DAQ digital output default to low Earth

Hello

This probably a basic question, but I have found information on this subject.

When I connect a data acquisition unit (usb6009) to the USB port I get any output defaults to a high (+ 5V) until I run my program and I put them to lower ground.

Is it possible to configure default data acquisition always at 0V?

I enjoy any comments or links on how to change this.

Thank you

Cristian

The default value for the 6009, is that they are all set for entries. What you see is the pull up. But, no, you cannot set the power of the State to the 6009.

Tags: NI Hardware

Similar Questions

  • 6008 daq digital outputs to control relays

    Hi all, I'm looking to help create a VI to send out digital to a daq 6008 to control relays. What I'm trying to do is when you press start and a condition is met send a digital output to control a relay for 30 seconds or so to take a measured voltage to be taken an analog voltage. After 30 seconds, I want the first relay to switch off and the next relay lights for the same amount of time. I want to continue this sequence to 7 readings, blood for every step and send the data to an excel file. I know it's basic stuff, but my experience with labview is limited! Any help would be greatly appreciated.

    Thank you

    Paul

    Hi Paul,.

    I looked on your problem this afternoon and I agree completely Fan Ravens that the state machine is in fact the most appropriate architecture for such a task of data acquisition. A state machine architecture is one of the most commonly used in LabVIEW design patterns and is especially suitable for any program where you have clearly defined the steps that can be represented by the States and rules for the transition between these States.

    There is a model of Machine of State Standard contained in LabVIEW which should give you an idea of the underlying architecture and is a good starting point. To give you a better idea of how this architecture can be applied to a data acquisition task, I would recommend that you look at This example. Although States will be slightly different in your case, this should provide you with a good understanding of how you can architect such a request.

    I hope this helps.

    Best regards

    Christian Hartshorne

    Technical sales engineer

    National Instruments UK

  • take the digital output USB-6001 always high or low in c

    Hi all

    I am new to the NI DAQ interface. I have a USB-6001 and I am trying to use this device to control some flowchart in C. What I want to do is:

    * set digital output lines with high and low intensity and change their status as needed (in C).

    I tested the device NEITHER Max--> Test panels and found that the device is capable to do that. Then I try to do in C. I have checked hace examples and function I use is one called "DAQmxWriteDigitalU32". I have problem in the understanding of its input parameters. I tried something with my own knowledge, but it does not work as I expected. Here is a test I did:

    data uInt32 = 1;

    Int32 wrote;

    TaskHandle taskHandle = 0;

    DAQmxErrChk (DAQmxCreateTask("",&taskHandle));
    DAQmxErrChk (DAQmxCreateDOChan (taskHandle, "Dev1/port0/line7", "", DAQmx_Val_ChanForAllLines));
    DAQmxErrChk (DAQmxStartTask (taskHandle));
    DAQmxErrChk (DAQmxWriteDigitalU32(taskHandle,1,1,10.0,DAQmx_Val_GroupByChannel,&data,&written,));

    taskHandle = 0;
    DAQmxErrChk (DAQmxCreateTask("",&taskHandle));
    DAQmxErrChk (DAQmxCreateDOChan (taskHandle, "Dev1/port0/$line0", "", DAQmx_Val_ChanForAllLines));
    DAQmxErrChk (DAQmxStartTask (taskHandle));
    DAQmxErrChk (DAQmxWriteDigitalU32(taskHandle,1,1,10.0,DAQmx_Val_GroupByChannel,&data,&written,));

    I just want to set ' Dev1/port0/line7' and ' Dev1/port0/$line0"at a high level, but only ' Dev1/port0/$line0' answer me. The second parameter of the DAQmxWriteDigitalU32 function is numSampsPerChan. If I replace (currently 1) with a higher value, such as 100, I see that "Dev1/port0/line7" sends a number of 1 output, then back to 0. So I guess that the problem is just that I understand not all parameters for the DAQmxWriteDigitalU32 function. Is someone can you please tell me how I can set up a line of digital output 1 or 0?

    Thank you!

    Hongkun

    Hello

    I finally find a way to do it! The feature works very well, and my problem was not set the data value to write correctly. It seems that if I want to write a 1 to the port0/line1, I put "data = 2 ^ 1" rather than "data = 1", because by default it is the second bit of the port.» Similarly, "data = 2 ^ 7 ' high level to port0/line7. I find that this setting is surprising when you want to control an individual line. It seems more reasonable when you control the whole port. In any case, is to solve the problem!

    Thanks anyway!

    Hongkun

  • Low 6343 USB digital output

    Hello

    I use the digital output to USB-6343.

    Sometimes when I stop writing (clear the task) the rest at high output pin (I see it in the oscilloscope).

    Is it possible to set that after earasing task output pin will always be low?

    Thank you

    Leonid

    Thank you

  • digital output without DAQ Assistant

    Hello

    I can produce a digital output signal of some sampling rate 10 kHz with the acquisition of data-assist. Now I would like to implement the same functionality with normal DAQ - screw, as I have to synchronize serveral exits lateron. However, I failed get the normal screws so that they work as the DAQ assistant. The most important thing is out the arbitrary signal with 10 kHz.

    Thank you.

    Thank you very much. The idea of watching inside the acquisition of data-assist helped.

  • NEITHER 9205 digital output configuration with DAQ Assist

    Hello

    I have two NI 9205 Analog Input Modules which I have configured to read from each of their 32 channels.  I used the DAQ Assistant help generate the vi which contains the task out - DAQmx event and also the DAQmx Read vi.

    I used the Wizard twice, once for each module 9205 and then put the playback functions in a sequence structure so that only read would be carried out at the same time.  It all works very well!

    Now, I want to add in the code to wait in a loop before all loop containing playback functions, so that the user can press the GUI to send a logic 1 to the unit under test, and after it is sent immediately starts collecting data.

    The DAQ Assistant Help does not recognize the module 9205 when I try to set up a task to write a digital output.  9205 a 1 digital output so why is the wizard does not recognize this?  I also tried to create a task manually, but I got stuck.

    Someone please help.  I can reach the source if needed, but I thought that the descriptions above were sufficient.

    Thank you

    Gary

    Hello

    You are right it shows a line in this user manual, and in fact, you have found an instance where an important piece of information was left out of the documentation. This digital output line is actually only available when you use a cRIO chassis. It will not work with the chassis for the acquisition of data compact 9172. Here is a knowledge base that explains it. I'll also go ahead and file a request for corrective measures so that this note be included in the next version of this manual. Thanks for the comments.

    Chris

  • Maximum speed of digital output of the DAQ 6009

    Hi all

    I'm trying to generate a clock the digital output on my USB DAQ 6009 puse. The maximum frequency, that I was able to produce was 0.5 kHz, but I would like to generate at least 1 kHz. I HT wired port0/$line0 of the OID of data acquisition to the data acquisition ai0 and attempted to read the output via the input of an analog of the same device. I have attached the programs here. Don't know if it's right. You can help. Thanks in advance.

    150 s/s is the maximum rate of the analog output.  The 48kS/s is the maximum rate of the analog input.  Read a little more closely.

    This unit will not do what you want.  I recommend putting the hand of your representative local of NOR and discuss your needs with them.  They should be able to set you up.

  • How can I write a digital waveform to the digital output (traditional DAQ)

    Hello

    I use a NI 6023e, PCI, with 8 digital outputs. I generated a digital waveform. How can I write for a specific digital production line now?

    I only have Labview 7, so I can't use DAQmx.

    Thank you very much


  • Configuration of the digital output in the USB-6009

    I have a card for the acquisition of data USB 6009. It seems that him when DAQ card is turned on, it is always default to digital output of 'High' or 'floating '. I want to default to 'low '. Is there some setting I want to 'program' the hardware DAQ to have all the outputs low when it is powered on the value? Right now I have manually enter MAX and adjust the level 'low '. Thank you very much for your help.

    Sid05,

    Yes, it's low of 820 ohms. Unfortunately, the way in which the system is built, it is the only choice you have without having to build external circuits such as SnowMule suggested.

    AK2DM,

    Thanks for pointing the USB-6000. Finally a real, if limited, the DAQ hardware.  Nevermind, he was only 4 DIO lines.

    Lynn

  • OR PCI-MIO-16-1 & NI PCI-6070E digital outputs to scb - 68 to the electric circuit

    I am designing a circuit and trying to the NI PCI-MIO-16-1 & NI PCI-6070E DAQ from the interface and the SCB-68 to my small circuit. The outputs are 0 (low) to 5 (high) Vcc. I underdstand the NI PCI-MIO-16-1 & NI PCI-6070E are identical regarding the outputs Digital 5 volts? Also the current will be enough to light an LED on a PS2501-4 or an Octocoupler of 4N38? If this is not the case, what I need to do to amplify the current. Or if you have any suggestions on the resistances, capacitiors, etc. Please let me know. Thank you!!

    You are right in your way of thinking! From a source of 5 volts, resistance of 500 ohms in series with your LED will provide enough drive in math class of ohm's law you have already provided.

    Paul C

  • Analog input base changes during digital output in parallel

    Hi all

    We use a USB DAQ 6008 to run a simple system. AI6 is connected to a motion sensor; digital outputs control a pump. We noticed that if we read AI6 without performiong of additional tasks, the baseline is ~0.17V and when movement is detected it comes down to 0.13V. However, if we read AI6 while performing outputs digital at the same time, the base line comes down to ~0.14V and when activated he descends to ~0.1V, i.e. the line base modified by 0.03. We do something wrong or is this expected behavior?

    Thank you

    Danielle

    Danielle,

    You may have a ground loop. If the current of the pump or even digital control for the pump through the Earth wire similarly that the signal of the motion sensor, the voltage drop in the common driver because of the current pump could cause an effect like you report.

    Lynn

  • Digital output to a different voltage?

    Hi all

    I finally finished my program of data acquisition for the laboratory, which made the acquisition of AI voltage multi-channel digital outputs with timed to control the gas valves. My post-doc asked an additional feature, however, which must be able to have the analog output... It turns out that what he really wants to do with the feature is to have output digital, but with a different voltage value (he never hears on the sending of information higher or lower in a simple experiment).

    Now, the program is set up to run digital output now, it would be quite a bit of work to change in a zone of OCCUPATION. Is it possible to simply change the settings somewhere to have the ups to DO at a different voltage? I hope that this is the case from a card of 10 dollar Arduino can do 5V or 3, 3V.  Our DAQ card is the NI PCIe-6353. I looked through the data sheet and found nothing.

    Thank you.

    p.s. I suggested using additional circuits to set the tension for a given experiment. The post-doc said it was possible, but inconvinient. :/

    RaymondLo wrote:

    p.s. I suggested using additional circuits to set the tension for a given experiment. The post-doc said it was possible, but inconvinient. :/

    Well, it is just too bad for the post-doc.  Maybe he should give you better requirements next time

    OK without being in a bad mood, it is not a way to change the digital output voltage.  When you talk about DIO cards really aims to communicate only high or low and have a voltage. The only other option would be to use 4 AO ports on the card and treat them programmatically as DO.

  • Re: Satellite U400-145: no option to select HDMI as HDMI output default

    Hello

    I followed the instructions in the manual its leader to the topic on the HDMI output.
    However, when I go to the sound control panel, there is no digital output device (HDMI) listed only the Conexant Chairman and SPDIF outputs.
    I tried the default SPDIF output, which did not work.
    The HDMI works fine for the image, but the absence of the sound through it is driving me crazy.

    I'm running VISTA (fully updated) on a U400-145

    Someone at - it ideas how I can get the HDMI to appear as an output option?

    Andy

    Hello

    Go to the control panel-> Sound-> playback tab
    There, you have to click to the right.
    Then mark the options; Show all disabled devices
    Then you should see the HDMI interface.
    Now, you could activate and chose as the default value.

    Welcome them

  • No digital output on a Satellite P200

    I just bought a Satellite P200 and am trying to put it to the test. Today I tried recording on a digital audio recorder (Edirol R-4) connected to the output digital output jack but can get no signal. I can get a signal from the analog headphone jack.

    I have no problem with digital recording of my desktop PC which has an installed Sound Blaster card so it doesn't seem to be an incompatibility or a problem with the connecting cable.

    I tried to set the default output to digital mode in the setup of REALtek but no difference.

    Any suggestions gratefully received.

    David

    Anyone else had this problem?

    David

  • NI USB-6501 digital output problem

    Hello

    I use DASYLab v.11 and I'm working on an interface with the NI USB-6501 where I'm putting a digital high on four ports.

    With the module "NOR-DAQmx - digital input", I managed to read the digital inputs of the ' NI USB-6501 ".»

    It's only the "NOR-DAQmx - digital output" I can't go to work.

    Using 'NI MAX' of NOR I have easily can emmit my four LEDs in the way of my High/Low ports.

    But not with DASYLab. When you use DASYLab tension on the ports remains unchanged.

    Now, I have a switch module, generating 5/0, directly connected to the digital output module, which is assigned to my four output ports for my task.

    I also tried with a module of relay between the two without success. I also tried to use 1.5 above instead of 5 without success.

    I use the option 'Bus (0/5 supply) for the module "Digital output".

    "NI Max", I configured the ports as "active drive.

    Any suggestion of what I might be missing?

    Thank you

    Martin

    Hmm, four ports, or four lines?

    A port consists of eight lines. Each line can control an LED (ON / OFF ~ 0/5V).

    If you have created a task to dig-out to control a port, 5V to this port sending sets all lines of this port to 'high '.

    You need to 255 for each line one too high port (at the bit level: 128 + 64 + 32 + 16 + 8 + 4 + 2 + 1).<- eight="">

    Or, you can create a dig out tasks to control four lines of a specific port.

    Four lanes of the EEG DAQmx DigOut module.

    Each of the channels of the modul will feed a single line of the task/device.

    Four switches will then turn the lights, or turn off.

    Make sure, that the 'bitposition' is the number of correct line (see picture).