How constantly sample meter signal for quadrature linear encoder

Hi all

Trying to permanently change value of Schneeberger linear encoder quadrature at certain frequency as 10 Hz of the sample and the output to the text file. Everything I choose in the external clock that I can't run the DAQ assistant. No one knows how to fix? I don't know what kind of clock can be used.

I have read a few replies that ppl would connect to an another sample clock in parallel with the meter channel. But I don't know what is the principle of operation of this method.

I spent already as 10hours on it. Could someone tell me that some information?

Looking forwad to your help! Thank you very much!

Best regards

George

Hi George,.

You'll want to use the DAQmx API rather than the DAQ Assistant.

The task of the encoder itself would look like this.  You need not use the trigger options.

The problem is that you will need to generate a clock from another source that the card is unable to generate examples of clock counter directly.  The simplest is probably to use the second counter.  Here is an example showing how to use a meter to output as a sample clock (although the example shows the analog inputs, simply replace the task of analog input with the task of the encoder shown in the previous example, I linked).

So, you would use a counter (say ctr0) for the task of the encoder, and then the other (ctr1) meter for the generation of sample clock.  You must specify your task of encoder using the InternalOutput of ctr1 as the sample clock (if the internal counter output appear in the drop-down menu, click with the right button on the terminal and select the e/s name filtering > include Advanced terminals).  Start the encoder task before the output task (although given that both are running continuous it doesn't really matter that much).

In addition, 10 Hz is not too fast.  If the exact date is not as important and the above seems a little overwhelming, it might be easier for you to just run a loop of 100ms clocked by the software from example like this.  Note that the exact time loop pourrait will vary from cycle to cycle, so this method is not very good if you use sampled data to perform calculations of rates.

Best regards

Tags: NI Software

Similar Questions

  • Re-sample a signal for the least points of data using the value average

    Hello

    I'm trying to resample a large wave to a smaller wave form. I want to use the average value of all data points in a time interval of the initial wave as the new value of the new waveform. For example I have want to resample a waveform of 100 points from k to a waveform with 1 k points, in the waveform again the value of each data point is the average of 100 data points in the old waveform. Is a convenient way to do this in labview? The VI of waveforms resample seems only provide coerce/linear/spline/FIR interpolation modes. Thank you very much for your help.

    Just use build a waveform is your decimated table, dt is climbed by the decimation factor and t0 can remain the same.

  • Best practices for the implementation of META tags for content items?

    Hello
    The portal site, I am responsible for the management of our content (www.sers.state.pa.us) runs on the following WebCenter products:

    WebCenter Interaction 10.3.0.1
    WebCenter Publisher 6.5
    WebCenter Studio 2.2 MP1
    10gR 3 content services

    The I work for the Agency is one of many for the commonwealth of PA, who use this product suite, and I meet some confusion on how to implement META tags for content of our site, so we can have effective search results. According to the [explanation of META tag standards W3C website | http://www.w3schools.com/tags/tag_meta.asp], description tags, keywords, etc., should be between the cephalic region of the HTML document. However, with the configuration of the suite WebCenter is implemented, how the head section of the HTML is closed at the end of the model code to a common header portlet. I was advised to add fields to our entry presentation models and data for the content to add these meta fields, however, since they are then placed in the body section of the HTML accordingly, these tags fail to have a positive impact on search results. Instead, a lot of our content points, when searched, the description in the search results displays only the text that appears in the header and the left navigation of our model, which arrived earlier in the body section of the HTML.

    Please advise possible or methods that would be best to implement the use of META tags so that we can our pages containing content to come to the top in the search results with relevant data.

    Thanks in advance,
    Brian

    Basically, you want to add tags portals in the presentation model to move the meta tags of the element of the header of the actual document content. Check out http://download.oracle.com/docs/cd/E13158_01/alui/wci/docs103/devguide/apidocs/tagdocs/common/includeinhead.html

    email me if you have problems with thezmobiegroup.com andrewm

  • How to speed up my signal for Auntie plu photon strength?

    Original title: problem with Auntie photon more net connection

    How to speed up my signal for Auntie plu photon strength? My net is too slow & very dull.

    Hi shilpadarbar,

    Thank you for writing to Microsoft Communities. From your problem description, I understand that the internet connection is slow. Before you start the troubleshooting steps, I need the required information:

    ·         You get the error message?

    ·         Did you do changes on the computer before the show?

    Follow these methods to the problem of slow internet connectivity.

    Method 1: Follow these steps:

    Step 1: Start the computer in safe mode with network and check if the problem persists.

    Step 2: If the problem does not persist in safe mode with networking, perform a clean boot to see if there is a software conflict as the clean boot helps eliminate software conflicts.

    Note: After completing the steps in the clean boot troubleshooting, follow the section How to configure Windows to use a Normal startup state from the link to start the computer to a Normal startupmode.

    After the clean boot used to resolve the problem, you can follow these steps to configure Windows XP to start normally.

    (a) click Start, run.

    (b) type msconfigand click OK.

    (c) the System Configuration Utility dialog box appears.

    (d) click the general tab, click Normal startup - load all services and device drivers and then click OK.

    (e) when you are prompted, click restart to restart the computer.

    Method 2: Temporarily disable the security software.

    Note: Antivirus software can help protect your computer against viruses and other security threats. In most cases, you should not disable your antivirus software. If you do not disable temporarily to install other software, you must reactivate as soon as you are finished. If you are connected to the Internet or a network during the time that your antivirus software is disabled, your computer is vulnerable to attacks.

    Method 3: Follow the steps in the article.

    How to troubleshoot network connectivity problems in Internet Explorer

    You can also see this article for more information.

    How to troubleshoot possible causes of Internet connection problems in Windows XP

    For internet speed and signal strength, you must contact the Auntie Photon support for more help and information.

    http://www.tataphoton.com/customercare.aspx

    If please follow the methods and let us know if you need more assistance. We will be happy to help you.

    Thank you.

  • Count the edges of the 2 signals TTL (Heidenhain linear scale)

    Hi all

    This is my first post here. :-)

    Currently, I'm doing a VI to be used with a linear scale. The linear scale gives 2 TTL signals that have an offset of 90 °. The change in distance of the linear scale is given by counting the fronts and edges of the two signals. See the following image: Ua1 is signal 1 and signal 2 Ua2. You can ignore the other signals.

    Now, I want to count the 4 edges in order to translate the 2 signals in the distance. This means that I need advanced two counters for Rising-rising, Rising-Falling Falling Falling, falling on the rise for 2 signals. I tried to do 4 points two counters in LabVIEW but that of course does not work, because an acquisition of data can access the card TTL or I did it wrong.

    Once I have to handle this, I also want to understand the meaning.

    My card TTL: NI 9402

    My electronic Heidenhain interface: 100 IBV (http://www.heidenhain.com/fileadmin/pdb/media/img/598_160-23.pdf - also at the origin of the image)

    Hardware configuration: linear scale-> IBV (Elektronic Interface)-> NI 9402-> LabVIEW

    Signals: Analog 3-> 3 TTL-> OR 9402

    I hope I do not double post. Any help would be greatly appreciated.

    I used Heidenhain linear scales in many applications.

    As stated in the previous post, the output of your balance is as a quadrature encoder. Therefore, you must use an entry of the DAQ card counter to measure the position of the scale.

    The desired X 4 mode is done by the meter itself (not possible with some old maps of OR).

    As starting point, see measure angular Position.vi that comes with examples of LabVIEW. On your linear scale, change the type of the polymorphic DAQmx create channel VI CI linear encoder and etiquette of pulses per revolution at a Distance by pulse.

    Feel free to post back if you need further assistance.

  • How to demodulate a signal?

    I need to output a low frequency signal (between 2 and 30 Hz) between the output of a sound card at the entrance of another sound card, between two PC. Sound cards is not designed for frequencies below 30 Hz. That's why I need to use the Modulation of frequency or AM, (or PCM) on the outlet side to create a signal, for example, between 102 and 130Hz. How can I accurately demodulate the signal so I can treat it as the original 2 Hz to 30 Hz signal in my journal? I think since DASYLab support the AM, FM and PCM that it would also support the process reverses. How would I do that?

    Although I wasn't able to demodulate the signal, using the FFT, I was able to isolate quickly the frequency since my entry was a pure sinusoidal signal.

    Thanks to CJ to work up to an earlier example that brings me down the path to implement the solution!

  • How to sample multiple sensors in a vital task?

    I need to a thermocouple sensor and voltage sensor in a vital task of the sample. My hardware configuration includes a cDAQ-9172 chassis with a NI9211 TC module and a NI9221 voltage input module.

    I started with the example C code for thermocouple measurement found here:

    Instruments\NI-national DAQ\Examples\DAQmx C\Analog In\Measure Temperature\Cont Thrmcpl samples-Int Clk\ContThrmcplSamps ANSI - IntClk.c

    I then added a few lines of code for the measurement of additional tension.

    The task runs for about 20 seconds, until the buffer overflow with the error 200279.

    I can run task, thermocouple or voltage, indefinitely. It's only when I include them both in the same task as the buffer overflow.

    I have tried many combinations of sampling rate, the size of the buffer and DAQmxRegisterEveryNSamplesEvent nSample size.

    Below is the code I'm running.  Changes to the original file are marked in bold. I've attached a screenshot of the output of command line.

    Can anyone offer a C program example showing how to sample multpile sensors permanently in a task?

    #include 
    #include 
    
    #define DAQmxErrChk(functionCall) if( DAQmxFailed(error=(functionCall)) ) goto Error; else
    
    int32 CVICALLBACK EveryNCallback(TaskHandle taskHandle, int32 everyNsamplesEventType, uInt32 nSamples, void *callbackData);
    int32 CVICALLBACK DoneCallback(TaskHandle taskHandle, int32 status, void *callbackData);
    
    int main(void)
    {
        int32       error=0;
        TaskHandle  taskHandle=0;
        char        errBuff[2048]={'\0'};
    
        /*********************************************/
        // DAQmx Configure Code
        /*********************************************/
        DAQmxErrChk (DAQmxCreateTask("",&taskHandle));
        DAQmxErrChk (DAQmxCreateAIThrmcplChan(taskHandle,"cDAQ1Mod1/ai0","",0.0,100.0,DAQmx_Val_DegC,DAQmx_Val_K_Type_TC,DAQmx_Val_BuiltIn,25.0,""));
     DAQmxErrChk (DAQmxCreateAIVoltageChan(taskHandle,"cDAQ1Mod4/ai0","",DAQmx_Val_Cfg_Default,-60.0,60.0,DAQmx_Val_Volts,NULL));
        DAQmxErrChk (DAQmxCfgSampClkTiming(taskHandle,"",1000.0,DAQmx_Val_Rising,DAQmx_Val_ContSamps,1000));
    
        DAQmxErrChk (DAQmxRegisterEveryNSamplesEvent(taskHandle,DAQmx_Val_Acquired_Into_Buffer,1000,0,EveryNCallback,NULL));
        DAQmxErrChk (DAQmxRegisterDoneEvent(taskHandle,0,DoneCallback,NULL));
    
        /*********************************************/
        // DAQmx Start Code
        /*********************************************/
        DAQmxErrChk (DAQmxStartTask(taskHandle));
    
        printf("Acquiring samples continuously. Press Enter to interrupt\n");
        getchar();
    
    Error:
        if( DAQmxFailed(error) )
            DAQmxGetExtendedErrorInfo(errBuff,2048);
        if( taskHandle!=0 ) {
            /*********************************************/
            // DAQmx Stop Code
            /*********************************************/
            DAQmxStopTask(taskHandle);
            DAQmxClearTask(taskHandle);
        }
        if( DAQmxFailed(error) )
            printf("DAQmx Error: %s\n",errBuff);
        printf("End of program, press Enter key to quit\n");
        getchar();
        return 0;
    }
    
    int32 CVICALLBACK EveryNCallback(TaskHandle taskHandle, int32 everyNsamplesEventType, uInt32 nSamples, void *callbackData)
    {
        int32       error=0;
        char        errBuff[2048]={'\0'};
        static int  totalRead=0;
        int32       read=0;
        float64     data[1000];
     uInt32 i;
    
        /*********************************************/
        // DAQmx Read Code
        /*********************************************/
        DAQmxErrChk (DAQmxReadAnalogF64(taskHandle,-1,10.0,DAQmx_Val_GroupByScanNumber,data,1000,&read,NULL));
    
        if( read>0 ) {
     for (i = 0 ; i < 2*read ; i++ ) { printf("%.5f\n", data[i]); }
            printf("Acquired %d samples. Total %d\r",read,totalRead+=read);
            fflush(stdout);
        }
    
    Error:
        if( DAQmxFailed(error) ) {
            DAQmxGetExtendedErrorInfo(errBuff,2048);
            /*********************************************/
            // DAQmx Stop Code
            /*********************************************/
            DAQmxStopTask(taskHandle);
            DAQmxClearTask(taskHandle);
            printf("DAQmx Error: %s\n",errBuff);
        }
        return 0;
    }
    
    int32 CVICALLBACK DoneCallback(TaskHandle taskHandle, int32 status, void *callbackData)
    {
        int32   error=0;
        char    errBuff[2048]={'\0'};
    
        // Check to see if an error stopped the task.
        DAQmxErrChk (status);
    
    Error:
        if( DAQmxFailed(error) ) {
            DAQmxGetExtendedErrorInfo(errBuff,2048);
            DAQmxClearTask(taskHandle);
            printf("DAQmx Error: %s\n",errBuff);
        }
        return 0;
    }
    

    One more thing I've noticed is that the size of the data [] has not be increased to include additional channels. Wear in 2000 and also to increase arraySizeInSamps (parameter 5) and 2000 to the DAQmxReadAnalogF64 function call. Given that it was only the value 1000 samples, it will get only 500 samples per channel causing the overflow.

  • How to divide 3 signals which are produced by a data acquisition Wizard?

    I've only been using LabVIEW for a few months without any type of training, so I don't know what's ridiculous or wrong. I checked the forum and LabVIEW help for something similar to what I want to accomplish, but I use the wrong words, questions or methods in general. No matter what help options or directions.

    I expect 3 incoming signals that will be accessible by a DAQ Assistant. A signal is the complete signal, the other two are part of the complete signal. Partial signals each separately multiplied by the full signal. How to separate the signals that are in dynamic data produced by the wizard DAQ in a way that I can apply these processes?

    Or, as I have already said, use the Signal from Split.

  • How can I create delays for many digital outputs?

    I use a PXI 6537 DAQ card to produce four digital signals for each starting at a different time. In order to create this delay the appropriate number of '0' have been placed at the front of the boolean table before converting it into a digital signal. It works by creating the delay, however using this process, the delay may be discrete increments equal to 1/╔chantillonnage and my card can be up to 50 MHz, this means that I can produce only delays 20ns increments (check my calculations). For my application, I need to be able to control the time until the second nano. Delays will usually be more than 20 ns, but less than 40ns. I tried to use the entry 'to' the 'build waveform.vi' but it doesn't seem to do anything to manipulate the data. The sub VI used to create the waveform is attached. Sending the data isn't a problem. Thanks in adavnce for any suggestion.

    JS

    Hi JS.

    The 6537 allow placement of subsample edge on its i/o lines.  The best you can do is unique (like you the noticed) examples of resolution (20 ns @ 50 MHz).

    Have you looked at the SMU-6545?  It is a 200 MHz Board, which means that you will get 5 placement ns, as WELL as data delay feature which lets you place your samples with ~ 150ps resolution compared to the output clock.

    Are you having to change the location of a signal in the clock period each cycle?  Or is the position fixed by reference to the clock, but must be placed exactly?

    Thank you

    Keith Shapiro

    National Instruments R & D

  • Compound of button: how to return the signal of a component to another

    Hello, I am quite new in BB10 development. I have a question:

    I'm doing button image with text. My idea was that I will create custom component consists of two components:

    ImageButton

    Label

    Label will be the button Image and contain customizable text

    It basically works, but it seems that swallows Label click signal and it is not propagated to the ImageButton control. I tried to put

    setFocusPolicy (FocusPolicy::None); for the label, but it does not work.

    My questio is: how to do it correctly? Should I capture signal clicked() on the label and send it back to the ImageButton control?

    Thank you

    Tom

    Hello!

    This can help you!

    Refer to overlap touch policies in

    https://developer.BlackBerry.com/Cascades/documentation/dev/touch/touch_propagation.html

    OR

    See how to create the source for the Timer class

    https://developer.BlackBerry.com/Cascades/documentation/dev/signals_slots/signals_slots_create_timer...

    I think so

    The control signal can be connected to the signal of another control.

    -Rezé

  • How to sample the color gradient

    I have an object in Illustrator CC which has a gradient applied to it. I would use the eyedropper tool to sample a color from a specific point in the gradient, but when I click on the object with the gradient eyedropper everything is picked up.

    How to sample just a point in a broken down?

    I looked at my preferences of pipette. Everything is checked. I saw an option that would let me do what I wanted.

    Any suggestions?

    Phil

    Try this: draw a filled rectangle of gradient or make a copy of your gradient filled object.

    Go to object > expand. The slope becomes a mixture and you can choose any stage you like for sampling.

    Note that, according to fussy how you are, you can choose less than 255 objects (steps) case of extension.

  • How cancel a monthly fee for storage space I need is no longer? Does anyone know how to do this?

    How cancel a monthly fee for storage space I need is no longer? Does anyone know how to do this?

    You can change the storage volume by selecting a new plan directly from your iPhone, iPad, iPod, Mac or PC (for example, a plan of 5 GB to replace a 200 GB plan). The downgrade will take effect after the end of your monthly or annual subscription in progress.

    iOS


    1. If you use the iOS 8 or laters, go to settings > iCloud > storage. If you are using an older version of iOS, go to settings > iCloud > storage and backup.
    2. Click Change storage Plan.
    3. Press Downgrade Options and enter your Apple ID.
    4. Choose a different plan to use when the expiration of your current storage plan.
    5. Tap done.


    OS X


    1. Choose the Apple menu > System Preferences, and then click iCloud.
    2. Click on manage in the lower corner on the right.
    3. In the upper right corner, click Change storage Plan.
    4. Click Options to Downgrade.
    5. Enter your Apple ID password, then click on manage.
    6. Choose a different plan to use when the expiration of your current storage plan.
    7. Click done.
  • How many gigabytes of data storage has the new Apple Watch series 2 and how many GB are free for music?

    How many gigabytes of data storage has the new Apple Watch series 2 and how many GB are free for music?

    Hello

    Apple has not yet announced the capacity of storage or the music of Apple Watch series 2 storage limit.

  • How to choose an icon for a pinned tab?

    I created a two pinned tabs. We have a very generic icon and the other has no icon. How to choose an icon for the pinned tabs? Can it be done? Using a hack is fine too.

    You can probably do it with the code in the userChrome.css file.

    Here is an example that works for Gmail.

    Add code to the file userChrome.css below default @namespace.

    @namespace url("http://www.mozilla.org/keymaster/gatekeeper/there.is.only.xul"); /* only needed once */
    
    #tabbrowser-tabs .tabbrowser-tab[pinned][label^="Gmail"] .tab-icon-image {
     list-style-image: url("https://ssl.gstatic.com/ui/v1/icons/mail/images/favicon5.ico") !important;
    }
    

    The file userChrome.css (UI) customization and userContent.css (Web sites) are located in the folder of chrome in the Firefox profile folder.

  • How to recover my sign for email

    How to recover my sign for email

    Of your email provider. Your username and password are created and managed by your email provider.

Maybe you are looking for

  • WIN 7 32 BIT DRIVER CHART FOR PAVILION DV9823CL

    I DON'T THINK WIN 7 32 BIT DRIVER CHART FOR PAVILION DV9823CL

  • My Toshiba warm up and show off

    I can't fix this problem my toshiba become hot n demonstration outside someone can help me? pls

  • Some keys on my Qosmio G40-10 do not work

    Hello Some keys (y h and 6) on my Qosmio G40-10 do not work for some unknown reason and I'll have to use the keyboard for on-screen typing.I bought my laptop at pcworld.co.uk while I was in the United Kingdom, but now I'm in Mauritius (Africa). I hav

  • What is the function of password HARD drive physically?

    Acer Aspire E3 I put a password on the HARD drive in the BIOS, and I see that it is opposed to access F12 boot without the password menu, thus preventing boot from somwething like a Linux Live CD (good!). My question is, in addition to password lock

  • Memory card blackBerry Smartphones

    Hi all Not sure if this has been discussed before and I apologize if it has.  I'm looking on microSDHC memory cards for my bberry.  Someone had problems with one of these based on its size?  I thought to put in a 8 GB card, but the site says to be ca