cDAQ-9174 error-89137

I have a cDAQ-9174, I want to configure to use and DI, at the same time in two different tasks. It gives me the following-89137 error:

Possible reasons:

Specified route can not be satisfied, because it requires resources that are currently in use by another route.

Source device: cDAQ1
Terminal: / SampleClock
Target unit: cDAQ1
Destination terminal: Slot1/ConvertPulse

Resources in use by
Task name: _unnamedTask<32>
Source device: cDAQ1
Point source: di/SampleClock
Target unit: cDAQ1
Destination terminal: Slot1/ConvertPulse

Task name: _unnamedTask<33>

I understand that this is because some courses is in use, but I don't know what that means. I tried to reference the peripheral roads in MAX, but I don't really understand. I have attached my code. You can see that I want to read and write at the same time, which I think should be a fairly simple thing to do. I found a note on the site NOR on the definition of the clocks with the Timekeeping property node. I tried to do, but I couldn't make it work anyway.

Can someone tell me how I can use DI and PERFORM tasks on the cDAQ even at the same time?

I figured out that I can not do readings of equipment timed at the same time, I do timed written with this module, so that was the problem.

Tags: NI Hardware

Similar Questions

  • Simultaneous release of digital waveform at different frequencies - cDAQ-9174 with two 9477 maps error-50103

    Hello

    I'm working on a project that requires two outputs digital signals at different frequencies. Frequency of the phase 1 is about 1 kHz and is modulated on and off to a pace that will change during the execution of the program. Wave 2 passes from 1 kHz to about 6 kHz while needing to be pretty accurate to the tenth of a Hertz.

    Initially, I tried to manage the simultaneous output of signals at different frequencies using a single task on a single 9477 daqmx in a cdaq-9174 chassis. As far as I know, the best way to get a specific frequency in a waveform output is to set the sampling frequency up to 2 times the frequency of the wave that is generated and generate a waveform that is given to each clock cycle. This works very well when it comes to a gesture, but I was unable to get the frequency of the modulated wave (wave 1) remains constant when the frequency of wave 2 modified or vice versa.

    I have a few other modules lying around (another 9477, 9403 and 9476) and I thought I would try another task running on a separate module. I find myself receive error-50103 message if I add these modules to my cdaq chassis and run one of the waveforms of a task set to run on the add-on. Is there a way to bypass this error? I guess it would cause by running two digital output on the cdaq-9174 tasks at the same time, but it seems to me that this wouldn't be a problem with an additional module.

    How can I have two outputs digital signals, running at the same time, maintaining their independent frequency frequency of the clock of the other sampling rate changes. In addition, because the wave 1 is enabled and disabled at a defined frequency, I use it to set the number of samples to write when you write a range of waveforms in the module (when generating these two waveforms on the same module).

    I'm sorry if the explanation was difficult to understand. I've attached an example of error 50103 lifting. This isn't really a part of my project as a whole, but it is the easiest way to reproduce the error.

    Thank you

    Hi awol.

    I have a follow up for you. You encounter the error because the cDAQ chassis has only a timing engine of. You will not be able to perform simultaneous tasks of the call by the hardware. Please see the following knowledge base: http://digital.ni.com/public.nsf/allkb/5E0B829E50ADE1BC86257AC50062B2D2

    Mike

  • Difference between the cDAQ-9174 and cDAQ-9178

    I have several cRIO-9081.  Each cRIO-9081 running NOR-DAQmx 15.5.  On four of the cRIO-9018 we show you cDAQ - 9178 with 8 modules.  Everything just worked.  On the fifth cRIO-9018, we tried to connect a cDAQ-9174.  The aircraft appeared in the Windows Device Manager and MAX, but MAX has indicated that the aircraft was missing (error-88705) and that Windows can't find a driver for the instrument.

    We then plugged a cDAQ-9178 in the fifth cRIO and everything just worked.

    What is the difference between a cDAQ-9174 and cDAQ-9178 other than the number of the module support?

    Paul

    Hi Paul,.

    It seems it could be a driver problem! What DAQmx version you have installed on two computers? Reinstall the latest version may be worth it because it looks like there may be corruption. You can find the latest version here: http://www.ni.com/download/ni-daqmx-15.5/5901/en/

    You can manually associate the device driver, here: http://digital.ni.com/public.nsf/allkb/5B80CAAA64413B4C86256C5E0000C1A7

    There are modules connected to the cDAQ chassis?

    Also, sometimes, this can happen if you do not restart your computer after the installation of the NOR-DAQmx driver. Restart your computer after the driver installation will allow you to reserve the device successfully. Here are a few others not to try:

    • Check the charger to device OR to see if it is running

    • Reset the service OR charger

    • Reset the MAX Configuration database

    • The system power cycle

  • problem installing driver cDAQ 9174

    Hello

    I am a new user of LabVIEW and just got our first shipement material today of 2 x cDaq 9174 with analog i/o.

    I tried to install the driver for several hours without success. At the beginning of the installation, I get an error message that says:

    "This installer needs to access 'My images' However this place XXX is not available." Make sure that this directory exists and is available, and rerun Setup.

    First of all, the folder my pictures pointed to a NAS drive where all of our photos of the company are located (which was the N drive) - so I show it to the location of the normal C drive but it is still giving me an error message and try to find the folder in drive instead of C N (and Yes I tried to reboot).

    Secondly, the location was available, as I opened through Explorer without any problem.

    Thirdly, why a driver would fail to install simply because it could not access the folder my pictures?

    Any help would be appreciated.

    Hi Doug_1970,

    It may be a registry key which is always pointing to the NAS drive. Try to use the following Microsoft Fix it tool, reboot, then install DAQmx.

    http://support.Microsoft.com/kb/886549

  • cDAQ-9174 cannot use the counters of 3 or 4

    Hello

    I try to display the entry of a potentiometer and 3 encoders of Quad.  I have a CDAQ-9174 with a 9215 OR (for the POT) in the slot 1, 9411 or in slot 2 and a NI 9401 in the Groove 3 (soon to another NOR 9411 when it happens).

    If I run the example 'Measure angular Position.vi' on each individual encoder (by selecting either modules on counter 0), I can see all three encoders work properly (exit count increments and decrements as expected)

    If I run the Test panels able and automation, I see all three encoders switch their respective 2 channels each.

    But when I change 'measurement Position.vi"to have all three separate entries (I copied the task of creating and all its entries twice, then expanded the while loop to have three different DaqMX Read vi, then 2 clearer spots on the right side (output) at all in a loop), I see only 2 encoders increment and decrement.

    I do not use Index Z entries, I changed the type of decoding to X 2 for all three.  I tried to change that, but I especially get the same result (2 encoders working, is not 1).

    I changed the encoders are the use of counters to 0, 1, and 2.  Only the assigned to counter 2 encoder does not work.

    I tried to use meter 3 - same problem.

    Is there a way to confirm that the cDAQ-9174 counters all work?  I tried all three in the test panels and they seem to be good?

    A cDAQ-9174 can manage 2 Quad encoders using 2 signals of each (A and B signals - A for metering, B for management)?

    I have something configured incorrectly?  If you use two Modules of entry at the same time, is it some additional configuration that must be done?

    Really appreciate some help here.

    Thank you

    Sam

    Yes, we use the 9.0.2 version drivers and no there is no error message.

    Turns out we had wired it incorrectly for Counter 2 use.  Being on the second module, I thought we had wire the same way the first module has been wired - not realizing that counters need a provision of the specific pins.

    Thank you for looking into this.

  • Problem with multitasking, Amnesty International and: error-89137

    I have a PXI system with two modules in the series X-series SMU-6345. I need to create a task HAVE meter sampling, to acquire the data of channels in these two modules, that works very well. My application requires then to create a task and task object some time later, during execution of the task to HAVE it. But I'm getting following error:

    Error-89137 occurred at Test_DO_Write chn.vi

    Possible reasons:

    Specified route can not be satisfied, because it requires resources that are currently in use by another route.

    Property: RefClk.Src
    Source device: PXI1Slot2
    Terminal source: no

    Resources in use by
    Task name: _unnamedTask<53>
    Source device: PXI1Slot2
    Terminal of source: PXIe_Clk100
    Target unit: PXI1Slot2
    Destination terminal: RefClockInternal

    Task name: _unnamedTask<51>

    (unnamedTask<51> 's task and HAVE _unnamedTask<53> is in the post above)

    I get this error when I try to create a new task to DO or a counter in the same module 6345, during an acquisition of GOT underway. It seems that the task tries to access the line of Int.Ref.Clock, a 100 MHZ clock is already for the task to HAVE it. I tried to assign this clock of 100 MHZ, as clock Ref. for the task of MAKING, using hardware node. But it returns error saying "not supported; harware.

    I enclose a VI to explain this problem.

    The DIF has been resolved. "By defining the PXI 100 MHZ clock bottom of basket (PXIe_Clk100) as the reference clock for both HAVE and PERFORM tasks, the use of property nodes DAQmx Timing 'ReferenceClock.Source' and ReferenceClock.Rate ', things are working properly.

  • Generate a pulse train, NI 9402 modules in cDAQ-9174 chassis

    I have two modules NI 9402 in a cDAQ-9174 chassis.

    When I output a pulse train on a specific line of the PFI to a specific module, the pulse train is out on the right line of PFI, but on BOTH modules.

    I want that the pulse train out only one of the modules.

    for example, I select cDAQ2Mod1/ctr1 to output a pulse train on PFI3 of module one.  I hear the pulse module one PFI3 train, but I also get it on PFI3 of module 2.

    I am also a measurement of separation of two edges with a different counter, but I don't seem to have this problem.  (The measure only works when I have the signals connected to the module that I've specified.)

    -Paul

    This is the Vi I work with.

    Hmmm...  Looks like it's actually only after I exit a signal on that line.  Maybe I should try to clear the line.

    -Paul

  • synchronization and clock cDAQ-9174

    I have a cDAQ-9174 with chassis

    HAVE - 9205 OR 9205

    HAVE - NOR 9219

    DIO - OR 9401.

    "NOR-DAQmx allows the on-board clock routing synchronize acquisition or generation of data transactions. A single sync signal or a trigger can be shared between multiple operations on the same device to ensure that the data are synchronized. These signals is shared by the simple signals routing functions that allow connections integrated hardware DAQ. " (from http://zone.ni.com/devzone/cda/tut/p/id/6829)

    When I check the itineraries device in MAX tab' for chassis and c modules individually, I found many clocks with the Green blocks (direct Route). "Once a physical connection is determined, the NOR-DAQmx driver lets the user quickly and easily use this connection to synchronize operations".

    Now my question is how do you choose among these clocks available. What is the frequency of these clocks? The highest frequency the better? What is the difference between /Ctr0SampleClock, /Ctr0InternalOutput, /Ctr0Source, /Ctr0A?

    For example, why well/Dev1/I/SampleClock or Ctr0InternalOutput or of others, especially as a source for the .vi DAQmx Timing (sample clock)?

    Thanks a lot for your help

    Bing

    Concerning

    Hello NCLbingji,

    I think you mix instructions from the manuals of different material. The link you posted refers to the synchronization of several separate devices, while your cDAQ with its modules really is like having one device almost. You will find more relevant information in your manual of cDAQ chassis. For multiple devices, you indeed have a "physical connection", but with the cDAQ is not necessary. You set up these connections in the software in any way, and this should get the routing sorting at the driver level.

    Try to identify which channel will be your a 'primary', the one driving all others and synchronization, the value then asked SampleClock of this string as the rest (DAQmx Trigger.vi) trigger ('secondary') channels, then start all 'secondary' jobs and finally start your "primary" task This will trigger all tasks at the same time. Take a look at the attached picture, which is another application that I worked on, but I think this will help.

    I hope this helps.

    Kind regards

    Michael S.
    Technical sales engineer
    NEITHER UK & Ireland

  • DASYLab w / support NOR-DAQmx NI 9227, NI 9229 &amp; cDAQ 9174

    I am trying to use DASYLab 11 w / NOR-DAQmx 8.5 to my configuration material of NI 9227, NOR 9229 & cDAQ-9174. After decommissioning of my PC Win7 for XP that takes in charge OR-DAQmx 8.5 (the OR-DAQmx last, supported by 11 DASYLab), I discovered that NEITHER-DAQmx 8.5 does not support the cDAQ 9174. Right now, looks like I'm watered except so use LABView. Is there any other way at this time to use this hardware configuration with DASYLab or any other application?

    DASYLab V11 supports Windows 7 when you install the Service Pack (see www.dasylab.com).

    Where are we going to say that it does support NOR-DAQmx 8.5? I installed the most recent NI.com last week, and it seems to work very well. I use only simulated devices so far, but DASYLab is data collection and display correctly.

  • cDAQ-9174 with modules 9225 and 9239 for issue of power measurements...

    Hello to one and all.

    I do turn a cDAQ-9174 with modules 9225 and 9239 to measure the voltage (120, 208-240, 277 VAC) and current (via the clamp amp i400, Fluke - 10mA/exit A) respectively.

    I have the set scale scale Differential No. (1:1 I guess) for 100: 1 for current and voltage.  With the acquisition mode set to "N samples", 10 samples to read, at the rate of 500 m.  This feeds into 'EPM_Power.vi' (polymorphic VI of the palette of EMP... can be downloaded here, http://sine.ni.com/nips/cds/view/p/lang/en/nid/209826)

    I try to check my numbers with an Analyzer of power quality of 435 Fluke and supporting that upward is a Fluke 179 True RMS DMM with clamp amp i200 (for current playback)

    Now, I'm running is my numbers I take data look like they're following a curve of fishing... the averages are on the right, but when my blood pressure is expected to be relatively stable, it's over the map... same story with my current readings.  I tried several different filters without success.  I would like to have relatively stable values for voltage and current so my other values are correct.  Is there a signal conditioner or filter that will help with this?

    As always, a point in the right direction would be greatly appreciated!

    Thank you!

    Chad

    Hi, Will,

    Your request for a screenshot actuall lead me to the solution.  Not enough samples!  I took little data too slowly and only seeing part of the wave of fishing AC each iteration.  .. that became glaring when I tried to this graph.  So I increased my sampling of 100 to 5 Hz and things seem to have settled things down a bit.  Thank you for the good question!

    SH?

    C

  • synchronize 2 cDAQ-9215 modules on the cDAQ-9174

    Hello

    We collect 8 channels of the AI of the comments about 32KS/s, but they must be synchronized with precision. We intend to use 2 CDAQ-9215 modules mounted on a CDAQ-9174 chassis.  This system will do the job? And what kind of calendar and clock configuration should I use to improve the precision of the synchronization.

    Thank you.

    Hello

    I'm sorry I didn't know not the 9215 was a simultaneous module.  Instead of the S series card, you should be good to go with a 9174 and 2 9215 modules. If you use the same DAQmx task to run on all 8 channels acquisition they share all the same timing engine and snack at the same time.

  • Synchronize the pci-6254 and cDAQ-9174

    Dear all,

    I hope you enjoy!

    I intend to synchronize two cDAQ-9174 chassis for an application where I need to acquire data of 8-Thermocouple, voltage 3, 1-Full bridge (strain gauge) and channel 2-meter at the same time. This enforcement action is based encoder (quadrature with 1024 ticks per round) and I need to acquire data very fast and synchronized in all channels!

    I bought the material and I'm going to set up 3-voltage channels in NOR-9221, Thermocouples-4-NOR-9211, 1 strain of NOR-9237 and 2-meter gauge in NOR-9401 first cDAQ-9174 chassis Modules. I will then put remaining Thermocouples-4 in the second cDAQ-9174 chassis.

    The question is whether it will work?

    If Yes! Then, kindly help me on how to have synchronized it data from these two cDAQ-9174 chassis with all such modules as mentioned above?

    I hope someone can help me with that and can guide me to do this task! I'm using Labview 2009 with NOR-DAQmx 2009

    Kind regards!

    Tajim!

    A few comments after looking at your code:

    I don't understand what you're doing with cancel in the task of counter.

    You try to use the encoder as your sample clock?  You can do it for the module 9221 and NI 9211, but it will not work with the NI 9237 as I mentioned above.  You can move the NI 9237 in a separate task and from there, you have a few options depending on how you want to synchronize.

    How many times you expect to get your clock on cDAQ2Mod3/PFI0?

  • Download a cDAQ-9174?

    Hi newbie here. Just bought a cDaq-9174 equipped with a card of entry and exit. I'm a little excited to try the new unit out, but I can't seem to figure out how to download anything for her. I just want to try to shoot an entry and see an exit lights. MAX found the controller software and maps. I have labview and I created a small VI which works, but after that I have run out of knowledge. Can't find something on the compilation or download. Any help would be greatly appreiciated. Yes I know the cart before the horse type stuff, but what can I say its new.

    kmay, CompactDAQ is as measure USB Board. It contains no embedded CPU, you would be able to program. Therefore, all the code you generate in LabVIEW will run on the local computer. To try some measure, use DAQ Assistant to paletter DAQmx LabVIEW or examples of example Finder (help-> find examples).

    But as I said, the code runs on the local PC and therefore not of compilation and download is necessary.

    Kind regards

    s9ali

  • cDAQ-9174 WinXP hotfix usb

    I'll put up a new machine with a cDAQ-9174 running a 8.6.1 exe.  Everything seems fine, but I went to add a scale and noticed a strange message in MAX.  He said looking for "windowsxpusbhotfix" and to search the knowledge base for "5F81A72I."  Something to do with the USB data.

    I tried to search in each of these and came with a post from 2004, or almost, but the knowledge base article is different.  The computer is running XP service pack 3.

    Hi Bryan,.

    Here is the link to the KB for reference.

    There is a bug in the Windows XP USB driver (which was actually discovered and reported by NEITHER).  Microsoft has a fix for what is available here, but NOR does not directly offer this patch for licensing reasons.  Unfortunately, Microsoft does not push this hotfix through Windows Update, so NEITHER decided to try to proactively inform customers of the possible problem.

    A note linking to the Microsoft KB added first to the DAQmx 9.0 readme, then the importance in the readme 9.1and more.  Because there is always concern (valid) that many people do not actually read the readme files, a notification has been added to MAX in DAQmx 9.3.  If your development computer is on an earlier version of DAQmx, or if you already have the patch installed, then you wouldn't see the warning.  Note that the warning itself by design disable all the features, it is simply intended to inform customers of the behavior.

    Regarding the consequences of not having the hotfix installed, the KB pretty much sums up it.  You might see crashes or incorrect data in the form of a swap chain.  The behavior itself is actually quite difficult to reproduce (I remember runs tests during the night several times, start and stop tasks with random intervals and still not see all systems), but nevertheless the integrity of the data is of course a huge concern OR if we want to make sure that customers are duly notified of the bug and stressed the appropriate solution.

    If you have any questions/doubts on made me know and I can contact you directly to discuss.

    Best regards

  • Reset hardware cDAQ 9174

    I test station with two cDAQ 9174. We work without problem. But the second after a time more long every day, using problem. The second is not recognized by the MAX of NOR. When I connect it to the USB port, nothing will happen. No new hardware found... No refreshing of the hardware in the Windows Device Manager list... Nothing. :-(cDAQ power led lights, but no led lights the LOAN.

    It seems to be that the cDAQ is broken. Is no way to reset the cDAQ 9174?

    Thank you very much for your answers.

    I'm sorry to hear that this might be the case.

    Here is information on how you can begin the RMA process:

    https://www.NI.com/services/warranty_repair_policies.htm

    See you soon!

Maybe you are looking for

  • Equium A60: DVD - RAM, being not recognized correctly

    Hiya! I have an EA60 and my carpet * a DVD-RAM UJ-820 s is currently not recognized by the laptop properly. When I go to device drivers, he said the driver has been loaded but returns the error code 41. I reloaded the Driver and the device, and it ma

  • How to reword scope of global variables on a per instance basis?

    I have a legacy application of this unforutnatly makes a lot of use of global variables (not the functional kind).  If we want to run two instances of the first level VI simultaneously without using global variables collided with another, which is th

  • How to navigate and send photos

    There are two pictures in my email that I need to send in a magazine contest. They do not accept emails - I type a form online and browse to send photos with the application. I do not know how to navigate - thanks for help-

  • Can I download videos from my computer to my DCR-SR200?

    I deleted the video on my DCR-SR200 after that he was transferred to my computer. I would have rather on the handycam, can I put it from my computer to the camcorder?

  • Can I make in the light of the screen stay longer?

    I have my laptop connected to the TV to watch a game on the internet. The screen light turns off every few minutes. Can I change this?