E3631A Agilent

Hi all

I am relatively new to labview and trying to control my diet E3631A Agilent.

So I bought a usb series adapter and downloaded the driver from the website of OR:

http://sine.NI.com/apps/UTF8/niid_web_display.download_page?p_id_guid=0475216F9FCA5335E0440003BA7CCD...

I unzipped the folder and open the examples, but I keep getting errors...

I have attached the error I get when I open the basic vi example.

It takes about 45 seconds, until what this error pops up, so I guess that's something, but in execution to highlight mode I do not see what exactly...

Thank you very much in advance,

Keksbold


Tags: NI Software

Similar Questions

  • How to control the current program view lab for a power E3631A agilent

    Hello

    I wrote a program to view lab for agilent E3631A triple output power supply in constant voltage mode, and I want to update to operate in constant current mode also. But I can't find any function in the Instrument of e/s, which is related to the current but only the current limit I already used to write my programs to the constant voltage mode. How can I do this?

    Thank you

    When the current exceeds the current limit, the power supply will be running constant.  The food isn't really have one running constant set.

  • The IVI Driver download for E3631A Agilent expects a C: drive - solution

    Exe self extractor for the Agilent E3631A IVI driver expect drive c to exist.  I don't have drive C: self extractor opens then C: drive is no reports and they closed the extractor.  Does anyone know a work around for this?

    How can you not have a c:\ drive?  What kind of PC are you running?

    Try to rename the .exe is a zip file.  Then see if you can open it using winzip or another zip program to be able to drag files manually.

  • Connection of agilent E3631A DC power supply to the computer via a RS 232 for labview

    Im trying to access the computer via a cable RS 232 for sully power DC E3631A agilent use labview to operate the instrument. But the instrument is now detected by the computer and when I check instruments instruments connected tools-tool-find, told me no device not connected. I tried to use max by changing the configurations, but so far it doesn't work anymore. Please can someone tell me what to do

    I don't know if MAX is the instruments of automatic detection on the serial port.  You can contact the instrument using Hyperterminal or something like that?  Also make sure you serial port settings are all correct (it is the most common reason).

  • Connection of GPIB USB HS TWO agilent instrument (power)

    Hello

    I'm fighting to control two power supplies (E3631A) Agilent using GPIB USB HS connection.

    But they are not detected as TWO different power supplies regardless of STAR or linear connection.

    However, when I try to set up a power source and a Tektronix oscilloscope, they are considered TWO different instruments.

    Single power supply control is also OK.

    So, I'm wondering how to set up TWO power supplies.

    I enclose the code VI to not control than a single power supply and want to extend to the version of the two instruments.

    you have configured the power to two addresses different gpib?

    This is needed!

  • Delay between two operations in the same visa write buffer

    Hello

    I'm new to labview and have tried to find a way to introduce some delay in milliseconds between two operations in the same buffer of writing a block of single entry visa. Here's what I want it to be like:

    CURR: LEV 1.5

    wait about 50 ms

    CURR: LEV 2.0

    wait once again

    In fact, I must apply an alternating signal to a coil with a certain frequency through a source of supply of power e3631a Agilent. So if there is another alternative for her, so please let me know.

    Thank you

    Concerning

    Here are two approaches. First being the easiest, simply put a delay between written VISA. The second being a more robust because it allows you to develop a series of commands by simply adding the string array. You might forget the case of the loop if you don't mind the 50mS delay before e first command is sent.

  • Visa read stops execution

    Hi all

    I have a little problem with the control of my power supply E3631A Agilent via Labview:

    I downlaoded all drivers and with the help of you guys I even persuaded that it is running properly.

    At least that's what I thought.

    Turns out he didn't really. At some point the VI just stops.

    Using the execution mode of nail I knew where the problem was:

    The VISA read error query of the configure output has stopped working.

    I have no explanation for it. I get unpredictable and I have not yet managed to influence by changing the different settings.

    (And I tried it for a while now..)

    I built a very simple example VI to show you where the problem lies.

    Explanation:

    I'm basically feeding the output a voltage variable and at the same time measuring how long the vi fonctionnee already has...

    I thank you very much in advance,

    keksbold

    OK it looks like as the time to copy this section of the FAQ: (posted in a place not yet public ~ ~ ~ Hey Jordan? is still FAQ on the right way?)

    "

    In this topic, we will discuss some of the common problems that have been observed using devices with LabVIEW on Windows operating systems.  Many of these points apply also in other environments, but the examples will use the Windows 7 operating system.

    FAQ 1: USB device stops working unexpectedly.

    The first thing to look at is the OS power save options.  There is a global trend towards "Green" electronic development and energy the coast to stars become fairly common.  "If his is used not to turn it off" is not new.  Cave men learned how to Bank a fire in order to preserve the energy that would otherwise be wasted.  Similarly, the Windows operating system has an energy saving function to stop power to the USB hubs when no user activity is present.  In automated systems, this feature can cause problems since remove power USB hub will stop the USB device.   Solution: use Device Manager to change the Power Options USB hub.

    FAQ2: I put the power options and my device connection is still not reliable: Don't forget, the USB ports on the computer are often the least expensive that can be mounted on the chassis and system share the PC power supply USB-powered. Most of the uses of USB are temporary connections such as a USB key or a camera.  These connections don't require high reliability because there the user interacts with it.  Mounted sudden power and fault tolerance at worst cause the operator to retry the transfer of data.  Automated systems require a bit more robustness.  Solutions:

    1) ALWAYS use a free external powered hub. Perform your due diligence genius and to inspect devices specifications too - if you do not find them for this device which should clue pick you a product from a vendor who published their specifications.

    2) high noise environments require the use of ferrites on the USB cable- and do not buy the cheapest cable either! The good markets are poorly protected.

    3) PROTECT connections HUB-if you have a USB 2.0 device and Joe User plugs in a 1.0 device into an open slot, managed by the same hub-Bingo all ports on the rear hub can convert USB1.0.  WORSE, there are a lot of damaged or marginally engineered USB devices out there.  Joe User should cause power fluctuations when it is inserted or removed from the hub just don't let it happen!

    FAQ3: I test USB devices and the operating system can not find more.

    It is a Plug-and-Play feature that deserves exposure.  When you connect a device P - n - P BONE remembers his serial number in an entry in HKEY registry (hive-Key).  This is useful when (for example), you want a specific instrument, say a NOR-USB-6008, to appear as a device DAQmx with VISA Alias "MyDAQ1" whenever it is plugged.  On the other hand, if you want to test a range of USB-serial converters that can be problematic because the pilot of P-n-P will mount the first serial number like "COM3" and the other as "COM4" Add infinitum until the controller of the enumerator in the registry and the VISA recognized alias adjust upward.  Solution: use the Windows registry API and the API of hardware in LabVIEW to clear unused VISA alias and HKEY entered.   Speak with your professional COMPUTING staff structure HKEY and possible side effects before developing a plan to modify registry entries.

  • Applying a mathematical operation of string data

    I am a PSU (E3631A Agilent) interfacing via GPIB. When the power supply is turned on to 12v and a device consumes current 104mA, vi reading gives a data string + 1.04485600E - 01. Now shown data are right, but I want it to appear on the front panel as 104 or 0.104 instead of exp + 1.04e format - 01. One way is to multiply by 10 and one way is to divide by 100 and maybe way to convert a simple decimal number. Any body can help me to get simple 0.104 or 104 out of the string above please.

    If this is the answer of the exact string you can use the Exp/Frac string to number function to interpret the content of the string in Double. You can then use a digital sign with the defined display format, but you want to ensure you get a rounded number. I think a '%.3f' format specifier should give you 0.104

  • Question of sampling E3631A

    Hi all

    There is a program to set vary the output voltage and read it on table and write it to the file, but the data is too long (about 4 sec for each cycle), how can I take as small as possible. Thank you.

    Hi nvnewbie,

    In fact, your problem has nothing to do with your structure of data acquisition, although on both of your diagrams of block, you do not have the entry in the file cable write to the file of the measure, if you log all the data. It must wire your table to the input string to write on a file of measures.

    The delay you see is expected; set the output current limit to set it up and screw out of reading in the Agilent E363X Instrument Driver Library all call query VI error, which is also within the Agilent E363X instrument Driver Library. Error query VI has a built in 1 second delay to ensure that there is not a timeout when communicating with the E3631A during the series. You can see this delay by yourself if you open your SubVIs in your block diagram.

    Given that you call the exit configure, configure current limit, and screw out of reading with each iteration of the penalty loop, which is 3 seconds delay guaranteed each time the while loop iterates. It is also integrated into these screws themselves, so this delay adds up to 4 seconds of delay. If you want to reduce this time, you can change the error query VI in order to reduce the delay, but I think this causes time-out errors if you slow forward. Series is not made for communications high speed. Alternatively, you can rewrite your code to call only the exit set up and configure current limit live when you need to adjust the settings.

    Best,

    Dan N

    Technical sales engineer

    National Instruments

  • What is "agile basic error" and why I can't get rid of?

    I tried to place an order with the company of Melitta coffee. Everything is going well until I hit the "place an order" and then comes back the gibberish "agile basic error" and I cannot cross it, get rid of it, and I am never able to place my order.

    This is an error with the server of the Web site. Please contact website assistance.

  • low frequency with labview 2010 auto and agilent 35670a

    Hello

    I work on the automation of low-frequency measure a 35670A Agilent. I currently did a program that allows a measurement with an average of FFt given a frequency bandwidth and frequenct from beginning, see picture attached, but I try to measurement of spectrum using this power VI ( ), but unfortunately I can't seem to find how to convert a signal from my block diagram below in order to send the input signal power VI above spectrum X no indication or help will be appreciated.


  • Problem with driver with Agilent 6031 A feeding

    Hello and goodmorning,

    I tried to develop my first application of cvi in which I need to use a power supply (model Agilent 6031 A). I first use direct command of GPIB and I managed to make it work, but with some difficulty, due to the fact that sometimes orders are missed.

    So I downloaded the driver from the website of OR, and I tried to implement in my code. I found a problem with the first statement, the init it. Here is my code:

    ViPSession InstrumentH;
    ...
    ...
    ...
    All = hp6xxxa_init ("GPIB0::5:INSTR", VI_ON, VI_ON, 1, & InstrumentH);

    After running the above code, I found in 'All' value '-1073807339 "and InstrumentH is equal to NULL."

    Any idea and advice would be much appreciated.

    P. S.

    There is a way, as in Labview, to find the explanation of the error Code?

    Found the problem. Food must be set of SCPI commands on the old controls ARP using "SYST:LANG".

  • Agilent ENA s3p file

    I'm looking for something fairly simple to generate a file s3p for my Agilent E5071C Network Analyzer.  I guess my greatest need is to find a way to get scanned data.  At the moment I have just set up the 9 tracks (S11, S12, S13, S21, S22, S23, S31, S32, S33) and then configure my markers and take individual measures.  There must be a way to collect a data table on a bandwidth of a trace.  Anyone know how?  Thanks in advance.

    The problem has been resolved relatively easily on mine as follows:

    Use the VISA Write block to send the following strings:

    • : MMEMorytorusNP:FORMat AUTO

    • : MMEMorytorusNP:TYPE3 P 1,2,3

    • : MMEMorytorusNPATA 'C:\filename.s3p '.

    The Agilent ENA Series.lvlib:File transfer from storage Device.vi allows to recover the file from the Network Analyzer and store it in a local file.

    Again, I have no problem to find a solution on my own (obviously), nor do I have problems with a lack of support or answer, however, I have a problem with the comment ironic and demeaning.  Please keep to yourself if you do not have any positive suggestions for those seeking honest suggestions and help on this support forum.  Thank you.

  • Agilent 33500 B configuration of duty cycle

    I use the function generator to Agilent/Keysigt 33511 B to produce a square wave with offset, amplitude and frequency specified. However, I'm unable to set the duty cycle. Is there a VI to control this setting?

    Hi jmountney

    If you use this driver , I think that there is a function named .vi configures signals Standard Advanced (square) and it has an entry for the duty Cycle, so I think this could help you.

  • communication with agilent 33522A

    Hi all

    Please can someone help me communicate with my instrument named below (agilent 33522A)

    What do I need exactly please (drivers, hardware)

    Thank you

    Here is an article in the knowledge base that helps solve your problem of GPIB communication. Have you seen this before?

    http://digital.NI.com/public.nsf/allkb/A80DBFCCAC36EBDE862562C80058856E

    Best,

    Doug

Maybe you are looking for

  • iPhone WiFi problem

    Hey, my iPhone 5s has a Wifi problem. I'll send text messages, but will not work iMessage. I can receive emails and receive phone calls. But, when I open my apps video/social media (Youtube, Amazon, Instagram, Snapchat, video etc.), the wifi does not

  • my hp deskjet 5740 does not print more of my change of ink cartridges.

    Today, I changed the ink in my HP Deskjet 5740 printer cartridges; I can't go to my printer.  Someone knows why?

  • Download of Windows 10 delay

    I downloaded Windows 10 & he gave me a few options for installing the download that forced a choice of days / hours.  I would wait & need to know how to stop the scheduled time. Thank you.

  • my computer crashes, I press ctrl alt delete and it thaws

    original title: don't know what's wrong my computer crashes, I press ctrl alt delete and it thaws, this happens very often, he does not have it for weeks at a time and then it again. I'm a paint shop pro user, I uninstalled and reinstalled psp more 2

  • When I try to open Skype, yahoo, live mail a small box prevents them from working

    Hi from opening Skype, yahoo messenger or live mail a notiufication little box comes up immediately after the program trys to open it. This also happens on fix that he downloads Microsoft and rarely on the internet explore. I downloaded such product