State of relay by default USB-6525

Hello

When the NI USB-6525 enclosure is powered on the default state to solid state relay is closed.  I'm trying to use it to control a 4 position RF switch.  Obviously we do not want to be on as this disrupts the 4 switch inputs.

Is it possible to change the default state of the relay, so that when the device turns on the relay are all open rather than closed?

Hi SeanJ,

Sorry for the delay on this point - I just noticed this post by searching for clients who may have encountered this problem.

From DAQmx 8.9, programmable start-up States have been added to the USB-6525.  However, flash on these cards was not initialized correctly to 'open' State, so the check for PPS caused tips have added a default PPS to 'closed '.  The solution is to simply set the States start to their desired state.  This can be done through MAX as follows:

All the councils that are currently on the market have their flash correctly initialized to the open State.

To reduce the risk of this question anyone adversely affecting during updates, we also add a check for the firmware in DAQmx 9.2.2 and later charger to initialize the flash if you are upgrading a version of the driver without PPS.  If the upgrade to a version of the driver with PPS, the startup States will not be affected.

More information can be found here:

Behavior of the start state Programmable NI USB-6525

I apologize for any inconvenience that you please do not hesitate to let us know if you have any questions.

Best regards

Tags: NI Hardware

Similar Questions

  • measure the State of the output to USB-6525

    Hi, is it possible to check the status of microsatellites in a USB-6525 using a multimeter.  I hooked up a meter to measure the resistance through the P0.0A and P0.0B output connections.  I wanted to use the automatic test routines to check the SSRs were closing.  As expected, there was an open circuit when LIFE expectancy was 0, but the circuit seemed not close when I changed the TTL of 1.  Do I need to connect to a power source and load to be able to see the closing SSRs?  Is not which is what a multimeter?

    Thank you, Nick

    Nick,

    Multimeter will not provide the source of tension and support you need.  Check out this link.  http://digital.NI.com/public.nsf/allkb/B0FF01B2AFFA52FC862573A2005A9570

  • I can parallel two solid state relay ports for a maximum current of 1 AMP USB-6525?

    Hello

    I plan use USB-6525 to change a PSU 7VDC/1 AMP DC.

    USB-6525 having only 500 my maximum current for each channel, I can two parallel ports of relay to solid state for a maximum current of 1 AMP?

    Is it safe to use in this way?

    If not, is there any other USB device can provide 1 channel AMP relay?

    Thank you

    Ben

    Hi Ben,

    Yes you can do this, please refer to the KB, it contains the information you need

    Can I use several relay on one device to allow power to pass?

    Kind regards

    Margasan

  • Watchdog USB 6525

    I need to return my USB-6525 to state results course in case there is a problem of communication software.

    Basically, I have sent close the relay for a short period s.5 a.25 seconds then the relay open. I can do this by writing a loop for close and then delay for the milliseconds and then open the 2nd iteration.  It works fine except the relay must the software or computer in bulk communication unit would remain shut down and cause a dangerous situation.

    The literature indicates you can do this by setting a watch dog timer but I can't find how to configure it. The way I understand, once the watchdog is set if it does not receive a software reset within the time limits, it will define the output for which is determined the safety position.

    Someone at - it an example of configuration of this device.  Either way, I would also do this with a SCXI-1163R also module.

    Another solution would be to send a pulse on a specific length of time, but I do not think that the 6525 allows execution of timed material.

    Thanks for your suggestions.

    Randall

    There are DAQmx watchdog timer screws if you have the driver installed DAQmx (installed device drivers), you can easily configure and use this timer. Their screws is very explicit.

    http://zone.NI.com/reference/en-XX/help/370469AF-01/lvdaqmx/mxcreatewatchdogtask/

  • Turn it on and an Opto SWITCH with USB-6525

    The attached photo, a composition of a circuit consists of: Mitsubishi PLC FX series, USB-6525 and OPTO Switch (PNP OTBVP6). In the code, I put under tension of the OPTO then cutting the voltage; power on/off, using the USB-6525.

    My problem comes when that at some time in my code, the OPTOMETRISTS remain without electricity (immediately after programmatically, I just

    They shall ensure with outputs USB-6525 high level). I already checked the code, in my opinion its seems fine. I have in mind the 6525 isn't a fast DI / DO peripheral, that's why I use delays above 200ms.

    The last hope is to change the USB6525 for a brand new, it's what I'll do. But if anyone see a problem with the code or has another explanation why the OPTO not to receive or be (somehow) not powered, I'll really appreciate your help.

    I also enclose the .VI

    I really need help.

    Thanks to all in advance.

    I found what was going on. Connectios, common, everything was fine. In the code, between the States (I use a state machine structure) missed me a delay of 200ms. This because I change the outputs to reading entries, so between them missed me a delay, that's all. in any case, thanks for your help!

  • USB-6525

    Hello

    At first, I want to control a USB-6525 in labview. I want to creat 3 switch control relay as if switch1 is out of relay 1 is open, switch2 is narrow so relay is close

    Can I control switch using the DAQ (Labview8.5) Wizard, I choose relay wich I want to open/close, but I don't know how to connect with a switch for failing to enter the DAQ assistant.

    Secondly, I would like to control them by timers but I first want to understand how it works with the simple switch

    Thank you

    Kind regards

    PS: I'm french so sorry for typos

    Please answer me

    I'll try that

    Kind regards

    Rodolphe

  • USB 6525 6501 digital for output to the step motor

    Hello

    I try to use USB 6501 or USB-6525 out of step motor signals which command the stepper motor. My questions are

    (1) do I 6525 USB, I'm not sure the function of it (perhaps as a relay).

    (2) now I connected input 5V for USB-6501 "+ 5V" pin and GND to pin "GND". On the other side (output side), I connected ' enable '(from motor drive) to P0.0. 'direction' to P0.1 and GND to GND.  Can I use the express signal to test, the error says "lack of entry."

    (3) I guess the next step is programming labview. Does anyone know of similar examples?

    Any help would be appreciated!

    Melody,

    If the engine must input external logic level I advise to use the USB-6501, which is just a digital I/o card. The USB-6525 housing does not have the digital outputs to control your motor drive. If I understand correctly you just try to turn the motor on and off with a digital signal. It seems that you also provide your drive motor + 5 v and GND. The USB-6501 has channels for + 5V and GND. I've attached an example of navigation that controls the outputs digital using DAQmx and LabVIEW. This specific programme allows to control 8 digital output lines, but it looks like you don't have one. If the engine waiting for you just a strong to put logic in operation and a logic low to shut down this program example will be able to turn on or turn off your engine. Just connect one of the USB-6501 digital output lines and then use the program to this line of control.

    I don't really know any reason, you need to use the USB-6525 it seems to me that the USB-6501 run action you need. I hope this helps.

  • USB-6525 NiDaqmx how to detect where the device interrupts the communication

    This could be considered a more generic question for NiDaqmx, but I write a custom control for internal use with the USB-6525 and possibly other devices of i/o using c#.  I use DigitalChangeDetectionEvent to warn me when an entry has changed and everything seems to work very well.  In an effort to be complete, I decided to unplug the USB cable and see the result and found that no error has been detected until I tried a writing output.  This is not acceptable because under normal conditions abstraction entry begins a process that will change output - given that admission never happens, the system is basically hung.

    I tried many things and have spent more time on what I want to admit so I would like to hear suggestions on how to address this problem.  Here are my attempts that I am not satisfied:

    (1) periodically switch (pulse) output that allows to detect the problem, but I may not be an available exit, so this isn't a good situation.

    (2) periodically trying to read the input byte with the same entry workflow, this does not work because I believe that the other player is still active with the DigitalChangeDetectionEvent.

    (3) attempted to take the message of Windows to change hardware, but seems too complex.

    (4) I read on the guard dog, but not sure that it is available on the box USB-6525.  So before I start with this I'd hear rather suggestions of others with more experience.

    Any suggestions please?

    THX, Ed

    Hi IntegroEd,

    Those are all very innovative way to potentially address the problem, and I'm sorry you haven't seen success with them.  Fortunately, the typical way to capture a break between a device of data acquisition and monitoring system is indeed a timer, and the 6525 has the innate ability to use the feature, as seen on page 5 of the user manual.  I suggest you go that route, if another user does not post a peripheral solution specific to your case.  This is a documentation on the implementation of the monitoring-

    http://www.NI.com/product-documentation/14616/en/

  • NI USB - 6525 Labview lib

    Anyone know where I can get the labview for the NI USB-6525 device drivers. I searched but Idk if the machine is new and the driver do not yet exist. If someone has a link please let me know.

    You have installed DaqMX?

    http://www.NI.com/download/NI-DAQmx-14.1/4953/en/

    http://www.NI.com/dataacquisition/nidaqmx

  • How to use the NI USB-6525 digital filtering feature?

    Hi can someone tell me how to use the 'digital filtering' feature of the NI USB-6525 device? How can I set up the device to turn on? Using LabVIEW or NI MAX?

    It is configured using a DAQmx property node:

  • control solenoids with USB-6525

    Hi, I'm new to Labview and try to use a USB-6525 control six independent valves.  They'll be turning on and off at different frequencies to control a cell infusion system.  I drew a diagram that allows me to program schedules that the CRDD is closed and open.  The routine is inside a while loop to repeat until a given time has elapsed.  I can duplicate this six times to allow each valve must operate independently.  But I would like to trigger all tasks at the same time.  How do I synchronize loops so they all start at the same time?  Simply put all the loops inside a conditional loop, or it will run the tasks in the order?

    Thank you

    your synchronizeyour can while loops with the help of tax filers, please see the attachment.

    Barros (Certified Developer partner)

  • Meaning of the States LED tricolor on device USB-232

    It's more curiosity, rather than a problem.  I noticed that the devices USB-232/4 most recently bought that we have received seem to now have the status of Panel before LED that show * 3 * States (4 If you include "LED off"!) - yellow, green and Red now.

    I have not found anything whether in literature online to describe something as mundane as what led.  I always assumed that yellow and green are the indicators of mark/space.  Sometimes now they light up briefly in red, but I have not found to correlate to any abnormal operation.

    What is happening with this?

    Dave

    Hi David,

    I found the following page in help OR Serial.

    I hope that answers your question!

    Thank you

    Scott M.

  • Need help with USB-6525 change detection


    6525 support change detection is added in DAQmx 8.9.

    You can download the latest DAQmx since http://joule.ni.com/nidu/cds/view/p/id/2034/lang/en

  • Change the audio output default usb headset speaker 3000 Microsoft livecaht

    Yesterday, when I ran my Rosetta Stone language software that I heard noise coming from my newly installed Microsoft Live Chat 3000 USB headset, today, when I run the software, the microphone still works. But the sound comes through the speakers of the laptop. I looked through Windows and cannot find a place to tell windows to use the headphones to the audio output.

    I saw a similar problem on this site and followed the recommendation to remove the Chuck and download the latest driver on your part and re - install but still the sound comes through the computer speakers.

    Can my children re-configured something but he does not confess to it. But I do not understand why remove then re-loading the drivers does not work. I use Windows XP SP3 entirely under license and updates.

    I have them just a couple of days ago take their back and try something else?

    Thank you

    Try Control Panel, sound, properties, change the headset speakers, also, R.click

    the President of the clock opens, properties, adjust.

  • HELP and warning: Exchange ActiveSync email account disables the WiFi, Bluetooth, Cam and default USB!

    You establish a new Exhanche Business Account Provisioning System of the TPT Lenovo ActiveSync connection turns off WiFi, Bluetooth, camera and USB connections, even it is not caused by the commissioning of the company server. I trink Lenovo has to fix this emitiatly or the opportunity to randomly the profile settings of the Provisioning Manager in TPT.

    LENOVO IF YOU PLEASE FIX THIS OR LET US UNINSTALL YOUR SECURITY SYSTEM WHICH DISABLES THE FUNCTION NEEDED TO WORK ON TPT.

    By removing the Exchange account, the functions are turned on again, but it id not not possible TO use you business EAS account as it should!

    There is no solution the TPT us will BE useless...

    Best regards
    Pasper

    Quick solution myself:

    you will need to edit in the file following with the name "yourname.lenovoconfig" in the second line the uuid, you can see under settings / location & Security / Configuration profiles. The uuid is the alphanumeric code long of the parameters that was installed by the exchange server. You are able to also change th eDevice settings above, I blocked Dataroaming for example.

    
    
    http://www.w3.org/2001/XMLSchema-instance" xsi:noNamespaceSchemaLocation="lenovoconfig.xsd" uuid="xxxxxx-xxxx-xxxxx">
    
    
    Sample policy Manual yes no
    Allow Allow Allow Allow Block Allow Allow Allow Allow Allow Allow Allow Allow Allow

    What to do step by step:

    1 configure the Exchange account

    2 Note the uuid of the Exchange Server in the settings

    3. delete the Exchange account

    4 adjust the uuid and your settings at the top of the source code and send it to your Lenovo over e-mail - for example your googlemail.

    5. click on the lenovoconfig file and install the Lenovo-Profilmanager

    6. new to connect to your Exchange Server.

    7. now, everything should work how implement you the Lenovo Support contains a sample file with all the setting options which could be very useful. My settings are only a small facility to come around the problem. Hope to train here all I've been able to help.

Maybe you are looking for

  • Not compatible with the anniversary 10 Windows Update software Dynadock

    Anniversary Edition 10 Windows tells me that the Dynadock software is not compatible, and windows was freezing since I've updated.Y at - it a fix for this?

  • Time Machine broke my office

    My story I had just to edit a picture on my desk at the right size. "Not much". I thought "I'll just use time machine." I want to use time machine, but my macbook pro crashed and had to restart. I went back to time machine and tried. Files said they

  • bluetooth HP envy touchsmart problems - can't activate bluetooth

    So I recently did some updates on my computer and even update my Itunes and rebooted my computer as soon as I could no longer my bluetooth to appear or work (so I have no mouse) it has already happened many times and I've managed to start this time t

  • Impossible to find in LabView DAQ assistant

    Hello! I installed the driver NOR-DAQmax 9.3.5 after installing LabView 8.5 Student edition and I'm not able to find the DAQ Assistant, in the function palette. Could you please let me know what else I could do? Thank you!

  • How can I activate the infrared remote on my Canon Rebel T5 trigger

    My awesome wife bought me a starter kit complete with a Canon Rebel T5 (not a T5i) for my birthday... I'm so happy! In the starter kit is an infrared remote that looks like this. It is said that I just need to set my camera to accept a remote trigger