GPIB Garland address help

I'm working on the implementation of new equipment, a second facility that we have.  In both configurations, we use a single GPIB-ENET/100 and a GPIB-232CV-A with LabVIEW.  On the original configuration, which works very well, here's the order of connection:

Computer--> GPIB-ENET/100, connected via an ethernet cable

Power supply high voltage--> Lock-In Amplifier--> GPIB-ENET / 100--> GPIB-232CV-A, Garland with GPIB cables

The GPIB-232CV-A has a dip switch address 1 and also runs a cable DB-9 to a series of commands of stepper motors.  In my global addresses VI, these stepper motor controls receive a global address of 1, what I understand.  What I don't understand, is why the high voltage has an address of 3 and the Lock-In has an address of 4, which my other VI scream in order to communicate with them.  That this has something to do with the order of Garland?  Yet once, there is only a single use GPIB-232CV-A, and it has an address of 1.  Thank you.

Too bad, my devices have internal GPIB address

Tags: NI Hardware

Similar Questions

  • I have firefox 9.0 - when I opened a 'new tab' there is no task to enter the web address - help

    I have firefox 9.0 - when I click on 'new tab' get me a... but no place to enter the web address. I tried to delete 9.0 and installed the previous version. The same problem. I can't get it to open the tabs correctly in safe mode, but that disables my flash drive. Can anyone help?

    You did it ensure that you have the visible Navigation toolbar?

    If the menu bar is hidden, then press the F10 key or hold down the Alt key to make the call of menu bar.

    Make sure the toolbars like the toolbar 'Navigation' and the 'personal bar' are visible: "view > toolbars".

    See also:

  • All stuck in 169.254.106.192 network adapters IP address help!

    Hello

    I tried everything to fix this.

    First of all it is just the wireless card for my laptop that got the question, so I went to an Ethernet directly between a router and that possessed the same auto config ip. Then, I used my tp link TL-WN722N usb wireless adapter and it worked fine until today. Wi - Fi points are not stored in manage wireless networks since I can't access Internet and sometimes I can't even connect to them. Those can connect to stay in the "unidentified network". First I tried to manually release and renew the ip address in the cmd with no luck, it is said that it is impossible given that the media is disconnected or a timeout error when the survivor crashes just there for a few minutes. I am 100% positive, that this is a problem with the computer and not the access points as I traveled the past few days and it worked very well the first days of my trip. First, when this happened, I let windows troubleshoot it and automatically, he knew that there was an invalid ip configuration and fixed it himself. Now he recognizes that it is the same problem, but "has not set" next to him and don't let not the dhcp on the network assign the ip address. I tried manually as well as on all NICs IPv4 static addresses and nothing works. I think that it is the cause of a new update of windows... I am running windows 7 Enterprise on a dell latitude e6400.

    I just need to reset the IP address and it keeps auto configuration EVEN when it is connected to a network. It drives me crazy. Help, please

    IP address with 169.254.xxx.xxx means that the DHCP (Dynamic Host Configuration Protocol) delivery is affected. The DHCP server to allocate IP address to computers, to the local IP in your case.

    There are three chances for the problem to occur:

    1 wired AutoConfig (for wired connectivity) or autoconfiguration WLAN (for wireless connectivity) service is not running.

    2. bad network resources on the network configuration.

    3 cable bad network.

    Solutions:

    1. the following link will be very useful to start the Wired AutoConfig service (find the WLAN auto configuration service in the same list and repeat the process involved in from depending on the link):

    http://computerstepbystep.com/wired_autoconfig_service.html

    2. the following link will be useful by resetting the TCP/IP settings. Make sure that the options are set to automatic:

    http://Windows.Microsoft.com/en-us/Windows/change-TCP-IP-settings#1TC=Windows-7

    3. check with another ethernet/network cable. In addition, check the power cable in some other computer for the work of the cable. Try to replace the cable, then check.

  • VXI by GPIB on LV 2009 + Windows 7 help

    Hello.  I have no experience with VXI.

    I have a client who got hired to encode a VXI product that has been marketed as controlled by GPIB (the chassis contains a GPIB interface).

    I was informed late last night I have to use Plug and Play VXI drivers to make it work.  (Long story and I am not happy about).

    I'm coding LV2009 SP1 on Windows 7.

    I installed NI-VISA and NOR-VXI.  I can see the GPIB instrument in MAX.  I can even run the SCPI via Max and LabVIEW controls.  (The problem is that they have apparently not applied specific GPIB commands that they published).

    They provided a VXI - PNP driver and very vague instructions.  I installed this driver and I have imported their .fp file in LabVIEW.  When I run the init code with the right GPIB resource passed into the function, I get the same error when I call a non-existent GPIB device - even if it takes more time to get back.

    I searched for VXI tutorials on the site of NOR and there are a LOT of things to go, and I'm not even sure that I have all the necessary drivers.

    Anyone know a good place to start to read up?

    (And Yes, I'm dogging people instrument to help me too)

    Thank you

    Hi Jed,

    From our conversation with your service request, I wanted the documents a link to this post to help with future issues.

    Why don't My GPIB-VXI system appears correctly in MAX?

    Getting started with the GPIB-VXI/C

    OR GPIB-VXI/C

    Configuration of T & M Explorer use a GPIB-VXI address different GPIB?

    * NOTE: The T & M Explorer is the predecessor of MAX

  • address fails to TNT4882 gpib

    I can not get the TNT4882 chip addressed correctly. I use onechip mode.

    I run according to the configurations (constant of the ESP-488 header files).

    TNT_Out (R_cmdr, F_softreset);            / * Reset memory FIFO * /.
    TNT_Out(R_keyrg,0);
    TNT_Out (R_hssel, F_onechip);             / * Set TNT on a single chip mode WINK * /.
    TNT_Out(R_misc,0);                      / * Disable HS mode * /.
    TNT_Out (R_hier, B_dga |) B_dgb);            / * A deglitching circuits slower value * /.
    TNT_Out (R_auxmr, F_chrst);               / * Reset TNT * /.
    TNT_Out (R_auxmr, F_pon);                 / * Clear power on * /.
    TNT_Out (R_admr, F_normalprimary);        / * Set address mode * /.
    TNT_Out(R_adr,13);                    / * Set primary address * /.
    TNT_Out (R_adr, B_ars |) B_dt | B_dl);         / * capabilities * /.

    Just after this configuration

    TNT_In (R_adr0) and TNT_In (R_adr1) return

    R_adr1 = 0x01 and R_adr1 = 0x00

    According to the documentation R_adr1 should never return 0x00, the value must be 0 x 80 or higher.

    After the chip TNT4882 gets addressed as a listener (to the wrong address 0).

    TNT_In (R_adr0) and TNT_In (R_adr1) back:

    R_adr0 = 0x0d and R_adr1 = 0x01 or

    R_adr0 = 0x0d and R_adr1 = 0 x 21 or

    R_adr0 = 0x0d and R_adr1 = 0xa1

    Now R_adr0 seems to be OK, but R_adr1 once again is not going and again, TNT4882 gets directed to the gpib wrong address 0.

    I used the very conservative calendar. I used two boards with similar results. However, I get the * IDN? Quite correctly from MAX scan device.

    I also used this "setup anymode" in the beginning of the installation without modification for response.

    Someone at - it suggestions or find something wrong in the Setup?

    -Matti

    You Steven for your valuable advice. I found the problem connected to the PIN PAGED. This question is left connected to the Earth of the initial approach using the compatibility mode 9914. PAGED mode to use the registry I materialized does not. Now things work very well.

    Thank you

    Matti

  • ADDRESS BOOK NOT BE DISPLAYED WHEN YOU TRY TO SEND E-MAIL,

    address book is not displayed when you send an email. Normally, when I send an email, I click on the front, and the address book appears on the left side of my screen.
    also, when I bring my address book, it takes half of the screen, when I click on the arrows to shrink the screen, it takes away the names and the addresses.help please!

    When you open a window of writing and click VIEW in the menu bar is framed verified contact?

  • benefits of VISA by GPIB

    I want to know the benefits of the VISA by GPIB in labview when we connect our pc to a device that has interface GPIB

    What could be the reason behind the use of the VISA program

    I want at least 5 advantages of visa by GPIB.

    Hello

    1 VISA Write equals a GPIB entry unless the entry VISA returns a VISA Out reference which is much easier to use. You who go to the next function VISA instead of a GPIB cabling address at all. This will make your cleaner blockdiagram.

    2.portability:-VISA supports GPIB, series, Ethernet, USB, PXI, VXI, and PCI. So, with a single piece of code, you can support all the interfaces and the same GPIB controllers from different vendors.

    3 VISA supports aliases. You can assign logical names to resources. Instead of trying to remember the GPIB address in a rack full of instruments, you can reference "DMM", "Scope", etc.

    4. If you program multiple devices communicating on more than one type of bus, it might be easier to use VISA for your entire system.

    5. in the VISA, you must always call viEnableEvent() before being allowed to receive events, it is necessary for the VISA to avoid the race condition to try to wait for events for which the equipment may not be activated.

    6. the control of VISA resources allows you to choose from a list of available resources.

    Enough

  • Verification of GPIB devices

    Hello

    I have a code that sends a "* IDN? GPIB addresses 1 to 31, if I get a response to the motion, I guess there is a GPIB device address. This approach gets watered until when I have a GPIB instrument does does support/recognize the * IDN? query.   However, MAX is able to recognize that there is a GPIB device to the address of the instrument I have. Ideas of how recognize/detect if a GPIB device exist if it does not meet a * IDN?

    Concerning

    Kaspar

    Use the function of the same name on the Instrument i/o > GPIB > palette 488.2.

  • Establishing connection between PC through GPIB-USB_HS and AQ6315-E

    Hello

    I couldn't create connect with spectrum AQ6315-E max. For "scan for instrument" request Max returns "Instrument not found."

    GPIB-USB-HS about what - it's 100% working properly. Some problems with GPIB analyzer - port could be? In addition, the Analyzer have 2 GPIB port. But it is indifferent for MAX - port which is connected (it shows that "instrument not found" in both directions). What should be the reason for this?

    I thank the of for attention. I found my miss. The GPIB-USB-HS and analyze the same GPIB card - address. When I specified one for parser - MAX reveal properly connected Instr.

    Thank you.

  • Address book does not load in OutLook Express

    Operation of a limited account, Outlook Express cannot open the address book with
    the following messages:
    (1) the address book failed to load. Outlook Express s wrong
    configured, please re-install.
    (2) unable to open the address book. The address book cannot be
    installed correctly.

    However, if the limited account is changed to an administrator account
    are no problem.

    There are two accounts on the PC; HP_Administrator and User1. User1 is the
    the problem account. HP_administrator has administrator privileges
    and if User1 receives administrative privileges, and then I noticed that the two accounts
    use the same address book. In other words, Outlook Express
    "-> Addresses--> help on address book" in the two accounts show

    C:\Documents and Data Data\Microsoft\Address Book\Administrator.wab

    used as the address book.

    This seems incorrect. I would expect Admin to be replaced by HP_Administrator
    or User1 in the path above, according to the account that Outlook Express has been
    Run from. It seems that the operating system is configured to use the address book above,
    regardless of the account running Outlook Express. If User1 is a limited account
    It is not possible to view the location of the address book due to the error that
    is happening, but I suspect that Outlook Express attempts to access the address book
    above and a problem arises due to try limited permissions User1 account
    to access the address book in the folder of the administrator.

    Windows XP with service pack 2 is the operating system and of its Outlook Express 6.

    What do I do so that Outlook Express can operate successfully in the
    limited account User1.

    The problem has been resolved. I wrote the solution for people
    who, like me, have no experience with the Windows operating system.

    The sites of these links must be read. For Outlook express
    http://www.insideoe.com,

    more precisely

    http://www.insideoe.com/files/WAB.htm.

    For more information about the registry and how to change the following two sites
    provide the necessary information and should be read carefully.

    (1) http://www.techsupportalert.com/content/learn-how-use-windows-registry-editor-regedit-one-easy-lesson.htm

    (2) http://pcsupport.about.com/od/termsr/p/registrywindows.htm
    (You will find other useful links on this site)

    A single note; There are different sets of keys "HKEY_CURRENT_USER", 1 set for each
    account. To change a "HKEY_CuRRENT_USER" key for the user account, you must be
    logged in as user.

    In the registry, the key to "HKEY_CURRENT_USER\Software\Microsoft\WAB\WAB4\Wab file name.
    contains the path of the address book used by Outlook Express, for
    the account that you are logged.

    When logged on as a user and by examining the registry with regedit

    "HKEY_CURRENT_USER\Software\Microsoft\WAB/WAB4\Wab file name.

    the value

    C:\Documents and Data Data\Microsoft\Address Book\Administrator.wab.

    Given that the user is a limited account, user doesn't have the privilege to access the address book in this
    location, which led to the errors listed in the original post. The solution is to
    change this path. Using regedit path has changed to

    C:\Documents and Settings\User\Application Data\Microsoft\Address Book\User.wab.

    C:\Documents and Settings\nom being the folder root for the user account, the user has
    the privilege to access the address book to the new location. Outlook express has been
    then be able to access the address book.

    Upon reflection, it seems to me that the above solution is just a symptom, and not the cause.
    The original question, in fact, should be:

    How the system is configured so that when a new account is created the new account
    will use an address book to

    C:\Documents and Account\Application Book\New Account.wab Data\Microsoft\Address CAB.

    Instead of

    C:\Documents and Data Data\Microsoft\Address Book\Administrator.wab,.

    It should not be necessary to change the registry for each new account.

  • Hotmail won't let me before an email to my list of fried of 36 addresses.

    HOTMAIL WON'T LET ME BEFORE AN EMAIL TO MY LIST OF FRIENDS OF 36 ADDRESSES HELP PLEASE

    I RECEIVED AN E-MAIL WITH A YOU TUBE LINK IN IT AND I WANTED TO SEND IT TO MY FRIENDS IN MY FRIENDS LIST THAT IS LONG OF 36 ADDRESSES AND A SIGN CAME SOMETHING. SPAM AND TOO MANY ADDRESS IN MY FRIENDS LIST.

    I CAN'T FIND ANY HELP IN WHAT MY LIMITS ARE AND WHY.

    THANK YOU FOR YOUR HELP

    PATRICK BENNETT

    E-mail address is removed from the privacy *.

    Hi Patrick,

    Since you are facing problems with your hotmail account, you must validate your application in the Windows Live Forum to get help.

    In the Windows Live Forum

  • How do you reset security issues where there is a link at the bottom of the page?

    When I was younger, I did my security questions and do not remember them. I tried to reset, but it isn't a link at the bottom of the page click on reset. Any help is welcome!

    Hi Kayla_8

    This link appears only if you have a rescue address put in place on the e-mail account.

    If you don't have an e-mail address (different from the secondary e-mail address) help set up on the account, you will need to contact iTunes Support Online: http://apple.com/emea/support/itunes/contact.html (select security account as your help topic) or by phone: Numbers of Support from Apple

  • Agilent 34401 react is not the * IDN? query

    I connected an Agilent 34401 via a GPIB cable. I saw the instrument in MAX. Her having a GPIB 9 address. But the device does not respond to the * IDN? query. What could be the reason?

    Implementation of command language? Page 165 of the instruments user manual. If you don't have it on paper, you can download it at http://cp.literature.agilent.com/litweb/pdf/34401-90004.pdf

  • latest version of the driver for keithley 2400 sourcemeter instument

    Hi all

    I'm unable to connect with sourcemeter Keithley 2400 with GPIB / Rs232. (currently using the driver from keithley Web site)

    I did a search preliminary Forum and learned that there is some bugs in drivers.

    Then comes the last avilable pilot site OR the corrected version?

    http://sine.NI.com/apps/UTF8/niid_web_display.download_page?p_id_guid=25B255F3AA83660EE0440003BA7CCD...

    The one above IS suitable for the two abd GPIB, RS232.

    concerning

    Lorris Somasundaram

    I think there has been a problem in the GPIB bus address. I used MAX and tried triubleshooting and reattach the GPIB and finally the connected got intrument, sorry for the inconvenience.

    The driver loaded since NEITHER website works very well.

    Thank you and sorry for the inconvenience

  • Why USB-6501 does not install on Win7?

    We are in a process of migration of the old XP machine test system to the new system of Win7. All but got properly installed USB-6501. After you install LabView / DAQmx and connection USB-6501 windows could not find the drivers for this device (poster the exclamation sign on the icon of the device in devices and printers). MAX 4.7.4 detects either of the device. I looked for similar scenarios on the Forum OR and abroad and tried the solutions proposed. Nothing worked for me. I know the issue isn't with the USB-6501 material because it is functional on the old system. 4.7.4 MAXLE Win7 system detects a very well (GPIB-USB-HS) different USB device, but not for USB-6501.

    I hope someone can help us identify the problem that we have struggled with for some time. Here are the differences between shipments of software on two systems:

    Old system:

    OS - XP SP3

    LabView - 7.0

    DAQmx - 8.0

    VISA - 4.4.1

    MAX - 4.5 (detects the GPIB-USB-HS and USB-6501.) Status on 6501 LIGHT flashes continuously)

    New system:

    OS - Win 7 Enterprise

    LabView - 7.0

    DAQmx - 9.2.2

    VISA - 5.0.3

    MAX - 4.7.4 (detects only GPIB-USB-HS. Status on 6501 LIGHT flashes 3 times and then road)

    Germano,

    The device has been appear in Manager devices, but with the yellow exclamation mark for no drivers found/expenses. Update drivers on the property page does not help.

    So yesterday I spent most of the day removing all facilities OR according to the instructions, I found in this forum. Then I reinstalled the software OR in the order following, restarting after each installed:

    (1) LabView 7.0

    (2) DAQ 8.0.1

    (3) NEITHER-488. 2 (v2.30)

    (4) connected USB-6501 first. The installation of the device has gone well and is now detected to the MAX!

    (5) connected GPIB-USB-HS. Scanned for changes on the hardware, peripheral GPIB presents itself to the MAX, but with a weird name announcement its INF file on his behalf. See the screenshots below:

    GPIB device incorrectly installed:

    GPIB device properly installed:

    The strange thing is that this GPIB device was fully functional on my previous installation (see attachment max_report_3-23 - 11.pdf), now reinstall after its unusable (see attached max_report_3-24 - 11.pdf). Looking for GPIB instruments to help Max returns no result, but if I reconnect GPIB-USB-HS to the old system of PC all instruments are detected (for instrument scan error message see attached MAX after reinstall.jpg).

    That my new problem is not related to USB-6501, maybe I should open a new forum thread.

Maybe you are looking for