Measurement & Automation Explorer crashes while loading

Hello

I recently bought a DAQ (PCI-6239) card. I installed the NOR-DAQmx driver that came with it and follow the installation steps. When I was at the time of the opening of the Explorer Measurement & Automation to detect my Board, it crashed during the loading of the "plug-ins loading" and invites the error message "M & A E has encountered a problem and needs to close. We are sorry for the inconvenience. If you were in the middle of something, the information you were working on might be lost. »

I tried to repair, reinstall and update all software OR related. Nothing has worked. I uninstalled LabView in case there was a mistake with it and reinstall DAQmx again. None of this has worked. I wonder whether it is a known issue or if there is some kind of problem with my system/OS. I use Win XP SP3.

I have not tried the software to open after you have uninstalled the PCI card, but that would really help me if the software works without the card.

Thank you

--

Patrick

Adam,

I had the operating system reinstalled with LabView 8.2 (it seems that it is the license available at the moment at our University). I downloaded the latest version of the drivers and now everything seems to work very well. From this point, our tech lab should be able to guide me on how to make the "connection" between my DAQ in MAX card and my LabView VI.

Thanks for your help!

--

Patrick

Tags: NI Software

Similar Questions

  • LabVIEW Measurement & Automation Explorer very slow

    Hi all

    Here is my setup of the system and the version of LAbView:

    • [LabView professional Development System - Version 11.0 - 32-bit]
    • OPERATING SYSTEM - windows 7 Home Premium 64 - bit SP1 [MS]
    • CPU - [Intel Core i7-2600 processor 3.40 GHz]
    • MEMORY - [16.0 GB RAM]
    • GPU - [NVIDIA GeForce GT 530]
    • OR PXI-7951R & NOR 5761

    Everything works absolutely fine until yesterday. All of a sudden, everything about LabView is very slow. Send commands to run the stepper motor and even check on Measurement & Automation Explorer. There have been major changes in the updates or removing software from National Instruments. Everything is intact. My antivirus is AVG (which is a recent change after I removed mcaffee, but I don't think that's maybe the reason).

    any kind of advice is appreciated.

    Thank you.

    Add or change your antivirus/firewall can have a major effect on LabVIEW and related products, because they can potentially block your network connections. MAX scans your network for remote devices.

    I also had a problem a while back where my Firewall blocked access to shared variables. The following instructions helped me: http://zone.ni.com/reference/en-XX/help/371361G-01/lvhowto/createditlogosxt/ (I don't think this affects the MAX, but I mention this in case you encounter a similar problem)

  • problem w / measure automation explore or/o Inst Asst

    Hello

    I'm new to Labview and I work my way through the tutorial. I loaded 3 disks for measurement Automation Explorer such that I can complete getting started with Labview Ch.4, using a DAQ hardware. The part where I place an e/s of Instrument on the block diagram Wizard is taking me to the top. When I double click the sub VI, it appears an error message telling me that "measurement automation explore or e/s Instrument wizard is not installed correctly. I loaded the disc twice up to now, the same questions. No one around the Office can understand it, either. I use discs of Labview dated February 2007, NOR MAX 4.2, 3.0, NI DAQmx 8.5 compliance record. Any help is greatly appreciated. Thank you.

    Hey,.

    Try the solution that is presented in this knowledge base article.

  • How to uninstall corrupted NI Measurement & Automation Explorer?

    Hello

    I have problem to install LabView 2012 and all components, cos does not properly uninstall my version previous (2012).

    I have LabView_Fall_2012 and NI_Device_Drivers_August2012 and Win7 x 64. (try to install it on a partition not OS).

    Now when I try to install new I get "OR Measurement & Automation Explorer 5.3 (superior version already installed)" and installation stops there.

    My goal is to set everything up if possible to properly uninstall and then reinstall it. Now I have managed to remove most of the files, reg entries and stopping services, but some things are still left.

    What can I do for that Setup to deal with?

    Its possible to remove all the files/reg manually?

    There is a file that cannot be deleted ' C:\Program Files (x 86) \National Instruments\Shared\mDNS Responder\nimdnsNSP.dll ", its possible to remove it?

    Thanks a lot for your help

    IMHO, there are two options:

    1. try to reinstall 5.4 MAX, download here

    2. the STI DLL used by the LV you update tool can remove it using this tool: Unlocker

    Stefan

  • Cannot open Control Panel Test VISA in Measurement & Automation Explorer

    With the help of Visa OR Measurement & Automation Explorer. Select the devices and Interfaces and select my device. The Test Visa Open Panel is grayed out and is not available. Some tools-> NI - VISA-> VISA Interactive Control, and the following error message appears-"unable to start the program because it lacks cvirte.dll form your computer." Try reinstalling the program to fix this problem. "I reinstalled the driver NI-VISA and the problem persists.

    Hi David,

    I'm glad that you solved. It could be that the library file has been installed, but that it was somehow lost and unable to be found by measurement and Automation Explorer. A massive compilation could have located this file and placed in the expected folder. I'm curious to know if you did a search system for the library file? I know you have talked about doing one for the file in the document I linked to you. Whatever it is, I'm glad that you were able to get this working.

    Have a great day!
    Jackie

  • NEITHER Measurement & Automation Explorer 5.0 (superior version already installed)

    I have downloaded the driver NOR-DAQmx 9.4, but when I try to install it I get the message:

    Cannot install

    o OR Measurement & Automation Explorer 5.0 (superior version already installed)

    What does that mean? I have LabVIEW 2012 (Windows), version 12.0, 64 bit.

    Thank you

    Neil

    I think what is happening is that DAQmx is installed correctly, but the installer also try to install an older version of Measurement & Automation Explorer (MAX), which is what causes the error.

    You can check that DAQmx has been properly installed by opening MAX and check if it is listed on the software tab on the left.

    One thing to keep in mind is that 9.4 DAQmx is not compatible with LabVIEW 2012, so you can just upgrade to the latest version of DAQmx, which is 9.6 and is compatible with LabVIEW 2012.

  • How to set the Date and time on a remote target in Measurement & Automation Explorer (MAX)?

    Hello

    I'm trying to follow the steps described in the following link:

    How to set the Date and time on a remote target in Measurement & Automation Explorer (MAX)?

    It is said

    The software on the target currently supports only the time zoneand deletes the old files?

    I need set the date and time

    and remove all the old unwanted files

    the calendar was gray does not allow me to change the current date.

    Anyone know how to change

    Hi all!

    According to this the below Knowledge Base article, Remote Configuration System Support must be installed on the target for time settings to change in MAX.

    http://digital.NI.com/public.nsf/allkb/FAB0EC4D6E5EE0F386257738005733A3?OpenDocument&MeTC=max_MAXDat...

    If you don't know what version of the system supported remote Configuration software that you have installed on your device target, it is likely that it is not yet installed. Add this software by clicking below your target to the MAX software option and selecting "Add/Remove Software". This is where you will install the System Configuration Remote Support software.

    Please let us know if you have any questions!

  • How capture and save a sequence continues to images using NI 1422 LVDS card capture, through Measurement & Automation Explorer?

    It is possible to capture and record a sequence of images with a LVDS OR 1422 EZMaker using Measurement & Automation Explorer?  Everything is installed correctly and running and I am able to use the "Snap" command to save an image, as well as the 'Grab' command, but it seems that you must stop the dump command before you can save anything, and it only allows you to save the final image.  Is there a way to record a sequence that is being caught, or should I use LabVIEW or other additional software to do this?

    Kind regards

    James S.

    Hello James,.

    You will need to LabVIEW or another environment development in order to record a sequence of images. If you have LabVIEW and the Acquisition of Vision software, you will be able to capture and save images in an AVI file. There are examples that come with the Acquisition of Vision software that can do this.

    I hope this helps.

  • My MS Office Communicator opens when I click on 'Connect with instrument' on Measurement & Automation Explorer.

    Hello

    Recently I am having problems using the option "contact instrument" on measurement and Automation Explorer.  When I tried to click on 'communicate with instrument' button, my MS Office Communicator starts to open, then an error message appears indicating that "specified address is not recognized or does not exist. With this, I can't use MAX Explorer

    for the purpose of debugging. Apart from this problem, communication between my PC and my instruments are okay during the VI run. Hope you could help me with that. Thank you!

    I see this on two machines in my group too. Bringing this back to the top to see if we can get an answer. What do you think of gurus OR?

  • Firefox crashes while loading content and me must move mouse even though I tried all the fixes, including reboot in safe mode, see Add-ons for V33.1.

    It does not matter what I am trying to load to open e-mail messages to commercial sites to watch cbs programming. The problem started a little over a week and I tried the measures indicated in my question. I also took my computer to a national computer fix - it decision-making and that they didn't know what to do. The problem does not occur when I use Internet Explorer, but I really prefer Mozilla. Can you help me?

    Hello mozart99, several other affected users have reported that it was a problem caused by malicious software on their pc.
    Please run a scan of your system with adwcleaner and the free version of malwarebytes.

    Fix Firefox problems caused by malicious software

  • How to specify the sampling frequency? Must use "measurement & Automation Explorer '?

    I use to measure the input current analog OR cDAQ-9171 (chassis only location USB) and NOR-9207. I have 2010 NOR-installed DAQmx and LabVIEW.

    How can I specify the sampling frequency?

    If I use M & A Explorer to create the task, I can specify the flow rate (Hz) on the Configuration tab-> sync settings.

    For the acquisition of data NOR, it is mandatory to use M & A exploring?

    If I don't want to use M & A Explorer, how can I specify the rate (Hz)?

    Hello

    You can specify the sampling frequency with "DAQmx Timing.vi" located in the function palette DAQmx (read context-sensitive help on how to use wisely).

    You do not have to use M & A exploring (MAX) to create a task.

    A simple and quick way is to use DAQ Assistant (same configuration as in MAX) to configure your measurement.

    Another is to use blocks of DAQmx function to manually build your application code.

    In my experience Assistant DAQ is ideal for simple tasks (one measure), with regard to the more complex measures (synchronized the analog and digital inputs).

    I tend to use function blocks because they give you more freedom about code execution.

    Note: You can also build DAQmx code from a wizard configured DAQ task.

    Best regards

    Matej

  • F1 2009 crashes while loading a race, race test works very well

    I can make the day of the event but why I can´t race? In the middle of loading the game crashes.

    Try Codemasters support (although it is unclear if they will support the 2009 version more)--> http://www.codemasters.com/us/f12012/pc/support/

  • Premiere CS4 crashes while loading a specific project

    I have looked everywhere for the answer to that and I tried everything, I just want to be fixed / without alteration if it is this project file.

    Well so I first cs4 on Windows 7 and its working fine it opens new projects and especially all my "old" in addition to this one project I have. This project is 14.9 MB (I know it's big enough, but I have worked on this for a while now) and each time I have try and open the project to load window comes up and it freezes about 95% of the way. I can still move the project window load but if I press Cancel then first pale and unresponsive.

    It is a common problem I've found, but I have not found a solution that works for me.

    Here is what I tried:

    Restarting my computer

    clearing preferences

    Clear cache

    changing file names (so it brings offline media, but never until now)

    relocation of first (almost worked, now it happens about 95% instead of like 75%)

    Disconnect the external hard drives (one more time to get the window currently non-media, but it doesn't show)

    using a xml key (I do not know to this program but I tried to use it)

    Y at - it a hope? I have no backup automatically usable either I just want to fix this corrupted project that I've been working on for more than a year.

    In case anyone is wondering, I ended up going in a friends House who has CS5 and USB 3.0, and it ended up working. So apparently cs5 or usb 3.0 or the combination of both is the solution. Thanks for all those who have contributed to

  • Firefox crashes while loading images for the first time after windows starts upward

    every time I log on windows, Open firefox and load an image or a site with content firefox flash stops responding for 2 or 3 seconds and then returns to normal after that and everything is fine... it happens only when I open firefox for the first time after starting windows to the top...
    thing I tried
    Start windows and Open firefox in safe mode (does'nt fix the problem)

    updated 13 solved the problem...!

  • PXI-2576-200055 by Measurement & Automation Explorer error code

    Hallo,

    I have a problem on the PXI-2576 switch.

    When I open Explorer - MY, I can see the PXI-2576 devices and also the automatic test and reset to zero, they are OK, no problem.

    But when I click on the button "Testpanel", there is always an error message "error-200055 occurred at DAQmx Get relay Position .vi:1 (multiple)" Details on the Photo.

    Is there anyone know what I need to do?

    Thank you very much!

    Hello

    You can just get rid of this error by removing the special characters in our devicename.

    Find below I have a screenshot where I changed the name of the device for example PXI2576Multiplexer. This works.

    I also tried with the name that you used. ->, it causes the error. -200055

    I hope that the info solved the problem on your machine.

    Andreas K

    OR Germany