Is the DAQ Assistant compatible with LabView7?

I am equipped with a LabView7 and a NI DAQ - 6015 Pad and am trying to monitor the temperature with two thermocouples. I searched some tutorials online, but most suggest using assistant DAQ, which is not listed in the section of my functions palette entry. I have downloaded the NOR-most recent DAQmx driver who is supposed to be compatible with LabView 7.x. I don't know if this means that it is compatible with LabView 7.0. I still do not see the DAQ assistant appear.

Any help or suggestion (about DAQ assistant or how to program in LabView7 to solve my problem) is greatly appreciated!

Thank you.

It looks like the DAQmx latest version is 8.1 which will work with LV 7.0

See here http://digital.ni.com/public.nsf/allkb/97D574BB1D1EEC918625708100596848

Tags: NI Software

Similar Questions

  • New Daq with the Daq Assistant in the filtering code

    Hei,

    I have a NI USB-6225 DaqMx I used a couple of years. When I started with LabVIEW, I found the Daq Assistant to the best way to measure the voltage with my Daq etc. My company has purchased another DaqMx NI USB-6225 and now I have a big problem: the Daq Assistant in my old Vi does not work with the new data acquisition. I understand why there is this problem, but I do not know how to solve. I found this code on the forum who finds that Daq is connected:

    The problem is that Daq Assistant do not have an entry for it, and it gives me an error if I try to run the code with a different device than the original, I used when I created the code.

    Is there a way to solve this, so I don't have to convert all the Assistants Daq normal code?

    Hello again,

    two options:

    (1) as the old software is related to 'Dev1' you must rename your new device to this alias and skip/rename the old device (and lack).

    (2) rewrite your old software does not become is not dependent on the name of the alias for the data acquisition card...

    It's your choice!

  • Calendar and the problems of data collection with the DAQ Assistant

    Hello NOR Developer area,

    I am a Novice of LabVIEW and have seen how helpful you all can be, and if I come to ask for your help.

    I'm having some trouble with a VI I built that specifies an input voltage, a SCB - 100 connected to a PCI-6031E and converts this tension in a temperature displayed on a waveform table. The goal is to give a constant reading of the temperature and display it in a chart for as long that the VI is running (and to reset the chart the next time the tracks of VI).

    The problems I've encountered currently are:

    -After a few minutes of the VI running, I get an error message 200279: tried to read samples that are no longer available. The requested sample was already available, but has since been replaced. (to the DAQ Assistant express VI).

    -I don't know how to change my chart so that the minimum value X is both during which the VI was launched and have the maximum X value increases with each iteration of the loop. Currently, I have the VI get the time system and contributing to the property node X scale. This worked for the graph of the voltage, but not for the temperature chart

    I appreciate those of you who took the time to read my post.

    Thank you all for your help.

    Sincerely,

    Ethan A. Klein

    SB candidate in Chemistry & Physics

    Massachusetts Institute of Technology

    Class of 2015

    PS I enclose my VI to give you a better understanding of my current situation.

    E A Klein wrote:

    Thanks for writing.

    What property node is talking?
    I do not understand that many different data types. How can I go on the treatment of all the data?

    (Did you mean I should wire 'blue' data for mathematical functions rather than using the node property tension?)

    Sincerely,

    In fact, one of the nodes property.  I mean specifically the tension property node.  But in reviewing, I noticed the other nodes in property for the chart.  Just set auto-scaling to the X scale and that should take care of two of the nodes property (right click on the graph, X scale-> AutoScale X Scale).  I also recommend placing your mathematical functions in a Subvi to make things easier to read.  Attached, that's what I think you're after.

    I hope that these small tweaks will speed things up enough to avoid your error.  If this isn't the case, then we should begin to look at the design of producer/consumer model or take readings at the same time.  It might also be worth looking away the DAQ Assistant and DAQmx real screws.  But one step at a time.

  • Replacement of simulate signal VI with the DAQ assistant

    Hello

    I have a VI fucntioning, where signal comes 2 simulate VI signals and is being recorded.

    I replaced that with the DAQ assistant. I want to know if the wiring is correct and if it will give me the same result as the simulated VI.

    Thank you.

    Hello Andy,

    Thanks for the reply, yes I did that as well, another way to do it is the 'split signal VI' that automatically separates the signal into 2 channels, most of the high sons being Ch 0

    S

  • 201003-error occurred in the DAQ Assistant

    Hello. I use "cDAQ-9178" and "NI 9215" and "NEITHER 9402" are added on. "

    However, when I run Labview code, "Error-201003" occurs.

    {

    Device not available. Possible causes:

    Device is no longer present in the system / device is not powered.

    Device is turned on, but was temporarily without electricity / device is damaged

    }

    (Error appears as the 1st and 2nd figures below).

    (Plans of logic is the figure below).

    Thank you.

    I could be something with the pilot

    Check this box:

    Error 201003 to the MAX test panel or all by running the DAQ Assistant

    http://digital.NI.com/public.nsf/allkb/5413F392D88326148625746B006745C5

    In this forum, they speak the same error:

    Spontaneous error code 201003 for acquisition of data PCI configuration

    http://forums.NI.com/T5/SignalExpress/spontaneous-error-code-201003-for-PCI-DAQ-Setup/TD-p/830707

  • Simulate signals wired to the DAQ assistant for USB-6009 device

    Hello

    I'm trying to send a signal to the DAQ Assistant Express VI.  I watched the movie "Generating a Signal" on the Web site of NOR (www.ni.com/academic/students/learnlabview/generate.htm) and I have my Signal simulate connected directly on the DAQ Assistant, as shown in this film.  In my case, the DAQ Assistant sends the signal to a device USB-6009.

    However, I received this message:

    Error-200077 occurred to the DAQ Assistant
    Possible reasons:

    Requested value is not supported for this property value. The value of the property may be invalid because it is in conflict with another property.

    Property: SampTimingType
    asked the value: Sample clock
    You select: On-demand

    If I select 'On Demand' in my DAQ assistant and run the vi everything works beautifully.  However, I need my DAQ assistant to be configured to generate a waveform AC continuous, not output a single alternating current rippling.

    What happens here?  I did not have this problem before on other devices of NOR. I am using LABView 2010.

    Please answer.

    Thank you.

    double post

  • Multiple entries to the DAQ Assistant

    Hello

    I'm doing my DAQ Assistant, in several (formed of an array) Boolean inputs where there is 1 digital output. (see attached software folder)

    Physically, I want a valve to open and close at a certain pace, where the user can install/control this pattern until the program starts.

    I think that the best way to do it is to have multiple Boolean values that the user can press or unpress.

    Before that, I started, I tried with only Boolean 1 where it worked perfectly.

    As seen on the attachment (error), it is possible to an easy problem to solve, but I just can't figure it out, I'm stuck at my already made solution.

    I use USB6008.

    I hope that there is a gentle soul who can help out me.

    Best regards

    Kenneth G. Vejen

    Hi Kenneth.

    When the output to the generation mode is set to "sample 1", which means that whenever you call the DAQ Assistant will generate 1 sample. In order to generate 5 samples, you must therefore call 5 times.

    I have attached a modified version of your VI, which shows a way to archive it. However, be aware that the samples will be generated fast and not at 100 ms note your loop runs. It depends on your application, if it is as you want samples to be issued.

  • Calibration of thermocouples in the DAQ Assistant, using data from spreadsheet?

    Hi all

    For my first application LabVIEW, I'm looking to automate the calibration of thermocouples by measuring their response at different temperatures in a dry well Sizer. I get temperatures of thermocouples six by SCXI 1303/1102/1600 and have six channels put in place in one of my subVIs in the DAQ assistant.

    I compare these values to temperature calibrator that I am acquiring by VISA series in an other Subvi. All these values are written in a .csv file.

    Can I import these data into the DAQ Assistant to use for calibration? Is there a simpler way to associate with the channel calibration data? Currently, I could manually copy - paste the cells on the worksheet in the calibration sheet, but that seems just silly.

    If there is everything that I could provide to help solve the problem, let me know!

    Thank you!

    Hi Zoysiamo,

    It is possible to automate the calibration screws DAQmx at a lower level, in particular the DAQmx channel property node. Using you can specify advance nationwide and the values on the scale for your channel. I recommend you take a look at this example of the community. The property node configuration will be similar to, as illustrated below:

  • In Labview 8.5, what happens if the input signal exceeds the scope of the signal defined by the DAQ Assistant?

    Hi all

    This should be a pretty simple question, but I can't seem to find the answer online and currently do not have the functionality to test this:

    I'm using LabVIEW 8.5 and have a VI that imports data from sensor through the DAQ Assistant. In the configuration tab, there is a range of signal input. What happens if my sensor exceeds this range? I get a warning? The default value is the maximum (or minimum)? I was interested in writing a code to display an error that I approach the limits of this range, but did not know if I also need to include code to display an error if the scope is exceeded as well.

    Thanks for the help,

    Tristan

    Hello, Tristan,.

    The behavior depends on the selected range and the device you are using.

    If you are using a device with a single input range is valid, we will use this range, even if you set a smaller minimum and maximum in the DAQ Assistant.  So, if your device only supports ±10V and you set the range to ±8V, you will still continue to get valid data after your top sensor 8V until what you approach 10V.  When you reach the limit of the extent of your device, the output will be 'rail', and simply return the maximum value until the signal is less than the maximum value again.

    Note: A device that is nominally ±10V usually has a go-around (such as ±10.2V) which are usually specced in the manual.

    However, if you use a device with several ranges of entry then things become more complex.

    NOR-DAQmx player will choose the smallest range that entirely covers the interval you choose.  For example, suppose that your device supports the following input range: ±0.2V, ±1, ±5V, ±10V and you choose 0V - 3V as the range in the DAQ assistant.  The NOR-DAQmx driver will focus on the input range and the list of the entry lines that your hardware supports and choose the smallest encompassing the entire range that you set.  This would be the ±5V, because this is the only beach that contains up to 3V.  Thus, all between ±5V input signal is returned and none outside this range will be 'rail' to the maximum or minimum value.

    We do this because using small beaches make more efficient use of the resolution of the ADC.  So, we try to use the most effective range based on what you ask without picking up a range that will make you miss data.

    Let me know if I can clarify it more.

  • Precise triggering voltage input and output generation in the DAQ Assistant

    Hello

    I wonder if anyone has come across a simular problem with the synchronization of input and output voltage. I use a box 11 LabView and NI USB-6259. I have been using the DAQ Assistant to configure the input and output channel. In particular, my task is to generate a single rectangular "pulse" as the output voltage to drive a coil and once the pulse went to get a signal from a sensor of magnetic field and get a power spectrum. This means that the order and the time during which the DAQ Assistant is used is extremely important. For example, the output voltage channel must be opened first for 2 seconds. Subsequently, the channel of input voltage must be open for 1 second, in which the sensor signal is obtained and post-processed. Only after these tasks are performed in this order he can can be repeated in a loop until the experiment is over. I don't know how to trigger data acquisition assistants (one for entry) and the other for the voltage output correctly. Y at - it a trick?

    See you soon

    Michael

    Hi Dave,.

    Thank you that I wired the error strings but the timing issue was unrelated to it. In the DAQ assistant, I simply had to choose the continuous aquistion of the 'samples' methods 'N-switch' for input and output voltage and all works fine now.

    Thanks again

    Michael

  • No device supported in the DAQ assistant

    Hi all

    I use the scanner high speed (NI 5922). I installed Labview 8.5.1, then the software scope OR the digitizer NI 5922 high of speed, then the hardware driver. I tested the DAQmx with the MAX and it passes the test. But when I started to use the labview with the DAQ assistant, as described in the manuals I have not found all devices supported! Should I log a physical device to the hardware driver channels or what should I do?

    Thank you

    Ibrahim

    The 5922 is supported by NO-Scope http://www.ni.com/digitizers/software.htm and not by DAQmx. This page should help you get started.

  • How to find the DAQ Assistant in LabVIEW student 2009 Edition

    I got 1 Core and Core 2 in LabVIEW 2009 prices in June. When I went to use the student edition I have LabVIEW 2009, I couldn't find the DAQ Assistant. This will present a problem when I come to use it in class this fall. It works fine on my version 8.5.

    You must install DAQmx, just like you had to do with 8.5.

  • USB-8451 does not appear in the menu of the DAQ assistant

    Hello

    I use a USB-8451 with LabVIEW 8.6 device. I installed the 1.1.4 pilots of this device and windows and MAX detect. I also renamed the alias and the status of the device is "this device works correctly.

    The problem comes when I try to add a block DAQ Assistant to my project. I select the entry type, I want to have (digital line, although I've tried them all) and LabVIEW doesn't detect any device.

    I tried to make a diagnosis with Panel Test VISA. Here, I assume that something is wrong, because when I run a viRead I get status of return 'BFFF0015h' in red letters... but I don't really know what that means. The LED on the device is green and flashes.

    Thank you in advance.

    It is not supposed to appear in the DAQ Assistant (and certainly won't NI-VISA) or if you have not read the Manual?

  • is daq bit resolution applied to the area defined in the daq assistant or well to the limits of physical blood of the daq hardware.

    Hello, this is my first post here,

    I've been wondering about this:

    When you configure a new daq with the daq Assistant, is the resolution applied to the voltage / number of entries you define, or is it a fixed value per volt by channel.

    For example:

    I have a 6015 usb data acquisition, it has 16 inputs analog, 16-bit resolution 250 kech. / s 09:50 volts.

    so is this as (2 ^ 16) / (16 * (-10-10)) = my resolution through volt?

    or that this has nothing to do with the amount of channels,

    If I want to measure more than 0 to 5 volts, my 16-bit apply from 0 to 5, or even more scale of data acquisition (which is 09:50 volts)

    Thank you

    Marco

    You will get full-resolution 16 bits on each channel regardless of the number of channels is configured.  Renault most are multi-plexed if your specification of sampling RATE in usually a global significance that you can only scan channels with a sampling frequency = MAX RATE / number of CHANNELS.  Also, most NI Renault have a Programmable Gain amplifier on the front-end server, so if you specify a range of voltage smaller in your task configuration the amplifier will automatically increase the gain to use most of the possible BIT ADC.  You see care device and for more details...

  • Recovery of the DAQ Assistant data acquired

    Hello

    I'm currently dealing with a continuous data using NOR cDAQ-9174 proposed acquisition and recording of analog input signals of a built-in three-measuring probe.

    I built a simple vi using DAQ Assistant to acquire data and write to an output .txt - rather than .tdms using Signal Express.

    On a day 10 cycle of data acquisition computer was mistakenly turned off - leaving the empty output .txt file.  LabView recovered the VI cut and I wonder if there is a way I can access the data that has been saved by the DAQ Assistant which can be saved in temporary files etc..

    I have no idea where that might be, since you cannot delve deeper down into the 'levels' of DAQ assistant as you would a sub - vi.

    Just as a note aside to apologize my stupidity - I realize that all the data at the end of the writing task is stupid and completely avoidable... but I worked for a date limit.

    Thanks in advance for any help you can provide.

    Dan

    The most likely answer is not, unfortunately. It looks like you were a table of data at every point of the construction and then measure he writes at the end. In this case, unless you have explicitly recorded data in a temporary file, it is located right in volatile memory, waiting for you to do something with it.

    I realize that this isn't what you want to hear, as it comes to the time of submission of draft / year...

    If you post your VI (preferably version LV2012 or below), I can have a look to see if there is anything obvious.

Maybe you are looking for

  • Hello am installed without my permission

    Hello app or the "Start a Conversation" icon appeared in my browser Firefox/interface without my permission tab. I remember getting an email saying to start a Conversation, I downloaded from the app store (itunes) and then as quickly as the open emai

  • HP recovery new hdd

    I have a HP dv6 3260es Nightingale that came with his "original HARD drive. I recently bought a new hard drive 250 GB SSD and replaced the old HDD by this news. I kept the old HDD and placed in the CD/DVD slot. Now, I have 2 hard drives and no CD/DVD

  • How to reset to factory setting

    I want to reset my HP Touchpad to the factory setting.  What should I do?  Thank you.

  • I lost my serial number and that you can't find on any Adobe

    I lost my serial number for CS6 and have lost half a day to research on Adobe web sites that keep back me to the same home page. I can not type in one of the toolbars of the editable field. I need it's resolved quickly as I am a teacher and you need

  • If / then syntax

    I'm fighting to understand how the "column" is identified by the syntax error. Regardless, this is what I have in the script and can't identify the script errorIf LoanAmount.rawValue = (FranchiseFee.rawValue + Buildout.rawValue + EMR.rawValue) thenPo