Recommendation of USB data acquisition

Hello

I'm looking for a digital acquisition of data NOR-USB 12v e/s of at least 8 channels. He would be grateful if someone recommends me.

From what I can tell, the USB-6525 should work for you.  It can accept a digital input of up to 60V (anything over 3, 2V is regarded as a MAXIMUM).  And then there 8 SSRs to serve as exits Digital.

Specification of the

Tags: NI Hardware

Similar Questions

  • 6361 versatile usb data acquisition does not automatically detect the SCXI chassis/accessories

    versatile usb data acquisition 6361 will detect automatically SCXI chassis and accessories?

    Hello kdCMC,

    As far as I know that the SCXI-1600 USB Module is able to auto-detect SCXI Modules.
    This is also mentioned on page 11 and 12 of the following SCXI to start document:
    http://www.NI.com/PDF/manuals/373236m.PDF

    All data (including USB-variants such as the NI USB-6361) have only a 68 pin shielded cable (IO) between the DAQ hardware and the SCXI chassis.
    This cable does not auto-detect opportunities since it basically just "transfers analog and digital signals" between SCXI chassis and DAQ hardware.

    Is there a specific document that created the confusion on this subject?

  • Exchanging USB data acquisition systems

    Hello

    I'm looking to use the same program in labelled while sharing the acquisition system of USB-6210 data with the same program. I use labeled 8.5.1 and have a voltage reading program. What I find, however, is that the .vi program does not seem to acquire on the software side, when I disconnect the USB-6210 I use and plug in another, but the same hardware system model. The Green LED on the material will begin to Flash, but there is no entry in the charts of the program. To measure the voltage, I use the function block "DAQ Assistant" to connect to the analog channels in this system. To work around this problem, change the DAQ Assistant function for the device interchanging at work (delete and use the wizard to generate a new DAQ Assistant). It's quite heavy but doable, however, the hope is to apply this program and equipment to several sites, each with their own system of fittings. Is there a way to exchange the hardware systems, but keep the same program so that it might be a way around this? I also used the Application builder to generate a stand-alone program for it but the same problem occurs. Is there any recommendations for multiple devices, or am I missing something? Thank you very much.

    Whenever you connect a new device, it gets its own device ID.

    from Dev1, Dev2 etc..

    So either rename in solution of measurement and of Automattion (MAX) Explorer or use another parameter in the daq assistant.

  • How to synchronize the analog input and the output of two different USB data acquisition boards

    Hi all

    I have two tips very different USB NI USB 6008 case, which I use to acquire the data (analog input) and a USB of NI 9263 is a output analog only site I use to route a signal (in this case a square pulse). The reason why I use the outputs analog 6008 is because I need to deliver negative tension and need the full +/-10 v range.

    Looking at similar positions, I'm pretty sure that I can't use an external trigger or a common clock, I also tried to use the timed synchronization of the structures but no cigar.

    I'm including a quick vi I whipped showing how the jitters because of the lack of synchronization signal. The OD of the 9263 connects to AI in the 6008 in this example.

    I talked to a specialist in the phone and tols me that's not possible.

  • Size of USB data acquisition 2008/USB 2009

    We have several USB 2008 Renault in my company, and we plan to go to a USB 2009. We use them in some regions of space is counted rather however, and I was wondering if there is a difference in the physical size of 2009 compared to the 2008 USB USB.

    Thank you

    There is no distinction between 2008 and 2009. Could you be confusing a with a 6 2 and talk about the 6008 and 6009? If so, the dimensions are in the User Guide , which is something that comes with the units.

  • How to know the number of USB data acquisition device

    There are 2 devices and user can plug in different port each time.

    But the program uses the same string of device and may have errors.

    They can auto detect and generate the device chain, thank you.

    Is this something similar to this (using) you need? If so, please go from the link to Andrew post (a 2nd). Post if you need something else.

    Thank you

    Mathan

  • I am building a system and I need to help experts to choose the data acquisition card.

    Hello world

    I am currently working on a design project I need to take a data acquisition card that responds to the output of the amplifier. I'll build a force detection system which I have taken the following as detector and amplifier:

    http://content.Honeywell.com/sensing/Sensotec/pdf_catalog08/008834-1-EN_Model_UV%20Bridge.PDF

    http://content.Honeywell.com/sensing/Sensotec/pdf_catalog08/008638-1-EN_Model_53.PDF

    Please recommend me which data acquisition card I could take. The only requirement I ask is to have less than 0.1% global uncertainty.

    Thanks for anyone who can help

    Since your amp has - 3db at 7 kHz and you need only 10 bits 0.1% will do almost any card with.   USB-6009 is the lower end. USB-6210 the next step.

    Then take a look at some other boundary conditions: do you need it calibrated? (Is there a QM adj. who could ask questions?) Space?, EMC/noise, entry, differential

    Take a look at the USB-9237 with a conditioner bridge construction...

  • USB multi channel data acquisition

    I use the VI to arduino HC-06 bluettoth of data acquisition. I would like to acquire data multicahnnel

    Could you suggest me how to acquire multichannel data from USB? To do this, how I have to program the arduino?

    The crux of your property doesn't do anything since you not write anything.

    Why you set the number of bytes to read at 8? You send more than that. You should be able to set it to a high number, such as 1000. Why do you use a dbl for this control?

  • USB 6008 data acquisition: automatically turning in a port

    Hi all someone could please help me.

    I need to autoamtically send output didgital since my data acquisition based on a value. the bottom of my little project. I read in a database if a value is equal to a certain value, I want to send a digital camera of the signal of acquiring data in a relay, so for example if the value 1 database generates a digital signl 0 7 port line.

    If someone could help it would be much appreciated

    So what's the problem with sending AUTOMATICALLY one. You simply compare the value of the database. A comparison function returns a true or false, which is quite ideal to display a true/false signal.

    Fix the code you wrote. This is not a service of howework and see some code real goes a long way to get detailed help.

  • Z600 Workstation: workstation Z600 missing of pci data acquisition and signal processing controller driver

    I just installed Win7 Pro x 86 on my Z600 workstation.  It was an upgrade to Windows XP Pro.  I'm missing the driver for PCI Data Acquisition and processing of the Signal and Hardlock USB 1.02 controller in Device Manager.  Can someone help me find these drivers.  I looked on the page of Support/Drivers for this model, but could not find anything.

    Hello

    You can get assistance on the HP Enterprise Business Forum since you have a professional worktsation.

  • motion control for vertical actuator and data acquisition

    Hello

    I am a researcher (a branch of civil engineering) geotechnical engineering and I have very little knowledge about the acquisition of control and data motion, so would need a lot of help from the experts OR. I have only knowledge base on these 2 aspects based on my reading of some materials on the Web site of NOR and youtube videos, so I hope that you bare with me . Here are my questions:

    I am trying to build an actuator which will be used to push a probe (a penetrometer with a load cell to measure the resistance of a soil sample), resembling the concept, photography in the attached file. I need to have these criteria for my system:

    (1) actuator, which can push the probe at speeds between 0.01 mm/s - 300 mm/s with precision and move the probe cyclically (upwards and downwards) in the vertical direction
    (2) load expected on the probe into the ground range: 0.02kN - 6 kN.
    (3) necessary to get the load cell load data and the speed of the probe.

    4) able to control the actuator to a PC (speed and posotion) and monitor data from transducers and data log time even the transducers.

    Guess my beginners is that I will need:

    For orders:
    (1) software - LabVIEW and NOR-motion assistant

    (2) controller - NI PCI-7342

    (3) driver/amplifier - analogue servo AKD Drive

    (4) motor - motor brushless servo AKM

    For the acquisition:
    (1) software - based LabVIEW development systems

    (2) amplifiers or other device - no idea what type on the conditioning of signals

    (3) data acquisition device - no idea what type

    Since I'm a beginner, is - that someone might recommend components (hardware and software) for the control and data acquisition. I'm on a tight budget, so I thankful if someone could help me to recommend components good enough to build my system.

    Thanks for your help.

    At these rates, you will need to run the sensor for the cDAQ.  You can configure the analog output on the Tritex nationally on the position.  There is an adjustable filter that you can set in order to get a clean enough to 300 Hz signal.  When you learn about the Tritex, make sure that let you them know what comms and e/s that you want to use.  If I remember, not all options have worked together.  The analog output may need to be my, but you can put a resistance through the acquisition of input data to get the voltage instead.  I don't remember all the details.  You should really not too much on the Tritex/LabVIEW side.  You will send your movement parameters (beginning of end of race, speed, position, accel, cut), and if you cycle (I believe you) or simply running in a loop.  You could also just be able to use the functions of jog.  When you get close to knowing exaclty what you need, PM me and I'm sure we can work something out with the drivers.  You need only the basics.  In fact, you could probably do this all your movements via digital and analog i/o.

  • LeCroy Waverunner 640Zi - Data Acquisition

    Hello... I'm trying to set up my oscilloscope waverunner with LabVIEW SignalExpress for data acquisition.

    I took the steps so far:

    1 pulse generator hooked to scope of signal generation

    2 USB scope to the installed computer with LabView

    3 downloaded lecroyscope driver 3.2.9 - x 64

    I turn on the scope and plug in the USB to the computer and SignalExpress begins.

    a. start by using data acquisition

    b. Add step/aquire signal / IVI aquire / IVI brought aquire

    c. create new IVI session... resources descriptor (I choose my USB device ' USB0::0x05FF:0 x 1023: 2812N61507:INSTR '), I select the right driver (lcscope), and I do not click enable simulation data, press ok
    d. I still receive configuration errors

    e. did the research... some forum said goto MAX, find drivers and uncheck the Cache and the exchange of check

    f. attempt to initialize... always get config errors.

    g. return to MAX... change to simulate with specific driver.

    h. initialization works... NO errors, BUT no data are acquired.

    Help, please!

    Hello

    Sorry to jump in if I was out of the country for a while and am still catching things in my office.

    I think you are looking for someone to say yes, "you can connect to the scope with NOR-MAX and VISA, and here's how interactive tool do"

    A few things:

    LabVIEW for XStream extended driver is the right one.  It works with all the TeledyneLeCroy Windows based scopes.

    As I see has already been noted. (I'll give Kudos soon), the scope of application must be configured to use interface USBTMC.  To do this, go to the drop down Utitlites on the scope menu and select "utilities configuration... '. "in the tabs that appear at the bottom of the screen, select the 'Remote' tab and make sure that the interface type is set to USBTMC.  This will also show you the VISA resource (I see it in the title of the image of VISA interactive tool indicated in a previous post).

    Once this field is selected, the PC should detect the USB connection and install the device. (you can see in your device manager as a Test of USB and the measurement device):

    Once this is done, you can then enter the NOT-MAX and it will detect and display resources.  You can now communicate with the device:

    If you have problems, do not hesitate to give me a call and I'm happy to walk through it over the phone.

    Kind regards

    Leonard Brown
    Technical sales engineer
    Teledyne LeCroy
    1-800-553-2769

  • Data acquisition tool NOR-DAQmx with Matlab R2012a

    Hello

    I'm trying to control NI USB-6211 of Matlab 2012 using NOR-DAQmx Data Acquisition tool:

    http://zone.NI.com/DevZone/CDA/tut/p/ID/3005

    I'm working on win7 64 bit. And I see the device AND Measurement & Automation Explorer.

    The tool does not work: DAQ_Demo_Browser do nothing. And I got the error "unexpected or unbalanced parenthesis or support" of AcqNUpdates_nonUI.m

    What is the problem?

    Thank you and best regards,

    Arthur Shulkin

    Hi Arthur,.

    Tools OR DAQmx for Acquisition of data with the Software Inc. MATLAB® from The Mathworks, supports up to the 2008 version of the MATLAB® software.  In order to use our products DAQ Multifunction with MATLAB® software, you could get back to 2008 or earlier, or instead use the Data Acquisition Toolbox provided by The Mathworks, Inc.

    Another option would be to import your ".m" files in a node MathScript in LabVIEW and use the functions of NOR-DAQmx everything in the LabVIEW development environment.  For more information on the Module LabVIEW MathScript, you can consult the information available on this link:

    Inside of the LabVIEW MathScript RT Module

    MATLAB® is a registered trademark of The MathWorks, Inc.

    Katie

  • Configuration of the two of the same model data acquisition

    Hello, community of NOR.

    I'm an intern in mechanical engineering with experience in base with LabVIEW.

    I would like to speak to OR directly on this issue, but I don't have a service contract and my company wants me to understand this before you buy LabVIEW.

    I hope that someone has experience about my question, and I would be very happy to help.

    We intend on purchasing an expansion card for our acquisition of data (OMB-USB-2416), but unfortunately it is offline and no custom would not happen in time.

    So, I need LabVIEW to read voltage HAVE two of the same model of data acquisition, which would amount to about 30 channels.

    Is this possible with LabVIEW?

    Thank you.

    Measurement computing says that the "physical channels" dropdown list is automatically filled in once both devices are configured.

  • When I send a 5v of data acquisition, the voltmeter indicates 5V, but when I connect a "normally open" valve Parker for the acquisition of data, the voltage decreases automatically a.14 V. It seems to be grounded. Why is this happening?

    We send 5v data acquisition using a voltage generator. Hook us it up to a voltmeter and see 5V. When connect us the generator voltage to a valve "normally open" parker, the voltmeter indicates .14V. It seems that when we connect the two sons of the valve for the voltage generator, the son act as pattern. We want to control the voltage flowing to tap through Labview. We checked the wires to the valve and they work very well, because if we send a constant 5V since the acquisition of data and put ashore, she, the voltmeter indicates 5V. Someone knows why the son act as pattern and low blood to .14V?

    nsatpute wrote:

    Our data acquisition is NI USB-6259. The valve requires only a 5V max and our DAQ provides up to 5V. However, after connecting the valve to the acquisition of data, the grave tension to almost 0. We start from the principle that the son somehow act as the reason, but we are not sure if this is the case.

    The question here is not how much voltage the valve wants, it's the current needs of the valve.  The 6259 can put only 5mA via an analog output.  Your very likely tap needs much more than that.  If you need to add in an amplifier circuit that can supply more current to operate your faucet.