visa resource can't find ports

Hi all

I'm having some problems with access to my COM-Ports: they are not only apperar in the drop down "Resource Visa..."

There should be a USB adapter to series, which has been installed correctly (appears in devices and printers as COM4) and has so far worked correctly with labview on other computers.  Also, I added just a Dev series, but also who is not recognized...

I've just updated my computer from windows xp to 7 and labview 2011 to 2012 if it is of no relevance.

Before updating the series on the same computer USB adapter worked very well... (The card series however was not recognized backthen either..)

I am very grateful for any idea you may have!

Thank you very much in advance,

Keksbold

In MAX, under 'Software '. NI-VISA would be on your hardware driver DVD you got with the distribution of LabVIEW. Or download it (click on the help link at the top of the page).

Tags: NI Hardware

Similar Questions

  • How can I find Port Proxy?

    My printer requires a "proxy port". Where can I find?

    If your network is not on a proxy, you need not enter a proxy. Here are the steps to solve this problem.

    1. find the IP address of your printer by printing a Network Setup page.

    2. Enter the IP address in the address bar of your web browser.

    3. click on the tab network at the top of the page.

    4. look on the left hand side, then click on IPv4 under wireless.

    5. change the IP settings in manual.

    6. Enter the IP address of the printer, your subnet mask and gateway routers IP address or by default in the spaces.

    7. switch the DNS preferred 8.8.8.8.

    8. adjust the auxiliary DNS to 8.8.4.4.

    9. click on apply. (You may need to refresh the page if it says the treatment longer than a minute and click on apply again.)

    10. now try to activate Web Services.

    Although I am an employee of HP, I speak for myself and not for HP.

    If I have solved your problem, do not hesitate to provide laurels and make sure that you mark this thread as a solution provided.

  • driver problem Keithley 2700: VISA resource not found

    I use Keithley 2700 DMM with two 7702 card scanner with card PCI GPIB Keithley and GPIB cable connecting the DMM and the PC.

    I downloaded the driver of the Keithley 2700 project style and loaded on my PC base.

    I created an executable file for the process of scanning running in another PC (client).

    I installed on the client, the runtime of LV 2011, Keithley GPIB PCI card driver (appears in the Manager of devices with errors), the VISA RT 5.0 engine.

    Yet when I get the drop down menu to contact VISA, I can only find the COM ports. The GPIB0::X:INSTR cound not be found.

    Please help me how I can make this resource name VISA in the client computer.

    Thank you in advance.

    Dear Dennis

    We had card GPIB Keithley stock and where it is used.

    The problem asked above, and the solution has been found.

    Solution: in the Keithley 2700 GPIB option turned OFF. I allowed him and

    the system has worked successfully.

    Thank you for your support.

  • Can not find the name of resource VISA

    I am trying to run the example code, provided by the manufacturer with a 3a Keithley 2420 SourceMeter.  It is turned on and put in place for 488,1 protocol GPIB.  It is connected to my computer via a GPIB KUSB - 488A cable.  On the front panel of the example program, there is a menu drop down titled "VISA resource name."  The two options at my disposal are LPT1 and COM1, but I don't think one of this is the name of the resource on the right.  Does anyone know what the name of resource VISA should I use with the Keithley, or how I can find this? If it helps, I have provided the example program I use. Thanks in advance!

    As a general rule, only the material NOR will be recognized in MAX and you wouldn't get any VISA resource for Auditors not - NOR. If there is a VISA for the Keithley driver, you can try to install as secondary visa driver and enable Tulip MAX. Keithley support support is the best place to ask questions about the installation of their hardware to LabVIEW.

  • With the help of VISA can not find the com port

    I use visa read but cannot find any com port in Windows 7.

    How can I find the com port in Windows 7?

    Windows 7 is not released. A version of LabVIEW for windows 7 is not released. A version of NI-VISA for windows 7 is not released. Wait until MS releases windows 7. Then wait tile view deleted OR made available versions of LabVIEW and NI-VISA that work with windows 7.

  • visa resources Finder will find PXI-4070

    Hello

    I use LV 2015, 32-bit on windows 7 x 64.  I have an SMU-1065 OR connected to the PCI computer bus.  I have several instruments in the PXI chassis included some switch cards, a scope and a PXI-4070 DMM.

    When I run the VISA resource finder, the result will show me all PXI cards in the chassis (as well as the serial port of the computer), but it won't find the PXI-4070 module.

    MAX shows the map I can run self test and open the interactive panels in MAX I saved (no alias) NAME of the PXI-4070 as "PXI-4070.

    I can also use the DMM OR-DMM pilots very well.

    the PXI - 4070 Max address is: PXI24::15:INSTR

    I must be able to detect programmatically PXI/SMU cards are in my frame so that I can put them up as system in the configuration it is in.  can someone help me understand why the finder resources VISA see the PXI-4070 module and report?

    Hey Reece,

    Have you tried to use the Configuration API of the system NOR? It allows you to programmatically query the hardware on your system. Here is some additional information on the API:

    http://zone.NI.com/reference/en-XX/help/373107J-01/nisyscfg/introduction

    https://decibel.NI.com/content/docs/doc-13216

    https://decibel.NI.com/content/docs/doc-29606

  • How a cRIO can access remote systems VISA resources?

    I have a NI 9871 RS485 module on an ethernet 9148 RIO.

    I have also a laptop and a cRIO 9073.

    I can access each of the 9871 serial ports on the computer but not the 9073.

    When you try to access it from the cRIO, I get error:-1073807192, VISA: (Hex 0xBFFF00A8) is denied access to the resource or to the remote machine. This is due to the lack of sufficient privileges for the current user or the machine.

    How the cRIO can access the serial ports of the module 9871 on the ethernetRIO?

    Or how a remote access cRIO visa resources?

    Or how a cRIO can have sufficient privileges to access remote resources from visa?

    Thank you for your help.

    LabVIEW 2010 & VISA 5.1 on pc

    NOR-RIO 4.0 and 5.1 VISA & NI 9871 SERIES scan engine support on cRIO and eRIO

    Hi Woss,

    It is possible that your RIO Ethernet has some parameters of active access which is to prohibit the connection to serial ports.  You can change these settings in a position and Automation Explorer (MAX) for your device by going to the device under remote systems, expanding the category of software under it and choosing NI-VISA.  There will be a tab in the right pane of MAX called Options of VISA.  In this tab, select Server VISA > security to access these settings.

    From there, you should be able to add the IP address of the RIO device that needs to access the VISA on the RIO Ethernet server.  Once you have changed these settings, just click on register at the top of the window and allow the unit to restart.  Let me know if that helps!

  • I can't find GPIB to VISA.

    Hello

    I have pluged GPIB-USB-HS to Agilent Network Analyzer and Analyzer GPIB address is 17. I can find it in MAX. But I can not find it in Labview VISA. Please help me to fix it thanks.

    BTW, all information Setup in MAX GPIB is grey, I can't change anything. Is this good?

    Thank you very much.

    I see several listed in the VISA resource E5062A devices.  One of them is it?

    In addition, sometimes LabVIEW needs rebooted to get these updates from drop-down list boxes.

  • Find VISA resources

    I have trouble finding VISA resources properly. I have no problem detection instruments that are connected on the GPIB, but after that they have detected once LabVIEW is convinced that they are there for ever more, even if I unplug and disconnect the device! The only way that the list of the connected instruments is reset completely out LabVIEW and restart it. Why is this?

    I'm using LabVIEW 6.1 (but who can not make a difference) and I attach a VI with two ways to detect if one of the four possible instruments are connected. The Subvi is used to give a command to one of the instrument, and it is able to return an error if the instrument is not connected, regardless of the situation. How can it detect the instrument as expected, but the other methods not?

    In this VI, I need detect how much, if any, and that out of four possible instruments are connected. Simple, but amazingly difficult!

    Any help please?

    James

    Well, now that you say is starting to make a little more sense. I misunderstood what you were asking before.

    Jeff Bohrer wrote: what is the selection if you look at "tools > Options > various... > automatically close VISA sessions?

    Help: "automatically close your left VISA open sessions by the application when the first level VI becomes idle. This check box is cleared by default. "This was added at some point (don't remember when), and the default value is True. It caused a lot of sorrow and complaining from users (I was a), so NEITHER changed the default value false.

    As Dennis has pointed out, there is a distinction between the resource and the instrument. When you perform an initial analysis resources will be created. There is no mechanism that automatically detects that an instrument is no longer there. In fact, when you scan for instruments of MAX, she sends * IDN? queries to find instruments. So, you must check for the presence of an instrument by actually trying to talk to her.

  • Satellite A30: Can not find the COM ports when trying to query the modem

    Hello.
    I have satellite A30 and I reinstall windows xp at home, but I can't find any com port whenever I try to query the modem the display hardware confilct, but in Device Manager no sign of error/confilct, I did everything what I can.
    Add the port,
    Add more then 1 port
    disable the printer port
    change the printer port
    Reinstall derver modem,
    In short, I've done everything I can. its always show (! yellow) mark.if I add the port and if I remove the port and there is no sign of error.
    If someone help me to solve this problem a little,
    regared.
    MGK.

    Post edited by: mghouskhan

    The modem uses a virtual com port. It s not a real port.
    I wonder why it usually happens after the installation of the new operating system, you must install drivers Toshiba together in the right order
    Installing the chipset utility is important

    The modem driver should do the work and you n t need to activate com ports
    Please choose the right side of the page of the Toshiba driver modem driver after installing the OS correctly. I think that's the key

  • Photosmart HP 6510 - can't find the USB port to connect to the computer. It is right in front of my face?

    Photosmart IHP 6510 - can't find the USB port to connect to the computer. It is right in front of my face? I can't find a diagram online. do not have wireless capability. With the help of MacPro with Snowleopard.Thanks.

    Hello

    The usb port is not #22. as seen in the picture. It's in the back of the printer on the side right hiand.

    Best regards

    ERICO

  • HP ZBook 15 G2: I can't find driver for Serial Port PCI

    After re-installation, my HP ZBook 15 G2 is missing the driver for "PCI Serial Port (COM5)".

    This unit has the following ID: VEN_8086 DEV_8C3D

    I installed all the drivers for chipset for the HP ZBook 15 G2.

    Where can I find the driver for this device?

    Hello:

    You need this driver.  If the driver will automatically install the driver for serial port, and then manually install the device by navigating to the folder of the driver that was created when you ran the file.

    This file will be located in C:\SWSetup\sp68428.  Make sure that the include subfolders is selected and the driver should install.  Then restart.

    http://h20564.www2.HP.com/hpsc/SWD/public/detail?sp4ts.Oid=6978815&swItemId=ob_137806_1&swEnvOid=4059

    The hardware ID you posted is supported by this driver if you look in the mesrl configuration file.

    [Intel.NTamd64]
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_1E3D & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_1CBD & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_1C3D & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_1DBD & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_1D3D & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_8C3D & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_9C3D & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_9CBD & CC_0700"
    % SRL_DeviceDesc % = ComPort, "PCI VEN_8086 & DEV_8D3D & CC_0700"

  • can I use visa.vi to rs232 db25 port?

    can I use visa.vi to rs232 db25 port? because where the use of rs232 to the parallel port (db25).

    and also any vi to read the data of the series.

    IM new in labview I just learn the tutorial to youtube.

    If not can someone give me an idea how or can give me an example vi.

    TNX...

    Your front panel can be... catchy (and frankly annoying), but your VI does not make much sense. If you want that your 'start' code when the user presses a façade 'Start' button, then you have to sit and wait for the user to press start, doesn't depend on the user to press Start before you run your VI.

    In addition, you have:

    • Wired the value that configures 'Port' at the entrance to "Activate the tank of termination" of the VISA. Why? You should know whether it should be enabled based on what you're trying to talk. The unit will stop his messages with a specific character? If so, a real thread of this entry and a value of thread at the end tank entry matches the value ASCII of the nature of the instrument used to put an end to his messages.
    • Wired a zero for the number of bytes Read VISA. Why? This means that you do not read anything by definition.
    • Do you need to wait a certain amount of time after sending the order before data are available?
    • Do you need to terminate the order that you send with something?
    • You currently have the set loop in place as an infinite loop. This means that the only way to stop your VI is using the button abandon in the toolbar. IT OF BAD. Provide a stop button on your façade to allow your VI finish properly.

    See attached mod as a better starting point.

  • How can I find my password for port locker

    I'm trying to back up files on a USB device, and a window opens asking a password for windows port locker. I don't know the pass wod. Where can I find?

    Hello

    1. what operating system do you use?

    2. have you set any password for port locker?

    3. you are a system administrator?

    I suggest you to contact the software publisher to recover the password for port locker

    In addition, technical support engineers Microsoft cannot help you recover passwords files and Microsoft who are lost or forgotten product features.

    Microsoft's strategy concerning lost or forgotten passwords

    http://support.Microsoft.com/kb/189126/en-us

  • I had a hard drive crash and had to move all the data again. Now, I get an error saying that my Visa is not an authorized version. How can I find the product key to reactivate?

    Vista Product Code question

    I had a hard drive crash and had to move all the data again.  Now, I get an error saying that my Visa is not an authorized version.  How can I find the product key to reactivate?  I don't have the sticker on the back and the software I have does not have the code?  I bought this laptop DELL at Best Buy a little more than a year.

    Thank you.

    Hello

    This problem may occur if you make a significant hardware change. See the link provided below.

    Error message when you start Windows Vista: "your activation period has expired".

    http://support.Microsoft.com/kb/925616

    If the OS came preinstalled Dell you may need to contact Dell technical support for assistance.
    http://support.Dell.com/support/topics/global.aspx/support/win7_support/win7_portal?c=us&CS=04&l=en&s=BSD