Application errors of the RT with the PXI-4461 with Labview real-time 9.0.1, DAQmx 9.0.2

HI -.

I recently converted a PXI time system real OS (PXI-1042 q chassis, controller PXI-8187, DAQ, PXI-4461, DAQ, PXI-6259). I can write and run DAQmx applications in real time with the 6259 very well. Whenever I try to write a labview RT app to use the 4461, however, it will fail. Note that I can use two cards through MAX I tried switching card slots, just in case it was a problem. Both cards worked with labview, the PXI chassis was before Windows.

Attached are pictures of the screw base demo I built to show the problem. Since I was a mistake (-200758) if I started from the raw strings, I tried to create a MAX task and use it. The task, but he complained of a buffer is too small. I explicitly put the buffer to work around this problem and still get the same error (-200608). I've also attached a screenshot of the software currently on the Max MAX RT PXI system is version 4.6.2 btw. (I installed the most/all this from DS1 Dev Suite 2010 version).

Please let me know if I hurt something installed, versions if 4461 just don't play nice with the new BT or RT software, or if something is wrong. Thank you.

Kregg

The first error you see is (details in the help-> error explain in LV)

Tags: NI Hardware

Similar Questions

  • How is it you can't use more than 126 GB of a 1.5 TB with LabVIEW Real-time hard drive?

    How is it you can't use more than 126 GB of a 1.5 TB with LabVIEW Real-time hard drive?
     
    It's only 8.3% of disk space!

    http://digital.NI.com/public.nsf/allkb/E7E2F5657B6026358625750B007201A6?OpenDocument

    The national Instrument is reformat that you drive less than 126 GB or replace the drive with a less than this size.

    Please OR tell me you go fix your LabVIEW Real Time soon, and that we will not have to wait for LabVIEW RT 9.2.1 in 2011 for this one to fix.

    http://upload.wikimedia.org/Wikipedia/Commons/9/90/Hard_drive_capacity_over_time.svg

    Readers as high as 126 GB or more have been around since mid-2002.  It's about 7 years now and LabVIEW Real-time STILL cannot use greater than 126 GB drives!  Please OR.  Do something.

    We deserve the patches for older versions of LabVIEW RT.  What about updates to LabVIEW RT.  As 7.0.2 7.1.2 7.6.2, 8.0.2, 8.2.2, 8.6.2.

    We don't need new features or versions of LabVIEW RT as 9.0 until we have some of the bugs fixed in existing versions.

    Kevin.


  • details on the test of suspension of vehicle (car) with labview real-time

    Hello friends,

    I'm doing phd in the suspension system active semi auto. Here, I need to study and measure all the setting of suspension of the complete vehicle experimentally. . So far, the experimental cost is so high.that too will give the approximate result.

    In the experimental method, we need create the profile of the road using jacks. Its cost too much. I don't want to do that.

    I would like full of model.is of vehicle it is possible to time real suspension test using labview? Let me tell you what I want.

    I have to fix four sensors on four wheels to measure the road disturbance. Then I will join four more sensor on the car body.

    This difference in two reading will give the perfromace of my suspension. Whenever I drive my car, I can get this place profile and the road to the deformation of the car body. Finally, I can measure my performance of car suspension.

    is it possible to get the upward movement of the tire by using sensor in labview? That's what I want exactly. In the experimental method, we spend a lot of time and money to create the profile of the road. but it is already available in envirment. So, I just want to eliminate the road profile. If we eliminate we can easily measure the performance of the suspension.

    The measured value recorded in the vehicle itself using labview computer. So we can better the performace.

    If you do not understand my question pls ask me once again.

    looking for your answer.

    Spengler

    I saw an article that uses labview for quarter car model. but want to do it for the full car model.

    Hello

    I hope that I can understand what you mean, because I did several tests on vehicles and I hope that I can tell you something good for your topic.

    First of all, I am interested in what kind of upward movement? Speed, displacement or same acceleration? You can have a lot of possibilities for a test in the vehicle using LabVIEW, CompactRIO or CompactDAQ, but they are all software problems. If you know which parameter you will measure, we can simply fix the sensors on the suspension and run the test program.

    Also you can collect the State of the car, for instance a speed of ROTATION of the wheels and so on, through the CAN bus.

    If you are you doing PhD on this topic, you may need to find a company to present a better solution. Well, in a Word, it is a part of a doctoral thesis.

    Wilbur

  • Robotics/control system with LabView real-time + OR DAQ

    Hello

    I am building this real time application in which you can try control methods for a particular system (construction of the algorithm of control in LabView on Windows) and then with a DAQ board (I have the E-Series SCB-68) communicate with the hardware (actuators).

    Is this DAQ card enough or should I try with any other advice? For the application considered, it is important that the price of the freq is at least 1 kHz.

    Thanks in advance

    The 6220's card perfectly fine for your current application.  It is true that windows is not deterministic and cannot compete with the hardware/software in real time, but your DAQ will be responsible for the management of timing and synchronization you will be fine.  If the advance until you need a very quick response in real time of your application, you can always buy a RT LV system.

  • using NI USB 9201 with LabVIEW Real-time

    Hello

    I try to use a USB of NI 9201 on a remote real-time target. Here's what I did: 1) tricked out of a desktop PC to serve as target RT using the instructions provided on the Web site of NOR. He now starts in an RTOS with LabView RTOS 9.0. (2) I communicated successfully with the help of a host computer that has a windows with LabView 2009 BONE and I installed the software from the computer remote RT target host using ethernet. However, my problem is when I plug NI 9201 USB on the target RT that I do not see the material listed on the devices tab in MAX and interfaces on remote systems. I tried to get the manuals for the HW 9201, but it seems not much explained here, or maybe I'm looking in the wrong places. Any help is greatly appreciated.

    PS: I am attaching 2 JPEGs for additional explanations (tab network settings and the settings of the system).

    Thanks in advance,

    Benoit.

    AVS,

    9201 USB is not supported on RTPCs please visit this link. Currently no usb-DAQ devices do not support an environment CRPC. Only the PCI DAQ devices are supported.

    http://zone.NI.com/DevZone/CDA/tut/p/ID/6913

  • Svchost.exe - Application error. The instruction at "0x7d4caa9b" reference memory at "0x00000010". The memory cannot be: "read"

    Hi, I get the error message after my system boots - svchost.exe - Application error. The instruction at "0x7d4caa9b" reference memory at "0x00000010". The memory could not be "read".

    And then after a few minutes it stops and restarts.
    I'm also not able to use the mouse.
    I need help with this problem
    Thank you and best regards

    Hi GeorgeMotaung,

    ·         Did you do changes on the computer before the show?

    ·         Are you able to boot to the desktop?

    Follow these methods.

    Method 1: Start your computer in last known good configuration.

    How to start your computer by using last good known Windows XP Configuration

    http://support.Microsoft.com/kb/307852

    Method 2: Follow these steps:

    Step 1: Start the computer in safe mode and check if the problem persists.

    A description of the options to start in Windows XP Mode

    http://support.Microsoft.com/kb/315222

    Step 2: Perform a clean boot to see if there is a conflict of software like the clean boot helps eliminate software conflicts.

    How to configure Windows XP to start in a "clean boot" State

    http://support.Microsoft.com/kb/310353

    Note: After the boot minimum troubleshooting steps, follow section How to configure Windows to use a Normal startup state of the link to return the computer to a Normal startupmode.

    After the clean boot used to resolve the problem, you can follow these steps to configure Windows XP to start normally.

    (a) click Start and then click Run.

    (b) type msconfig and click OK.

    (c) the System Configuration Utility dialog box appears.

    (d) click the general tab, click Normal Startup - load all services and device drivers and then click OK.

    (e) when you are prompted, click on restart to restart the computer.

  • Acquire more than 2047 samples with the PXI-4461 instaled in SMU-1073

    Hi all, I would ask you for help with the buffer limit.

    I intend to buy digitizer PXI-4461 and he instal in SMU-1073 chassis, namely control via MXI Express of Labview installed on a separate computer.

    What I need:

    -to acquire data of a single channel of AI, but at least a sequence of 20 kS by a acquire task, in some situations until 200kS by a task to acquire.

    The question:

    - I can gain more than 2047 samples in a single sequence, like 200kS, with the PXI-4461 installed in SMU-1073?

    Internal buffer of the PXI-4461 is reserved to 2047 samples. So I'm not sure if Labview can download remotely via MXI Express the data in the buffer of the PXI-4461 via MXI Express fast enough without any affection of the sampling program.

    -in the case, this PXI-4461 with SMU-1073 isn't the right combination, what chassis and a controller can do?

    Thanks much for the reply

    Jan

    It will work for you.

    The on-board buffer 2047-sample is used only as a backup if the flow of data to the PC host (via MXI Express in this case) is not fast enough... that it will be (explained below).  DAQmx transfers data from the buffer of the device to the host PC as fast as he can and, in ideal conditions, should not save the buffer 2047 much at all.

    Let's just say you get 110 MB/s (randomly from a MXI data sheet) flow on your MXI connection.  The 4461 has 2 analog inputs, which will be at 24 bits, we just round 32-bit in case it transfers the data in this way.

    4 bytes/sample (32 bit) x 200,000 s/s x 2 (channels) = 1.6 MB/s, which is well below the 110 MB/s, which will make the MXI link.

    clear as mud?

    Germano-

  • "VI a error 1002288.The full version of LabVIEW Development is necessary to correct the errors."

    Hello

    I had created an executable in LabVIEW 8.6 on a PC windows XP 32-bit. The application uses some screws of Vision. The executable is running in most of the PC except one. the PC is a windows 7 64 bit. I first got the error "Labview support 3 error code: could not load front". "" I fixed this by changing the settings of the source file in the application builder. Now I get the error message "VI a error 1002288.The full version of LabVIEW Development is necessary to correct the errors"and the exe displays a run button broken.»

    But if I build the exe in the same PC I get no error and it works very well. Moreover, I tested this exe in another PC with the same configuration, it works. I tried to reinstall labview in the PC, the problem persists.

    Hi Nitrox,

    There is no official adapter for LabVIEW 8.6 in Windows 7. See link below.

    www.ni.com/LabVIEW/OS-support/

    Upgrade to SP1 LV2009 should solve your problem. We had this problem with another client and update SP1 LV2009 solved his problem and should solve your problem too. If you have a problem after the update let me know.

    Kind regards

    Christophe P

    Technical sales engineer

    National Instruments

  • The PXI-4461 connections and text based programming

    Hello

    I looking for a guide-schedule based on text for the PCI card, / AO PXI-4461 24 bit.  Currently, I am able to program it similar to the PXI-6259 but could not find a way to set it up as (differential, Pseudo-differentiel) in order to obtain the correct output impedance.

    Is there a hardware (user) for PXI-4461 guide?

    I use Visual c# to encode the PXI-4461.  How can I determine the location of the classes and the specific methods for the PXI-4461.

    Thank you.

    Anand


  • How to generate a single pulse using PXI-4461 in Labview

    I need to generate a single positive pulse is 100ms using an OD on the PXI-4461 with DAQmx Labview

    I have trouble getting the exact time of the pulse.

    Help, please.

    Thank you.

    He works, see attached pattern obtained.

    Thanks to NI Applications Engineer ally Finney for example.

  • & quot; An error occurred while installing LabVIEW Real-time 2009 (in English). ...

    While trying to install the real time of the third quarter of 2009 DVD module, I get this error message.

    "An error occurred while installing LabVIEW Real-time 2009 (in English). Do you want to continue installing remaining products? »

    After removing the program OR series (USB) 1.1 installation completed up to the end.  If you answer no to continue to install the above question, the following error message refers to abduction OR Serial.

  • "error-200621" trying to generate 20 kHz sinusoidal aboard the AO0: 1 of the PXI-4461

    Hello!

    I am getting a-200621 error trying to generate a sine wave of 20 kHz the signal output of a Council of PXI-4461 0 and 1. The signal is generated for a few seconds, then the error «on-board device memory overflow negative...» "happens! Why? and how can I avoid it?

    If I'm generating 1 kHz or using a single channel, all is well!

    I joined the vi as well as the error message.

    Here is the configuration of my system:

    LabVIEW 8.5

    PXI-1044 chassis with 14 locations and MXI-4 Council (optics) to communicate with the PC:

    -Connector 2: OR PXI-4461

    -Slot 3: OR PXI-4461

    -Slot 4: OR PXI 2561

    Computer:

    -Intel Core 2 Quad Q9400 2.66 GHz CPU

    -3.49 GB of Ram

    Thank you in advance.

    Frédéric

    Hi Peter,.

    The issue is resolved in fact to increase the number of samples that are written in the buffer!

    Thanks to you and Laurent!

    Frédéric

  • SearchFilterHost.exe do memory Application error at the < address to spell > unreadable

    Windows XP SP3, search V 4.0... continually get application error memory to unreadable, terminate application?

    Tried to uninstall and reinstall search V 4.0, unless there is a fix package?

    Really a PITA.

    I ended up uninstalling Search v4.0 and the problem never CMAF on the rise again.

    I ended up uninstalling Search V 4.0 and the problem disappeared.

  • What is the maximum voltage of the members of the PXI 4461 gain 20 dB?

    I use a Board, PXI-4461 sample sine wave with an amplitude of slightly greater than 1 V peak (1.0005 V). The gain of the Board of Directors is set at 20 dB and the input range should be from-1 to + 1 V V! That's why I expect that see a saturation of the converter ADC when the input signal is greater than 1 V. However, I do not respect this saturation of the output value, and I don't understand why?

    Could someone explain that to me?

    Thank you in advance.

    Frédéric

    Hello Frederic,.

    almost all NI DAQ devices I know has a small "margin of safety" in each range given acquisition (usually about 0.5-2%), allowing you to accurately measure the voltage specified limit.

    The exact width of this margin is not explicitly mentioned, probably to prevent users to use this line on a regular basis.

    In short: there's a small safety margin, but do not count on it when designing measurement applications. ;-)

    Best regards
    Sebastian

  • How can I display an analog input for the PXI-5105 on LabVIEW?

    Hi all

    I am very very new to LabVIEW and I started to tinker with it. I use the version of LabVIEW 2010 SP1 on Windows 7 OS. I also have the chassis NI SMU-1073 with SMU-6361 and PXI-5105 modules and the chassis is connected to my PC via PCI. I became familiar with the devices and trying to see some analog signals to one of the channels on the PXI-5105 module in a graph in LabVIEW.

    I would appreciate your help.

    Hello Henokview!

    I would like to read through these tutorials to understand the steps of programming of the NOR-SCOPE, NOR-DAQmx. After reading these links below, you will be able to understand how to connect the output of a readfunction to a chart or table.

    DAQmx

    http://www.NI.com/white-paper/5434/en

    OR-SCOPE

    http://www.NI.com/white-paper/3382/en

    Best regards

    Jonas