SCXI-1102/SCXI-1303 undetected in SCXI-1000 chassis

I use an installer with the SCXI-1303 module in an SCXI-1102 module in a chassis SCXI-1000 measures voltage and thermocouple.

Until recently, everything worked well. but now it seems that I am unable to communicate with the interface. I've tried two different computers and hardware is not detected in MAX. The power light and fan on the SCXI-1000 chassis light again, it just seems to be no acknowledgement of the connection to the computer.

A few more specific error codes:

* When I double-click on the DAQ Assistant, in the block diagram window and click on 'Test' one of my entries, I received the following error message:
"88705 error: the specified device is not present or is not active in the system."

* I also got the error code "error-89130: not available in NOR-DAQmx device" in Control Panel test DAQ Assistant

* During the closure of the DAQ Assistant test Panel, I see the error "error-89130: device not available for routing.

* To the MAX, if I try to 'test' the chassis or the module, I've seen errors either saying that the device is not detected, or that the device must be reset or is being restored, or the traditional DAQ drivers must be reset (I tried the reset suggestions without result)

At this point, I wonder if the Commission or the cable is pulled. Unfortunately, I don't have an other 68-pin cables to test at the moment.

Thanks in advance...

Thanks guys, I was able to solve the problem. To all those who might find this,

* I noticed in the Windows Device Manager that two PCI devices have been listed as being not installed correctly.

* After trying to update the drivers, I couldn't start the computer; got BSOD after the BIOS loading.

* Some research led me to go out two PCI cards that I was not using, and I moved also to reinstall my data acquisition on an another slot card to force material.

* Starts up, load drivers automatically and I'm gone.

Doesn't explain not why it did not work in the other computer, but in the words of Spinal Tap, the authorities have said it was better left unresolved at this point.

Tags: NI Hardware

Similar Questions

  • SCXI-1600 vs term of mass OR-6251 (SCXI-1000 chassis)

    I just need a mental health check using an SCXI-1600 instead of a NOR-6251 module.

    I have a SCXI-1000 chassis with 3 cards: (w/1300) module SCXI-1100, SCXI-1104 (w/1300) and SCXI-1126 (w/1327)

    I'm sampling 62 channels of these cards to 1200 samples per second, using LabVIEW 2011 on Win 7 (32 bit).

    We previously had a PCI-6040E to connect to the chassis, but we must move on to a USB interface.  I tried the 6251, but we have serious space and mounting issues since that a portable Setup.  I would use one SCXI-1600 module instead, but I'm afraid of all the severe warnings recommending a 6251 instead.  Are there problems I'll meet?  This seems to be such a basic configuration.

    The 1600 allows me to save space the 6251, its power supply and cable 68 pin linking the two.

    Hi CoryM,

    Since you are using Win 7 (32 bit), I plan no compatibility problems. We strongly encourage our customers to use the 6251 because it is a new product with more features, but if you need the SCXI-1600, that should be good as well. I hope this helps!

    Carisa - L.

  • To put the module chassis SCXI 1000 with sinusoidal input name

    It is imperative to turn the module chassis SCXI 1000 with a pure sinusoidal nickname of entry sector rather than sinusoidal waveform?

    This will affect the performance and reliability of the DAQ module to provide power with a sinusoidal waveform of nickname, he entered input sector?

    Concerning

    Jamal

    Hi Jamal,.

    If you look at page A-2 of the 1000 users manual found here you can see that the power needed for the 1000 is not too strict. So I guess that if your hands are almost sinusoidal at all about these tensions ca then you won't have a problem.

  • Error 200320: SCXI-1001 Chassis - Mod 1502

    Hello.

    I'll put a SCXI-1001 chassis with several modules, 3 of them 1502 with their 1314 terminal blocks for the first time. I used this chassis before reading temperatures and worked fine, so I'm sure that the problem is me.

    I get this error 200320 on simultaneous sample and hold trying to acquire data of deformation, but I think that the configuration is fine. I'm still fairly new to LabView and have tried to check the installation by the Express data acquisition module and MAX; but I get the same result.

    Please help me to implement as soon as possible, I need your knowledge.

    Thank you

    usuario

    Hi user,

    You can learn the development of LabVIEW applications and data acquisition as shown in the following links:

    http://digital.NI.com/public.nsf/allkb/7D5F60ADBFD9CDC2862572BA005DD96A?OpenDocument

    http://zone.NI.com/DevZone/CDA/tut/p/ID/2835

    The latest links explains what a global virtual channel and a task.

    Regarding the shunt calibration, you can consult the following links:

    http://digital.NI.com/public.nsf/allkb/8D81F07FBCCB6DA186256BE6000691B3?OpenDocument

    http://digital.NI.com/public.nsf/allkb/286AC9A8DC0B689986256C2C0064B183?OpenDocument

    http://zone.NI.com/DevZone/CDA/tut/p/ID/3642

    You can open a new thread and forum to do not mix topics, that's my advice.

    Kind regards

    Jesus.

  • Information SCXI

    This is the first time I work with platform SCXI. We have a SCXI-1000 chassis, a 64 mio pc HAVE, a 32 entry SCXI 1102 and a block of endings 1303 for reading signals from thermocouples 32. Now we have the opportunity to purchase a module SCXI-1100 and 1300, one block of endings. Could they to Thermocuple read? What is the difference with 1303 1102 and 1100, and 1300 respectively?

    Francesco

    Thank you very much for your answer. I should say: "that's all".

    Thank you once again

    Francesco

  • SCXI-1600 is not communicating with MAX

    All-

    Maybe that someone else has come across this.  I have connected an SCXI-1600 module in an SCXI-1000 chassis to a computer with a USB cable.  I can not get NOR-MAX to recognize the device.  I ' v tried to restart and reconnect in different orders.  I traded with a well-known USB cable.  I even took it on another computer.  On the new computer, when I plugged it first of all, I have a message "new hardware found" Windows (XP) that the-1600 had been identified, but I still couldn't it appears in the tree, even if I tried to add devices manually.  I removed the other modules in the chassis and reconnect the - 1600, still no luck.  Any advice would be great!  Thank you.

    Reinstall the suite OR-DAQ solved the problem.  Didn't try it before posting because I had the same problem on several computers and so I thought of something else.  Please do not take into account post.  Thank you.

  • SCXI and acquisition with Windows 7 x 64

    Our just updated IT to update our phones and we have received Windows 7 x 64.

    Data acquisition, we use the SCXI hardware in the SCXI-1000DC chassis and an acquisition of data SCXI-1600 USB card.  Doing my gear OR set up on the new laptop, I discovered that neither does not support Windows 7 x 64 for the SCXI-1600. The other options to connect the SCXI-1000 chassis to a laptop are the DAQcard-6062E and DAQcard-6036E, nor which supports Windows 7 x 64.  According to the information below, there are no plans to support x 64.

    What are my options to continue the use of the material, we have invested with the laptop running Windows 7 x 64?

    You will need the NI SCXI-1349 connection block and SH68-68-EP cable.  The SCXI-1349 plug at the rear of your SCXI module and cable connects the 1349 the USB device series M mass termination.  The rest of the configuration is identical to any PCI/PXI E Series device.  The USB M Series device will be the analog-to-digital converter for the system instead of the SCXI-1600.

  • Calibration of the scxi-1520 and scxi-1314

    I have a SCXI-1000 chassis with the scxi-1520 module and the scxi-1314 terminal block and I was wondering how to do a calibration on it?  What I have to send back them to OR for calibration or I can send it to a local company to perform the calibration?  If I can send it to a local provider, what the procedure would be to calibrate?  According to the manual of the module scxi-1314, everything we do is check the resistance and for the scxi-1520, it performs internal calibration.  Is this correct?

    Ron

    Hi Ron,

    You can send them to OR get these calibrated, or you may have a local company. Manual calibration procedures are available herefor many of our products online. Here are the procedures of calibration of 1314and 1520.

    Other issues on the calibration should be directed to ni.com/calibration.

    Chris W

  • SCXI and USB 6210? Getting started...

    Hello

    I'm looking to get an online data collection system and had a few quick questions of SCXI signal conditioning high-performance...  First of all I have is 1 chassis SCXI-1000, SCXI-1121, SCXI 1180 and 1100 SCXI modules assorted.  I also have a few blocks of different endings, SCXI-1300 block and a 1320.

    (1) by section "what to start' in"the SCXI chassis user manual"everything only I need is hardware of SCXI chassis (in my case the SCXI-1000 chassis), a 120 VAC power chord and a computer.  However in the ' what you need to get started section "of the manual"Getting started with the SCXI"I need also a SCXI-1200 or"Hardware DAQ"module  So my question is this, can I set the SCXI-1000 chassis directly to my laptop without any additional hardware?  I feel, I need something else, but want to make sure...

    1 (b) assuming I need a material additional DAQ to start, I can use a NI USB 6210 somehow?  I checked the manuals and little is said about SCXI.

    (2) neither manual mentions a controller, but the SCXI Advisor shoot one.  They should say if all I do is collect data, not active control?

    Thank you

    Mike

    Mike,

    There are two options for you to get your SCXI system operational. You will need either an M series or a card of E-Series you connect to your SCXI system or you could focus on the NI SCXI-1600, which is a controller for your SCXI chassis and allows you to connect your laptop to the controller via a USB cable. This controller will operate similar to the E-Series and M-series cards.

    We also offer a couple of PCMCIA cards which allows you to integrate the SCXI system with your laptop. It comes from the NI DAQCard-6036E, and NI DAQCard-6062E. Unfortunately, your NI USB-6210 is unable to be used as a device for the Acquisition of data for your SCXI chassis. The list DAQ cards which are possible to use with your system can be found here on the tab 'controller '.

    http://ohm.NI.com/advisors/SCXI/pages/SCXI/modules.XHTML?conversationContext=1

    In addition, we have an old manual SCXI of starting you can find here:

    http://digital.NI.com/manuals.nsf/WebSearch/C195895E8C9CAE3D8625691F004DB285

  • Error 1600 SCXI 200329

    Hello world

    I used my scxi-1000 chassis with the scxi-1600 daq card. Today, I get this error when I try to get data from several modules. Error number is 200329 and I have attached a screenshot for this error. Do you have an explanation for it?

    Thank you

    Bertrand

    Hey Berkayaga,

    It is possible that a single module, once damaged, which can affect three chassis.  There are a few steps that you can do to try to fix your equipment.  You might start by checking the fuses on the chassis.  Information on this may be found at Troubleshooting an SCXI chassis.  As well as specific information on the fuses themselves, there is also a documentation on where you can find them.

    If this does not work, you can consider the process of authorization to return goods (ARM) located in a National Commission of Instruments sent in for repair or calibration.

  • pilots and SCXI?

    Hello

    I have a SCXI-1000 chassis, with a 1100 unit, a unit of 1520 scxi and a 1600 scxi. Now when I connect, I get only a "USB not recognize unit."

    I installed 9.1 OR DAQmx and LabVIEW 9.01 32bits, need more? I have never tried to use before SCXI.

    I got Windows7 64 bit.

    Thanks in advance.

    For these devices, it is not currently planned to add support for 64-bit operating systems.  This is due to incompatibilities between these older devices USB design and implementation of USB in 64-bit version of Windows.

    Kind regards

  • Two measures of different thermocouples 1303

    Hi all

    I use a SCXI-1303 module and a module SCXI-1102 and SCXI-1000 chassis. The plan is to measure the temperature using thermocouples type T. 1303 modules (I have 3 of them) are completely new, and I pulled out of the box and put them on my 1102. I was testing my thermocouples to ensure they all worked when I noticed a difference in temperature between some of the thermocouples. When I started using AI16 (0-15 was already in use) temperature readings began to appear about 4 degrees below all other temperatures. Most of the thermocouples read 21 c, which is good, but those channels AI16-19 (I'm not plus far only 19), have been reading of 16 C. So I swapped the thermocouples, who had read the right temperature (from AI15) set AI16 and he showed the same offset of 4 degrees, so should not be a thermocouple problem. I put the 1303 on another module of 1102 and I swapped the 1303 modules, I had the same problem each time. The polarization resistance came in 1303 modules, so I have not tested with them. I get the same temperature difference using "integrated" and "constant" for the source of the CJC. I can hardly believe that it's a hardware problem, scxi-1303 1102 both had the same result. So I was hoping there is something else here I'm missing. I looked in the forums and not found anything quite like my problem. I read the manuals but have not encountered anything that would explain this.

    Please let me know if you need more information,

    Thank you

    Jim

    Well, I feel stupid. When you use the wizard daq, you must specify the type of thermocouple for each channel. I've had several channel sent to the type of thermocouple wrong without realizing it. Switching the appropriate type, made them of course work.

    Heh, sorry for the useless post

    Jim

  • How to use HX85BA sensors with NI DAQ system?

    Hello

    I'm trying to connect two sensors of Omega HX85BA moisture to my NI DAQ system. The sensors detect moisture, barometric pressure, and temperature, using three 0 to 10 VDC positive outputs @ 10 mA or less, with a common thread of output. I have a SCXI-1000 chassis with two modules SCXI-1100 cards, 1102 b and a single 1102. Map of 1102 uses a block to connect SCXI-1303, while the other three cards use some 1300 years. The system will mainly use thermocouples for temperature measurements, but there will be other types of sensors, such as the two HX85BAs.

    The manual of the product for the HX85BAs is not very clear on how to connect the sensor to the system. Terminal blocks each have 32 channels with the positive and the negative screw terminals, it doesn't connect three positive sensor outputs to individual channels, with the municipality of output connected to each channel? It will work installation with SCXI cards I have, or I will have to purchase different cards? The system will be expanded to include at least 4 cards in a separate chassis SCXI 1000 more, is not a problem for the purchase of different components to adapt to the HX85BAs.

    Thanks in advance!

    Hi BBalmforth,

    You can connect the sensor to the SCXI-1100 1300 TB. Connect the positive output of CH + and CH - CHSGND. Common output wire must be connected to GND.

  • thermocouple reading limited to a value

    Hello

    I'm new here and don't know much about Labview.

    T type thermocouples are connected to the SCXI-1000 chassis with 1102 module and 1303. I use 20 channels and 12 are left open. For open channels, the Labview program is around 139oC.  However, the temperature measured in these busy channels cannot exceed this temperature, no matter how hot I heated the test surface.

    Is it because of the Labview program or the resistance of the inside of the 1303 or any other possible reason?

    Thank you very much.

    Hi Johnrally,

    By '139oC' do you mean 139 ° C, 1390 ° C or something else?

    1390 ° C is well outside the range of type T temperature thermocouples according to NIST. However, it is close to the upper limit for the thermocouples of type k are you sure that your program is to specify the correct thermocouple type?

    If this does not answer your question, thanks to post a code snippet to show how your program. Also, I recommend double-checking the SCXI configuration in MAX, to make sure the SCXI-1303 module is configured correctly (and it is not configured as an SCXI-1300 module).

    Brad

  • Determination of entry-level PCI-6034E

    Hello, I have a few questions. I use K thermocouples to measure temperature using the following data acquisition system:

    • SCXI-1000 chassis
    • Connection block SCXI-1102 b
    • SCXI-1303 terminal block
    • Controller PCI-6034E

    With the SCXI-1102 b PCI-6034E, which of these two determines the polarity and the gain of the signal?

    PCI-6034E offers only bipolar mode? (- 10V + 10V /-5V + 5V /-0, 5V + 0, 5V /-0, 05V + 0, 05V)

    What are the earning opportunities in the PCI-6034E (even if it's a programmable software)?

    Thank you