Noise thermocouple with PCI-6230

What level of noise should be expected for a K type thermocouple used with an acquisition of data PCI-6230 and a map derivation CB-37F-HVD? I read the data with LabVIEW and Daq Assist module on the setting of the thermocouple and the noise see of +/-1 ° C. For a type thermocouple K 41uV / ° C is 3 times the 6230 noise expected (13 uVrms) runs around 0.2 full scale. I think it should be more stable measures or is this a normal behavior?

http://sine.NI.com/DS/app/doc/p/ID/DS-31/lang/en

Thank you

Dan

This level of noise is expected. To improve, you can use a card with a noise filter. A dedicated thermocouple module would also work, but this includes a noise filter as well.

Tags: NI Hardware

Similar Questions

  • DI/O cards such as the 651 x are compatible with PCIe 1 x?

    Hello

    This can be a bit of a stupid question, but it's my first outing with NEITHER and yet I somehow received some amount of responsibility for a very strange procurement process.  I hope the question is easy to answer and that will mitigate its possible idiocy.

    Are maps such as the NI 651 x (http://www.ni.com/datasheet/pdf/en/ds-105) compatible with PCIe 1 x?  Of the image, it looks like they are designed to run on PCIe 16 x, but since the speed is not a major concern, that I wondered if she settle and operate normally on PCIe 1 x.

    I am limited in contracts with a default value, extensive construction of tower PC organization, whose mother has two PCIe 16 x and two PCIe 1 x.  Since I will probably need three cards, I was hoping that it would be compatible with the PCIe 1 x slot and we would save the hassle of special purchase order form.

    Thanks for any help!

    N ° it is even not a PCIe connection.  It's PCI.  I recommend that you call your local representative of OR discuss your needs and they can help steer you in the right direction.

  • docking w520 with pci-e slot, but few likely?

    Hi, is there such a thing? I have a w520 I want to put a raid or hba cards. My t61 had a dock with pci-e slot and I'm looking for something similar.

    no such thing with the ThinkPad W510, W520 W530, new.

  • Acquire thermocouples with 6062E DAQ

    Hello world

    I wonder to acquire a bunch of thermocouples with a map of 6062E DAQ (and SC2345) but I wonder if it of possible or not. The smaller input range is 100mV and with a resolution of 12bits, it gives a resolution in volts of 0.0244 mV or 24uV. If I compare this value with the K type thermocouple NIST table (in degrees celsius):

    1 ° C: 0,039 mV

    2 ° C: 0,079 mV

    So I should get a resolution of less than 1 ° C. Even if this is not the best card for this application...

    Thanks for your help

    Christophe.B

    Hello

    The map got a resolution finer than the sensor itself.

    Your problem must come from your sensor.

    A K-type sensor resolution: 41 microVolts per degree Celsius

    http://digital.NI.com/public.nsf/allkb/8166630BD6965EEA86256FDD0052E89C?OpenDocument

  • IMAQ extracted buffer is suspended in an ongoing acquisition with PCIe-1429

    I acquire images of 500 x 500 pixels at 1000 FPS from a camera of the Basler A504k using a PCIe-1429 (LabView 8.6, x 64 Vista). The images are processed online, and all works - some time. From time to time the program and out of just to extract vi buffer never gets. No error is produced. I need to kill Labview and restart it to leave. I read http://forums.ni.com/t5/Machine-Vision/BUG-with-PCIe-1429-Cameralink-full-config-Basler-sprint-lines... and implementation of the proposed solution, but no luck. I followed the lost images and buffer number called and can't find any problems. My VI is attached. The images, go to a queue and treatment by another part of the program. Any help will be much appreciated.

    Thanks Vinod,

    Another thing is that the structure of your program looks a little broken. You are extracting tampons from the ring, put a reference to the image in your queue, and then going back and extract a different buffer. This means that your previous buffer is no longer checked and could be replaced, but your loop else may not have finished processing this image yet.

    While this should be causing your current problem, but I thought it is interesting to point out. I think I have two queues of the source images. One that contains free images and one that contains the populated images. You producer loop of who you've given here would consume the free queue, take in one of these images, and then place it in the populated queue. Your consumer loop would be remove from the populated queue, process the image, and then put the image in the free queue.

    Eric

  • CAN I USE PXI-8330 WITH PCI-8331

    CAN I USE PXI-8330 (CHASSIS) WITH PCI-8331 (PC)?

    CAN I USE PXI-8331 (CHASSIS) WITH PCI-8330 (PC)?

    CAN I USE OR PXI-8110 (CHASSIS) WITH PCI-8330 (PC)?

    IS CONTROLLER MUST BE THE SAME MODEL?

    Hello

    MXI controllers have a pair use, as you can see the manual on page 1-4.

    http://www.NI.com/PDF/manuals/370840b.PDF

    OR PXI-8110 is a PXI controller, so it's a computer doesn´t need to be connected to a PC using a card-MXI.

    Kind regards

    Abel

    OR Brazil

  • I use pci-6221, I need her to interface with thermocouple with voltage up to 5v

    I use as my pci-6221 or data acquisition card and card 8.2.this labview version gives the constant 10.5 volt signal in at the entrance to analog channel AO on pin 68 and 34. why it shows 10.5 although I did not connect any input.i use type k thermocouple and after signal conditioning with tl0804 I need it interface with AI 0.i channel unaware aware of off the road on the output pins this Card.i need to operate an electric rod that needs 24 volt DC.i give entry to the pins HAVE with variable dc power block after reaching the limit I set(eg:2v) it jumps instantly to 10.5 volts.

    You have your task to acquisition of data configured for the mode differential or asymmetric acquisition for the analog input?

    I don't understand your comment about to connect the pins WITH a DC power supply.  Why is that you connect a DC power supply to the analog input?

    Using an analog output or digital output to operate the electric rod?  I'll assume that you are looking for on/off control.  A digital output is not the voltage or current to drive something that big.  You may be able to find a relay for coil 5VDC.  Check current requirements.  With which you can have the relay to connect or disconnect a power supply of 24 VDC is the actuator.  Make sure you have a protection diode across the relay coil wired, so that the magnetic field of the coil does not damage the analog output of your card.

    Another possibility is to have the 5 VDC output transistor circuit switches the 24 VDC circuit.

  • Synchronization of the signal with PCI-6602

    Hello

    I need to generate three signals that must be synchronized.

    To do this, I use a PCI-6602.

    The first signal is used to trigger a device, the second signal is used to simulate the noise.

    The third signal should wear the noise so that it starts before the trigger signal and stops shortly after the trigger signal.

    The signals are then combined with an external logic circuit.

    (To see what I mean, see the first PDF).

    I use a fourth timer to synchronize all other clocks, but the three signals are beginning to run before starting the timer task. I have this evaluated using not to no and a oscilloscope.

    The trigger and gate (first and last signal) are below 100 Hz, the noise signal is about 200 kHz.

    At least the trigger and the door must be started synchronously, otherwise the waveform is not what we want. The noise signal didn't need to be synchronized.

    I am sure I am missing just a little detail here, maybe you can show me what I missed.

    Please check my screenshot of LabView to see what I've done here.

    If you want more information about the VI, I can take apart a little and send it, too.

    Try using start to the outputs of the 3 meter (see attached VI)

  • Voltage corresponds not counter in the terminals on SCB68 with PCI-6254

    Setup: I have a PCI-6254 measuring 10 channels, differential all connect. 0-4 and 16-19, with SCB-68 2s and armoured and other cables.  Entered 0-4 have pressure sensors that turn off 0-10 V air-conditioned and equipped with the same commune.  The seams of the configuration to work, but I have a strange phenomenon.  When I put 30 lb/PO2 three input I get 6.149 VDC measured with a meter on the SCB-68 and something very close when I watch the raw tension out of Labviw (BTW, 2010).  If I provide 30 lb/po2 at the two entrance he see the voltage according to LV to 6.04 VDC on channel 3.  Tension with the meter is always 6.149 VDC on channel 3.  LV then reflects the pressure went down on this station, but I checked with an external meter does not have.

    I have two things I'm curious about:

    1: why the channels interfere with each other?

    2: why not make the same tension when measured with a multimeter?

    Could there be noise in my cable OR between the CBS and the card?

    Hi MU123 -.

    It seems that what you see is due to ghosting.  Here's a great knowledge base article which addresses this phenomenon and how to avoid it.

    Let us know how this helps out.  Good luck with your application!

  • problems syncing multifunction with PCI-6229

    Hi, I use PCI-6229 and LabVIEW 8.0 for synchronized of the thermocouples and voltage measurements. The diagram is attached here. Here are my questions:

    (1) with the analog inputs of TC is ranked very first and followed with inputs analog voltage, as shown in the diagram, it works fine. But once the tension AIs are placed first and followed with AIs for thermocouples, thermocouple reading makes no sense, for example telling 25 deg C, the program shows strong fluctuating temperature 10-80 deg C.

    (2) the delay time using DAQmx Trigger, here I put the Start.Delay = 2, Start.DelayUnits = ticks. And sampling rate = 1000 Hz for sample clock VI. Then, is equal to 2 X (1/1000 second) delay time = 2ms? If no deadline is set, which is the time between the trigger signal and actural reading? Will it be the time device resolution 50 ns?

    All entries would be very appreciated.

    Hi Matt11,

    (1) the order that you specify strings in your task will affect the order of scanning.  In other words, when you add the channels voltage before channels of temperature, the material can enjoy the tension or s channels followed by the string (s) temperature.  Since the 6229 is multiplexed, it comes to each channel sequentially switching in a single ADC.  In your case, it seems that the tension on the ADC to measure voltage not had time to resolve completely prior to taking the measure of temperature.  We refer to this sometimes as ghosts in extreme cases where the value measured on channel b mimics the tension on the channel.

    I think that you have found that you can get this by adding the TC task first.  You could probably also get the same result by allowing more break-in between channels on your scan list.  You can set the time between the channels with a Timing DAQmx property node (you can set the property is the clock frequency convert, which is the opposite of a waiting time).  For slow sampling rates, sets DAQmx convert a default clock rate up to a minimum of 1/14 US (on the x 622).  If your overall sample rate is less than 1/14 us (about 71 kHz) then you should have room to lower the clock rate to convert more of this if you wish to allow more time to settle between the channels.

    I'm a little surprised that you see errors because of it (generally, the problem comes from a signal source of high imepdance measure or when sampling at higher rates), but then again, the temperature measurement would be very sensitive to small voltage fluctuations since you deal with thermocouple voltages at room temperature.  Given the order of scanning affects your measure the problem is almost certainly introduced leaving not enough time for the ADC to settle after reading your string tension.

    (2) ticks means actually timebase tick rather than the graduations to sample.  The minimum is always 2, but this match 2 graduations of 20 MHz (100 ns) time base from which the sample clock (unless you choose otherwise or specify a sampling rate less than 20 MHz / 2 ^ 32), instead of 2 graduations of the 1 kHz sample clock (2 ms).  In fact the trigger is probably performed asynchronously at the base of your time, if you are anywhere between the graduations of 1-2 the basis of delay time (50-100 ns) according to the relative phase of the trigger to the time base.  If no time limit is set, the default value according to me is still 2 ticks which is the minimum that allows hardware (you can check on your own by reading the property node if you wish).

    These points should be compatible between different driver versions, with the exception that DAQmx chose by default convert rate differently in earlier versions of 7.4.  Select the conversion rate is a compromise between the settling of deadlines and the channels are sampled how "simultaneously" on the MULTIPLEXED Board.  The final result in point 7.4 of DAQmx was a compromise apparently arbitrary add 10 US for the minimum conversion period and use this as the maximum by default between the channels.  The user can of course always set the clock to convert to all what best fits their application (in your case, you might want to allow a settlement more between each channel).

    Best regards

  • R310 Dell: What do the unknown fans with PCIe cards?

    Hello

    I'm looking to replace my aging modified PE2950 gen 3 with a server that is more effective, while maintaining 2 power PCIe slots.

    My use case is not particularly normal in that it's for a home installation and the server will be install re-usable to serve my ad and TVServer for home.  Therefore, I need to be able to add two DVB - S2 cards on the chassis to receive satellite service.

    I know that with some other manufacturers, installation unknown PCIe cards cause their fans go nuts as they assume that the maps are overheated.  How Dell servers in this situation, in particular the R310?  The R310 seems to satisfy all my needs and I think it would be perfect.

    In addition, until I removed the Perc 6 / I of my R410 it was quite noisy, how the R310 Dell price for the fan noise?  The system, with all remains will be in the basement in a rack Assembly away from the living areas. I think I could deal with what up to all about 60dba but especially because it would be too noisy... I understand that 1u rack systems are not designed to be quiet around but a few real comments here would be great.

    Any help that you can provide, or direction is appreciated.

    Thank you very much

    Chris

    Hello

    Bernos
    I know that with some other manufacturers, installation unknown PCIe cards cause their fans go nuts as they assume that the maps are overheated.  How Dell servers in this situation, in particular the R310?

    PCIe cards unknown/3 rd party causes servers 13 g and likely all generations run fans to its maximum SPEED, regardless of the temperature. To my knowledge, it is new with our servers to 13g. I don't think previous generations like 11g detected what exactly was settled in the PCIe slots. They changed fan thresholds based on whether or not the slots have been filled not based not what was in the slot.

    I can't say with certainty what will be the decibels. We do not have a list of all the possible configurations and the decibels. The highest specified configuration in my Datasheet does not include additional PCIe cards, but the dBA listed under stress with this population is 43. I can't guarantee that the system will operate less than 60 dBA, but I suspect that it will do so.

    Thank you

  • Incompatible NB550D with Advanced - 6230 N?

    Hello
    I replaced with an Advanced - 6230 N Intel Wifi card. After that the computer does not start.
    If I remove the Intel card, then it boots without any problem. It seems that my NB550D doesn´t as
    This new card at all.
    No clue on how to make them work together?
    If it is not possible, any Wifi card job with 5 GHz to recommend?

    Thank you
    P

    It seems that the BIOS has a whitelist. Can someone, especially Toshiba, confirm if the NB series computers have a BIOS with whitelist? In this case,.
    It would really help to know what are the WIFI cards are allowed to be used.

    Thank you
    P

  • Satellite C750-A103 made a lot of noise - problem with the cooling system

    My toshiba machine made a lot of noise... like a plane. And it gives me the message "a problem with your cooling system has been the detection.

    Turn off immediately and come back for the service "."
    What should I do? It has been recently formatted, also?

    + The message was edited: mode number + added

    You did not provide details on your laptop model and system that is installed on the laptop.

    Even his I don't know how old is the laptop, but according to this short message, it looks like there is a problem with the cooling system.

    Probably notebooks cooling fan doesn t work properly and the unit cannot be cooled as it should this can lead to cell damage through an internal temperature longer.

  • With PCI-6229 NMI Parity error

    Currently, I'm trying to get a PCI-6229 to work with one of our lab computers. We'll use LabVIEW 8.5.1 with DAQmx 6.7.1 to program the card. Using our measurement VI owner, the computer will crash at some point (not really in reproducible way) with a blue screen saying "NMI: Parity Check / Memory parity error. All the tests of auto and examples of NOR (delivered with LabVIEW) perform well however.

    Most likely, this problem is not a RAM issue (MemTest I tried, different memory chips, different memory banks). Also, the PCI computer bus works very well otherwise (for any other acquisition cards for example). Annoying enough, the card seems to do a great job in an old and slow computer. Which is maybe something to look at? Or are there other ideas?
    Our workplace is a Fujitsu-Siemens of Celsius M460 with BIOS version 6.00 rel. 1.09 running Windows XP (32-bit version). The software to use with the card is LabVIEW 8.5.1 with DAQmx 6.7.1.

    Dear Lutz,

    I could reproduce the blue screen, and
    changed your program in order to avoid the race condition. What is
    actually happens is you want 2 tasks over time the
    AOSampleClock as a source before the task of AO started.

    When
    programming as in the example I sent you (see attachment VI), you
    shouldn't deal with this blue screen problem more. However, I recognize
    that such a thing should not happen even when it's 'bad' programming... I'm going to
    inform the development so that it will be corrected with the next DAQmx
    version of the driver.

    Merry Christmas,

    P. Lawrence

    National Instruments

  • Triggers the analogue output with PCI-4461

    Hello

    I'm trying to generate a signal of analog output triggered with a card PCI-4461. First I tried to use the feature OR DAQmx 'start analog edge' with the way analog input AI0 as the source and the channel analog output AO0 as task. After it gave an error that I tried to use the NI DAQmx 'start digital dashboard' function with PCI0 as source and channel of analog output AO0 as task. It ran, but did not produce any output. Now I wonder if I can use the trigger analog or digital of the PCI-4461 to all of the output.

    Thanks for support you,

    Pribislav

    Pribislav salvation,

    you still have this problem? I did exactly the same configuration (power play) and it works fine on my system. The PCI-4461 does not support analog triggering, that's why this error occurs.

    Kind regards

    Michaud

Maybe you are looking for

  • Cancel the current order number

    I am struggling to cancel 3 order delayed due to no any payment method suitable for me. Therefore, I need rgarding Skype Customer service team to help me cancel these 3 what order does not, [withdrew privacy], [removed for privacy] and [removed for p

  • Service Pack One with questions by e-mail

    I installed the service pack successfully & now I can't access my outlook mail.  I get an error from the pop3 server.  My provider says that they are not to blame?  The last time I got the email was just before the service pack was installed.  Any he

  • Green freeze screen when loading apps on Installer

    A new E5-571P-tl55 received yesterday and going through the installation process I got all the way where it says download apps. It displays a message that he dealt or I forgot what he said exactly. Then the words went away after about an hour. I reme

  • password reset disk problem

    OK, guys, no one responded to the question.  Can created a password reset disk in drive "D" on a CD?  The wizard suggests he can as he lists a CD as a removable storage device, but try as I might I can't get it to create a CD.  Sorry, but I do not ha

  • How to copy CD to computer

    HELPPPPP how can I copy CDs to the computer? for some reason, I don't have the opsion to do...