Information SCXI

This is the first time I work with platform SCXI. We have a SCXI-1000 chassis, a 64 mio pc HAVE, a 32 entry SCXI 1102 and a block of endings 1303 for reading signals from thermocouples 32. Now we have the opportunity to purchase a module SCXI-1100 and 1300, one block of endings. Could they to Thermocuple read? What is the difference with 1303 1102 and 1100, and 1300 respectively?

Francesco

Thank you very much for your answer. I should say: "that's all".

Thank you once again

Francesco

Tags: NI Hardware

Similar Questions

  • SCXI-1581 reading bad weather

    Hello

    I'm an intern summer and I'll try to find some help with the SCXI-1581 module. We had a major failure during the test of an engine and it damaged some of the test equipment. We have replaced all that appears to be damaged and tried to restart the test system. Everything worked fine except for this add-on. I use a 100ohm RTD to test the different channels and I get a constant temperature of evil. Instead of the ambient temperature from 26 to 27 degrees Celsius, I myself am-243 degrees Celsius-244. It seems as if there could be a scaling factor that is causing temperatures to record these low values.

    As part of my process of troubleshooting, I took the module SCXI-1581 the machine damaged and replaced by the same module from another machine. Temperature readings read correctly for all channels that I tested. Then I took the right module and put it back in the machine to work and he still read appropriate temperatures. With this small made check I took the bad module and placed in the machine to work and ended up reading the bad weather. At this point, I had concluded that the module must be bad and replaced the good module back into the machine to work. As a final check, I tested again temperatures and good module is reading the same bad temperatures as another module.

    I have trouble believing that mearly placing bad module into the working machine was sufficient to damage the good module as well. So my question is, is it possible to adjust the scale factor or the SCXI-1581 module needs to be replaced? Is there something else I can do for these modules work?

    I can not provided enough information, and for that I apologize. I'm not familiar with this equipment at 100%, but I'll do my best to provide additional information that may be required.

    Thank you

    Thanks for the reply.

    My boss and I looked into it more and found the two cards to work very well. The error was in the RTD that we use to test the card.

  • How can I filter the noises of thermocouple using a TC-2095 and SCXI - 1102c?

    Hi all

    I currently have four TC - 2095 s connected to the SCXI - 1102c inside an SCXI-1001 chassis cards. These thermocouples are floating and I get a good amount of noise (400 Hz 3 phases and same 60 Hz noise) when we turn on some equipment such as fans, and fans. Using this current set up, how can I filter the noise? Or I would need to put in place a new hardware configuration (e.g., the SCXI-1102 cards)?

    Hi RALoo,

    The 1102 C has a lowpass filter with a cutoff at 10 kHz frequency, so you will not be able to filter into the material with your current configuration.

    Your first option is to use a software filter.  LabVIEW has several functions that do this. More information can be found here: http://zone.ni.com/reference/en-XX/help/371361J-01/lvanls/filter_vis/

    The second option would be to change to SCXI 1102 cards as you suggest.  The low-pass filter of the 1102 was a cut-off adjustable to 2 Hz frequency, which will stop the noise of 400 and 60 Hz don't affect your readings.

  • SCXI and acquisition with Windows 7 x 64

    Our just updated IT to update our phones and we have received Windows 7 x 64.

    Data acquisition, we use the SCXI hardware in the SCXI-1000DC chassis and an acquisition of data SCXI-1600 USB card.  Doing my gear OR set up on the new laptop, I discovered that neither does not support Windows 7 x 64 for the SCXI-1600. The other options to connect the SCXI-1000 chassis to a laptop are the DAQcard-6062E and DAQcard-6036E, nor which supports Windows 7 x 64.  According to the information below, there are no plans to support x 64.

    What are my options to continue the use of the material, we have invested with the laptop running Windows 7 x 64?

    You will need the NI SCXI-1349 connection block and SH68-68-EP cable.  The SCXI-1349 plug at the rear of your SCXI module and cable connects the 1349 the USB device series M mass termination.  The rest of the configuration is identical to any PCI/PXI E Series device.  The USB M Series device will be the analog-to-digital converter for the system instead of the SCXI-1600.

  • Parameters of Configuration of No (gain) of SCXI-1313 able and Automation Explorer

    Hello

    QUESTION:

    Help, please.  What should I do for my system to be able to define the 1125/1313 to re-read a +/-50v range?

    DETAILS OF THE PROBLEM:

    I have a SCXI-1125 module with a 1313 mitigation module which is be digitized by a PXI-6251.

    I create my daqmx inside my program tasks.  I want to put the max/min range for signals entering the 1125/1313 to +/-50v.  However, I get an error of daqmx saying that the mini/maxi is limited to +/-10v.

    Furthermore, I see no to the MAX so that the attenutaion of the 1313 or the gain of the 1125.  MAX actually displays a message saying that there isn't any configuration required for the 1313.

    Installation program:

    LabVIEW 8.6

    MAX 4.6.2f1

    DAQmx 9.0.2

    Windows 7 Pro - 32-bit

    Best regards

    Chris

    Dear Chris,

    You may find it useful to review the following Knowledge Base so that you can find more information on what I will explain:

    http://digital.NI.com/public.nsf/allkb/B138E2AA836B5354862570380073EA60?OpenDocument

    The SCXI-1313 is an attenuator module which serves to reduce your levels of input voltage so that they are in the range of input for your DAQ card voltage. When you use DAQmx in LabVIEW, you don't need to worry about setting the +/-50v rank because it will be scaling automatically. Simply choose any given voltage range is valid for your specific DAQ card (+/-10 v) in this case.

    It is for this reason that when you go to MAX to try to configure the attenuation, you get a message telling you that you need not do this from the descaling is done automatically.

  • PCI-6052E and SCXI-1520

    Hi all

    I had problems to understand the program to someone. Could someone of you help out me?

    These programs are written in Labview 6.x and integrated in a software, which can detect the DAQ (PCI-6052E) card, a series of input modules (SCXI-1520) and then collect the data from the input modules simutanously.

    Issues related to the:

    1. how to make a menu drop-down list all the modules detected, and then the customer can choose between them? Looks like the former programmer do that with property node...

    2. now, we have a new input module (SCXI-1530). However, the software says it can only suppor SCXI-1531 and cannot detect SCXI-1530, although MAX can do. The difference between SCXI-1530 (4 channels) and SCXI-1531 (8 channels) is the number of channels. It is possible to modify the software and make it compatible with SCXI-1530?

    All ideas are appericiated.

    Thank you

    Hi oly.

    You're right, aid for SCXI Information.vi does not have the types of module for SCXI-1530/1531. The function C (SCXI_Get_Module_Info) in the Start Menu > all programs > National Instruments > NOR-DAQ > traditional NOR-DAQ function reference Help contains 48: SCXI-1531; 50: SCXI-1530.

    Brad

  • My PCI-6250 does not see my SCXI box

    Hello

    I want to install a station using a PCI-6250 connected to an SCXI chassis cable 68 pin.  We have a couple of these formations, so they gave me a Setup disk that have our configuration is already done.  However, I had to install a software, that they neglected to give me.  I got the latest driver for the latter of NOR.   Well, maximum running, my tree is distinguished by a previous configuration.  And mine is not communicating with the SCXI.  My cards are "x" ed out.  The Communicator 'chassis' in the SCXI chassis Configuration is grey (no options).  I checked the wiring and disconnected all the blocks for the cards.

    Here is my tree.

    NOTE: Mine also does not have the "NOR-DAQmx devices" under the heading "devices and Interfaces, only in the software section.  Other configurations have this list.  I don't know how to get "devices OR-DAQmx" shown.

    Any help would be appreciated.

    My system

    • Commune
  • Editor Script OR 1.3.1
  • NEITHER Spy 2.7
  • NOR-488. 2 2.73
  • NOR-DAQmx ADE support 9.0.2
  • NOR-DAQmx Device Driver 9.0.2
  • NOR-DAQmx MAX Configuration 9.0.2
  • NOR-FGEN 2.5.1
    • FGEN Soft Front Panel
  • NOR-HWS 1.4.5
  • NOR-PAL 2.5.2
  • NOR-TClk 1.6.1
  • NOR-USI 1.0.2
  • NI-VISA 4.6
    • NiVisaServer.exe
    • NIvisaic.exe
  • Duration of NI-VISA 4.6
  • NOR-DAQ traditional 7.4.4
  • IVI drivers
    • Advanced

      • Instrument Driver software modules
      • Simulation software driver modules
      • Pilot simulation sessions
    • Pilot sessions
    • Logical names
  • Hi mz6yx6,

    First of all, don't worry about trying to get to the ' devices OR-DAQmx "appear. 4.6 Max, which was first included with NOR-DAQmx 9.0, has a revised UI layout which displays the devices inside "Devices and Interfaces" rather than display them in subfolders like ' devices OR-DAQmx.

    Also, thank you for the view a report of the configuration of the system MAX. The report shows that the SCXI chassis and modules are configured as simulated devices, which I think is the problem. DAQmx (> = 8.7) does not support configurations where a simulated SCXI chassis and modules are connected to a real communication hardware DAQ. Since your system has not simulated hardware DAQ (except the SCXI chassis and modules), the "Châssis Communicator" menu has no options to display.

    To resolve this issue, remove the SCXI chassis in MAX and create a real one (not simulated). How to do this depends on if you manually configure the chassis or you import a configuration from a file .nce, .ini or .txt:

    • Manual: right click on "Devices and Interfaces", then select «Create a new...» ». In the window "new...". "which appears, open the ' Chassis SCXI of NOR-DAQmx ' subtree and select the chassis type from there. Do not select "Simules NOR-DAQmx device" or modular Instrument The modules must be detected automatically, but you will need to manually reconfigure the accessories.
    • Import from a file .nce, .ini or .txt: in the dialog box 'Assistant Import Configuration', check that the 'Warnings and conflicts' column displays "XYZ will be replaced" or "Create device" and does not "create a simulated device.
    • If the Setup disk makes the SCXI configuration programmatically or an installer for the application, then more information about what is on the Setup disk would be useful.

    Brad

  • How to replace the SCXI-1600?

    Since our IT obliges us to switch to Win7, 64-bit edition, our SCXI-1600 becomes unnecessary because there is no 64 bit driver (Sidenote: the device is still available in the online catalog without obvious by the lack of such a driver warning). in an another tread on this subject, Seth_B advise

    For example, the USB-6251 device mass termination that provides 5 x faster sampling with 7 x better accuracy, USB 2.0 transfer speeds of data transfer more reliable and is only about $300 more including the cable between the DAQ hardware and the SCXI chassis.

    I need more information - I don't understand how exactly who will replace the SCXI-1600 as a connection to the SCXI block, for me, it looks like an ordinary e/s USB device must connect directly to several types of e/s analog and digital through a CBS. What cable should I connect the chassis SCXI containing other maps, how she will turn upward in MAX - in short: I need to decide if the replacement is more trouble with respect to particular software rewriting labview and other efforts that fight OUR guys to keep the System 32 bits for this testbed.

    Hello MBohnet,

    To avoid work or acquisition of new equipment on your part then I would recommend against your IT Department to maintain 32-bit system.

    If you need to upgrade to Windows 7 64 bit then the 6251 is an option at your disposal to communicate and scanning information of you modules SCXI.

    There are steps for setting this up

    1. Remove the 1600 chassis.
    2. Fix a module SCXI-1349 in the back one of the SCXI modules. Then use an SH68-68-EPM cable (the cable is listed on the pricing for the device 6251 tab) to connect the SCXI-1349 module to the PCI-6251.
    3. Your configuration in Measurement & Automation Explorer would be alike in what concerns the addition of your device, but you will point to the 6251 as the digitizer.
    4. The quote above regarding the (5 x faster sampling rate) applies only when using the USB-6251 in parallel mode.
    5. You wouldn't have to make changes to code with switching on the material element.
    6. Also, if you are not connected to the USB take a look at the PCIe-6320 or PCIe-6341 with the SHC-68-68-EPM cable. They are less expensive than the USB-6251 housing alternatives.

    Kind regards

    Izzy O.

    Technical sales engineer

    National Instruments

    www.NI.com/support

  • The SCXI-1530/1531 devices support

    Hello

    I found confusing information on the SCXI-1530 support software. Now it is entirely supported by the Traditioanl DAQ and DAQmx?

    I found that some document said that "DAQ traditional only when you are using an SCXI-1530, SCXI-1531 or SCXI-1540 device with accelerometer, linear voltage differential transformer (LVDT) or the virtual channels Rotary variable differential transformer (RVDT), which are not supported by NOR-DAQmx.» It should be outdated information?

    Thank you

    Hi Oly,

    To be sure, it would help if I knew what document you look. However looking at the Readme for DAQmx looks the SCXI-1530 is supported only for accelerometers and microphones, which means that the LVDT and RVDT task are not available.

    NOR-DAQmx 9.0.2 ReadMe

    http://FTP.NI.com/support/softlib/multifunction_daq/nidaqmx/9.0.2/Readme.html

    Hope that helps,

    Thank you

    Scott M.

  • Error 1600 SCXI 200329

    Hello world

    I used my scxi-1000 chassis with the scxi-1600 daq card. Today, I get this error when I try to get data from several modules. Error number is 200329 and I have attached a screenshot for this error. Do you have an explanation for it?

    Thank you

    Bertrand

    Hey Berkayaga,

    It is possible that a single module, once damaged, which can affect three chassis.  There are a few steps that you can do to try to fix your equipment.  You might start by checking the fuses on the chassis.  Information on this may be found at Troubleshooting an SCXI chassis.  As well as specific information on the fuses themselves, there is also a documentation on where you can find them.

    If this does not work, you can consider the process of authorization to return goods (ARM) located in a National Commission of Instruments sent in for repair or calibration.

  • measures of thermistor using modules SCXI-1102 and SCXI-1303

    Hello

    I have a 1303 SCXI front-end module connected to an SCXI-1102 module

    I read the article detailing how to remove bias and pullup resistors to activate measures of RTD/thermistor on the 1303. What did not mention article was the case plug the excitation source, and what is the best source of operating voltage (normal etc., high precision)

    I was wondering where I can find this information, also, I could do several types of measure on the same 1303? I have an analog voltage to a flow meter, and I would like to put it on the same module as the Thermistors

    Thank you!

    Kevin

    Kevin,

    The pulling up of resistors to connect to + 5 V internally.  If the values are not suitable for your application, you can use those.  It does give you much flexibility but.  The 1303 does not give you any other access to any source of current or voltage through screw terminals.

    Pull upward and pull down resistors are networks of resistors Series 9-pin SIP. So you can edit them in blocks of 8 channels for joint measures.

    Lynn

  • Name of the traditional model DAQ and SCXI-1530

    Hi all

    I'm using Labview 6.1 and 7.1 of DAQ traditional. I'm working on a system with a data acquisition card (PCI-6052E) and several SCXI (1520, 1530) cards. The SCXI cards are installed correctly because MAX can detect all.

    I want to do a program with traditional DAQ that allows to detect all SCXI cards and Show names of model on the front panel.

    I found in Tradtional DAQ, there is a VI entitled "Get SCXI information", which returns the types of models of SCXI cards, something like, 50, 51...

    I found in the help file on 'Get SCXI Information' VI there are some kind of configuration file saying 51 stands for SCXI-1520. What does the acronym for 50?  Where can I find the full configuration file or can somehow I change it? As 51 means something else...

    Any ideas are much appreciated... Thanks to you all!

    This is a duplicate of the last post in this thread: http://forums.ni.com/ni/board/message?board.id=250&thread.id=53087

  • I need to change my security issues and said we do not have enough information to reset your

    I need to change my security but said Questions we have insufficient information to reset security of your apple ID questions.my is [email protected] I want to slove this problem please help me

    You should contact the account of Apple security team. To join, click here and choose a method; If this page does not list one for your country or if you are unable to call, complete and submit this form.

    (145174)

  • Slide to open, how to inform apple we want it back

    How to inform apple that we want the Slide-to-open back function?

    See the following: http://www.apple.com/feedback/iphone.html

  • insufficient information,

    What should I do if I am trying to reset my security questions but not enough information?

    You should contact the account of Apple security team. To join, click here and choose a method; If this page does not list one for your country or if you are unable to call, complete and submit this form.

    (145026)

Maybe you are looking for

  • MacBook Air 13 vs MacBook Pro 13 Retina Display

    Hello! I worked as a teacher in a dental school & I'm an orthodontist by Profession. So far, I was using a Windows laptop. Now, I am planning to buy a MacBook. I did some research and I found two best options: MacBook Air 13 inch or Macbook Pro 13 in

  • Safari no longer works on my iPad Mini

    This morning, when I tried to surf the web using my iPad Mini, I received a message that reads "Safari cannot open the page. "The error as"WebKit has encountered an internal error. My internet connection works fine and all Apple products in the house

  • HP officejet 6500 biased paper feeding of small paper (4 X 6)

    I have alignment issues when I try to use the 4 X 6 photo paper. I tried the printer head alignment and it seems to be makes it worse. I am running Windows 8.1. I see a little oblique when I print a test page for full size. I watched it pull into a 4

  • OfficeJet Pro 8620: Print borderless with Officejet Pro 8620

    Hello I recently bought a hp OfficeJet Pro 8620 and I'm testing now a few important functions for my business. One of them is printing without margins of leaflets in A4 format. But the problem is: even if I chose "A4 borderless" paper size and "photo

  • can I change the graphics memory shared in a dedicated graphics memory?

    I just bought the 3 battle field, his does not work well. It told me that I have dedicated memory graphic need more, is it possible to change my graphics memory shared in dedicated graphics memory more since I got 0 in dedicated graphic memory right