I don't think that you currently deal with one of them.

    2. your general concepts on timing & sync are satisfactory.  You will be able to share and to route signals that help ensure synchronization and determinism between the timestamps for your various tasks.  Note that for meter entry tasks, you need set up the trigger 'Arm Start' rather than the regular start trigger.

    3 is not authoritarian, but I believe that the flow of signal # will correspond to the tasks #.  For you, it would be 1 task of HAVE and tasks CI 1 or 2.  (Not clear if you have 1 Encoder with 2-channel quad that would require 1 task of CI, or if you have 2 encoders with 4-way quad).

    4. pay attention to the hall effect signals that are not virgins.  Digital filtering is available and probably better on the X-series, the series M.

    5. strictly speaking, edge detection is a type of digital input task that produces samples but no timestamps.  Ideally, I would like to parallel wires on the two digital inputs for the entries of detection and counter change to position quadrature decoding.   Then I would sample the counters Encoder 1 or 2 using the internal pulse 'event of detection of change '.  I would create another counter timestamp change detects pulses as well.

    -Kevin P

  • USB-6210 on Linux - is DAQmxBaseExportSignal supposed to work for this unit?

    I have two devices USB-6210 I need to synchronize so that they
    both collect data exactly at the same time. I want to trigger data
    collection for each device by sending a trigger off Dev1/PFI6 and
    have two devices to receive the signal on PFI0.

    The input analog trigger doesn't seem to work, but the input digital trigger seems to be accepted. Unfortunately, I can't get the clock signal to export.

    Here is the sample program that I use:

    /*********************************************************************
    *
    * ANSI C program example:
    * contAcquire-ExtClk - DigStart.c
    *
    * Example category:
    * HAVE
    *
    * Description:
    * This example shows how to acquire a continuous quantity
    * the data using an external sample clock, launched by a numerical advantage.
    *
    * Execution mode:
    * 1. Select the physical channel is the place where your
    * a signal came on the DAQ hardware.
    * 2. Enter ranges of minimum and maximum voltage.
    * Note: for best accuracy try to match ranges of entry of the
    * waits the voltage level of the measured signal.
    * 3. Select a sample clock source.
    * 4. Set the approximate rate of the external clock. This allows to
    * the internal features of the acquisition to be like
    * effective as possible. Read also assign samples.
    * 5. Select a source for the beginning of digital edge trigger.
    * 6. Select the edge, rising or falling, to trigger.
    *
    * As follows:
    * 1. Create a task.
    * 2. Creating an analog voltage input channel.
    * 3. Set the parameters for an external clock Source.
    * In addition, set the sampling mode to be continuous. The
    * external clock frequency is given to allow internal
    * characteristics of the acquisition to be as effective as
    * possible.
    * 4. Set the parameters of a beginning of digital edge trigger.
    * 5. Call the start feature to start the acquisition.
    * 6. Read data from waveform looping up to 10 seconds or a
    * error occurs.
    * 7. Call the function in the task to clear to clear the task.
    * 8. Display an error, as appropriate.
    *
    * Overview of e/s connections:
    * Make sure that your input of the signal Terminal corresponds to the physical
    * Control I/o channels. Also, make sure that your digital results in the application
    * signal is connected to the terminal that is specified in the command Source.
    *
    * Recommended use:
    * 1. Configure and start calling features.
    * 2. Call loop function.
    * 3. Call Stop function at the end.
    *
    *********************************************************************/

    #include "NIDAQmxBase.h".


    #define DAQmxErrChk (functionCall) {if (DAQmxFailed (error = (functionCall))) {goto error ;}}

    int main (void)
    {
    Task settings
    Int32 = 0 error;
    TaskHandle AITaskHandle = 0;
    tank errBuff [2048] = {'\0'};
    Int32 i;
    time_t startTime;

    Channel settings
    char [] = "Dev1/ai0" chan
    float64 min = - 10.0;
    float64 max = 10.0;

    Sync settings
    char [] = clockIn ' / Dev1/PFI0 ";
    char [clockOut] = ' / PFI4/Dev1;
    uInt64 samplesPerChan = 1000;
    float64 sampleRate = 10000.0;

    Trip parameters
    char [] = triggerIn "/ Dev1/PFI2";
    char [triggerOut] = ' / PFI6/Dev1;
    uInt32 = DAQmx_Val_RisingSlope triggerSlope;
    uInt32 = 100 triggerSamples;

    The data read settings
    #define bufferSize (uInt32) 1000
    float64 data [bufferSize];
    Int32 pointsToRead = bufferSize;
    Int32 pointsRead;
    float64 timeout = 10.0;
    Int32 = 0 totalRead;

    printf ("calling CreateTask - HAVE... \n");
    DAQmxErrChk (DAQmxBaseCreateTask("",&AITaskHandle));
    printf ("calling CreateAIVoltage - HAVE... \n");


    DAQmxErrChk (DAQmxBaseCreateAIVoltageChan(AITaskHandle,chan,"",DAQmx_Val_Cfg_Default,min,max,DAQmx_Val_Volts,));
    printf ("calling CfgSampleClkTiming - HAVE... \n");
    DAQmxErrChk (DAQmxBaseCfgSampClkTiming (AITaskHandle, clockIn, sampleRate, DAQmx_Val_Rising, DAQmx_Val_ContSamps, 1000));
    printf ("calling CfgSampleClkTiming - HAVE... \n");
    DAQmxErrChk (DAQmxBaseCfgSampClkTiming (AITaskHandle, "OnboardClock", sampleRate, DAQmx_Val_Rising, DAQmx_Val_ContSamps, 1000));
    printf ("calling DigEdgeStartTrig...\n");
    DAQmxErrChk (DAQmxBaseCfgDigEdgeStartTrig (AITaskHandle, triggerIn, triggerSlope));
    printf ("calling ExportSignal - sample on %s...\n clock", clockOut);
    DAQmxErrChk (DAQmxBaseExportSignal (AITaskHandle, DAQmx_Val_SampleClock, clockOut));
    printf ("calling StartTask - AI\n");
    DAQmxErrChk (DAQmxBaseStartTask (AITaskHandle));

    The loop will leave after 10 seconds

    startTime = time (NULL);
    While (time (NULL)
    printf ("loop - calling ExportSignal - Trigger...\n");
    DAQmxErrChk (DAQmxBaseExportSignal (AITaskHandle, DAQmx_Val_StartTrigger, triggerOut));
    printf ("loop - call ReadAnalogF64...\n");
    DAQmxErrChk (DAQmxBaseReadAnalogF64(AITaskHandle,pointsToRead,timeout,DAQmx_Val_GroupByChannel,data,bufferSize,&pointsRead,));
    totalRead += pointsRead;
    printf ("acquired %d samples. %D total \n",pointsRead,totalRead);
    Print only the first 10 points of the last data read
    for (i = 0; i< 10;="">
    printf ("data [%d] = %f\n", i, data [i] ");

    }
    printf ("%d \nAcquired total samples.") "\n", totalRead);

    Error:
    If (DAQmxFailed (error))
    DAQmxBaseGetExtendedErrorInfo (errBuff, 2048);
    If (AITaskHandle! = 0) {}
    DAQmxBaseStopTask (AITaskHandle);
    DAQmxBaseClearTask (AITaskHandle);
    }
    If (DAQmxFailed (error))
    printf ("Error in DAQmxBase % d: %s\n", error, errBuff);
    return 0;
    }

    And here is the result of the execution of the example program.

    $. / contAcquire-ExtClk-DigStart
    the CreateTask call - HAVE...
    call CreateAIVoltage - HAVE...
    call CfgSampleClkTiming - HAVE...
    call for DigEdgeStartTrig...
    Calling ExportSignal - sample on/Dev1/PFI4 clock...
    DAQmxBase error-200428: value passed to the control task/channels is not valid.

    Someone can tell me what I am doing wrong or is ExportSignal not supported for devices USB-6210 on Linux?

    Thank you

    -Tom

    Hey Tom,

    Thanks for the update, I should have tested my suggestion before posting but I had no useful material at a time.  It turns out that, DAQmx Base supports not the outputs of the internal sample clock routing counter HAVE.

    Workaround will be physically wiring the meter output terminal (default is 4 PFI for counter 0, PFI 5 to 1 meter) until one of the PFI input lines (PFI 0:3) and use the PFI input as SOURCE for your sample clock (for example "/ Dev1/PFI0").

    Best regards

  • SCXI and USB 6210? Getting started...

    Hello

    I'm looking to get an online data collection system and had a few quick questions of SCXI signal conditioning high-performance...  First of all I have is 1 chassis SCXI-1000, SCXI-1121, SCXI 1180 and 1100 SCXI modules assorted.  I also have a few blocks of different endings, SCXI-1300 block and a 1320.

    (1) by section "what to start' in"the SCXI chassis user manual"everything only I need is hardware of SCXI chassis (in my case the SCXI-1000 chassis), a 120 VAC power chord and a computer.  However in the ' what you need to get started section "of the manual"Getting started with the SCXI"I need also a SCXI-1200 or"Hardware DAQ"module  So my question is this, can I set the SCXI-1000 chassis directly to my laptop without any additional hardware?  I feel, I need something else, but want to make sure...

    1 (b) assuming I need a material additional DAQ to start, I can use a NI USB 6210 somehow?  I checked the manuals and little is said about SCXI.

    (2) neither manual mentions a controller, but the SCXI Advisor shoot one.  They should say if all I do is collect data, not active control?

    Thank you

    Mike

    Mike,

    There are two options for you to get your SCXI system operational. You will need either an M series or a card of E-Series you connect to your SCXI system or you could focus on the NI SCXI-1600, which is a controller for your SCXI chassis and allows you to connect your laptop to the controller via a USB cable. This controller will operate similar to the E-Series and M-series cards.

    We also offer a couple of PCMCIA cards which allows you to integrate the SCXI system with your laptop. It comes from the NI DAQCard-6036E, and NI DAQCard-6062E. Unfortunately, your NI USB-6210 is unable to be used as a device for the Acquisition of data for your SCXI chassis. The list DAQ cards which are possible to use with your system can be found here on the tab 'controller '.

    http://ohm.NI.com/advisors/SCXI/pages/SCXI/modules.XHTML?conversationContext=1

    In addition, we have an old manual SCXI of starting you can find here:

    http://digital.NI.com/manuals.nsf/WebSearch/C195895E8C9CAE3D8625691F004DB285

  • I need help to change a VI of the NI PXI-1042 q to USB-6210

    Hi all.  This is my first post, and I apologize in advance if it is in the wrong place or is not suitable... I was instructed to come here to help.

    I'm working on the evolution of a VI that works perfectly with an NI PXI-1042 q on a much smaller unit, the USB-6210.  Supposedly the USB-6210 can handle input/output same I need (a read analog at 30 kHz, 2 outputs digital input, written at 10 kHz), however I am having a little trouble to make it work properly.  The exits seem to go where they are supposed to, but the new system is orders of magnitude slower than the original, and for the system, I use this is unacceptable.

    I'm sure I'm doing something wrong, but I'm about as new to the LabView you can get, so I have no idea where to start.  Any help would be appreciated.

    More info you might need:

    LabVIEW 2010

    The VI is attached.

    Respectfully,.

    Emrys Maier

    University of Texas at Arlington Research Institute

    Research Assistant

    The device is unable to produce the frequency required by the output digital, but the outings of the counter do it perfectly.

    Emrys-

  • Can the USB-6210 output signal square 10 kHz?

    I scoured the internet/forums to find out the maximum data rate for a USB-6210, but nothing helped.  I know that the device can handle 250 kech. / s, but this does not appear to be linked to output speed.

    By experimentation, I estimated the maximum rate 2500 kbps, but I fear that this figure may be inaccurate because I don't really know what I'm doing.

    I would like to know:
    (a) is the USB-6210 is able to produce a square wave of 10 kHz?

    (b) if not, is there a USB DAQ, which is?

    Respectfully,.

    Emrys Maier

    University of Texas at Arlington Research Institute

    Research Assistant

    Let's see here.  No analog outputs.  Only software timed DIO, so it does not work.  But there are 2 meters which should be able to output a square wave at the speed you need.  I have not done it myself, but if you use the finder to the example and looking for something like 'pulse train' or 'exit meter', you should find some good examples of how to do it.

  • sensor pressure and USB-6210

    I am installing a USB-6210 to read pressure sensors (specifically: http://www.omega.com/pptst/PX309-100mv.html), with labview2013.

    Its one task relatively simple and should not be this hard...

    I used a similar system to take similar measures in the past, but its been awhile.

    I don't know if it's just a wiring problem or incorrect installation of labview (program), but I can't measure the tensions that are remotely close to the sheet of calibration of the probe.

    Thoughts?

    It turns out that my diet was not come out correctly. Passed to a diet different and everything works as it should.

  • USB-6210 on Linux - is DAQmxBaseCfgAnlgEdgeStartTrig supposed to work for this unit?

    I have two devices USB-6210 I need to synchronize so that they both collect data exactly at the same time. I want to trigger the data collected for each device by sending a trigger off Dev1/PFI6 and have two devices to receive the signal on PFI0.

    I "ve tried using the example AI program that receives a left analog trigger, and it doesn't seem to work. Here is the code I tried:

    /*********************************************************************** ANSI C Example program:*    acquireNScans-AnlgStart.c** Example Category:*    AI** Description:*    This example demonstrates how to acquire a finite amount of data*    using the DAQ device's internal clock, started by an analog edge*    condition.** Instructions for Running:*    1. Select the physical channel to correspond to where your*       signal is input on the DAQ device.*    2. Enter the minimum and maximum voltage range.*    Note: For better accuracy try to match the input range to the*          expected voltage level of the measured signal.*    3. Set the number of samples to acquire per channel.*    4. Set the rate of the acquisiton.*    Note: The rate should be AT LEAST twice as fast as the maximum*          frequency component of the signal being acquired.*    5. Set the source of the start trigger. By default this is*       analog input channel 0.*    6. Set the slope and level of desired analog edge condition.** Steps:*    1. Create a task.*    2. Create an analog input voltage channel.*    3. Set the rate for the sample clock. Additionally, define the*       sample mode to be finite.*    4. Define the parameters for an Analog Slope Start Trigger.*    5. Call the Start function to start the acquistion.*    6. Read all of the waveform data.*    7. Call the Clear Task function to stop the acquistion.*    8. Display an error if any.** I/O Connections Overview:*    Make sure your signal input terminal matches the Physical*    Channel I/O Control. In this case wire your signal to the ai0*    pin on your DAQ Device. The analog start trigger must be connected*    to PFI0.** Recommended Use:*    1. Call Configure and Start functions.*    2. Call Read function.*    3. Call Stop function at the end.**********************************************************************/
    
    #include "NIDAQmxBase.h"#include 
    
    #define DAQmxErrChk(functionCall) { if( DAQmxFailed(error=(functionCall)) ) { goto Error; } }
    
    int main(int argc, char *argv[]){    // Task parameters    int32       error = 0;    TaskHandle  taskHandle = 0;    char        errBuff[2048]={'\0'};    int32       i;
    
        // Channel parameters    char        chan[] = "Dev1/ai0";    float64     min = -10.0;    float64     max = 10.0;
    
        // Timing parameters    char        clockSource[] = "OnboardClock";    uInt64      samplesPerChan = 1000;    float64     sampleRate = 10000.0;
    
        // Triggering parameters    //orig    char        triggerSource[] = "PFI0";    char        triggerSource[] = "/Dev1/PFI0";    uInt32      triggerSlope = DAQmx_Val_RisingSlope;    float64     triggerLevel = 1.0;
    
        // Data read parameters    #define     bufferSize (uInt32)1000    float64     data[bufferSize];    int32       pointsToRead = bufferSize;    int32       pointsRead;    float64     timeout = 10.0;
    
        DAQmxErrChk (DAQmxBaseCreateTask("",&taskHandle));    DAQmxErrChk (DAQmxBaseCreateAIVoltageChan(taskHandle,chan,"",DAQmx_Val_Cfg_Default,min,max,DAQmx_Val_Volts,NULL));    DAQmxErrChk (DAQmxBaseCfgSampClkTiming(taskHandle,clockSource,sampleRate,DAQmx_Val_Rising,DAQmx_Val_FiniteSamps,samplesPerChan));        printf("Calling CfgAnlgEdgeStartTrig with %s...\n", triggerSource);    DAQmxErrChk (DAQmxBaseCfgAnlgEdgeStartTrig(taskHandle,triggerSource,triggerSlope,triggerLevel));    printf("Done with StartTrig.\n");    DAQmxErrChk (DAQmxBaseStartTask(taskHandle));    DAQmxErrChk (DAQmxBaseReadAnalogF64(taskHandle,pointsToRead,timeout,DAQmx_Val_GroupByChannel,data,bufferSize,&pointsRead,NULL));
    
        printf ("Acquired %d samples\n", pointsRead);
    
        // Just print out the first 10 points    for (i = 0; i < 10; ++i)        printf ("data[%d] = %f\n", i, data[i]);
    
    Error:    if( DAQmxFailed(error) )        DAQmxBaseGetExtendedErrorInfo(errBuff,2048);    if(taskHandle != 0) {        DAQmxBaseStopTask (taskHandle);        DAQmxBaseClearTask (taskHandle);    }    if( DAQmxFailed(error) )        printf ("DAQmxBase Error %d: %s\n", error, errBuff);    return 0;}
    

    and here is the result of the execution of the program:

    $. / acquireNScans-AnlgStart
    Calling CfgAnlgEdgeStartTrig with/Dev1/PFI0...
    DAQmxBase error-200428: value passed to the control task/channels is not valid.

    The file README.txt NIDAQ-mx says this is taken in charge for USB-621 X. Can someone tell me why it does not work?

    Tom,

    Analog start and trigger reference are supported only on the PXI, PCI, and the NOR-6210 PCIe form factors.  The Readme file has a list of the functions that are available for USB.

    Kind regards

    Justin

  • Clock and hw external trigger with USB-6210 on Linux with NOR-DAQmx Base?

    I have two devices USB-6210 I need to synchronize so that they both collect data exactly at the same time. I was told by support OR I can send the clock off Dev1/PFI4 and have the two USB-6210 s read the clock in through their own PFI0. I also want to trigger data collected for each device by sending a trigger off Dev1/PFI6 and have two devices to receive the signal on PFI2.

    All my attempts to try this are filled with error messages and my research online seem to say that's not possible with USB devices on NOR-DAQmx Base 3.4.0f2 on Linux.

    I "ve tried using example AI programs and those who do not seem to work either for external clocks. Here is the code I tried:

    #include "NIDAQmxBase.h"#include 
    
    #define DAQmxErrChk(functionCall) { if( DAQmxFailed(error=(functionCall)) ) { goto Error; } }
    
    int main(void){    // Task parameters    int32       error = 0;    TaskHandle  taskHandle = 0;    char        errBuff[2048]={'\0'};    int32       i;
    
        // Channel parameters    char        chan[] = "Dev1/ai0";    float64     min = -10.0;    float64     max = 10.0;
    
        // Timing parameters    char        clockSource[] = "/Dev1/PFI7";    uInt64      samplesPerChan = 1000;    float64     sampleRate = 10000.0;
    
        // Data read parameters    #define     bufferSize (uInt32)1000    float64     data[bufferSize];    int32       pointsToRead = bufferSize;    int32       pointsRead;    float64     timeout = 10.0;
    
    printf("Calling CreateTask...\n");    DAQmxErrChk (DAQmxBaseCreateTask("",&taskHandle));printf("Calling CreateAIVoltageChan...\n");    DAQmxErrChk (DAQmxBaseCreateAIVoltageChan(taskHandle,chan,"",DAQmx_Val_Cfg_Default,min,max,DAQmx_Val_Volts,NULL));printf("Calling CfgSampleClkTiming...\n");    DAQmxErrChk (DAQmxBaseCfgSampClkTiming(taskHandle,clockSource,sampleRate,DAQmx_Val_Rising,DAQmx_Val_FiniteSamps,samplesPerChan));printf("Calling StartTask...\n");    DAQmxErrChk (DAQmxBaseStartTask(taskHandle));printf("Calling ReadAnalogF64\n");    DAQmxErrChk (DAQmxBaseReadAnalogF64(taskHandle,pointsToRead,timeout,DAQmx_Val_GroupByChannel,data,bufferSize,&pointsRead,NULL));
    
        printf ("Acquired %d samples\n", pointsRead);
    
        // Just print out the first 10 points    for (i = 0; i < 10; ++i)        printf ("data[%d] = %f\n", i, data[i]);
    
    Error:    if( DAQmxFailed(error) )        DAQmxBaseGetExtendedErrorInfo(errBuff,2048);    if(taskHandle != 0) {        DAQmxBaseStopTask (taskHandle);        DAQmxBaseClearTask (taskHandle);    }    if( DAQmxFailed(error) )       printf ("DAQmxBase Error %d: %s\n", error, errBuff);    return 0;}
    

    When I run the resulting program, I see this:

    $. / acquireNScans-ExtClk
    The CreateTask call...
    Call for CreateAIVoltageChan...
    Call for CfgSampleClkTiming...
    Error-89136 DAQmxBase: route specified cannot be satisfied, because the hardware does not support it.

    For example, a clock and a trigger can be imported via one of the PFI lines by using a USB-6210 on Linux with NOR-DAQmx Base? A clock and a trigger exportable via one of the PFI lines?

    If so, does anyone have the code example illustrating how to do this, or can you at least tell me the names of the lines ("PFI0/Dev1" or other) so I can try again?

    Clues or suggestions would be helpful.

    Thank you

    -Tom

    The clockSource in the example specifies an output rather than an input channel channel. Change source "/ Dev1 / PFI0" solved the problem.

    Please close this post.

  • voltage measurement usb-6210

    Hello

    I'm looking for to accomplish three parallel en measures of tension by using a USB-6210 card. Regardless of the means of control used (vi in LabVIEW + acquisition of data SignalExpress or measurement and Automation Explorer), the read signal is not one I send on the map. Attached an example from the control panel. Has someone it already faced this type of results? I suspect a gross error on my part the signal car didn't Exchange read regardless of the voltage that I send on the map (using a generator). Yet it seems to have respected the wirings and my drivers are up to date.

    Small additional difficulty: I'm working on German envorinnements.

    Thanks in advance for the help

    Hello Saiko,

    Thank you for having posted the Forum of National Instruments.

    In your screenshot, you perform a differential measurement. Well connected the negative potential (mass) at the AI9 entrance? Maybe you connected the mass to the GND pin of GOT it, that she only works when a CSR measure.

    I invite you to download the manual of your card: here and here.

    Kind regards

  • -error message 200284 when calling DAQmxReadAnalogF64() with a USB-6210

    I have tried the code a CVI application to communicate with a USB-6210. I want to start the acquisition to gain 30 samples and then collect them. My problem is when I call the DAQmxReadAnalogF64() function, I get this message error of-200284. The error message means that the data has not yet been acquired in a nut shell. He said I should do to ensure that my startup trigger is configured correctly. I tried to use the internal clock to start the acquisition. I tell myself that I have not configured something correctly.

    In the DAQmxCfgSampClkTiming() function, I set the sampleMode to DAQmx_Val_FiniteSamps I thought I would collect 30 samples (number_of_scans = 30) when I called DAQmxStartTask(). It does not work. When I put sampleMode to DAQmx_Val_ContSamps I can collect data. Here is a stripped down most of my code:

    returnAcq = DAQmxCreateTask ("myAcqTask", & acq_taskHandle);
    returnAcq = DAQmxCreateAIVoltageChan (acq_taskHandle, physicalChannel, "acq_channel", DAQmx_Val_Diff, minVal, maxVal, DAQmx_Val_Volts, "");

    returnAcq is DAQmxCfgSampClkTiming (acq_taskHandle, "OnboardClock", sweepsInterval, DAQmx_Val_Rising, DAQmx_Val_FiniteSamps, number_of_scans);.
    returnAcq = DAQmxStartTask (acq_taskHandle);
    returnAcq = DAQmxReadAnalogF64 (acq_taskHandle,-1, 5, DAQmx_Val_GroupByChannel, Data_readings, number_of_scans, & read, 0);
    Disable the task
    DAQmxClearTask (acq_taskHandle);

    I guess my question is: am I set up so that the internal clock will begin the acquisition when I call DAQmxStartTask()? I thought when I put the second parameter in DAQmxCfgSampClkTiming() to 'OnboardClock', this would use the internal clock.

    Thanks in advance for your help,

    Don Pearce

    Don,

    You're right, your task is configured to use a sample clock internal with a trigger for starting the software.  The description of the error pointed out several reasons why people might see this error.  What is the value of 'sweepsInterval '?  I think that as configured DAQmx should be pointing at the speed (in Hz) specified by sweepsInterval.  By the way-1 in the entrance to numSampsPerChan, I think DAQmx will try to read all samples for a finite acquisition (30 in your case).  If you run below 6 Hz, the lira could be the time-out period before these data were available.

    When it is configured as a continuous, what are the values is passed in the sampsPerChanRead parameter?

    Dan

  • USB-6210 is loaded as Firmware Loader instead of correct device

    When I plug my new USB-6210 windows it loads as a "Firmware Loader" rather than the device. As a result, MAX does not recognize the device. Yes, I have the latest drivers, and Yes, I have already read through the other posts on the support sites. Charger to the unit service and the configuration service manager are both "started." Please advise!

    Solved.

    The cause was that McAfee had blocked the 'NI PAL Service Manager', (NIPALSM. (EXE). Apparently, this caused "OR loader Service device" stop running properly, which caused to turn the failed system to correctly recognize the USB-6210. Other symptoms included:

    1. a failure to expose a chassis cDAQ correctly installed or it is plugin in MAX. The chassis has appeared in Device Manager on the computer as a data acquisition device OR, but will not appear in the list of devices in MAX.

    2. the device loader Service would not start automatically. If the service has been started manually it would not work or it would just 'Stop' itself.

    These questions were tested on a laptop Dell Latitude running Windows 7 proffessional, developers Suite 2013, MAX 5.4 and McAfee version 6.0.0.

  • Open the NI USB-6210 collector

    Hello

    It's just a small matter. I have two Renault uses a NI USB 6009 and the USB-6210. I use the 6009 as outputs open collector. I can't find any information on collector/drain open for the 6210.

    This configuration can be configured?

    Thank you

    Cristian

    Ibanez,

    The digital output channels on the USB-6210 are sourcing outputs.  This module does not change this setting in the software.

    Good luck with the rest of your application!

  • USB-6210 voltage divider Calibration questions

    I'm trying to calibrate 14 tensions that I read from a voltage divider in 14 channels of a device USB-6210. Attached is an Excel spreadsheet of data I'm trying to understand. On the bottom of the sheet, I applied the same voltage to all channels and draw the raw tension for all channels and the differences between the expected voltage and voltage of raw. I find that these differences are not constant but linear I do not understand. I then use the basic linear function to find the factors in the top of the worksheet. Therefore, in the upper part, I read the input voltage by using a probe high voltage and voltage of Labview which has been adapted by the linear function of base in the background.  Most of the multiplying factors are ~ 1000 as expected, but the tensions for the last 4 channels are significantly different and are not linear for the last two channels. I mounted the two worst voltage channels with polynomial which is ok but not make any sense. Does make sense to anyone. I'm using Labview 9.0.1 and Windows 7. I ran Diagnostic Utility for the acquisition of data on the device and it passed all tests. I looked into the issue of ghosting, but the criteria seem not valid here because of tension in the neighbouring country of the channels is small. I also enclose the VI that supports voltages raw to show how these functions are used. According to me, the results are stable and precise, but I can't explain the above mentioned discrepancies.

    The problem was solved by changing the sampling frequency in the DAQ assistant. I think the default frequency is 1 kHz and I had given up to 750 Hz. By lowering the frequency 150 Hz, the capacitor is allowed time to discharge so that no ghost image occurs. In summary, the question always was ghosting and solve the I: question

    -Removed in the wizard DAQ, channels that have been disconnected and therefore not used.

    -Change the wiring to the DAQ assistant by sending the low voltage in channel 1 and the second lowest voltage on channel 2, etc.

    -Change the sampling frequency of 1 kHz to 150 Hz to let the capacitors discharge time.

  • The streaming of multiple channels directly to CSV USB-6210?

    I would like to stream up to 16 channels of data to a CSV file. My sample rate can vary from a file in a file, but I can reach 1 kHz. I understand the disadvantage of using text rather than PDM or binary files, but I have to be able to read these files later with programs such as Excel (Excel 2007 can make 1 million points).

    I have no trouble writing to TDMS files as seems very straighforward. I wonder though, it would be more efficient to use the TDMS files as the data are acquired and then after I stop, read the PDM and convert to CSV or just directly save to CSV? I only say this because I don't know if Labview can broadcast on a text or CSV file at 1 kHz.

    The scenario. The user will begin acquisition (using the USB-6210) to start playback 16 channels, applying multipliers for some data, and then send to a drive. Don't worry about the file size, the operator will be told that if it samples fast it must limit its total sample time. He has the ability to run slower to achieve longer sampling frequencies.
    Bottom line is I want good data for all 16 channels timestamped. No header information, nothing other than data raw beginning to end.

    I've seen several posts where the notice has been using TDMS files. Yet once again, I am not opposed to that, but I need to end up with text files.

    Jeff

    LabVIEW can handle this, if you do it right.

    1. If you are not already, use the design data of producer/consumer model.  It should be a template provided with LabVIEW.  This will allow you to take continuous data because your data treatment/backup of the routines take place in a separate thread.  If the hard data recording longer than reading the data, LabVIEW will keep only the size of the queue more and more with new data and I hope that you will not run out of memory system.

    2 write. you data in "chunks".  If you expect to write each data point to the disk as soon as it is available so that will probably be very tax on performance due to the excessive amount of 'read/write' header to your disk controller information and other things.  It is better to save 1000 samples at once, for example.  This way you are 'hit"your disc 10 times per second instead of 1000 times per second.

    3 ASCII files are very large, but I too tend to use them due to their convienance.  If you think about it, a 32-bit value can be "12332.0123" which in the ascii code is rather 80-bit (1 byte per character, 8 bits per byte, 10 characters is 80 bits or more than twice the amount if stored as a number 32-bit floating point).  It is therefore logical that if you have problems with performance (even after the application of 1 and 2) you write to a binary file.  Then, you can easily write a program that reads the binary file and writes a text file.

    4. If your computer has enough memory, you could put everything in memory and save it in a file at the end of the data collection.  You will need to add more controls and other things if you were writing this LabVIEW application for someone else.