1073807339 Timeouterror lecroy wavemaster 806ZI-A

Hi, I have a new driver for labview programming and connections using ethernet cable.
 
I use the driver for labview 2012 for lecroy wave series, and the model I use is as the title.
I keep the error timeout of the 'expectation' attached Subvi. I don't know what causes it.

Connection could be a problem, but I tested it in the test as attached Panel and it seems OK. (I use a connection LXI, I tried to use the TCIP connection, but it couldn't work)

Any help is apperciated, thanks a lot!

Hello Funan,

Also, I got your message from privat but will post it here for other institutions.

You are already connected to the scope.  You have found an error of the VI of WAIT timeout.

The VI queries one registry to see when data are available for you to read.  If the bit is not set within the deadline set on the buckle, it generates the error.

The scope has the arm, you see the event trigger, then process and visualize the wave before data can be read from it.

Here is a link to the manual of the remote control: http://cdn.teledynelecroy.com/files/manuals/wm-rcm-e_rev_d.pdf

There is a section beginning on page 27 of synchronization that will help you.

I use the WAITING; * MUTUAL FUND? the method that is described at the bottom of page 29.

You mentioned that you use LXI because ACIP did not work.

ACIP work of LabVIEW, it does no more than the interactive tool of NI-VISA. (This has stopped working when NEITHER changed the GUI shortly after the introduction of the 64-bit versions of LabVIEW).

Kind regards

Leonard Brown
Technical sales engineer
Teledyne LeCroy
1-800-553-2769

Tags: NI Software

Similar Questions

  • How to configure statistics Px in a LeCroy DSO?

    Hello

    I am trying to automate a series of tests. One of the requirements is to save all the screens with the relevant information. To do this, I need to configure programs P1 to P6 according to my needs.

    I already understood all that I need (save the screen, change the trigger, the scales...) but how to change through LabVIEW, Px configurations.

    To be as clear as possible, I attached and highlighted two images. In the first it is written: "previously set", I want to know is exactly how to do this configuration.

    In another image, the Pxs parameter accompany the red circles: How can I turn on/off, choose from 4 channels and what information (Amplitude, frequency...) through LabVIEW?

    Thank you

    PS: I use LabVIEW 2015 and a LeCroy WaveRunner in the 6000 series, connection via TCP/IP

    Hello Tiago,

    For custom settings, you must use the command of PACU to be implemented, then you can use the query to get results.

    The driver can set those up with all scopes of basic, but there are many parameters that can be added or which are not available on all scopes.  For those, you will need to use the command of the USPA.

    The quick way to find the settings for the command (a good thing for all parameters) is to configure the scope manually then send PACU? : CUSTx (where x = parameter, i.e. CUST3 P3).

    Then read back the response.  The scope will show you the settings you would send to him implement the same measure.

    The etc controls are located in the remote control and documentation of automation for the scope.

    Here's a link to a FAQ that contains links to documentation etc. that will help you get: remote control controls and automation

    Kind regards

    Leonard Brown
    Technical sales engineer
    Teledyne LeCroy

  • Error time-out random - 1073807339 using pilots of series Oscilloscope Tektronix TDS 200 1000 2000

    I am trying to make measurements of waveform of an oscilloscope Tektronix 1052 B using LabVIEW. My program is intended to be run for hours at a time, and it works perfectly as expected for the first few hours. However, I noticed that after a few hours, there are apparently random of chances that a VISA to read or write the block function will produce the error - 1073807339. After that happens once, all read or write functions also produce this same error for a while. This happens even using subVIs new driver provided by OR (http://sine.ni.com/apps/utf8/niid_web_display.download_page?p_id_guid=047216EC20B66FABE0440003BA7CCD... Meanwhile if I stop the program, I can still bring up the oscilloscope in MAX and read and write commands him. The error seems to occur within a VI. I already read this page here: http://digital.ni.com/public.nsf/allkb/874B379E24C0A0D686256FCF007A6EA0?OpenDocument, but none of the solutions helped. I know that this is not a matter of actual timeout because I tried increasing the timeout of 30 seconds, but the error always occurs immediately as soon as the function of reading or writing block is called.

    I would like to know why this is happening and what can be done to remedy this. The Subvi in question is provided. This VI is called at various points in the program do acquire a waveform. The inititialize and narrow subVIs are called outside the Subvi.

    Then check your powersettings. Windows may unexpectedly turn off usb power

  • How to use my 2012 LabVIEW for run the LeCroy Waverunner 610Zi?

    Hi all

    I'm a fresh starter on this topic. I have the 2012 version of LabVIEW and a scope of 610Zi LeCroy WaveRunner. How can I use LabVIEW for control the scope?

    Hello sqrtT,

    I think you and I just had a phone conversation about this, and you're all set!

    I would like to know if I'm wrong and I'll give you instructions for you.

    See you soon,.

    Leonard Brown

    Technical sales engineer

    Teledyne LeCroy

  • 104mxi folder, a transfer of PC-LeCroy to Lecroy HDD using GPIB

    The difficulty to send files to the 104mxi-a lecroy.  There is a file.vi of transfer which takes a file of the scope and saves it on the PC.  I want to reverse that and send a file from the PC to HDD scope. LeCroy has a manual which mentions the TRFL command that should do it, but I think that my syntax is incorrect.

    TRFL DISK, disk, FILE,'' string goes to the vi LCWAVE write out an error.

    Here is the link to the manual of Lecroy:

    http://CDN.LeCroy.com/files/manuals/WM-RCM-e_rev_d.PDF

    Hello

    Try the attached vi.

    You need to replace the LCWAVE write.vi with vi of your driver if it is not already. (depends on which driver you are using, the .llb or the .lvproj).

    If you are having trouble, feel free to contact me at 1-800-5LeCroy, post 6088 (08:00-17:00 Pacific time) and I'm happy to help you!

    See you soon,.

    Leonard Brown

    LeCroy sales engineer

  • Acquisition of data GPIB with oscilloscope TDS 210 tetronix should change to LeCroy wavejet oscilloscope 324

    Hello

    I used Tetronix TDS210 oscilloscope to the data of the Squire. But I need to change the oscilloscope LeCroy Wavejet 324 now. The program is not even read the signal now. What are the things I need to change?

    Thank you

    Quite frankly, I suggest a complete rewrite. You use stacked sequence structures, GPIB functions instead of VISA, a large number of locals, the hidden wires and structures, in a first stage. You should know that the command set varies from one provider to another and often from instrument to instrument. If you had used the instrument to the extent of tek driver, you could perhaps have an easier time to replace common functions. As it is, if you want to use the existing code, you need to replace each of your Scriptures GPIB hardcoded with writing appropriate for the LeCroy.

    As a side note, this is the kind of situation where IVI drivers would have made a big difference. If the IVI drivers exist for the two instruments, and if you used class drivers, you would have not had to change anything.

  • Exception during initialization of Lecroy Ivi driver using c#

    Hello!

    We bought a Lecroy WaveRunner 64MXI-a and tried to control remotely, we have already created a resource of Visa using MAX, tested using the Panel to test VISA and everything seemed ok, we interviewed without problem.

    The problem came when we downloaded and installed the driver of the ivi of lecroy http://teledynelecroy.com/support/softwaredownload/labview.aspx, we have added references as literature has declared and initialized in this way using c#.

    Mylec ILeCroyScope = new LeCroyScopeClass();
    MYLEC. Initialize ("TCPIP0::LCRY0621N67647:inst0:INSTR", false, false, "simulation = false");

    It throws the following exceptionSystem.Runtime.InteropServices.COMException Exception de HRESULT: 0x8000FFFF (E_UNEXPECTED).

    that the exception is not thrown when we simulate = true. And we already tested it using a session simple message in function, but no luck with the ivi driver.

    Has anyone encountered this problem and found a solution? any help is useful, thank you.

    PD: I have already contacted lecroy but no response atm, I'm looking for advice here.

    Hello Nelson.

    Just a tip, I just send you an email response to your support request by email TeledyneLeCroy.

  • LeCroy Waverunner 640Zi - Data Acquisition

    Hello... I'm trying to set up my oscilloscope waverunner with LabVIEW SignalExpress for data acquisition.

    I took the steps so far:

    1 pulse generator hooked to scope of signal generation

    2 USB scope to the installed computer with LabView

    3 downloaded lecroyscope driver 3.2.9 - x 64

    I turn on the scope and plug in the USB to the computer and SignalExpress begins.

    a. start by using data acquisition

    b. Add step/aquire signal / IVI aquire / IVI brought aquire

    c. create new IVI session... resources descriptor (I choose my USB device ' USB0::0x05FF:0 x 1023: 2812N61507:INSTR '), I select the right driver (lcscope), and I do not click enable simulation data, press ok
    d. I still receive configuration errors

    e. did the research... some forum said goto MAX, find drivers and uncheck the Cache and the exchange of check

    f. attempt to initialize... always get config errors.

    g. return to MAX... change to simulate with specific driver.

    h. initialization works... NO errors, BUT no data are acquired.

    Help, please!

    Hello

    Sorry to jump in if I was out of the country for a while and am still catching things in my office.

    I think you are looking for someone to say yes, "you can connect to the scope with NOR-MAX and VISA, and here's how interactive tool do"

    A few things:

    LabVIEW for XStream extended driver is the right one.  It works with all the TeledyneLeCroy Windows based scopes.

    As I see has already been noted. (I'll give Kudos soon), the scope of application must be configured to use interface USBTMC.  To do this, go to the drop down Utitlites on the scope menu and select "utilities configuration... '. "in the tabs that appear at the bottom of the screen, select the 'Remote' tab and make sure that the interface type is set to USBTMC.  This will also show you the VISA resource (I see it in the title of the image of VISA interactive tool indicated in a previous post).

    Once this field is selected, the PC should detect the USB connection and install the device. (you can see in your device manager as a Test of USB and the measurement device):

    Once this is done, you can then enter the NOT-MAX and it will detect and display resources.  You can now communicate with the device:

    If you have problems, do not hesitate to give me a call and I'm happy to walk through it over the phone.

    Kind regards

    Leonard Brown
    Technical sales engineer
    Teledyne LeCroy
    1-800-553-2769

  • LeCroy connection

    Hello.

    I downloaded the instrument drivers, but when starting from the block diagram, I'm going to find the instrument Drivers and then I double click on lcwave (so I guess that the drivers are installed correctly), it shows me a few examples of VI... the problem is that when I try to run these screws in the name of resource VISA on behalf of the LeCroy is not displayed. I have properly it connected with the ethernet connection.

    Any tips?

    Thank you

  • Need to c# for extended Lecroy Wrapper

    Hello developers,

    is - anyone can generate a c# Lecroy IVI Driver 3.1.0.0 Wrapper with

    .Net Assistant driver of instruments, please?

    Kind regards

    v.m.1

    OK, I found my admin.

    It was installed Studio measures, so I could build it by myself...

  • Error reading the attached viSA-1073807339 using Subvi, but why?

    Hello

    First of all, thanks to all who respond!

    I read N9020A Agilent MXA Signal Analyzer data.  It is successful if a Subvi is used to playback VISA; otherwise "Read Error-1073807339 VISA' rears its ugly little head.»  The Agilent gives the message "REQUEST not COMPLETED" just as endearing.

    The on error-1073807339 OR Knowledge Base article talks about the need to put an end to the chain of command correctly and explains how to check it by double-clicking the control of chain of command on the front and changing the display of '-' Display Mode of the Codes.  I call the VISA Read VI in the block diagram; It does not have a chain of command of front panel control, so I don't know how to apply these tips.

    In any event using a Subvi to perform the Read VISA takes care of the problem, but I have no idea why.  The VI image that does not work is "Agilent MXA Sig Gen vi.png."  A big thank you to Mike on the Agilent forum which provided workaround.

    Nina

    Mark, Fan of the crows and any other interested person:

    I found the error in Agilent MXA GIS Gen.vi.  There is typo in one of the VISA WRITE commands.  Originally, it was

    : MMEMORYATA:? "D:\TEST. PNG"

    It should be

    : MMEMORYATA? "D:\TEST. PNG' (remove: before?)

    Your comments have been invaluable.  Thanks a lot for your help and your patience!

    Nina

  • -Error 1073807339

    I have connected the Ocean Optics USB4000 detector and the program worked well.

    But now it displays a Visa Write error-1073807339 of the 'Initialize.vi' & other screws that I used as it is also for driver package.

    I read the previous discussions on this error but cannot get my program to run.

    All solutions?

    Thank you

    Have you unplugged the USB device and, after a few seconds, plugged in? I have USB devices work for weeks and lose their sanity. The only way to reset have disconnected them and reconnected them.

  • LeCroy WaveRunner 64Xi - using LabVIEW

    Hi all

    I want to contact Lecroy Waverunner 64Xi, I downloaded the driver from lecroy vi of our website or (http://sine.ni.com/apps/utf8/niid_web_display.download_page?p_id_guid=078549B4E9E85646E0440003BA7CCD...). I've been through these examples its really useful to know about the configurations.

    But I have a few details in the Communication part. Is it possible to communicate with the serial port Rs-232? If all these sick s VI support.

    And I have another option to contact Lecroy by using TCP/IP Instr. What are all the basic settings I need to do. ? What cable I would prefer. (Crossover cable or straight cable).

    Thank you

    SIVA

    Sorry for the delay...

  • LeCroy LT344L, MAX, IVI and the Ethernet connection

    Howdy!

    I installed lcltxxxx IVI driver of the NOR for the scope of LeCroy LT344L.

    IVI Compliance Package: 4.5

    MAX 5.4.0f0

    I want to use an ethernet connection, but the area of selection "Descriptor resources" does not offer a resource ethernet (see photo).

    With the help of the software ScopeExplorer of LeCroy confirms a working communication between the PC and the scope.

    I do not use the two program (MAX + ScopeExplorer) at the same time.

    Thanks for the help!

    Oh...

    In addition, the 'LeCroy ACIP passport' is necessary for communication via TCP/IP: http://teledynelecroy.com/support/softwaredownload/vicppassport.aspx

    As "Resource Descriptor' the text 'VICP::172.30.2.26' (IP of my scope) works for me.

  • Scope of Lecroy error 85

    I'm starting a project to automate certain measures using Labview 2009.  My Lecroy SDA 18000 is recognized by MAX VISA TCP/IP and seems to send the chains survey welcome back.  I downloaded the LCWAVE driver that is supported by NEITHER.  When I try to run the vi initialization or other I get an error 85 as shown below.  I don't know what to do as all Labview drivers, I used in the past has always worked.  Any help would be appreciated.

    Error 85 has occurred to lcwave Error.vi of playback control

    Possible reasons:

    Scan of string (arg 1) in LeCroy Wave Series.lvlib:Read command Error.vi-> Query.vi Series.lvlib:Error wave LeCroy-LeCroy Series.lvlib:Initialize.vi-> Untitled 1 wave >

    The format of the CMR? the answer is not as expected. Make sure that the command header is not being reprogrammed to the value configured in the default configuration VI Instrument, where it is set to OFF with the CHDR OFF command. (Default as Setup.VI Instrument is called from Initialize.vi.)

    «"" "String of full appeal:»»"»
    Wave of LeCroy Series.lvlib:Read command Error.vi
    LeCroy Series.lvlib:Error Query.vi wave
    LeCroy Wave Series.lvlib:Initialize.vi
    Untitled 1

    Hi, Alan LeCroy Technical Support here...

    The selection "control of the" East on the oscilloscope. in the scope, go to utilities > utility configuration... and then the Remote tab. There are several selections: 2 choices for connections LAN, LXI (VXI11) and TCPIP (ACIP), GPIB and outside.  (The ACIP choice has existed for a decade, LXI is very recent and a likely cause of the problem, given that the pilot is not completely calibrated for LXI. (Even if it is compatible).

    To connect to the scope you describe, you must use the type of connection (LXI).  When you use TCP/IP (ACIP), MAX has no way of knowing that the scope is there.  (MAX is not designed to connect with all types of instruments controlled by LAN)

    Here is a link to an application note that I wrote on how to use the LXI interface; It could be very useful.

    http://www.LeCroy.com/TM/library/AppNotes/LXI/LXI_Interfacing_AppNote.PDF

    The point here is that you must use a VISA connection string that corresponds to the type of connection selected on the scope.

    Best regards, Alan

Maybe you are looking for

  • MacBook Pro/Silver Keys/cleaning of the fans

    My "Silver key" Macbook Pro warms up and I see online there are a lot of videos to clean fan cooling, but not for money key Macbook pro units.  Is there a node where I can see how to do that my system?

  • Boots only to the second power up

    Sometimes I think that usually after a charge, my 8 Stream is not put on. I pressed the power button, then the "HP" logo comes on the screen and it's there, with no arrow whirly. If I pressed the power button again once it goes off, then the next tim

  • Pop ups in the Start Menu: how to eliminate them?

    When the mouse pointer is placed on something in the Start Menu, a popup blocks other elements of the Start Menu.  How can I stop this?  (What characteristic ways of XP!)

  • 15 - 1085wm: System disabled code 63509852

    FOR SOME REASION MY LAPTOP SUDDENLY WANTS TO NOW THIS BIOS PASSWORD WHEN I TURN IT ON. I THINK THAT MY GRANDSON IS THE CULPRIT. I DON'T THINK THAT HE HAS EVEN REALLY NOWS HOW HE DID AND EVEN LESS THE PASSWORD. PLEASE HELP THANKS SYSTEM DISABLED THE 6

  • sharing wireless with Windows 7 and Windows XP, mine is Vista.

    Periodically, I share a wireless connection with a computer laptop Windows XP and Windows 7.  When I connect mine to the internet (Vista), the other two are disconnected from the internet.  How do fix us this?