How interface bronkhorst RS232 communication flowmeter in DIAdem DAC?

Hello

I am Thiyagi.I am working in WABCO INDIA Ltd. I bought Bronkhorst Flowmeter. I use the tiara DAC Professional version. I want the meter to DIAdem.Flowmeter of the interface have a RS232 communication. In my pc, I have a COM1 port. DIAdem DAC which icon I want to use to connect the flow meter.

Hello hmongz

You can already succeeded, but still, here's what I used to do:

I could do this with flowmeters mass Bronkhorst and Brooks, and had a pretty good success using their method of DDE (dynamic data exchange). To do this, download their software from FlowDDE of Bronkhorst.com, start the DDE driver and he can communicate with the MFM. Once it works, insert a DDE lockk in CAD DIAdem, configure it appropriately and it should work.

Greetings

Michael

Tags: NI Software

Similar Questions

  • How to start a community Homekit (home app)?

    Homekit needs love. How to start a community of support for her?

    Excellent idea. I forwarded your suggestion to the moderators of Apple. Be aware, it takes time to get the results.

  • How can I cancel community support?

    How can I cancel community support?

    There is nothing to cancel.  You just stop asking questions here.

  • Chroma DC power supply RS232 communication (read problem of over-voltage and current)

    Dear all.

    I chroma programmable DC power. Based on the programming of the Instrument manual I develop using RS232 communication. Based on the program I can set the voltage, current, over-current protection, protection against overvoltages and make IT / OFF out put supply perfectly.

    But I have to read the measured values of the output power as current and voltage. Measured applications are the voltage and current of the output of the power supply. My problem is two of them read at the same time. Currently, the reading is only voaltge or current (if the first request is v? it is voltage read out but no reading for the current) and if the first request is CURR? the reading for the current, but not for VOLT? The status message is OK, even if it is to read values.

    Thanks in advance

    What I see in your program, it's that you do not use the stop for reading character. For your writing, you do the hard with all these concantanate string functions. You can set the stop character for all entries with a node unique propert - "ASRL end Out.

  • Can someone tell me how to run Office Communicator so that it is always on when a user opens a session? __

    Hello

    Can someone tell me how to run Office Communicator so that it is always on when a user opens a session? We want people to turn it off.

    Thanks in advance

    http://Office.Microsoft.com/en-AU/Communicator/default.aspx

    http://answers.Microsoft.com/en-us/Office/default.aspx

    Office at the above link forums

    See you soon.

    Mick Murphy - Microsoft partner

  • Cannot acquire with cDAQ in hardware mode timed using DIAdem DAC

    Hi all

    I use DIAdem DAC to acquire the cDAQ + 9225 + 9219. 9219 acquires the use of single point mode that works very well. However, I tried to acquire waveforms 50 kHz 9225 with the file attached dac. However, I found a problem on the block of the clock, as shown below.

    How could I solve this problem and gain the waveform with 9225 using DIAdem DAC?

    Sincerely, Kate

    Hello Kate,

    You must connect the NIDAQ-In1 exit the Save block block. In addition, you must change the parameters of the horloge1 block. The material is always measured in mode "DAC-kernel". "The software clock" is used for frequencies of up to 10 kHz clock (depending on the material and the PC). If you want to measure more quickly, you can use the "Clock" mode

    Greetings

    Walter

  • How can I set more than one character of endpoint for interface RS-232 communication?

    Hello

    I have to communicate via the interface RS-232 of the PC with a card microprocessor. I use the "Advanced Serial write and Read.vi ' off examples of LabVIEW and it works on the principle.
    "My problem is the stop character: it looks like I can put a stop character, IS 0xA (so LF / line break) OR 0xD (CR / carriage return), but I have the need to react on the different characters of endpoint as"OK"gold""; I want to set a control (this VI will be used in a TestStand sequence).
    Is it possible to do, so that I can put characters from end of Cork I need? The property node in "VISA set up serial port" just accept an unsigned integer, nothing else. Of course I can get it running with the time-out-, but it is not very pleasant and extends the duration of trial unnecessarily.

    Thank you best regards &!

    It would be for writing. The poster wondered about reading. To read it, you will need to write your own routine of reading. Basically, you will need to use a loop and inside the loop, use the VISA to the Serial Port bytes to see if there are all the bytes available. Read the number of bytes and add to a string that you build in a shift register. After verification of concatenation to see if you have the characters of endpoint. If so, the reading is done. Pretty simple. Don't forget to disable the option enable the termination character in VISA configuration.

  • Simple serial using RS232 communication

    Hello

    I did some research on how to read and write with the serial RS232 port. The only example I found was a test program that is really involved. All I do is pass 7 bits of information that are going through a digital communication kit then read the same 7 bits of information after the kit Comm. (bits, parity of 7 bit stop bits not included... ect).

    So I just need to

    -Desired TX 7 bits of information comm kit (in addition to the necessary stop bits)

    -Information cross kit comm (short delay)

    -PC receives the same bits of return.

    Any help would be appreciated!

    IM using CVI version 9

    The byte field should be the data you want to send the kit of comm.

    It's something you need to know.

    Status fields give you information on the question of if the function call completed successfully.

    Most of the time 0 means success and error average of negative values. But you should read the function for each function to be sure about this.

    When you receive an error code you can use the GetRS232ErrorString function to get a verbal description of the error.

    number of bytes written, as its name implies, represents the number of bytes transferred to the COM port. You wait for the value to be equal to the number of bytes to send to your application.

    Hope this helps,

  • How to solve the communication error series of HP34970?

    I want to use the serial port to commnucation between pc and hb34970. Using gpib, it's ok. But in serial mode, report the error. just like that:

    for com2 layout:

    16:11:47.359 26-8-2009 [Init_dq_Device] error:-1073807339 - HP34970A Initialize.vi-> Init_DQ_Device.vi-> main_app.vi VISA read.visa name: ASRL2::INSTR

    for the implementation of com1:

    16:13:25.921 26-8-2009 [Init_dq_Device] error:-1073807339 - HP34970A Initialize.vi-> Init_DQ_Device.vi-> main_app.vi VISA read.visa name: ASRL1::INSTR

    I don't know the meaning of 1073807339 and if I should together: ASRL1::INSTR for com1. Can someone help me on this problem.

    Thank you very much.

    fix my Subvi.

    MMM,

    I think that you can check your cable again.  Cable series come two types, Straight (patch) or null.  This is because the serial devices can be DCE (Data Communications Equipment) and DTE (Data Terminal Equipment).  You have proven the cable is good using it with another piece of equipment, but you have not proven that it is the right TYPE of cable frompg 273 of the manual you need a null-modem cable is the DTE instrument

    Consult the manual of user 34790A.  You must also implement the 34790A wait RS232.  From the front panel, you will need to access the menu of the interface and select RS - 232 and configure the port settings. Did you do that?  Verify that the port settings match the yourPC com port settings

    Another problem with Set - very common up is to connect to the interface of trigger/alarm by mistake.  The serial port is the lower connector 9 pins on the back panel of the 34790As

  • How to remove the community

    I want to completely remove a community including all data.

    How do I do that?

    I sent [email protected] , and they offered to turn it off and remove the community. Apparently this can be done from the user interface.

  • How to concatenate files of CT in DiaDem

    I saw the post on how to concatenate CSV files on tiara... but how do concatenate you files TDM together?

    I am surprised that this feature is not built into the software as it is a fairly common task, and the script is probably beyond the capacity of most of the user to maintain.

    Hello Colin,

    There are several examples that you can use to concatenate the data channels in DIAdem.

    This one is on the site:

    http://sine.NI.com/DevZone/CDA/EPD/p/ID/3870

    Another good example is included in the tiara and accessible from the help > menu examples. From there on, drill down to the folder: "examples > scripting > Scripts > add channels to each other.

    Between these two examples, we should hopefully take care of your issue.

    Best regards

    Otmar

  • Serial RS232 Communication VISA write does not

    I am trying to send a SPELL of 5003 01 command to my card that I am logged in via Rs232. This command should get my PCB to change play modes.

    I posted about this problem before thinking it was a problem with it not sent in HEXADECIMAL:

    http://forums.NI.com/T5/LabVIEW/RS232-serial-communications-visa-using-hex-values-seem-to-be/m-p/326...

    It wasn't the problem.

    I use the CTS line for flow control. I know that the CTS signal is reaching the computer of my PCB as I tested it and I know that the PCB can send data to the computer I get the answer correct 4 bytes in HEXADECIMAL when I use the reading visa to get my player to PCB data.

    I get a 1073807339 error when I try to write even when the read function is disabled.

    The only thing I can think is that LabView is not to recognize the signal CTS go down and, consequently, time-out.

    I use the good 9600 baud rate, 8 bits of data and no parity, 1 stop bit. I am using RTS/CTS flow control, but the RTS line is not connected. I don't think it should be a problem that RTS is used to indicate the Council when to send data to the PC to my knowledge and is not serious.

    FYI the written string section of the VI is unused. The string to write, that's where I'm seizes my "5003 01" command.

    Thank you

    Kyle

    I just thought of it. I had messed up the wiring of the PSTN line. She received the RTS line, I think. I thought I had it correct since the Tx line was correct. Thank you for your help on the problem of the HEXAGON.

  • RS232 Communication tracking

    Please is there a way to control the traffic of communication RS232 via VI?

    OT sounds like you want to 'see' what the series is then port that another program is using the port as a sniffer port?

    If LabView won't do that as LabView and Ni-VISA must have access to the serial port.

    If another program is using the port NI-VISA and LabView will be locked out.

    If you start a porgram LabView NI-VISA first will lock port so no other program can access it.

    I think you need a program like part of the Sysinternals Suite from MS Technet PortMon .

  • How to enable wireless communication between two clients?

    I've got a WRT54GL with the latest firmware.  I have two computers that are both connected to the router with the IP via DHCP addresses.  Nor has a firewall running.  I can't communicate with them - even ping does not work.  I can ping the router itself by using the IP address assigned to this side WAN by my ISP.   Both computers have no problem to reach the internet through the router.

    What the router settings will allow communication to occur?  I can't find anything in the user interface of the router, which seems to control it.

    Thank you.

    I just figured out the problem: Isolation AP (in advanced wireless settings) has been activated.  I turned it off, and the problem is solved.

  • How to change My Community email address

    How can I change the email address attached in my community?

    Hello Scooter46,

    You can change your details in your profile.
    To access your profile page, use the following link.

    http://h30434.www3.HP.com/T5/user/myprofilepage/tab/personal-profile

    Once there, find the 'HP Edit passport' and to the right of this 'coordinated '.

    From there, you will be able to change your email address.

    Kind regards
    Kyte

Maybe you are looking for