NOR-DAQmx subpalette is not available in my range of I/O of measure

I have a USB-6009 I installed in my computer before installing LabView (big mistake).

After I installed LabView 8.2 my camera would not work, he would be recognized by the computer, but whenever I try to read, I'd get an internal error message, so I re-installed the drive (DVD DAQmx 9.1.7) accompanying and now my subpalette function DAQmx disappeared.

My LabView still has the available controls DAQmx subpalette, I can create tasks and everything, but I can't read, write, etc. Anyone know what could have gone wrong and how can I fix it?

NOR-DAQmx 9.1.7 only supports LabVIEW 8.5, 8.6, 2009 and 2010.  You can check this KB for more information on which DAQmx versions are supported by each version of LabVIEW: NOR-DAQ and LabVIEW (for Windows) version compatibility.

The latest version of the driver NOR-DAQmx LabVIEW 8.2 support was NEITHER-DAQmx 9.1.1.  You can download it from our website here.

Tags: NI Software

Similar Questions

  • New installation on this PC, but no functions > e/s measurement > NOR-DAQmx subpalette; How to put there?

    Just installed LV8.6 on a PC and things seem fine except functions > range of measures of e/s does not contain a subpalette NOR-DAQmx. It contains only the Configuration MAX subpalette. Without creating a task, Read Channel, Start, etc etc.

    I go through stages during the installation with the DVD drivers and thought that I was installing NOR-DAQmx. MAX said that NIDAQmx is installed on the system. And I can run test MAX on a NI USB-6009 case panels.

    What should I do to get access to all NIDAQmx vi? I miss something during the installation or something?

    Thank you!

    I heard a similar problem as well... Here's the article knowledge base that describes what to do. Let me know if this corrects the problem.

  • NOR-DAQmx: nicrtsiu.dll not an Image Windows validates when you use DAQmx features

    Hi all...

    I had the problem using nor-daqmx in Labview...

    When not to use Ni-DAQmx, I have not had any problems using it

    So, try to use one of the NI Daqmx works.

    She appeared a message

    "The application or DLL C:\WINDOWS\system32\nicrtsiu.dll is not a valid Windows image

    "Please check this against your installation diskette."

    Here are versions of the programs that I use.

    -Win XP Pro 2002 SP3

    - NI Labview 7.1

    -Driver OR DAQmx 8.5.0f3

    - NO Max 4.2

    I am baffled by this error

    Please advise kindly.

    Thank you

    Clement

    Hello..

    I have the same problem, as your problem.

    If you solved your problem, I want to know your opinion.

    Please, I beg you. Send me your email address.

    My email address is "[email protected]".

    Nice day.

  • NOR-DAQmx compatibility with the system in real time

    I try to use a time system 11 with LabVIEW Real-time. When I try to connect with the driver OR DAQmx, an error saying that the version of the NOR-DAQmx drivers are bundles with the one on the real time. I tried to use different versions of the NOR-DAQmx drivers that are available in the table of compatibility, but not those who worked. I have an old computer working with the same real-time system that uses LabVIEW 9 and has the driver NOR-DAQmx 9.3.5 f0 driver installed on it but I can't find the drivers on the web. What should I do?

    I am using LabVIEW 2015 in my current computer.

    Thanks in advance.

    Looks like you can just have a lag in the versions of software installed on your host computer and the target.  In order to deploy the code from your host to your target, the two must have corresponding to libraries of functions.  See this link for instructions on updated versions of the software on your target match those installed on your host computer, the wizard of the LabVIEW RT software.  If you have already done this, and it did not help, please provide a screenshot or error code of the message that is to appear.

  • 5112 does not appear under devices NOR-DAQmx

    Why does not list MAX my 5112 under NOR-DAQmx devices, but it is displayed (undefinded) PXI system? Also, I can communicate with the map using the flexible Panel so I think that the device is correctly installed.  The main problem is that I can't use this card with my 5600 because MAX reports no digitizer.

    The 5600 must be coupled with a 5142 to form a DAA.  If you must use the 5112, you can use the 5600 on its own with the pilot DAMA downconvert only mode.  Then, it would be possible to manually use the OR-Scope driver with the 5112 to read in the data, but this is not recommended.  (It would be essentially the same using a third frequency step-down converter or digitizer).

    -Christina

  • NOR-DAQmx run options under additional installers are not displayed

    At halfway in this FAQ (http://zone.ni.com/devzone/cda/tut/p/id/3021) talks about the various Runtime options available when you include the NOR-DAQmx driver in a LabVIEW Installer - they are full and 5 different options of operation. I have 2011 LabVIEW and NOR-DAQmx 9.3.5; the only installation type I see is complete. No idea what I may have missed to install or configure correctly?

    Thank you

    Jorge

    Hi Jorge -.

    You make an excellent point raise this question.  The reality is that the documentation is a bit a type-o.  The latest version of DAQmx which allowed this ability multiple-install type was 8.7.1 that allowed full, TERM 3 and 5.  Version 8.6.1 allowed all full of Runtime and RT 2-5 selectable options.

    This was changed in the development of the installer because features such as Measurement & Automation explore (MAX) by omitting makes it much more difficult to program source files that use the DAQmx driver.  Limited choice DAQmx Installer problems in future development. SignalExpress is also omitted in some versions.   This change was made in order to avoid questions that were considered to be a problem when the installers have been created with these minimalist installs.  So I'll make this change to the Developer Zone confusion invites future readers.

    I really appreciate your message and bring it to our attention.

  • NOR-DAQmx 8.7.1 do not install

    I am trying to install my DAQCard-6024E card on an old laptop. I use the NOR-DAQmx 8.7.1 software supplied with the card. When I try to install the software, it says that the computer did not meet the minimum operating system requirements. The laptop running Windows 2000 and has an Intel Petium 3 Mobile to 800 MHz with 256 MB RAM. According to the file Readme for the NOR-DAQmx 8.7.1, for deployment software, recommended CPU is a Pentium 3 600 MHz or equivalent, 256 MB RAM and Windows 2000/XP. Can someone help me understand why this will not install. It seems to me that all my OS requirements are met.

    Matt,

    You have the latest Service Pack of Windows 2000?  I believe that the latest service pack was SP 4. Try to update to the latest pack service and then install the driver.

  • NOR-DAQ traditional vi not found in NOR-DAQmx

    Hello. We hope to control a detector via NI PCI-6533 (DIO-32HS). Our colleagues have implemented control programs in BT 7.1 to 7.3 of NOR-DAQ traditional. Now when we opened their programs on a PC installed with LV 8.6 and NOR-DAQmx 8.9, many vi was not located, including "DIO config.vi", "DIO wait.vi", "DIO clear.vi", "DIO read.vi", "Reset.vi Device", etc. We realized that they were supposed to come from vi.lib\daq\digital.llb and a few other .llb in the same folder. Apparently vi.llb\daq was installed by the traditional DAQ.

    We do not believe there is compatible with LV 8.6 traditional DAQ, which means that we need to replace those 'DIO *.vi' or 'Device reset.vi' with their counterparts DAQmx.  Could someone shed some light on how we should start working on that?

    Thank you very much.

    Xiaolan,

    Indeed, there is a traditional DAQ driver for LV 8.6. The installation can be found here.  I hope this helps.

  • Runtime error example NOR-DAQmx ANSI C in the Windows 7 virtual machine

    I am under a guest Windows 7 system with a host of Ubuntu 14, using VirtualBox and have encountered a problem running even the simplest examples provided by National Instruments, using the ANSI C API for NOR-DAQmx.

    The first time that the DAQ hardware is turned on, I can run a single measure, and after that, any other indicator displays the following error message:

    Attempted to read samples that are no longer available.
    The requested sample was previously available, but has since been overwritten.
    Increasing the buffer size, reading the data more frequently, or specifying
    a fixed number of samples to read instead of reading all available samples
    might correct the problem.

    Property: DAQmx_Read_RelativeTo
    Corresponding Value: DAQmx_Val_CurrReadPos
    Property: DAQmx_Read_Offset
    Corresponding Value: 0

    Task Name: _unamedTask(0)
    Status Code: -200279

    In order to start a new measurement, I have to restart the DAQ hardware, probably in order to clear the internal buffers that are in the process of substitution.

    I am interested in a measure of continuous tension in basic by using a callback function. (The example of ContAcq_IntClk provided by OR)

    The exactly the same setup, same version of NOR-DAQmx (9.7.5) and running on Visual Studio 2012 smoothly on a computer that is running Windows 7 directly.

    I suspect that the problem is with the internal buffer being somehow messed up because of the connection with the Virtual Machine, but do not find an elegant solution to fix it.

    The camera I use is NI USB-6289.

    Hello fromm8

    Thank you very much for your help, I managed to locate the problem.

    He was not the actual code, or something like that, it was a problem of communication between the VirtualBox machine and the physical device.

    Kept launches MAX test panels a comparable as error the program C.

    I fixed the problem by opting for VMWare, which seems to have the best compatibility with USB devices.

    See you soon!

  • Data acquisition tool NOR-DAQmx with Matlab R2012a

    Hello

    I'm trying to control NI USB-6211 of Matlab 2012 using NOR-DAQmx Data Acquisition tool:

    http://zone.NI.com/DevZone/CDA/tut/p/ID/3005

    I'm working on win7 64 bit. And I see the device AND Measurement & Automation Explorer.

    The tool does not work: DAQ_Demo_Browser do nothing. And I got the error "unexpected or unbalanced parenthesis or support" of AcqNUpdates_nonUI.m

    What is the problem?

    Thank you and best regards,

    Arthur Shulkin

    Hi Arthur,.

    Tools OR DAQmx for Acquisition of data with the Software Inc. MATLAB® from The Mathworks, supports up to the 2008 version of the MATLAB® software.  In order to use our products DAQ Multifunction with MATLAB® software, you could get back to 2008 or earlier, or instead use the Data Acquisition Toolbox provided by The Mathworks, Inc.

    Another option would be to import your ".m" files in a node MathScript in LabVIEW and use the functions of NOR-DAQmx everything in the LabVIEW development environment.  For more information on the Module LabVIEW MathScript, you can consult the information available on this link:

    Inside of the LabVIEW MathScript RT Module

    MATLAB® is a registered trademark of The MathWorks, Inc.

    Katie

  • Question about the Acquisition continues through NOR-DAQmx

    I'm a bit new to NIDAQmx methodology and I was wondering if someone could could give me some advice on accelerating certain measures of tension that I do with a case of DAQ NI USB-6363.

    I have a python script that controls and takes measurements with a few pieces of equipment of laboratory by GPIB and also takes measurements in the area of DAQ OR DAQmx via (I use a library wrapper called pylibdaqmx that interfaces with the libraries C native).  As I do with the data acquisition unit is 32 k samples at 2 MHz with a differential pair to AI0.  An example of code that performs this operation is:

    from nidaqmx import AnalogInputTask
    
    # set up task & input channeltask = AnalogInputTask()
    task.create_voltage_channel(phys_channel='Dev1/ai0', terminal='diff',       min_val=0., max_val=5.)
    task.configure_timing_sample_clock(rate=2e6, sample_mode='finite',
          samples_per_channel=32000)
    
    for i in range(number_of_loops):  < ... set up/adjust instruments ... >  task.start()  # returns an array of 32k float64 samples  # (same as DAQmxReadAnalogF64 in the C API)  data = task.read(32000)  task.stop()  < ... process data ... >
    
    # clear task, release resourcestask.clear()del task< ... etc ... >
    

    The code works fine and I can all the 32 k spot samples, but if I want to repeat this step several times in a loop, I start and stop the job every time, which takes some time and is really slow down my overall measure.

    I thought that maybe there is a way to speed up by configuring the task for continuous sample mode and just read from the channel when I want the data, but when I configure the sample for the continuous mode clock and you issue the command of reading, NOR-DAQmx gives me an error saying that the samples are no longer available , and I need to slow the rate of acquisition or increase the size of the buffer.  (I'm guessing the API wants to shoot the first 32 k samples in the buffer zone, but they have already been replaced at the time wherever I can playback control).

    What I wonder is: How can I configure the task to make the box DAQ acquire samples continuously, but give me only the last 32 samples buffer on demand k?  Looks like I'm missing something basic here, maybe some kind of trigger that I need to put in place before reading?  It doesn't seem like it should be hard to do, but as I said, I'm kinda a newbie to this.

    I understand the implementation of python that I use is not something that is supported by NEITHER, but if someone could give me some examples of how to perform a measure like this in LabView or C (or any other ideas you have to accelerate such action), I can test in these environments and to implement on my own with python.

    Thanks in advance!

    Toki

    This is something I do a bit, but I can only describe how I would do it in LabVIEW - I'm no help on the details of the C function prototypes or the python wrapper.

    In LabVIEW, there are accessed via the 'DAQmx Read property node' properties that help to implement.  One is the Mode "crush" which I'm sure must be set before performing the operation.  The other pair is known as "RelativeTo" and "Offset" and they allow you to specify what part of the CQI data buffer to read data from.   If you the config to "RelativeTo" = 'most recent sample' and 'Offset' =-32000, then whenever you read 32000 samples, they are the very latest 32000 which are already available in the buffer of data acq.  Between the readings, the task is free to overwrite the old data indefinitely.

    Note that you will need to do this continuous sampling mode and that you can explicitly set a buffer size smaller than the default which will choose DAQmx based on your fast sampling rate.

    An excerpt from LV 2010:

    -Kevin P

  • What is the latest version of the NOR-DAQmx driver? 9.7.0 vs 9.6.2

    Hi all... I'm puzzled. I found the version of NOR-DAQmx 9.7.0 released February 25, 2013 and NOR-DAQmx version 9.6.2 released March 4, 2013.

    Read the "readme" files does not lighten.

    What is going on? Two teams who does not speak with them?

    DAQmx 9.7 is always the latest version with the most recent device and the support of the API. The 'Release Date' refers to the post to date of web and it can not match the schema version.

    We'll push DAQmx 9.6.2 Update Service with the DVD of LabVIEW Driver 2012 SP1. DAQmx 9.7 will be available only for download web or included as a DVD with the hardware which requires for the purposes of support.

  • Maximum latency to read/write for PCI-DIO-32HS with NOR-DAQmx?

    I was asked to evaluate using the PCI-DIO-32HS and NOR-DAQmx on MacOS X to essentially close a control loop.  My first concern is whether or not it can run fast enough.

    So my questions are: what is the maximum latency from the moment of receiving a signal strobe reading data from external hardware, then read a single 16-bit sample and sending it upward by NOR-DAQ to a user space application?  And, similarly, what is the maximum latency of sending a unique 16-bit value from a space application user down through NOR-DAQ, then write a port and affirming a strobe signal of external-data writing material?

    Thank you.

    Hi AliasMe,

    Thanks for posting and welcome to the Forums EITHER!

    Do you need to process the data in the software and generate a value on this basis?  DIO-32HS (alias PCI-6533) offers in the buffer of inputs and outputs, but you'd have a latency in your control system relying on the OS to do treatment.  Since this period is dependent on the system, I can't give a specification for the amount of time it would take.  However, integrating data in memory, processing and writing back to the card would likely be on the order of several milliseconds.

    Generally, we recommend FPGA in situations of control like this, so you can do all the processing in hardware.  However, if provide you more details on your request we can look into all the options available.

    -John

  • Siri Option not available

    Just bought an iPad running IOS 9.2.1 3.

    In the General settings, I see no Option Siri nor I see the Option dictated in the keyboard settings.

    Restrictions is set to Off.

    I don't know this model iPad (MD-368C/A) IS compatible with Siri.

    Should Siri be downloaded and installed separately?

    It is not available in all countries. Here is the list where it is available.

    http://www.Apple.com/iOS/feature-availability/#Siri

  • cRIO9031 features not available

    I use Labview2015 to program the cRIO-9031. But when I put the project and after the device, there is a note that 'features not available' and the model cannot function normally. The problem is presented as the image below. Can someone give a little help on this? Waiting for the answer!

    It's interesting. I think you're on the right track. Given other missing dependency issues you mentioned, installing the software on your computer is damaged.

    You can try to fix it: go to the Windows Control Panel, open programs and features, double-click National Instruments software, select the Module Labview real-time and then click on repair. As a last resort, you can uninstall/reinstall LabVIEW RT and NOR-RIO.

Maybe you are looking for

  • error code - 40 in teststand deployment

    I just got a new laptop and installed all the products NOR including TestStand I had in my old laptop. Using the same deployment file I had on the old laptop, I run the deployment utility and get the following errors.  What I am doing wrong? : From n

  • error 101-1335 while installing the printer software

    I am trying to install my printer software and was the error message 101-1335 on a cabinet "data1.cab" being corrupted MSI so the install fails. I am using XP but noticed that someone who uses 7 becomes the same msg of error for its product from Koda

  • Contained in the factory of recovery failure

    3000 N100 Type: 0689 A42 Vista Basic Background: I pressed the F11 at startup the rescue and recovery program in order to have a recovery (content of the plant). Then, the system has begun to preload the customization. However, in State of play (the

  • Device-Motion Eye Webcam does not automatically appear my picture in Skype

    I tried to video calls on Skype, but my webcam could not automatically show my picture. I get the error message giving to think I close other programs that might be using my webcam. Up to now, I can say that there is no other program I know of its us

  • Queries on the mappings of DRM members

    Hi allPlease advise on the below.How is it possible to map a member in a hierarchy to a member in another hierarchy for example.Target source hierarchy member hierarchy member10000007 2000000410000008 2000000410000009 20000005And assuming I want my m