Missing DAQ Assistant

Hi guys,.

I have LabVIEW 2015 and 14.0 DAQmx.

Yet, the daq assistant is absent the express--> palette entry.

Any help?

Message 5 of this thread will give you the answer.  More precisely

DAQmx and LabVIEW version compatibility

Tags: NI Software

Similar Questions

  • problem with install two DAQ in a PC and missing DAQ assistant

    Hello

    I have two 6023E DAQ cards and 6024 and each card has its own DAQmx (DAQmx Vresion 8.7 for 6023E DAQ) and Version 9.1.7 for DAQ 6024

    and I want to install the two ofl these cards together in a PC with LabView 8

    After LabView installation and installation of cards if I install the drivers of DAQmx version 8.7 I don't miss assistant DAQ but NATIONAL & MEASURMENT > DEVICE INTERFACE AND I DAQ 6023E in traditional mode and I can not use traditional DAQ and the DAQ assistant in labview 6024 is also not recognized

    If I install DAQmx worm. 9.1.7 in NATIONAL & MEASURMENT > DEVICE INTERFACE AND I the two DAQ (not in traditional mode), but I've lost assistant in LabView DAQ

    What should I do?

    Please help me...

    Problem solved:

    I have install LabVIEW 2009 and then install DAQmx 9.1.7 and the problem resolved

    Thanks to all the bodies of their response.

  • Connection diagram missing in DAQ Assistant generate the signalling block

    This is my first post so please excuse the quality of my description.

    When I double click on the block of data acquisition - Assistant, there is no tab connection diagram I can access to see how things are wired to the top. I have a NI USB-6211 connected by USB and it is used to control many different sensors and a power supply. Currently, he works for everything and is hard wired correctly, but only blocks DAQ Assistant has a connection diagram available, the other are not. One who has a connection diagram is used to measure a voltage. Others who do not are used to generate a signal. I would really like to be able to see patterns of connection for each block.

    -Any help would be appreciated

    -Thank you

    You can always do like those who never use the DAQ Assistant and read the manual. Right click on the device in MAX and selecting "stitching of the device" works too.

  • missing 2013 LabVIEW daq Assistant

    I do not see the DAQ assistant in Labview 2013!

    LabVIEW 2013 + real-time + fpga

    Max 5.5.1

    Visa 5.4.0

    August 2013 device driver

    Why?

    Thank you


  • DAQ Assistant does not export the values on the scale

    Hello all-

    Potentially stupid question but here goes: I'm using the DAQ Assistant to read in 4 analog input voltages, continuous sampling acquires data at 10 Hz 1 point, using LabView 12 on a machine with an acquisition of data USB-6341 simulated device (because my office is more comfortable than the lab!). I want to change the first two signals of voltage to temperature and humidity, respectively. I used the «create a new...» "in the 'Custom Scaling' drop-down in the"Voltage configuration"tab for each of these channels, named gave the slope and the intercept at the origin for the respective linear scales and click OK."

    When I test the code - and yet once again, I'm not on a machine with a 'real' DAQ system, I use a simulated device, and it seems that NEITHER MAX generates a sine wave of long period with little noise on top for this - I do not get the results on the scale of my 'signal', I get the raw tension. (If you run my code, I will join, the Relative humidity must be between 0 and 100 and temperature-40 to 60, is not 0 to 5, for example.)

    So, what happens? Is there some flag or setting that I missed? The scaling only works on voltage data 'real' of a 'real' instrument DAQ, instead of a simulation (which is why I mentioned twice!)? I have to do something in NI MAX as well as Labview?

    Thanks for any help you can give.

    John Easton

    Simulations devices will not respond to custom scale.  They are just supposed to allow you to configure your device without errors when you do not have the unit on-site.

    "NOR-DAQmx simulated devices create a noisy sine wave to all the entered analog." Simulated data other set-up is not available at this time. »

    http://www.NI.com/white-paper/3698/en

    They generate a sine with an amplitude equal to half of your specified input range.  If you want to work with simulated data that would be more realistic for your application, you could write a VI to generate the data and have a business structure to manage both "simulations" and "real", then you could switch back depending on whether you have access to the material.

    I just checked this with a PCI-6254 I install and simulated a PCI-6254.

  • Conversion task DAQ Assistant

    When I try to convert my DAQ Assistant tasks with a right click - Convert to task NOR-DAQmx I get a name and a scrolly input output with all the broken wires. Similarly, if I create a task in MAX and then use a constant for the task on the block diagram name I get the same result of entry scrolly. Am I missing a step here?

    Thank you

    Hi Mirabelle

    Glad to hear that you managed to solve the problem! In response to your questions:

    -L'information that read you about the creation of tasks in MAX is correct; It is possible to create a global task (an available screw a lot) in MAX. However, it is not necessary to create all of the tasks in MAX. As you think, if you use the task within a single VI, there is no need to create the task in MAX - in fact, it's probably easier just to create the task in the VI himself, either with screws of Express Assistant DAQ or DAQmx screw person on the range of functions located under e/s measurement > DAQmx - Data Acquisition.

    -When tasks change in MAX, it will not automatically be changed in LabVIEW. To make changes in LabVIEW, you will need to edit the task properties in LabVIEW, or remove the old task and replace it with the change.

    -I'm not sure what you mean by the question in your "P.S." to 100%, but I'll try my best to answer. Is there a specific example you speak? If you ask the use of screws DAQmx as DAQmx Virtual Channel Create, task control DAQmx and DAQmx Read etc., as shown in the examples of the expedition, then there are a few reasons to use those. Basically, screws of Express DAQ Assistant consist of a combination of the screw in the DAQmx - Acquisition of data from the function palette section. In general, screws of Express DAQ Assistant moving slower than the standard SIE DAQmx. It may be easier to debug the code when using standard screw, as it is obvious which task runs each section of the code. And finally, there are some limitations to the use of the DAQ Assistant, so while they may be easier to use in the short term, long term, I would recommend familiarize themselves with standard screws, because they make possible more complex tasks.

    I hope this helps! Feel free to let me know if you want to bring me anything, or if you have anything else that I can help you.

    Best regards

  • Meilhaus 3106 MAX/DAQ Assistant questions

    Hello

    While I'm sure that you all have grown tired of my usual throat-clearing, I'm a relatively inexperienced developer, so apologies in advance if this question is incomplete or missing otherwise.

    I recently got a Meilhaus electronic RedLab 3106 DAQ.  Box manual is at: http://www.meilhaus.org/downloadserver/redlab/manual/RedLab%203106_en.pdf

    Attempt to use this device after completing installation 'Quick Start', I discovered that the data acquisition was not read by MAX (v5.0). In addition, by attempting to use the DAQ Assistant, in LabVIEW 2011 for this box DAQ (which I did successfully on this machine with a DAQ USB of NOR-6008), I discovered that the device is not listed with the text "no taken physical channel supported" that appears when the registration of the channel would be generally.

    Normally, I try just a reinstall of the drivers/software 3106 package, but the strange thing is that Windows (XP - SP3) gives me the bubble 'New hardware found' when the USB fuelling the DAQ cable is connected to the computer. In addition, the acquisition of data can be read by a program that came with the acquisition of data called "InstaCal", including an order to make the power led on flashing box. Finally, a universal library, which also comes with the DAQ hardware has itself installed successfully in LabVIEW under "User Libraries". Given this feature, I am convinced that it is simply not possible to use the DAQ Assistant in LabVIEW for this particular data acquisition or that there is something in LabVIEW/MAX in need of fixing.

    Someone knows a problem similar to this with the 3rd party hardware (ideally, Meilhaus) and software of NOR? If so, how do fix you it?

    Kind regards

    MG Wilkinson

    Material acquisition of third-party data is not supposed to be detected by MAX or used with the DAQ Assistant. Thus, simply use the functions you provided.

  • In Labview 8.5, what happens if the input signal exceeds the scope of the signal defined by the DAQ Assistant?

    Hi all

    This should be a pretty simple question, but I can't seem to find the answer online and currently do not have the functionality to test this:

    I'm using LabVIEW 8.5 and have a VI that imports data from sensor through the DAQ Assistant. In the configuration tab, there is a range of signal input. What happens if my sensor exceeds this range? I get a warning? The default value is the maximum (or minimum)? I was interested in writing a code to display an error that I approach the limits of this range, but did not know if I also need to include code to display an error if the scope is exceeded as well.

    Thanks for the help,

    Tristan

    Hello, Tristan,.

    The behavior depends on the selected range and the device you are using.

    If you are using a device with a single input range is valid, we will use this range, even if you set a smaller minimum and maximum in the DAQ Assistant.  So, if your device only supports ±10V and you set the range to ±8V, you will still continue to get valid data after your top sensor 8V until what you approach 10V.  When you reach the limit of the extent of your device, the output will be 'rail', and simply return the maximum value until the signal is less than the maximum value again.

    Note: A device that is nominally ±10V usually has a go-around (such as ±10.2V) which are usually specced in the manual.

    However, if you use a device with several ranges of entry then things become more complex.

    NOR-DAQmx player will choose the smallest range that entirely covers the interval you choose.  For example, suppose that your device supports the following input range: ±0.2V, ±1, ±5V, ±10V and you choose 0V - 3V as the range in the DAQ assistant.  The NOR-DAQmx driver will focus on the input range and the list of the entry lines that your hardware supports and choose the smallest encompassing the entire range that you set.  This would be the ±5V, because this is the only beach that contains up to 3V.  Thus, all between ±5V input signal is returned and none outside this range will be 'rail' to the maximum or minimum value.

    We do this because using small beaches make more efficient use of the resolution of the ADC.  So, we try to use the most effective range based on what you ask without picking up a range that will make you miss data.

    Let me know if I can clarify it more.

  • DAQ Assistant does not appear when starting a new project in Visual Studio 8 with Measurement Studio

    Hello

    I have installed a system with the following configuration:

    Windows 7 Prof

    OR MAX

    Visual Studio 2008 Standard

    Measurement Studio 2013 (evaluation)

    Device drivers OR Feb 2013

    When I start a new project in Visual Studio, I can choose a model with NEITHER Windows forms application. After that I get an overview of the classes, but after that, I get no DAQ assistant. What Miss me as is actual references in the Toolbox, with the exception of the buttons and graphics of NOR.

    Someone at - it a solution for this problem?

    Thanks in advance,

    Erik

    Dear Erik,

    I saw the same problem and I think that you can solve with the installation of the DAQmx 8.8 and make sure to include the Visual Studio integration.

    If you have any questions just reply to this email.

    Have a good time.

    Kind regards

    Hossein

  • Precison display digital DAQ assistant tank

    My tank is connected to a DAQ assistant. I want to change the digital display decimal places of the value of 6 to 2 (9.44223 to 9.44) tank, but no matter where I adjust in reservoir properties, none can do.

    I missed something? Thanks in advance.

    Hello sunflower.

    I also did not find an option in the Properties dialog box, but you can try this:

    (hmm, LV2009 change the property node of a tank to a generic reference/property node...)

    Or right click on the digital display-> create the node-> DisplayFormat or FormatString property

  • several channels in daq assistant

    Hi guys,.

    My first post on this forum, would appreciate some advice.

    I use labview 8.2 with PCI-6071E acq. card.

    I'm observing unexpected behavior when you try to read analog data on my card.

    If I define in the daq assistant, one way to read, everything is great, and the data is ready, as planned. I can see it in the preview window (explicit task)

    When I open the daq assistant.

    But, at the point where I have to add a channel more (from the same physical pci-6071e card), my measurements are totally fake - I get a strange signal...

    The sampling rate is more than enough (I enjoy with 64khz a 1 kHz signal arriving at one of my channels).

    What I'm missing here?

    Thanks for any advice you may have.

    Lenny

    With the generator of signals (output TTL) or the PCI-6071E AO the gsignla is NOT differentiated. Use one of the configurations of single entry completed.

    The Earth all unused inputs.

    Lynn

  • How can I include all of the necessary files required for the DAQ Assistant create an executable file?

    I create an executable file to run on a computer that does not have LabVIEW.  The problem is that I am using the DAQ Assistant.  The hierarchy of the VI is big enough, and I do not know how to include all the required files to eliminate the error "the version of LabVIEw full development is necessary to correct the errors."  A picture of the hierarchy of the VI is attached.

    I think I'm missing something in the installation.  I'm downloading now the DAQmx.  I knowticed that all 7 of my missing Subvi are DAQmx * .vi

  • Configuration of 1120D SCXI using the DAQ assistant

    Hello

    I tried to set up a channel on a module SCXI in 1120 D, as well as the rest of the SCXI system, using the DAQ assistant, but ran into a problem related to the range of the signal.

    According to this, http://sine.ni.com/nips/cds/view/p/lang/en/nid/1658, the maximum voltage is located between 10v - 10v, but when I tried to configure a channel with a range of 0v - 10v, it displays an error indicating that the range of the signal was invalid and that the available range is - 0.005v - 0.005v, could someone suggest what I'm missing or what I need to do to set up?

    Thank you

    Chris

    Hi Chris,

    Thank you for your message.

    Can I ask that you have the software gain set on the device? The table on page A-5 manual found here your device that the gain settings must be for some input range.

    I imagine you have the gain set to 100 while the available range is ±50 mV. For - 10V - 10V, please adjust the gain to 0.5.

    Thank you very much

  • There is no entries or exits on the DAQ assistant to connect to

    Hello world

    I'm new to labview and have encountered difficulties with the DAQ assistant.  I'm under v8.0, v8.7.1 OR-DAQmx labview and have an acquisition data PCI-6224 installed.  It seems that every time I have let down the DAQ assistant (assistant of the instrument and data acquisition simulated has the same question) on the block diagram and configure it, it don't y no input or output that I can wire a control or an indiactor to.  Y at - there a step I'm missing somewhere?

    Thanks in advance!

    Try to repair the DAQmx driver through control panels and see if that solved the problem.

    How to uninstall or repair OR software?

    http://digital.NI.com/public.nsf/allkb/AC6ED75D3D93375686256E8E00245F0D

  • DAQ Assistant is not in LabView

    Hi all

    I use NEITHER cDAQ-9174 and NI 9203.  I have already installed the driver for the NI 9174 cDAQ, which is NEITHER-DAQmx 9.8.0.  I can see the device when I opened NI MAX. However, when I open LabView 2015, I can't find DAQ Assistant in the function Palette. I noticed on the chassis, 'ACTIVE' light is not on while two other "POWER" and "READY" light is on.  I look in the forums OR but I can not find the solution.

    Any suggestion, please help!

    Thank you!

    DAQmx 9.8 is not compatible with LabVIEW 2015.  You must use at least DAQmx 15.0.  See here for more details: NOR-DAQmx and LabVIEW compatibility