Using SMU 6612 to measure PXI-6528 pulsewidth channel - channel is not available.

Hi all

I use SMU 6612 card counter to measure the pulse width of the signals to PXI 6528 DIO card. These two cards are in the same chassis PXI (NI-SMU-1065). I could measure the pulse widths using the example LabVIEW 2013 Counter - pulse width of reading and (over) frequency example of .vi. However not all channels of the PXI-6528 map appear in the drop-down list of channels on the pulse width can be measured. Try to connect any other channel that those which are available in the drop-down list returns the error. On the PXI card port 6528 0,1 and 2 are entered ports and port 3-5 are output ports. I can measure the pulse on port 0, 3 width and line 0 port 1 and 4.

Can someone explain to me why don't see port 1 or port 2 channels in the drop-down list or force the VI to measure the width of pulse on these channels?

I can plug PXI-6528 external input channels SMU 6612 counter input channels and measure the pulse width, but if possible I'd like to avoid the external wiring between the 2 cards.

Probably not.  Unless the routing plan is in fact reversed as it seems a bit sorta that.  As stated on my system, you can route * of * a port of entry * to * RTSI, or you can route * of * RTSI * to * one output port.  This does not make much sense to me, but that's what I see:

If the routing card * is * reversed, your only likely workaround without physical wire would be to generate impulses in question of port 3.  It's pretty clear that 1,2,4,5-tetrachlorobenzene ports have no ability to interact with the bus timing, physical wiring would be the only option.

-Kevin P