cDAQ 9191 and neither 9215

CDAQ 9191 or NI 9215 there a fpga module that we can program?

Is it possible to perform mathematical operations with the signals that we gain by using cdaq 9191 and NI 9215?

Once the 9191 is set up in MAX, you can read the channels on the 9215 and do what you want with the data.  The 9191 is used only to transmit data from the 9215 on a computer that is running LabVIEW, where you can perform mathematical operations.

If you want or need FPGA, you get a compact chassis of RIO, not a compactDAQ chassis.

Tags: NI Hardware

Similar Questions

  • How to speed up loop DAQ triggered using NOR cDAQ-9174 with NOR-9215 and NOR-9402

    Hello

    I use LV2010 and NOR-DAQmx 9.2.2. I have a NOR cDAQ-9174 with a NEITHER-9215 4 channel 100 k simultaneous ADC and NOR-9402 4 channel DIO module trigger and reset.

    We run WinXP sp3 on a Dell M4400 core 2 duo @2. 26 Ghz.

    I used the code example NI DAQmx for acquisition of tension with trigger HW. My goal is to try all 4 channels on the 9215 simultaneously when a trigger is received on channel 0 of the 9402, after data is read, I use channel 1 on the 9402 to reset the trigger of the target material.  I have a version of this work, however the maximum event rate is ~ 16/second. I have the Setup 9215 for finite samples / 10 samples per channel which is ~ 400uSec of conversion time and I realize he is above in the appeal of vi, but ~ 50mSeconds worth?

    The target detector can put out up to 1 k / event triggers / seconds.  Only, I received a rate of 8 per second and I added the NOR-DAQmx control vi driver and chose "commit" this did double the rate.

    My question is what is the maximum rate of loop for these devices (trigger/conversion/reading device / reset) and start over? I noticed that just let free the 9215, carried out using the 'Acq & chart internal strain Clk' raised only the rate of events up to 20 Hz.

    Thank you

    normbo663

    Hi normbo663,

    You can get this works far better assuming you have an available counter (there are 4 on the backplane of the 9174).

    DAQ Compact supports the tasks of meter output "redeclenchables" that can be used to generate a finite pulse train.  You can set a task of finished meter redeclenchables output to be used as sample for your task of analog clock.  The task of the meter output will be re-Army (less than 12, 5-25 ns) as soon as it's finished out the last pulse.  The task of analog input would be configured to run continuously, but it would only sample based on the output of the meter triggered.  For an example, see here.

    You can reference the internal counters on the cDAQ without signals through a routing module using: cDAQ1/_ctr0 (right click on the chain counter control, then select i/o name of filtering and check channels internal to add these options to the drop down).

    Thus, with the tips above, you should be able to immediately re - arming your analog acquisition on the 9215 using one background basket counters.  It seems that the second half of the application is to use a second channel on the 9402 to reset the trigger of your DUT.  You can deterministically generate this signal so by configuring a 2nd redeclenchables meter out task (single pulse, but this time).  All you need to do is the initial delay on the appropriate value for your analog acquisition.  Trigger this counter on the same PFI line that trigger you your analog task from.

    Using counters to generate the signals you need in a deterministic way, the loop becomes is no longer a problem (as long as your input buffer does not overflow).  You may need to re-read several triggers at the same time for the loop to keep (for example to read 1000 samples each, which would correspond to 100 triggers 10 samples).

    Best regards

  • Problem of discovery of the cDAQ-9191 in MAX

    I have a new cDAQ-9191 with 9208 module that I have connected to my PC with an Ethernet cable.  I use DAQmx 15.1.1 in Windows 7 Pro SP1 64 bit and MAX 15.0.0f0.  The cDAQ-9191 appears under devices and Interfaces > network devices.  There is no IP address and its status is "disconnected".  There is no found network card.  There is no Firmware button, and add device only returns an error message.  See the two photos attached for details of 9191 properties.  I have tried to reset MAX configuration, restart the 9191, exchanging several cables, restart the PC, turn off all the other networks of PC, turning the firewall, but without success.  I searched the forum, but no one else has reported a solution to a similar problem.  The port LAN PC has a static address, and I've used the port to connect to other devices.  As this is obviously not a normal state for this 9191 I'm totally stuck with no idea how to proceed.  Is it a hardware failure?

    In addition, on my main PC (MAX 15.0.0f0, DAQmx 15.1.1, 15.0 IMAQdx, win 10 Pro 64-bit), MAX cannot find network devices.  I tried with an Axis Network camera and the 9191, which both appear on the other PC.  I can't find a solution to either but it's a real pain that MAX is not working properly when necessary.

    Hi Dr_M,

    If your LAN port has a static IP address, it must run a DHCP server, which can affect the 9191 an address on the same subnet.  If no DHCP server is found, the 9191 a local IP address of the default link.  Very likely, you met the latter, and the static IP address that you have assigned to your LAN port does not link local beach.  This means that your computer and your cDAQ are on different subnets, so MAX can communicate with the 9191.

    You can change the IP address of your LAN port to one in the local area connection (169.254.x.x, subnet mask 255.255.0.0), or you can configure your LAN port to automatically obtain an IP address.

    Once you access the 9191 in MAX, you can adjust the parameters IP back at your convenience, as long as the chassis and the computer you end up on the same subnet.

  • Difference between the cDAQ-9174 and cDAQ-9178

    I have several cRIO-9081.  Each cRIO-9081 running NOR-DAQmx 15.5.  On four of the cRIO-9018 we show you cDAQ - 9178 with 8 modules.  Everything just worked.  On the fifth cRIO-9018, we tried to connect a cDAQ-9174.  The aircraft appeared in the Windows Device Manager and MAX, but MAX has indicated that the aircraft was missing (error-88705) and that Windows can't find a driver for the instrument.

    We then plugged a cDAQ-9178 in the fifth cRIO and everything just worked.

    What is the difference between a cDAQ-9174 and cDAQ-9178 other than the number of the module support?

    Paul

    Hi Paul,.

    It seems it could be a driver problem! What DAQmx version you have installed on two computers? Reinstall the latest version may be worth it because it looks like there may be corruption. You can find the latest version here: http://www.ni.com/download/ni-daqmx-15.5/5901/en/

    You can manually associate the device driver, here: http://digital.ni.com/public.nsf/allkb/5B80CAAA64413B4C86256C5E0000C1A7

    There are modules connected to the cDAQ chassis?

    Also, sometimes, this can happen if you do not restart your computer after the installation of the NOR-DAQmx driver. Restart your computer after the driver installation will allow you to reserve the device successfully. Here are a few others not to try:

    • Check the charger to device OR to see if it is running

    • Reset the service OR charger

    • Reset the MAX Configuration database

    • The system power cycle

  • Problems with NI-VISA 5.0.3 detect a 9474 cDAQ chassis and modules via USB

    Hello friends.

    I'm running Windows 7 x 64, I installed MAX 4.7.4, NI-VISA 5.0.3 and the other software that comes with them.

    I have a 9174 cDAQ chassis and 3 modules.

    My problem: when I connect the hardware to my PC via the USB port, it presents itself to the MAX, but NI-VISA does not detect sub "USB" (or elsewhere).

    Purpose: What I'm trying to do is communicate with the NI 9223 module, using VISA through MATLAB commands.  The MATLAB data acquisition Toolbox supports not the NI 9223, so I would try to acquire help from VISA.  However, I don't find the VISA of the device address.

    I would like to know your opinion on why modules are not displayed in NI-VISA and if my ultimate quest is valid without doing something nasty like writing C++.  I have not found a good source for VISA OR orders either, but I'm sure it exists.

    Thank you!

    Wes

    Hey eccentricorbit,

    Unfortunately, you can't talk about a module using VISA or cDAQ chassis. Your options are DAQmx and low level of coding (something nasty like C++). If you do not get DAQmx, you can find excellent examples by accessing:

    "" "Start menu" National Instruments "NOR-DAQ ' text Code support ' ANSI C examples

    I hope this helps!

    ~ kgarrett

  • Increase the frequency of sampling on the cdaq 9174 with neither 9212

    Hi all!

    I am trying to increase my sampling rate on my nor 9212, that I use with a cdaq 9174 and LV 2015. Apparently, the sampling frequency depends on the time of conversion and can be changed between the high speed, the Best 60 Hz rejection, best 50 Hz rejection, and high resolution. (see http://www.ni.com/pdf/manuals/374358a_02.pdf, page 6). My maximum sampling frequency is 1.8 Hz, which indicates the use of the mode of synchronization of high resolution

    I found this documentation explaining how to change with a crio:

    http://zone.NI.com/reference/en-XX/help/373197D-01/criodevicehelp/cRIO-9212_rt/

    However, I can't change it when the module is not in my project with my cdaq. My question is: How can I change the conversion without a crio time?

    Thank you very much!

    X0N0

    Hi X0N0,

    There is a DAQmx property for this:

    You are welcome!

  • My NOR cDAQ-9191 is over the display of the module installed in MAX.

    I just bought a 9191 cDAQ chassis with the NI 9219 AI module.  MAX initially showed the long-9191 9219 under network devices.  However, after several hours of messing around with the wireless settings and by pressing the reset button several times, MAX no longer displays the 9219.  Nothing dramatic happened to the hardware (that is, it was not dropped or mishandling), so I suspect a problem of self-inflicted free software.  The technician OR I have recommended 'Reset Configuration data' Max, I did, but that didn't fix it.  Has anyone else had this kind of problem?  Any suggestions?

    problem solved... several network services OR were stopped.  The module will appear once I restarted the service.

  • NEITHER 9215 delay in phase between two channels

    Hello
    I use Ni9215 with ENET-9163 to measure the phase between two sinusoidal incremental signals delay. First signal connected to AI0 and, secondly, to AI4 at 100 k sample rate. I know 9215 simultaneous ADC, but it seems to me that NI 9215 gives additional time between the channels. Is it possible, or l mistaken?

    Alexandr.

    Hi Alexander,.

    Please let us know how you get on,

    Chris is correct - with the idea of using a right signal to both channels, I tried to suggest this earlier...

    I hope this helps with your problems with delays,

  • NEITHER 9215, 9263 analog operations

    Hello

    I'm doing a simple entry that and power: I generate the signal using a Tektronix (30 Hz, 2 V p - p) signal generator and consider it as an analog input of 9215 using DAQ Assistant-1. Multiply by 2, and then send it as an output analog by 9263 using DAQ Assistant-2. Later, I see this singal on an oscilloscpoe. Here are the settings for the acquisition of data Assistants:

    Assistant DAQ-1:

    Mode: continuous

    Price: 1 K

    N ° of samples: 1000

    Clock type: internal

    Assistant DAQ-2:

    Mode: continuous

    Schedule use of waveform data

    Clock type: internal

    When I run the VI, it stops after about 25 seconds, which gives the error:

    Error 200279 occurred at DAQmx Read (analog 1-d Wfm NChan NSamp) .vi:1
    Attempted to read samples that are no longer available. The requested sample was already available, but has since been replaced.
    Increase in the size of buffer, most frequently the reading of data or by specifying a fixed number of samples to read instead of reading all available samples would correct the problem.
    Property: RelativeTo
    Corresponding value: current playback Position
    Property: Offset
    Corresponding value:
    Task name: _unnamedTask<6>

    Why are they bad?

    By test error, I found that by the acquisition of data Wizard-1 mode to "N samples", I was no longer getting this error. However, then a big delay time has been introduced. that is if I change the frequency of the signal (say 15 Hz to 30 Hz), then this change is reflected on the oscilloscope 10 seconds later. Why is this happening? And how can I solve it?

    I enclose the dig that for reference.

    Thank you

    Mandar

    Hi Mandar_K,

    Thanks for the reply and I hope that your well today!

    It is an example that I did with a synchronized AI read, some treatment, then AO. This seems to be what you need.

    If please take a look (LV8.2) and let me know what you think,

    Kind regards

  • My Favorites are gone and I can't make new bookmarks. I tried to restore their by clicking on the backups and downloading Backup and neither will not work.

    This is as far as I know. I followed the instructions to restore and nothing works.

    See also:

  • TAB GROUPS die after closing the window, restoration of session does not apply and neither is history! I do not think that I have been warned!

    On leaving firefox, update, then restart, I lost the ability to restore sessions. I restored the Group of tabs in the past; It's different. My tab groups are lost forever and I don't remember even be warned about losing when the window is closed (as often happens and then I avoid close it.)

    I sync my tab groups, but she must have sync'd their disappearance because I lost. Tab groups really should have always been separate from the window (as a form of records) - I don't know why I found myself never adopt this powerful feature knowing how bad it has been supported. I had actually named groupings with 2dont reading lists etc... the year value of groups... and I shared it between devices...

    In fact, I watched the sessionstore.js file, it seemed ok - I do javascript professionally. I think I can close this because there seems to be a lack of functionality/design of browser that we can solve a problem. my other computer did the same thing with clamping my tab groups and the closure of the window... it's a "feature" not a bug. I thought that maybe someone knew something that I do not have as to why this is so wrong.

    I'll just have to live with attention restore or restore backups the sessionstore.js if for some reason, he gets lost. Including me accidentally close the window and press on enter without thinking (Yes, I get that I can restore; but the problem is that I wouldn't restore groups of tabs because they are so easily killed in multiple ways.)

  • My 8 Firefox Menu button does not work and neither does the right click in the window of firefox

    It seems that all drop-down buttons in the browser, including the bookmarks dropdown in the toolbar and the main menu button (firefox), no longer work. This happened after that I upgraded to firefox 8. I tried uninstalling and reinstalling several times all also the removal of the preferences user and all that but that does not work. But everything seems to work properly in safe mode. So I disabled all my addons in the normal mode, but cela still does not resolve the problem. My right click does not anywhere in the browser window either. And also, although the menus are not displayed, it seems that they are still there because if I click in the area that a the function button active, that's how I was able to disable addons.

    Safe mode disables extensions in Firefox 4 and later versions, and disables hardware acceleration.

    • Tools > Options > advanced > General > Browsing: "use hardware acceleration when available.

    If disable hardware acceleration works then check if there is an update available for your graphics display driver.

  • Firefox opens, and neither does Profile Manager

    So yesterday some site to search Google (Kilauea photos search string) done something wacky, and now Firefox 4.0.1 will not start. Safe mode does not start. Profile Manager does not either. Uninstalling Firefox, by removing the Firefox of Program files, running a registry cleaner to clean the remnants of Firefox, running a full MS Security Essentials scan who find Trojans villain, and then reinstalling Firefox works only for the initial session. A replay of Security Essentials and a race with IObit Malware Fighter is nothing. Once I got close and restart, same as at the beginning. Internet Explorer 8 and Firefox 3.7a5pre minefield version seem to work normally, other mines fails to remember the passwords even if it is checked. Any ideas on what to try?

    Beta version 5.0 is not really get updates of security by saying as with point versions minor for 3.6. * for example. The 5.0 Beta recent 5 (which will be soon in Beta 6, may even be a b7) could be even more up-to-date then 3.6.17 and 4.0.1 even in a sense. Unlike with the latest versions with release candidate RC for the exit, the last last beta deemed good will become final.

  • The output on my Macbook Pro is not giving me the drop-down list, &amp; happens either in the port of the helmet or digital port and neither one is what I need. I tried turning off the computer and hold "Shift"' Control', 'Option' with the power-o ".

    The output on the sound output box has no drop downs. The audio output suddenly disappeared and when I went to the 'exit' sound in 'sound preferences' the only thing that would show was the headphones.  After that I tried a couple of things suggest that 'Helping Apple', sometimes in digital form & everything that happens you can't change for something else.  Things I've tried suggested by Apple press help were the "Shift control option buttons & button / stop at the same time."  or press the Option, the command, the P & R key as soon as possible after you have pressed the button On.  Hold down until after the second gong during a manoeuvre.   This did cause the digital option, but none of these 2 is the one I need.  Help

    Perhaps, it would be useful to know what that is exactly the one you need. Are get you an external speaker output or anything like that. Also, it would be useful to know the operating system you are using.

  • 8640A Kistler and NEITHER 9234

    I have probably a very simple question.

    Is it possible to perform the Kistler PiezoBeam accelerometer 8640A 5 http://www.kistler.com/mediaaccess/8640A_BP__000-842d-03.11.pdf

    to the NI 9234.

    If not, what materials would you use to operate up to 16 8640A 5?

    Thank you!

    You can and would also be able to use TEDS

    http://sine.NI.com/NIPs/CDs/view/p/lang/en/NID/208802

Maybe you are looking for

  • Why not download and read .pdf files

    I never had no problem download and playback of .pdf files to what I 'upgraded' Adobe reader version 11. Now all the downloads of attempts fail because "I don't have the necessary helper or same C:\DOCUME~1\Will-o\LOCALS~1\Temp\Back channel II - betr

  • Toshibia e-studio200s start-up of the printer/copier problems

    Whenever I start the Toshiba E-studio 200 s printer/photocopier, he has started making a rattling noise that continues for 7 minutes. After this, then warms, it keeps turning off for a split second only to 'warm up' again, and he continues to be on.

  • Re: Satellite L300-1BV: I want to use Draft N - how to disable Wlan in BIOS

    I think trying a new network adapter, which has the N Protocol. But I was told its best to disable the network adapter in the BIOS first realtek wireless and not Device Manager as I thought. So basically what the correct way to do it in the BIOS inst

  • DVD/rw rather dvd/cd

    Hello! I own M30 354 for laptop and I want to change my dvd/cd combo with dvd/rw burner burner! is there a compatible device? TNX

  • Tecra 9000 - sound problems

    Good then I obviously have my tecra 9000 series PC laptop and im trying to get the sound to work. Nothing don't is going on so I thought he might have a sound deseased... So I replaced it with a brand new sound Board in which cost $107 *. Anyway, so