VISA: (Hex 0xBFFF0015) timeout expired before the operation is complete.

Hello

We use LabView to control a sequencer of high-voltage (HVS448LC 3000 D) LabSmith using LabView drivers from the manufacturer, power supply. Everything works well in our program (attached) until we try to use another program on the computer at the same time. For example, if we try to open a window for our camera, or even a folder in Windows, the program gives the following error:

1073807339 error has occurred in the CITATION read in FromHVS448.vi-> CommandHVS448.vi-> RegulateVoltage.vi-> Gated timed injections version 06.the vi

Possible reasons:

VISA: (Hex 0xBFFF0015) timeout expired before the operation is complete.

We would like to be able to control the camera software and the power of LabView by using the same computer, but we do not know what could cause this problem. We would appreciate any suggestions! Thank you!

You should really put a lot of orders of pre-instrumentation in parallel.  Force an execution order by using the wire of the error.  Serialize all commands to the instrument.  You may have an order interfere with one another.

Tags: NI Hardware

Similar Questions

  • VISA: (Hex 0xBFFF001) timeout expired before the operation is complete. I get this error when you run a VI (instrument is set up to be controlled remotely using the TCP/IP Protocol)

    VISA: (Hex 0xBFFF001) timeout expired before the operation is complete. I get this error when you run a VI (instrument is set up to be controlled remotely using the TCP/IP Protocol)

    Please suggest a solution.

    Thank you very much, it worked!

  • error: timeout expired before the operation ended

    Hola a todos buenos dias

    Estoy trabajando con UN medidor el cual pressure is communication via USB micro USB, are not already marcarme este error (timeout expired before the operation ended)

    Yahe probado changing el timeout, delay el, config bytes to port y nada. ES mas, todo esto samples are:

    http://digital.NI.com/public.nsf/allkb/874B379E24C0A0D686256FCF007A6EA0

    I hope that I can help, thanks to photos

    End of the line constant is dependent on the platform.

    Choose one of these constants as string endpoint of your!  and seriously check all your constants in the VALUE chain.  showing the display style only helps you identify which are false

    Then: save that attachment to C:\Users\\Documents\LabVIEW Data\Probes\

    When you select and right-click on a wire string, you will have a new selection under probe custom > StretchyString! It looks like this:

    Believe me, you really want to use for debugging the channels of communication and messages it will allow you to completely inspect strings with nonprinting charaters and whitespace.  It resizes too big because you might need it and you can change the display format.

  • VISA: (Hex 0xBFFF0015) Timeout management

    Hi all

    I'm trying to commuincate with a patented device.  In full operation, it is supposed to send keepalive type

    data periodically, the first thing to do is so empty the receive buffer.

    My program is, (attached):

    1. set baud rate to 15200
    2. open communication of VISA.
    3. flush the receive buffer.
    4 write \r\n
    5. wait 100ms
    6. read the bytes to the value of the port
    7. read messages until the number of return is zero.

    The number of bytes to the serial port is 360

    The length of the message table is the following:

    Message 0 item has a string length of 87
    Message 1 item has a string length of 90
    Message 2 has a string length of 90 item
    Message 3 item has a string length of 90
    Message 4 has a string length of 3 item
    Message 5 has a string length of 0 item

    Concatenated string length: 360

    The element error 0a error: no
    The element error 1 error: no
    The element 2A error error: no
    Item 3A error error: no
    The element 4A error error:-1073807339
    The element error 5A error:-1073807339

    I can easily rub on the mistakes, but I was wondering if there is a method that does not generate an error?

    Also, why not make one answer rather than separate 5 responses, which results in an error?

    I mean, I love not waiting, is it better to interview a number of bytes > 0 or x number of Foundation that does not change.

    Is there a method that recommend NOR for this simple query series?

    TIA,
    Sean

    Some VISA basics.

    You have the stop activated with the VISA character set up the Serial Port. Then use the VISA to the Serial Port bytes. Using both is not recommended. VISA playback will end automatically when the stop character is detected. If you have a character of endpoint, you can place some high number for the number of bytes to read. If you have no character endpoint, disable it.

    You do not send a return trolley/line with the \r\n. You send each character separately. In order to send '-' codes, you must enable them. Simply right click on the control of the chain and select '-' code display.

    The warning is expected and normal, whenever playback VISA does not detect a termination character.

    Placing the bytes VISA to the Serial Port outside the loop is just incorrect and using a while loop with the VISA Read is just making it worse. You are guaranteeing a mistake.

  • Streets & Trips has expired before the date of the trial.

    Streets & trips has expired before the date of the trial

    Hello

    I suggest you post your questions on the forum of Streets and Trips is here:
  • The operation cannot complete because of an unknown error. [SExi]

    Whenever I launch Illustrator, an error: the operation cannot complete because of an unknown error. [SExi]
    In this case since I installed the latest version of CC.

    I uninstalled and reinstalled. It has not changed anything.

    Any ideas?

    Hi interactivemartin,

    As mentioned by Mike, to check the steps mentioned by me, I put here once again for your comfort. Please follow the steps below:

    1 activate the root (Mac) or hidden (Windows) administrator account user activation and using the user 'root' on OS X - Apple support

    2. connect to Root/Hidden Admin account

    3 uninstall AI with the help of the CC cleaning tool: use of the Adobe Creative Cloud cleaning tool to solve installation problems

    4. If you're on a Mac, go to Applications > check if you have any folder left for Illustrator, if so, trash

    5. If you are on a Windows PC, navigate to the C drive > Program Files > check whether you have any folder left for Illustrator, if yes, delete the.

    6 re HAVE with the help of Adobe Creative Cloud app in the root/Hidden Admin account to settle and then check.

    Thank you

    OM

  • The operation cannot complete b/c of a unknown error [!] MP]

    Whenever I'm going to delete something in Illustrator CS3 or just go to even close the program I get this error:

    The operation cannot complete b/c of a unknown error [!] MP].  And then he follows with this window immediately after:

    An error occurred during the processing of the appearance of an object.

    It seems that in this one file.  I tried to copy the above work in a new file, but the same problem occurs.  I don't want to have to recreate this object again.  But do not know how to solve the error message.  Anyone know?

    Thank you

    It will take more time but try to copy the objects one layer at a time (assuming you used diapers, if no, then one at a time) and try to make the backup process. When you get the error again, start addressing the last items added to the new file.

  • vCenter 5.5 evaluation licenses expire before the 60 day period Eval

    I recently downloaded the vSphere, ESXi 5.5 ISO assessment the Client vSphere and vCenter Server .exe file 5.5.0b and ISO modules.  I was able to install ESXi on my HP Proliant DL360 G5 server in my lab at home without any problems.  I was also able to install vSphere Client.exe on my computer of client without any problems. Via vSphere Client, I was able to connect to my ESXi host and create VMs towing without problem. At this point, I am able to access the host and two virtual machines by using the client vSphere with no problems.

    I then install the vCenter Server on a virtual machine, I then use the vSphere Web Client to connect to vCenter without any problem.  I then create a data center with the Web Client without any problems.

    My problems begin when I try to add the host and two virtual machines I get a message saying that the licenses are expired and so I can't add my host to vCenter.  I don't understand why vCenter isn't allowing me to add the host because of the license, because I thought that evaluation licenses are good for 60 days. I just installed the vCenter server today and should have at least 50 more a few days before its expiration date.

    I go through the process of adding a new host, I can see where the discoverers of customer Web host, but it is not added to the Web Client of stocks due to the problem of the license has expired.

    Anyone have an idea how I can solve this problem.  Thank you very much.

    Decided to download and deploy the vCenter device. Everything seems to be running now. Problem solved.  Thank you very much.

  • Operations are underway. Please wait, the machine turns off automatically when the operations are completed. The machine does not turn off.

    Windows VISTA.  All updates applied.

    When the computer is restarted after some microsoft / windows updates, the step 1 and 2 are completed.  After the second stage, the above message appears.  I waited up to 12 hours, the remains of msg.  After putting off manually, turn on the machine and then restarted with the start/stop button, sometimes it restarts in safe mode (as is expected of manual stop), but not always.  Sometimes, it brings up the ordinary sign on the screen.  This has happened only after windows updates.

    Neal

    Hi Neal1209,

    The "powersuite" should have an option to restore certain registry errors that it finds. Is it possible that you can restore them? It may be possible that in this case it is actually deleting files that are required for Windows and believes are redundant or contains errors.

  • The operation cannot complete because of an unknown error. (CANT)?

    I tried to run the diagnostics, it did to the #2 test, seems to be locked up to 50%. In safe mode, a box comes up saying it as disabled plugin services.ip. And to try to update or uninstall. I tried to uninstall, did not work, I tried to check box is disabled, what has not worked. Does anyone else have this problem? Anyone know a fix?


    There was also a time when a list of plugins fighting 25 appeared in an error box.

    I'm on Mac


    Thank you

    camdude,

    You can try to reinstall using the way complete three steps (the otherwise strange things may persist):

    (1) uninstall (check the box to delete the preferences),

    (2) run the cleanup tool, see the link below,

    (3) reinstall.

    http://www.Adobe.com/support/contact/cscleanertool.html

    The cleaning tool works for all versions of CS3 on (you can see the top of the page).

    Here are the instructions for use, including individual applications:

    https://helpx.Adobe.com/creative-cloud/KB/CC-cleaner-tool-installation-problems.html

  • App Manager (Windows) installed Illustrator CS6 16.0.3, now I get the error "the operation cannot complete because of an unknown error [CANT] now what?

    Any suggestions? This error alternates with an error message saying that "some plug-ins have been updated." You will need to re-launch Illustrator so that the changes are taken into account. »

    Photoshop CS6 boots fine after the update.

    I knew I shouldn't have installed this update. Everything worked well.

    Thank you!

    Dean

    FIXED IT, happy happy, joy joy.

    I've changed preferences for D:\ disks and no Scratch, and error messages are all gone.

    Back to your regular programming.

    Dean

    PS-Trivia, I have used to be a support technician from phone to PC Freehand working at Aldus, in 1990. I learned a lot during this hell.

  • Interactive Control VISA & VISA Read timeout expired

    VISA control gurus,

    I am trying to contact my wireless AGILENT N8300A Test set.

    I am able to do the same thing via NI VISA Interactive Control...

    (Attached is the captain)

    But if I tried to do the same thing with a sample program.

    (Also attached below)... program shows a

    VISA: (Hex 0xBFFF0015) timeout expired before the operation is complete.

    I tried to use Async mode and synchronization in the VI... Reading VISA.

    But it always shows error timeout.

    I'm not able to understand what could be the difference, if it works well

    in an interactive control NI VISA, but not in my program

    Any help or pointers would be appreciated...

    Kind regards

    Diakité

    Hello

    very good that you have added the vi.

    You add a CR instead of a line break.

    You can do that in the string by selecting "------" mode for the string constant, you can use any string appearing in your comment as the display string constant.

    Good luck

  • Time-out occurred before the event waited for SequenceRunner simple

    I'm unable to these tests. I always get Timeout occurred for my sequence in the testOnLogout test.

    Any ideas?

    Time-out occurred before eventTimeout expected occurred before the EXPECTED event
    Error: Timeout occurred before the EXPECTED event
    at org.flexunit.internals.runners.statements::ExpectAsync/handleAsyncTimeOut() [E:\hudson\job s\FlexUnit4-Flex4.1\workspace\FlexUnit4\src\org\flexunit\internals\runners\statements\Expe ctAsync.as:319]
    at flash.events::EventDispatcher/dispatchEventFunction()
    at flash.events::EventDispatcher/dispatchEvent()
    at org.flexunit.async::AsyncHandler/handleTimeout() [E:\hudson\jobs\FlexUnit4-Flex4.1\workspa ce\FlexUnit4\src\org\flexunit\async\AsyncHandler.as:169]
    at flash.events::EventDispatcher/dispatchEventFunction()
    at flash.events::EventDispatcher/dispatchEvent()

    package tests.view
    {

    Import flexunit.framework.Assert;

    Import mx.events.FlexEvent;


    Import org.flexunit.async.Async;
    Import org.fluint.sequence.SequenceRunner;
    Import org.fluint.sequence.SequenceSetter;
    Import org.fluint.sequence.SequenceWaiter;
    Import org.fluint.uiImpersonation.UIImpersonator;

    view of the import. LoginForm;

    public class TestLoginForm
    {

    private var view: LoginForm;

    [Before (async, ui)]
    public void setUp(): void
    {
    view = new LoginForm;
    Async.proceedOnEvent (this, opinion, FlexEvent.CREATION_COMPLETE, 600);
    UIImpersonator.addChild (view);
    }

    [After (async, ui)]
    public function teardown (): void
    {
    UIImpersonator.removeChild (view);
    view = null;
    }


    [Test (async, ui)]
    public function testOnLogout (): void {}
    var passThroughData:Object = new Object();

    passThroughData.username = "test";
    passThroughData.password = "test";
    passThroughData.selectedIndex = 0;

    var sequence: SequenceRunner = new SequenceRunner (this);
    sequence.addStep (new SequenceSetter (view.username, {text:passThroughData.username}));}))
    sequence.addStep (new SequenceWaiter (view.username, FlexEvent.VALUE_COMMIT, 400));

    sequence.addStep (new SequenceSetter (view.password, {text:passThroughData.password}));}))
    sequence.addStep (new SequenceWaiter (view.password, FlexEvent.VALUE_COMMIT, 400));
    sequence.addStep (new SequenceSetter (view.domain, {selectedIndex:passThroughData.selectedIndex}));}))
    sequence.addStep (new SequenceWaiter (view.domain, FlexEvent.VALUE_COMMIT, 400));

    view.onLogout ();

    sequence.addStep (new SequenceWaiter (view.username, FlexEvent.VALUE_COMMIT, 400));
    sequence.addStep (new SequenceWaiter (view.password, FlexEvent.VALUE_COMMIT, 400));
    sequence.addStep (new SequenceWaiter (view.domain, FlexEvent.VALUE_COMMIT, 400));

    sequence.addAssertHandler (handleOnLogout, passThroughData);

    Sequence.Run ();
    }


    protected function handleOnLogout (passThroughData:Object): void {}
    Assert.assertEquals)
    View.UserName.Text, passThroughData.UserName,
    View.Password.Text, passThroughData.Password,
    (view.domain.selectedIndex, passThroughData.selectedIndex);
    }

    }
    }

    So I take a guess here that you want is the call of onLogout() occur as part of the sequence. But it's not what your code. Sequences are defined, then run when you call the run() method.

    Here you call the method onLogout in the middle of set the sequence... then some time later (probably after the disconnection is made) your sequence starts waiting for things happen... and to expire because nothing happens.

    I think that what you want here is to use the SequenceCaller class to actually perform the onLogout as part of the sequence.

    Mike

  • Permissions insufficient to complete the operation after a clean installation of Adobe Bridge in Win7 Pro

    Three installed days ago Adobe photography CC bundle, which includes Photoshop CC, CC LR and bridge. I've never had this software on my computer before. I am running Win7 Pro with all updates installed. I'm on a single computer, only user with administrative rights.

    I wanted to use the extended features of the bridge to reorganize my D about 30,000 files raw drive photo library. I moved a folder to a new location without any problem. Unfortunately, now I am more able to move, copy, rename, or delete files in Bridge after moving only one folder. I got the message that I had insufficient administrative rights to complete the operation. I have administrative rights. Why could I offers a folder but not others?

    I tried to rename, move files with Windows Explorer, but got the message: you need administrator rights to complete this operation. I clicked on the button continue and the operation is complete.

    I spent two days several hours the Webchat and nearly five hours on the phone with Adobe support technology. We found in the 'Hidden' administrator I can complete all the actions file/folder without problem. However, I can not yet perform the same operations in my original user account, and I repeat, with administrative rights! This has been verified to the DOS prompt.

    Adobe technology had me reset the permissions in the folder Adobe Advanced Security/owner of the tabs, but nothing helped. I can not yet use this expensive software due to problems of Windows permissions. Also note that the UAC is in its default settings. Windows warns far from reducing the lowest.

    I've scoured the Microsoft Windows website for answers and found no that relate to my situation/scenario.

    Help, please!

    The unknown account is referred to an account that you deleted from this machine. You must now seize the property of the D: drive, the new permissions recursively apply.

  • Loading screen is not animated, during the operation, but after

    Hi all!

    Here's my problem: I am display a gif animated using the class presented here.

    It is used in a specific class loader that perform an action, while the gif is displayed (the gif is a spinner). Only once the operation is completed, the LoadingScreen came out of the screen.

    Here's my class loader:

    public class LoadingScreen extends MainScreen {
    
        /**
         * The action to be processed during the loading.
         */
        Runnable action = null;
    
        private AnimatedGIFField _image;
    
        /**
         * Default constructor.
         *
         * @param action the action to manage.
         */
        public LoadingScreen(Runnable action)
        {
            super(FIELD_VCENTER);
            this.setTitle(CommonTools.initTitleBar());
            this.action = action;
    
            displayAnimation();
    
            if(this.action != null)
            {
                UiApplication.getUiApplication().invokeLater(this.action);
            }
        }
    
        /**
         * Display the animation during the loading process.
         */
        private void displayAnimation()
        {
            EncodedImage encImg =
                GIFEncodedImage.getEncodedImageResource("loading.gif");
            GIFEncodedImage img = (GIFEncodedImage) encImg;
    
            _image = new AnimatedGIFField(img);
    
            getMainManager().setBackground(BackgroundFactory.createSolidBackground(Color.BLACK));
    
            LabelField label = new LabelField("Connect..."){
                public void paint(Graphics g)
                {
                    g.setColor(Color.WHITE);
                    super.paint(g);
                    }
            };
    
            HorizontalFieldManager containerSpin = new HorizontalFieldManager(FIELD_HCENTER);
            HorizontalFieldManager containerLabel = new HorizontalFieldManager(FIELD_HCENTER);
            containerSpin.add(_image);
            containerLabel.add(label);
    
            containerSpin.setPadding(Display.getHeight()>>2, 0, 0, 0);
    
            this.add(containerSpin);
            this.add(containerLabel);
    
            KasualApplication.isLoading = true;
        }
    }
    

    The problem is that the gif is not animated, while the action runs. But if I have a popup appears at the end of the action to let the loading screen, I see that the gif is animated. As well as the animation starts when the action is complete.

    I think something is wrong in the clock of my son, but I don't see that.

    Could you please help me?

    Thanks in advance.

    It's ok, the solution is described here.

Maybe you are looking for

  • How can I get my new password when I can't connect to the e-mail?

    I am unable to connect and they said they would send me a new password, but how am I supposed to get when I can't get into my e-mail address? I have had this for a long time and we went for a week and now I am unable to enter my e-mail address.

  • OfficeJet 6500 Wireless: Lost after upgrade Windows 10 Scan function

    During the upgrade to Windows 7 to 10, I received a message that the printer drivers may not be installed correctly.  I uninstalled and reinstalled the full features of HP software for my wireless Officejet 6500 several times.  No apparent problem un

  • You want to remove FUNMOODS?

    Chrome users... Go to the key in the upper right corner after the removal of Funmoods in the Control Panel Add/Remove Programs configuration Click on Define the Pages then highlight the page Funmoods and click on the X in the right to get rid of! I d

  • A ' complete engine, but nothing is there

    I have a "compaq laptop" that is not connected to the internet, and my wife uses to play games. I checked it had "drive and saw that it was almost full. I opened it and saw that the only thing in there is a tool from Compaq for purposes of recovery t

  • My screen turned to his side :(

    MY WHOLE SCREEN turned on ITS SIDE so now I'm watching my laptop as a book to read things on it... my dog jumped on the key board and then turned over to one side, I can't seem to get back to normal, it must have been a touch:-S... someone help pleas