Display of recorder in Signal Express

I want to display a chart of tape display a signal from slow speed (f = 0.25 to 1.0 hz) signal that represents the position that results in a resistive (1 k) position linear tranducer (IE Unimeasure LX - PA series).  This view must move left as new signals are acquired, rather than updating a display both as a field of application; IE more recent signal is on the right side, and there are N seconds are displayed on the left edge.

I'm a THE Signal Express 'newbie '.  I have the signal from the NI USB 6211 correctly and displaying as a scope, but may not know how to change the view of what I want.

Any ideas or suggestions?

You must pass your signal through a step amplitude and levels:

  • Add step > analysis > temporal measures > Amplitudes and levels.
  • Go to Set Up settings and deselect RMS, leaving only the DC value controlled.
  • From there just drag and drop this signal (the newly created DC signal) in the graph
  • Right-click on the graph, and then select the Update Mode
  • You should now have the opportunity to see in band, scope, or sweeping graphic.

Hope that's what you're looking for!

Tags: NI Products

Similar Questions

  • How to display the signal on a waveform graph in Labview Signal Express?

    Hello

    I want to display a signal in Signal Express 3.0 in a "waveform table", but don't know how to do it and I think that it may be impossible?

    In Labview, there are two ways to view data, a 'waveform table' or a 'waveform graph. The great thing with 'picture of waveform' is that it allows you to set a length of history and you can see the data move to the left (option graphic strip) that samples are recovered. It works perfectly.

    In Signal Express I can only select "Graph XY" or "Waveform curve" by doing a right-click of the mouse, but I can't select "table of waveform. Is it really true that it is not possible to view data in a graph of waveform "with Labview? (1 analog signal during streaming, 100 samples to read at the rate of 1 K)

    Thank you in advance,

    Enrique

    Hi Enrique.

    You are right that there is currently no graphic waveform in SignalExpress. The thing nearest you can join, is saved data under a log and then he looks one when recording (which I know this isn't quite the same interactive behavior).

    For your last comment, you wrote 'waveform curve', but I guess you meant 'picture' here as well.

    Sorry about that. We recently received this request, then perhaps in a future version.

    Phil

  • NI USB-6251 - Signal Express - Input Ports display signals idential

    Hello

    I plugged a DET50B (phototedector) at the port of entry Ai0 in my case NI USB-6251 and thanks to LabView SingalExpress 3.0.0 took a look at the Ai0 signals and Ai1 by clicking on the following points:

    Acquisition of Signal-> DAQmx acquire-> analog input-> voltage

    I was expecting the Ai1 signal or 0 Volt, because there is nothing connected to it and the Ai0 signal vary in what concerns the light I was shinning on the photodetector. Instead, the Ai0 amplitudes both Ai1 were identical and corresponded to the light I'm shinning.

    How is it that these ports appear to be connected? It's the same thing when I opened other ports for viewing by SignalExpress except the Ai8; This one seemed to be at zero (probably because he does not physically exist on the NI USB 6251 which Ai0 - Ai7).

    No, you will not read 0 Volts to a channel not connected. Do a search for 'ghosts' for a detailed explanation. If you want 0 voltas on a channel, you need to gnd.

    p.s. You do not have a signal generator Board. New questions should be mailed to multifunction DAQ and Signal Express cards

  • LEGO Mindstorms 2.0 crashes at startup when LabView Signal Express is installed first

    When NI Circuit Design Suite education edition is first installed in a computer and the LEGO Mindstorms software second, the LEGO Mindstorms software does not work, he reported the following:

    Error: A required file is broken

    Possible solutions: reinstall the driver of LEGO Mindstorms NXT (tried but does not work)

    Error code: 1003

    Ordinal not found

    The 18 ordinal not found in the library of dynamic links NIVISV32.dll

    Error loading "fantom.dll".

    The operating system cannot run %1

    On other computers the LEGO software starts, it displays the part of the initial screen and then a small window will open saying that the Mindstorms software has made an error and that it will close. The previous report does not display.

    If LabView Signal Express is uninstalled and reinstalled the software, LEGO, the LEGO software works.

    I downloaded and installed the new and old of the NIVISV32.dll versions and is not correct the problem. I went as far as version 4.20 of the dll because it is dated in 2007 and dll files installed by the LEGO software when it works is also dated 2007, but if I install the dll without uninstalling LabView Signal Express, the continuous LEGO software does not.

    Please advice how to fix this without having to uninstall first LabView Signal Express. If the LEGO software is installed first, LabView Signal Express will use and no not to replace the old version of the dll 2007 installed by the LEGO software? Y at - it an update of the LEGO software that corrects this problem?

    I repeat: I know a solution is to uninstall LabView Signal Express, install LEGO Mindstorms and then reinstall LabView Signal Express but I have 60 computers already cloned and deployed with this problem, so I don't want to go to this long process to solve the problem. Computers use Windows 7 and the problems occur on both 32 and 64-bit versions of Windows 7.

    Any ideas?


  • Double counters in Signal Express with a USB-6229

    I have a USB-6229, which I'm running via SignalExpress, and I am trying to simultaneously record two inputs of meter.  However, it seems that he has trouble to add the second channel.  When I add the first string via signals acquire > DAQmx Acquire > counter entry > Position > angular and select ctr0 channel, he (correctly) lists PFI8 and PFI10 as inputs of channels A and B.  However, when I add the ctr1 channel to the same task (via the blue and the button "Add a channel"), he still lists the same PFI8 and PFI10 as input.

    On the other hand, if I first add ctr1 then add ctr0 Secondly, it lists PFI3 and PFI11 as A and channel B for the two entries of meter.

    How can I get the two counters operating independently?

    Thank you

    Josh

    He realized that I can't run two meters of a single task in Signal Express (unlike, say, two analog inputs).  Once I added it as two separate tasks, everything works fine.

    Now I have problems of synchronization.  Looking through these tips and documentation, it seems that I will not be able to fix this via SignalExpress, so I continue in LabView.  I'm having some problems with it, but I'll post in a new thread here.

  • Turning off IEPE signal Express with 9234?

    I have a module 9234 and I can't understand how to disable IEPE while using the Signal Express.  A new tab named "Device" appears when you watch 9234 modules, but which has only a drop-down list to select the AC/DC coupling.  MAX displays a checkbox for IEPE, but I can't find that in the Express of the Signal.

    Any suggestions?

    Thank you

    Hi ChrisCW,

    In order to set the IEPE, you must choose a task of nonvoltage such as acoustic pressure or acceleration.  You can see in the photo I have both source IEPE and value available.  If a task of tension is the only one that has meaning for your application, you can always use one of the other and adjust the data.

    I'll go ahead and submit a product idea to have this added to the task of blood also.

  • How to record a signal all on *.wav

    I am trying to record a *.wav sound generated by the sinus function. I cannot record the signal of all, I can only get 1 second sound - I'm kind new to this and I have so I am struggling to get this to work.

    Due to the writing to the wav format does not allow to add (I guess I might need to use another way to save information), I put next to the loop so that the file save only one loop is over. I need to find a way to get all the information in the queue, so I use Mathscript

    Down while the loop has a tunnel that will display only the data of the last iteration.

    What you need to do is add data, but put the waveform in a registry change so that it is available for the next iteration to be able to add more data to it.

  • Application using Signal Express RTSI

    Hello

    I am interested in recording of analog signal synchronization (start time, acquisition of the sample, 3 cards PCI-6289 stop time) and possibly their synchronization with 3 to 5 more cards NI PXI-6221, that are connected to the same host PC via a PCIe-8361 interface PXI-8360.  For now, I will focus on the 3 cards PCI-6289.  I used a RTSI cable with VI recorder is no longer supported and guess that it works well with Signal Express.  I've set up a device cable RTSI NOR able & Automation, which includes 3 PCI cards.  3 PCI cards are the only devices connected to the RTSI cable.  When you check the properties of the RTSI cable device, I left all the boxes unchecked, that seems to be the right setting for my application.  In Signal Express, when you add a device in the Panel on the left of data acquisition (analog voltage signals), I always see each individual card (Dev1, Dev2, Dev3, etc.), but there is no device RTSI.  If I look into the advanced features of synchronization of each of the three panels DAQAcquire, I can apparently use a card as a master clock and the two other cards will have master card sync signals (is it automatically via the RTSI cable, or do I need to connect external clock out on the master card to the clock on the slave cards?) but once again There is no mention of the cable RTSI, or configuration of device RTSI in MAX.  Using the clock of a card in a situation of master/slave help, but it starts and stops all three cards at the same time, or no apparent synchronization between maps will simply be a coincidence?

    Any suggestions are appreciated.

    Kind regards

    Brian

    You really need your question to the SignalExpress Commission. Few people here use it.

  • How to analyze the data of the cDAQ and Signal Express, especially after analysis?

    In the first series of tests of my instrument, it took longer than expected for the race.  Thus, the data was saved in 6 days.  The file is too large for export to Excel.  At the beginning of the project, I was as ignorant as I could go ahead and add analysis and the scaling of measures.  By the scaling, I mean my data of switching current dew points or whatever it is that I record.

    How to evolve the data to read the output data as expected 4mA = point of dew of-20 C or 0 PSIG?  Can I pre program this to be recognized for each event?

    For real analysis I am doing – I would first analyze the data I recorded and choose different points to send to Excel to graph and analyze.  Is this possible?

    Secondly, I would like to know how to scale and analyze my data in the project without having to do this later analysis in the future?

    I have a cDAQ-9172 with LabVIEW signal Express 3.0 that uses four modules - 9211 2 modules of thermocouple, my 4-20 1-9201 module +/-10V module and 1-9203.

    Thank you for any assistance.

    Hi Patricia,

    "' You can do this by adding a step Load/Save signals ' analog '.  I hope this helps!

  • LabVIEW Signal Express 2012

    We have Labview 2012 Signal Express (version 6.0.0) installed on an XP computer and we want to upgrade to a Windows 7 machine.  Will be the last version of Labview 2012 to work on the Windows 7 machine?  The 2012 Labview Signal Express is used to control a cDAQ-9172 with module 4 channels thermocouple.

    It's the Commission of LabVIEW. Please, click on Options and ask that the moderator moving your question to the Board of Directors SignalExpress. LabVIEW itself runs on Windows 7, but they are two different products and not many people here use SE.

  • How do I capture the output of voltage full bridge with Signal Express NI9219

    Hello.  I'm trying to do and calibrate a load cell with the installation of full-bridge strain gage.  I use a NI9219 module with a cDAQ chassis.  Is it possible to capture the actual output voltage?  Signal Express gives me a value of strain, but I really need to know the output voltage.  Where to look.  I need only two channels for full-bridge.  I think that could connect the wires to the two remaining channels and read the output voltage of the strain gauges which would be connected as a tension of the 9219 entry, but I think that Signal Express could give me the voltage and output voltage directly.  Any input would be appreciated. Thank you!  P.S. I only use this equipment on occasion and am not the more familiar with it, so keep things simple for me.  Thanks again.

    Hi jgh@AET,

    The NI 9219 measures the ratio of voltage full bridge in hardware sensors, allowing any variation of the voltage to cancel. You won't be able to measure the output voltage of the sensor regardless of the voltage without additional channels, but you can measure the ratio of raw tension using the type of Bridge (V/V) . You can also use the type of measure of Force (bridge) measurement of load cell with engineering units (N, lb, kgf, no strain).

    This screenshot shows where the two Bridge (V/V) and Force (bridge) can be selected in the DAQ Assistant:

    These types of measurement were added sometimes around DAQmx 9.1, so if you have an older version of NOR-DAQmx, your DAQ Assistant maybe not them. The latest version is currently 9.4 of NOR-DAQmx. Front of NOR-DAQmx 9.1, the approach to recommend to measure the load cells was to use the custom with Excitation voltage type and a custom scale. However, Tension Custom excitedly can't Bridge of calibration in the DAQ Assistant.

    Brad

  • How can I use the USRP to record a signal using its two RX ports simultaneously?

    Hello.

    I am trying to record a signal using two antenna cone. The reason that I need two antenna to cover the bandwidth (DC - 6 GHz). a single antenna covers DC - 300 MHz and the other covers 300 MHz to 6 GHz. so I need to use two RX port of USRP at the same time to record the signal. I have two questions:

    1. is this all USRP market capable of covering this frequency range?

    2. is it possible to use the two RX port at the same time to the signals of the records I described? If this is not the case, how can do?

    P.S. I have two NI2920 USRPs and two USRPs N210 in my lab.

    Thanks in advance for your time.

    Sam.

    Hi Sam,

    To answer your first question, the USRPs you can reach the bandwidth you want. There is not a USRP, to my knowledge, that can reach this range in a single device.

    Also note that you can only use RX convened for two different ports at the same time using LabVIEW and the pilot of the USRP. If you want to use the two lines of RX, you will need to run a session with a single line, close the session and then start a different session for your second RX line.

  • Signal Express version unknown problem with Labview VI

    Hi all

    I am trying to run 2010 Labview Signal Express VI. It gives me "unknown version" problem. When I checked for Labview versions, he said 10.0f2 and Signal Express can read the version of LabView 2010. I don't understand this. 10.0f2 is not just a patch for version 10.0? I will appreciate if someone can help me with this.

    Kind regards

    Kanu

    Hi Kanu,

    I tried to write a simple VI and its importation to the Express Signal and I have not had any problems. I use 32-bit LV2010f2 and Signal Express 2010. I have attached the VI I used so that you can try it yourself. I followed the instructions of this KB to create a distribution of the source file. If you follow this KB and still can't make it work, can you please post the VI?

    Thank you

  • Questions about the Installation of Signal Express Version 2.5.0

    Hello

    I have a silly question on installation of Singal Express Version 2.5.0. Previous student installed on the computer in my Department, but I can't find the installation CD of Signal Express 2.5.0. The only thing I found is Signal Express Verstion 1.1.1. Is there supposed to be any extra CD to update. I need to install Express of Signal on my laptop. Thank you very much.

    Ningyu

    Hello

    Please post on the Forums of NOR. Unfortunately, we are unable to upgrade to version 2.5.0 of the 1.1.1 cd. However, if you have your serial number, you can download the evaluation version of 2.5.0 section and activate it with your serial number and the computer. The download is located here http://joule.ni.com/nidu/cds/view/p/id/981/lang/en. I hope this helps!

  • Analysis of frequency common time in Signal Express?

    Is it possible to do joint analysis frequency time Signal express?  I don't see in the section "analysis", so I guess that it is a separate tool that should be added, if possible still Signal Express.  Help?

    SignalExpress does not natively support joint time frequency analysis.  However, check out the Sound and Vibration toolkit that expands the SignalExpress.  It can do what you want.

Maybe you are looking for

  • more than 6 s gets very hot and the battery runs out in a few seconds

    Yesterday when I woke up I discovered that my phone was like on fire and after I got battery emptied in seconds, now I try to turn it on, but is not, and when I charge it it gets again so hot that I can't even touch. I took my phone to the apple stor

  • How can I open a 'CSV '?

    I Rec'vd my bank statement and instead of him being in a 'PDF' File Format it downloaded in CSV Format! I don't have Excel etc. Just Notepad + WORDPAD.  It is NOT of course the life or death, I'm just curious if this is something that I can fix it. I

  • How can I activate the scanner to the computer?

    HP Photosmart scan message button tells me to start the search from my computer. How do I do that? Not the same as others, I used.

  • Black screen after awhile

    What happened only a few times. It seems that when it is on for a while and I play maybe a Facebook (game) application or something else. After a while the screen just goes black. I waited a few minutes and nothing happened, so I knew nothing else to

  • Compaq Presario: updates

    Compaq Presario SR1950nx (converted to Windows 7) I tried to update my pc. It will not update. I went to "View history updated", each of them has failed. I tried to install my printer. It will not be installed. I would like to know what the problem i